Re: [time-nuts] Close-in phase noise measurements

2008-03-31 Thread Magnus Danielson
From: Richard (Rick) Karlquist [EMAIL PROTECTED]
Subject: Re: [time-nuts] Close-in phase noise measurements
Date: Sun, 30 Mar 2008 21:05:50 -0700
Message-ID: [EMAIL PROTECTED]

 The relationship between phase noise and Allan variance is
 a complex one and was described in papers at FCS in 1976
 and 1978 by my previous manager Mike Fischer (then of HP).
 I think these papers come closest to answering your question.

Add to that the added complexity that you would like to have a window function
before you DFT/FFT. Turning a FFT into Allan deviation or the reverse is among
the lost causes. Also, there is only a small overlap of interest mostly, so I
think they are best handled separatly, except when analysing spikes in the
Allan deviation.

Cheers,
Magnus

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Close-in phase noise measurements

2008-03-30 Thread Richard (Rick) Karlquist
The relationship between phase noise and Allan variance is
a complex one and was described in papers at FCS in 1976
and 1978 by my previous manager Mike Fischer (then of HP).
I think these papers come closest to answering your question.

Rick Karlquist

Jeff Mock wrote:
 This is a half-baked idea I've thinking about for awhile.  I wonder if 
 it might be possible to create a single measurement to combine allan 
 variance and phase noise in the same plot.  Allan variance usually plots 
 tau in seconds on the x-axis.  Instead, you might plot 1/s or frequency 
 on the x-axis.  This way, allan variance looks more like very close-in 
 phase noise.
 
 For example, a point where tau=1000s becomes the phase noise at 1mHz 
 (milli-hertz) from the carrier.  Combining this with more typical phase 
 noise measurements, you can create a single log-log graph covering 
 micro-hertz to hundreds of kilo-hertz.  The advantage of combining the 
 measurements into a single entity is that you get most of the 
 characterization parameters for a timebase in a single graph.
 
 Would this work?  Half-baked, I know...
 jeff
 
 
 Shane wrote:
 Do you know much about the RS FSUP50?

 http://www2.rohde-schwarz.com/en/products/test_and_measurement/product_categ
 ories/spectrum_analysis/FSUP-%7C-Key_Facts-%7C-4-%7C-966.html


 -Original Message-
 From: [EMAIL PROTECTED] [mailto:[EMAIL PROTECTED] On
 Behalf Of Bruce Griffiths
 Sent: Tuesday, March 25, 2008 8:22 PM
 To: Discussion of precise time and frequency measurement
 Subject: Re: [time-nuts] Close-in phase noise measurements

 Shane wrote:
 Wenzel has a setup you can purchase at low cost.  

 http://www.wenzel.com/pdffiles1/PNTS%201000/BP-1000-SC.pdf

 Phase noise test sets can be pricey... $200K
   
 Shane

 Their calibration method is somewhat problematic at the low frequency 
 end where the effect of the PLL and the audio amplifier low frequency 
 cutoff may be significant.
 The NIST calibration technique: 
 http://tf.nist.gov/timefreq/general/pdf/1000.pdf is far superior.

 Bruce

 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Close-in phase noise measurements

2008-03-30 Thread Bruce Griffiths
Richard (Rick) Karlquist wrote:
 The relationship between phase noise and Allan variance is
 a complex one and was described in papers at FCS in 1976
 and 1978 by my previous manager Mike Fischer (then of HP).
 I think these papers come closest to answering your question.

 Rick Karlquist

   
Essentially the mapping from phase noise to Allan variance isnt 1 to 1.

AVAR formula

Whilst it is possible to calculate the Allan variance from the phase 
noise spectrum using the above formula or a variant thereof,
the reverse isnt an unambiguous process as it is possible for different 
phase noise spectra to have the same Allan variance.
In practice the Allan Variance low pass filter frequency response isnt 
necessarily rectangular as assumed in the above formula.

Bruce

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Close-in phase noise measurements

2008-03-30 Thread Bruce Griffiths

Missing formula attached

bruce
inline: AVARCalc___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Close-in phase noise measurements

2008-03-26 Thread Jeff Mock
This is a half-baked idea I've thinking about for awhile.  I wonder if 
it might be possible to create a single measurement to combine allan 
variance and phase noise in the same plot.  Allan variance usually plots 
tau in seconds on the x-axis.  Instead, you might plot 1/s or frequency 
on the x-axis.  This way, allan variance looks more like very close-in 
phase noise.

For example, a point where tau=1000s becomes the phase noise at 1mHz 
(milli-hertz) from the carrier.  Combining this with more typical phase 
noise measurements, you can create a single log-log graph covering 
micro-hertz to hundreds of kilo-hertz.  The advantage of combining the 
measurements into a single entity is that you get most of the 
characterization parameters for a timebase in a single graph.

Would this work?  Half-baked, I know...
jeff


Shane wrote:
 Do you know much about the RS FSUP50?
 
 http://www2.rohde-schwarz.com/en/products/test_and_measurement/product_categ
 ories/spectrum_analysis/FSUP-%7C-Key_Facts-%7C-4-%7C-966.html
 
 
 -Original Message-
 From: [EMAIL PROTECTED] [mailto:[EMAIL PROTECTED] On
 Behalf Of Bruce Griffiths
 Sent: Tuesday, March 25, 2008 8:22 PM
 To: Discussion of precise time and frequency measurement
 Subject: Re: [time-nuts] Close-in phase noise measurements
 
 Shane wrote:
 Wenzel has a setup you can purchase at low cost.  

 http://www.wenzel.com/pdffiles1/PNTS%201000/BP-1000-SC.pdf

 Phase noise test sets can be pricey... $200K
   
 Shane
 
 Their calibration method is somewhat problematic at the low frequency 
 end where the effect of the PLL and the audio amplifier low frequency 
 cutoff may be significant.
 The NIST calibration technique: 
 http://tf.nist.gov/timefreq/general/pdf/1000.pdf is far superior.
 
 Bruce
 


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Close-in phase noise measurements

2008-03-25 Thread John Miles

 I am continuing my phase noise measurement quest. I gathered
 equipment (HP 8662a/11729C/8568B/multipliers) to measure
 100Hz+ from the carrier. I now need to get a grip on the
 0.1-100Hz range, which is where most of my applications are.

 What is the suggested measurement methodology for this range?

 My first idea would be to squeeze out the most of the above
 equipment...maybe add a dynamic signal analyzer like the
 35660/3561/3562? What is a good HP LF analyzer?
 This combined with multiplication (to 1GHz) and EFC
 locking, could take care of the 1-100HZ range?
 Anythings else needed?
 No ideas for  0.1-1... maybe by counting with the 5370B?

 I am sure this has been answered before but the archives are
 difficult to search...

So, I've started to look into the sub-100 Hz regime lately, but haven't had
time to get very serious about it.  Here's my take on the question:

1) The TSC-51xx analyzers are sure nifty if you have the budget and can live
with being limited to HF measurements.  They offer high performance with
less fiddling around than any other PN-measurement platform I've seen.

2) The next best thing would be a sound-card FFT option that works in
conjunction with an HF analyzer.  PN.EXE will do this at some point, using
the 11729's front-panel LF analyzer output.  It's really just a matter of
writing the UI code to support it.

3) The next best thing, after that, is a dedicated FFT analyzer, perhaps in
conjunction with an HP 3047A or 3048A system (i.e., an FFT analyzer plus a
35601A or 11848A interface).  Boxes like the HP 3561A and 3562A are neat,
but they are complete technological relics.  They are limited to about 80 dB
of dynamic range in a 100-kHz bandwidth, they take up quite a bit of space,
and they require either a lot of custom coding or some obsolete HP software.
The 3582A is in the same boat, only more so.  An 11729B/C plus a simple
sound-card interface will be the clear winner once the software support
issue is resolved.

Disclaimer: I do have a 3561A and a 3562A.  They are really cool boxes, but
if I ever write any code to support them in PN.EXE, it'll be because I think
they're cool, not because I think they're the right answer for any
FFT-analysis applications today.  Technology has left them behind in a
drastic way that hasn't happened with RF spectrum analyzers.

An FFT analyzer can still be used at offsets below 1 Hz, but at that point
people usually want to see Allan-deviation graphs rather than PN graphs.  It
becomes a different problem, since you most likely want to use a counter
rather than a spectrum analyzer.

-- john, KE5FX


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Close-in phase noise measurements

2008-03-25 Thread Bruce Griffiths
John Miles wrote:
 I am continuing my phase noise measurement quest. I gathered
 equipment (HP 8662a/11729C/8568B/multipliers) to measure
 100Hz+ from the carrier. I now need to get a grip on the
 0.1-100Hz range, which is where most of my applications are.

 What is the suggested measurement methodology for this range?

 My first idea would be to squeeze out the most of the above
 equipment...maybe add a dynamic signal analyzer like the
 35660/3561/3562? What is a good HP LF analyzer?
 This combined with multiplication (to 1GHz) and EFC
 locking, could take care of the 1-100HZ range?
 Anythings else needed?
 No ideas for  0.1-1... maybe by counting with the 5370B?

 I am sure this has been answered before but the archives are
 difficult to search...
 

 So, I've started to look into the sub-100 Hz regime lately, but haven't had
 time to get very serious about it.  Here's my take on the question:

 1) The TSC-51xx analyzers are sure nifty if you have the budget and can live
 with being limited to HF measurements.  They offer high performance with
 less fiddling around than any other PN-measurement platform I've seen.

   
The range can be extended with a couple of external mixers (plus low 
pass IF filters) and a low noise offset source.
The offset source is mixed with each of the signals to be compared to 
produce a pair of IF frequencies within the 0-30MHz range.
 2) The next best thing would be a sound-card FFT option that works in
 conjunction with an HF analyzer.  PN.EXE will do this at some point, using
 the 11729's front-panel LF analyzer output.  It's really just a matter of
 writing the UI code to support it.

   
Most, but not all, sound cards have a low frequency cutoff of 20Hz or so.
Some (but not all) sound card ADCs can dc coupled.
A high resolution dc coupled ADC may be more effective for frequencies 
below 20Hz.
 3) The next best thing, after that, is a dedicated FFT analyzer, perhaps in
 conjunction with an HP 3047A or 3048A system (i.e., an FFT analyzer plus a
 35601A or 11848A interface).  Boxes like the HP 3561A and 3562A are neat,
 but they are complete technological relics.  They are limited to about 80 dB
 of dynamic range in a 100-kHz bandwidth, they take up quite a bit of space,
 and they require either a lot of custom coding or some obsolete HP software.
 The 3582A is in the same boat, only more so.  An 11729B/C plus a simple
 sound-card interface will be the clear winner once the software support
 issue is resolved.

   
Sound card support appears to be something of a minefield, baudline 
thinks my 16 bit 48kHz motherboard sound system is a 24 bit 192kHz system.
This probably means that the frequency scale and consequently FFT filter 
noise bandwidths are unreliable.
However with a low frequency noise calibration source and set of marker 
harmonics derived from a crystal these calibration issues can be resolved.
Windows software fares little better and some crashes when set to sample 
at 192kHz (the windows machine has a sound system with a 192kHz 20 bit 
ADC system).
 Disclaimer: I do have a 3561A and a 3562A.  They are really cool boxes, but
 if I ever write any code to support them in PN.EXE, it'll be because I think
 they're cool, not because I think they're the right answer for any
 FFT-analysis applications today.  Technology has left them behind in a
 drastic way that hasn't happened with RF spectrum analyzers.

 An FFT analyzer can still be used at offsets below 1 Hz, but at that point
 people usually want to see Allan-deviation graphs rather than PN graphs.  It
 becomes a different problem, since you most likely want to use a counter
 rather than a spectrum analyzer.

 -- john, KE5FX
   
A finite beat frequency is required when using a counter.
With low beat frequencies a lower noise zero crossing detector than the 
counter input trigger circuitry is required to minimise system noise.

A dual mixer time difference system can have a lower noise floor than a 
single mixer system.

Bruce

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Close-in phase noise measurements

2008-03-25 Thread Shane
Wenzel has a setup you can purchase at low cost.  

http://www.wenzel.com/pdffiles1/PNTS%201000/BP-1000-SC.pdf

Phase noise test sets can be pricey... $200K

-Original Message-
From: [EMAIL PROTECTED] [mailto:[EMAIL PROTECTED] On
Behalf Of Bruce Griffiths
Sent: Tuesday, March 25, 2008 5:43 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Close-in phase noise measurements

John Miles wrote:
 I am continuing my phase noise measurement quest. I gathered
 equipment (HP 8662a/11729C/8568B/multipliers) to measure
 100Hz+ from the carrier. I now need to get a grip on the
 0.1-100Hz range, which is where most of my applications are.

 What is the suggested measurement methodology for this range?

 My first idea would be to squeeze out the most of the above
 equipment...maybe add a dynamic signal analyzer like the
 35660/3561/3562? What is a good HP LF analyzer?
 This combined with multiplication (to 1GHz) and EFC
 locking, could take care of the 1-100HZ range?
 Anythings else needed?
 No ideas for  0.1-1... maybe by counting with the 5370B?

 I am sure this has been answered before but the archives are
 difficult to search...
 

 So, I've started to look into the sub-100 Hz regime lately, but haven't
had
 time to get very serious about it.  Here's my take on the question:

 1) The TSC-51xx analyzers are sure nifty if you have the budget and can
live
 with being limited to HF measurements.  They offer high performance with
 less fiddling around than any other PN-measurement platform I've seen.

   
The range can be extended with a couple of external mixers (plus low 
pass IF filters) and a low noise offset source.
The offset source is mixed with each of the signals to be compared to 
produce a pair of IF frequencies within the 0-30MHz range.
 2) The next best thing would be a sound-card FFT option that works in
 conjunction with an HF analyzer.  PN.EXE will do this at some point, using
 the 11729's front-panel LF analyzer output.  It's really just a matter of
 writing the UI code to support it.

   
Most, but not all, sound cards have a low frequency cutoff of 20Hz or so.
Some (but not all) sound card ADCs can dc coupled.
A high resolution dc coupled ADC may be more effective for frequencies 
below 20Hz.
 3) The next best thing, after that, is a dedicated FFT analyzer, perhaps
in
 conjunction with an HP 3047A or 3048A system (i.e., an FFT analyzer plus a
 35601A or 11848A interface).  Boxes like the HP 3561A and 3562A are neat,
 but they are complete technological relics.  They are limited to about 80
dB
 of dynamic range in a 100-kHz bandwidth, they take up quite a bit of
space,
 and they require either a lot of custom coding or some obsolete HP
software.
 The 3582A is in the same boat, only more so.  An 11729B/C plus a simple
 sound-card interface will be the clear winner once the software support
 issue is resolved.

   
Sound card support appears to be something of a minefield, baudline 
thinks my 16 bit 48kHz motherboard sound system is a 24 bit 192kHz system.
This probably means that the frequency scale and consequently FFT filter 
noise bandwidths are unreliable.
However with a low frequency noise calibration source and set of marker 
harmonics derived from a crystal these calibration issues can be resolved.
Windows software fares little better and some crashes when set to sample 
at 192kHz (the windows machine has a sound system with a 192kHz 20 bit 
ADC system).
 Disclaimer: I do have a 3561A and a 3562A.  They are really cool boxes,
but
 if I ever write any code to support them in PN.EXE, it'll be because I
think
 they're cool, not because I think they're the right answer for any
 FFT-analysis applications today.  Technology has left them behind in a
 drastic way that hasn't happened with RF spectrum analyzers.

 An FFT analyzer can still be used at offsets below 1 Hz, but at that point
 people usually want to see Allan-deviation graphs rather than PN graphs.
It
 becomes a different problem, since you most likely want to use a counter
 rather than a spectrum analyzer.

 -- john, KE5FX
   
A finite beat frequency is required when using a counter.
With low beat frequencies a lower noise zero crossing detector than the 
counter input trigger circuitry is required to minimise system noise.

A dual mixer time difference system can have a lower noise floor than a 
single mixer system.

Bruce

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Close-in phase noise measurements

2008-03-25 Thread Bruce Griffiths
Shane wrote:
 Wenzel has a setup you can purchase at low cost.  

 http://www.wenzel.com/pdffiles1/PNTS%201000/BP-1000-SC.pdf

 Phase noise test sets can be pricey... $200K
   
Shane

Their calibration method is somewhat problematic at the low frequency 
end where the effect of the PLL and the audio amplifier low frequency 
cutoff may be significant.
The NIST calibration technique: 
http://tf.nist.gov/timefreq/general/pdf/1000.pdf is far superior.

Bruce

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Close-in phase noise measurements

2008-03-25 Thread Shane
Do you know much about the RS FSUP50?

http://www2.rohde-schwarz.com/en/products/test_and_measurement/product_categ
ories/spectrum_analysis/FSUP-%7C-Key_Facts-%7C-4-%7C-966.html


-Original Message-
From: [EMAIL PROTECTED] [mailto:[EMAIL PROTECTED] On
Behalf Of Bruce Griffiths
Sent: Tuesday, March 25, 2008 8:22 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Close-in phase noise measurements

Shane wrote:
 Wenzel has a setup you can purchase at low cost.  

 http://www.wenzel.com/pdffiles1/PNTS%201000/BP-1000-SC.pdf

 Phase noise test sets can be pricey... $200K
   
Shane

Their calibration method is somewhat problematic at the low frequency 
end where the effect of the PLL and the audio amplifier low frequency 
cutoff may be significant.
The NIST calibration technique: 
http://tf.nist.gov/timefreq/general/pdf/1000.pdf is far superior.

Bruce

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Close-in phase noise measurements

2008-03-25 Thread John Miles
 Most, but not all, sound cards have a low frequency cutoff of 20Hz or so.
 Some (but not all) sound card ADCs can dc coupled.
 A high resolution dc coupled ADC may be more effective for frequencies
 below 20Hz.

True; I'm assuming that anyone using a sound card for these purposes is
either going to bypass the coupling capacitor in front of the ADC, or
calibrate out the highpass response by adding an inverse function.  (A QEX
article gave an example of the latter technique not too long ago.)

I'm still hoping to get a 24-bit, 2.5-MSPS ADC chip hooked up via USB 2.0
fairly soon.  That will solve a multitude of problems, eliminating the need
for both a sound card and an HF analyzer.  I have C code on the PC that's
talking to the FPGA, but haven't yet tried to bring the ADC up with it.

 Sound card support appears to be something of a minefield, baudline
 thinks my 16 bit 48kHz motherboard sound system is a 24 bit 192kHz system.
 This probably means that the frequency scale and consequently FFT filter
 noise bandwidths are unreliable.
 However with a low frequency noise calibration source and set of marker
 harmonics derived from a crystal these calibration issues can be resolved.
 Windows software fares little better and some crashes when set to sample
 at 192kHz (the windows machine has a sound system with a 192kHz 20 bit
 ADC system).

Yeah, I think it'd be better not to even use the sound-card drivers if
possible.  At 10 million bytes per second (32 bits/sample at 2.5 MSPS) they
won't be an option for the hardware I'm looking at.

 A dual mixer time difference system can have a lower noise floor than a
 single mixer system.

I wonder if those are still covered by patents in various corners of the
world, the way the TSC's dual-ADC architecture appears to be...

-- john, KE5FX


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Close-in phase noise measurements

2008-03-25 Thread Bruce Griffiths
John Miles wrote:
 Most, but not all, sound cards have a low frequency cutoff of 20Hz or so.
 Some (but not all) sound card ADCs can dc coupled.
 A high resolution dc coupled ADC may be more effective for frequencies
 below 20Hz.
 

 True; I'm assuming that anyone using a sound card for these purposes is
 either going to bypass the coupling capacitor in front of the ADC, or
 calibrate out the highpass response by adding an inverse function.  (A QEX
 article gave an example of the latter technique not too long ago.)

 I'm still hoping to get a 24-bit, 2.5-MSPS ADC chip hooked up via USB 2.0
 fairly soon.  That will solve a multitude of problems, eliminating the need
 for both a sound card and an HF analyzer.  I have C code on the PC that's
 talking to the FPGA, but haven't yet tried to bring the ADC up with it.

   
 Sound card support appears to be something of a minefield, baudline
 thinks my 16 bit 48kHz motherboard sound system is a 24 bit 192kHz system.
 This probably means that the frequency scale and consequently FFT filter
 noise bandwidths are unreliable.
 However with a low frequency noise calibration source and set of marker
 harmonics derived from a crystal these calibration issues can be resolved.
 Windows software fares little better and some crashes when set to sample
 at 192kHz (the windows machine has a sound system with a 192kHz 20 bit
 ADC system).
 

 Yeah, I think it'd be better not to even use the sound-card drivers if
 possible.  At 10 million bytes per second (32 bits/sample at 2.5 MSPS) they
 won't be an option for the hardware I'm looking at.

   
That would be the AD7760 then?
 A dual mixer time difference system can have a lower noise floor than a
 single mixer system.
 

 I wonder if those are still covered by patents in various corners of the
 world, the way the TSC's dual-ADC architecture appears to be...

 -- john, KE5FX

   
John

Unlikely, the dual mixer time difference technique has been around so 
long that most patents like:

United States Patent 5128909 (issued  7 July 1992, expired on 7 July 2004).

which is an extension of the technique to multiple channels, have expired.

Bruce

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Close-in phase noise measurements

2008-03-25 Thread Bruce Griffiths
Shane wrote:
 Do you know much about the RS FSUP50?

 http://www2.rohde-schwarz.com/en/products/test_and_measurement/product_categ
 ories/spectrum_analysis/FSUP-%7C-Key_Facts-%7C-4-%7C-966.html


   
Shane

Only whats specified on the datasheet.

Its performance appears to be several 10's of dB worse than the state of 
the art (at low frequency offsets) when using the internal reference 
oscillator even when using cross correlation.
Not clear how much this is improved with an external reference oscillator.
It would be nice to know what the instrument noise floor (exclusive of 
its internal reference) is.

Bruce

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Close-in phase noise measurements

2008-03-25 Thread Bernd T-Online
Shane wrote:
 Do you know much about the RS FSUP50?
 
 http://www2.rohde-schwarz.com/en/products/test_and_measurement/product_categ
 ories/spectrum_analysis/FSUP-%7C-Key_Facts-%7C-4-%7C-966.html
 

This is a nice but expensive equipment.
Nice: can do PN measurement by phase quadrature(where you need to 
oscillators) _and_ can doe cross-correlation measuremnts (where you need 
only one external oscillator).
Expensive: depending on options abt US-$ 100,000 to $150,000

If you are ready to spend so much money, consider the Agilent Signal 
Source Analyser 5052B, who does use the cross-correlation method and has 
a lot of other features.

But both of them are no ham gear yet...

Regards

Bernd Neubig
__
AXTAL GmbH  Co. KG
Facility MOS
Wasemweg 5
D-74821 Mosbach / Germany
fon: +49 (6261) 939834
fax: +49 (6261) 939836
www.axtal.com

___