Re: [time-nuts] Prologix GPIB-NET, HP 5370A, HP5334A problem?

2008-08-29 Thread John Miles
I'm not familiar with any GPIB bugs on the 5370A, but I moved your code over
to C and ran it on my 5370B via a GPIB-LAN adapter.  The test app failed
after ~6 hours with a Winsock timeout error.  I'm thinking that was caused
by a power glitch, though, because I actually had to power-cycle the counter
(rather than the Prologix dongle) to get the program running again.  Was
that your experience?

Since then it's finished a 24-hour run without any sign of trouble.

-- john, KE5FX


If the 5370A problem is actually the 5370A gpib controller bug, then I
don't think it can be reproduced with the 5370B; I'm pretty sure that
was fixed in the B model.
However, if you want to give it a try, here's what I do (you'll have to
interpret, but it should be pretty clear)...

C version:

   GPIB_connect(atoi(argv[1]),
GPIB_error,
0,
2); // Set 20-second timeout

   GPIB_set_EOS_mode(10);
   GPIB_set_serial_read_dropout(2); // 20-second dropout

   GPIB_write(SS2);   // Sample size = 100
   GPIB_write(MD2);   // Lock out rate control, hold until MR
   GPIB_write(AR1);   // +T.I. arming only
   Sleep(2000);

   GPIB_write(MR);// Manual read (discard first reading)
   Sleep(1000);

   for (S32 h=0; h  24; h++)
  {
  for (S32 m=0; m  60; m++)
 {
 for (S32 s=0; s  60; s++)
{
GPIB_write(MR);
Sleep(1000);

printf(%d:%d:%d  %s,
h,m,s,
GPIB_read_ASC());
}
 }
  }


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Prologix GPIB-NET, HP 5370A, HP5334A problem?

2008-08-27 Thread Prologix

Bill,

How are you reading from 5334A? Is the read-after-write mode OFF?
I recommend the following sequence:

++auto 0-- turn off read-after-write
++read eoi  -- read until EOI is asserted by the instrument
++read eoi


(If your instrument does not assert EOI, specify a character to terminate
++read command. See manual for details.)

Yes, ++read will address the instrument to talk, read any output, and then
address it to listen.

We are testing an update that resets the controller if it detects a hung
GPIB bus. I can send that to you offline.

Regards,
Abdul



-Original Message-
From: [EMAIL PROTECTED] [mailto:[EMAIL PROTECTED] On
Behalf Of wje
Sent: Wednesday, August 27, 2008 3:12 PM
To: Discussion of precise time and frequency measurement
Subject: [time-nuts] Prologix GPIB-NET, HP 5370A, HP5334A problem?

I just got one of the new networked GPIB controllers, and I've been 
having some issues. I'm not sure if it's the Prologix or the 
instruments, or both.

With the 5370A, I can get samples for some period of time ranging from 
about 2 to 4 hours before the GPIB controller stops responding. I 
suspect this one might be the 5370A -488 controller bug I've heard of. 
However, there is no way to recover without power-cycling the Prologix. 
Even if this is the 5370A, the Prologix should really have an 
asynchronous reset capability.

With the 5334A, I can get samples for as long as I want, no problem. 
However, if I set the 5334A to 'WA1' mode, which is supposed to wait 
until the counter is addressed before taking each sample, what happens 
is that I can read one value. Every successive read returns the same 
value; the 5334A never triggers for a new sample. Either I don't 
understand the behavior of WA1, although HP's sample program indicates 
that what I've described is proper, or there is some incompatibility 
between the Prologix and the 5334A.

Unfortunately, the documentation for the Prologix is scanty at best. 
There is no description of the actual IEEE-488 handshake process that's 
going on. For example, does the Prologix do an unlisten after a read? 
The HP docs seem to indicate that in WA1 mode, a sample starts when the 
5334A is addressed. So, if it's never unlistened, perhaps it never takes 
a new sample?

Any thoughts or suggestions would be appreciated.

-- 
Bill Ezell
--
They said 'Windows or better'
so I used Linux.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Prologix GPIB-NET, HP 5370A, HP5334A problem?

2008-08-27 Thread John Miles

 I just got one of the new networked GPIB controllers, and I've been
 having some issues. I'm not sure if it's the Prologix or the
 instruments, or both.

 With the 5370A, I can get samples for some period of time ranging from
 about 2 to 4 hours before the GPIB controller stops responding. I
 suspect this one might be the 5370A -488 controller bug I've heard of.
 However, there is no way to recover without power-cycling the Prologix.
 Even if this is the 5370A, the Prologix should really have an
 asynchronous reset capability.

 With the 5334A, I can get samples for as long as I want, no problem.
 However, if I set the 5334A to 'WA1' mode, which is supposed to wait
 until the counter is addressed before taking each sample, what happens
 is that I can read one value. Every successive read returns the same
 value; the 5334A never triggers for a new sample. Either I don't
 understand the behavior of WA1, although HP's sample program indicates
 that what I've described is proper, or there is some incompatibility
 between the Prologix and the 5334A.

 Unfortunately, the documentation for the Prologix is scanty at best.
 There is no description of the actual IEEE-488 handshake process that's
 going on. For example, does the Prologix do an unlisten after a read?
 The HP docs seem to indicate that in WA1 mode, a sample starts when the
 5334A is addressed. So, if it's never unlistened, perhaps it never takes
 a new sample?

 Any thoughts or suggestions would be appreciated.

Bill, what commands and queries are you using with the 5370A?  I have a
couple of those new Prologix LAN dongles here and could try to repro it here
on my 5370B, if it turns out that Abdul needs to see it happen in person.

I don't have a 5334A, but it might help to send it a ++trg trigger command.
If not, you'd probably need to switch auto-read mode off for that one, and
use manual ++read commands.

-- john, KE5FX


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Prologix GPIB-NET, HP 5370A, HP5334A problem?

2008-08-27 Thread wje
YI don't use read-after-write, and I explicitly set it off before 
reading. However, I did try it once just for fun, and it does work as 
expected.

The 5334 doesn't seem to assert eoi, so I do a '++read 10', which works 
just fine except for the WA1 strangeness.

Bill Ezell
--
They said 'Windows or better'
so I used Linux.



Prologix wrote:
 Bill,

 How are you reading from 5334A? Is the read-after-write mode OFF?
 I recommend the following sequence:

 ++auto 0  -- turn off read-after-write
 ++read eoi-- read until EOI is asserted by the instrument
 ++read eoi
 

 (If your instrument does not assert EOI, specify a character to terminate
 ++read command. See manual for details.)

 Yes, ++read will address the instrument to talk, read any output, and then
 address it to listen.

 We are testing an update that resets the controller if it detects a hung
 GPIB bus. I can send that to you offline.

 Regards,
 Abdul



 -Original Message-
 From: [EMAIL PROTECTED] [mailto:[EMAIL PROTECTED] On
 Behalf Of wje
 Sent: Wednesday, August 27, 2008 3:12 PM
 To: Discussion of precise time and frequency measurement
 Subject: [time-nuts] Prologix GPIB-NET, HP 5370A, HP5334A problem?

 I just got one of the new networked GPIB controllers, and I've been 
 having some issues. I'm not sure if it's the Prologix or the 
 instruments, or both.

 With the 5370A, I can get samples for some period of time ranging from 
 about 2 to 4 hours before the GPIB controller stops responding. I 
 suspect this one might be the 5370A -488 controller bug I've heard of. 
 However, there is no way to recover without power-cycling the Prologix. 
 Even if this is the 5370A, the Prologix should really have an 
 asynchronous reset capability.

 With the 5334A, I can get samples for as long as I want, no problem. 
 However, if I set the 5334A to 'WA1' mode, which is supposed to wait 
 until the counter is addressed before taking each sample, what happens 
 is that I can read one value. Every successive read returns the same 
 value; the 5334A never triggers for a new sample. Either I don't 
 understand the behavior of WA1, although HP's sample program indicates 
 that what I've described is proper, or there is some incompatibility 
 between the Prologix and the 5334A.

 Unfortunately, the documentation for the Prologix is scanty at best. 
 There is no description of the actual IEEE-488 handshake process that's 
 going on. For example, does the Prologix do an unlisten after a read? 
 The HP docs seem to indicate that in WA1 mode, a sample starts when the 
 5334A is addressed. So, if it's never unlistened, perhaps it never takes 
 a new sample?

 Any thoughts or suggestions would be appreciated.

   

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Prologix GPIB-NET, HP 5370A, HP5334A problem?

2008-08-27 Thread wje
   If the 5370A problem is actually the 5370A gpib controller bug, then I
   don't think it can be reproduced with the 5370B; I'm pretty sure that
   was fixed in the B model.
   However, if you want to give it a try, here's what I do (you'll have to
   interpret, but it should be pretty clear):
   Prompt prompt='Enter file name' default='5370'/
   Save name='name'/
   Logger name='data' classname='FileLogger'
   Param name='filename' value='%recall name%-%date%.txt' /
   Param name='append' value='false' /
   /Logger
   Source classname='PrologixTCPSource'
   Param name='host' value='192.168.100.11'/
   /Source
SourceCmd cmd='device' param='6'/
SourceCmd cmd='++read_tmo_ms' param='4000'/
   Send msg='SS2'/
   Send msg='MD2'/
   Send msg='AR1'/
   Wait seconds='2'/
   Send msg='MR'/
   Wait seconds='1'/
   Repeat count='24'
   Repeat count='60'
   Logmessage name='data' msg='#date %fulldate%'/
   Repeat count='60'
   Send msg='MR'/
   Wait seconds='1'/
   Read timeout='4'/
   Log name='data'/
   /Repeat
   /Repeat
   /Repeat
   Sorry for the strange syntax, but that's for a Java control program I
   originally wrote to talk to my Solartron 7081 meter.
   The program is just sending out the literal string in the Sends,
   terminated with a newline.
   The Read command is sending '++read 10'. The timeout in the Read has
   nothing to do with the Prologix timeout, that's a timeout handled by my
   interpreter.
Bill Ezell
--
They said 'Windows or better'
so I used Linux.

   John Miles wrote:

I just got one of the new networked GPIB controllers, and I've been
having some issues. I'm not sure if it's the Prologix or the
instruments, or both.

With the 5370A, I can get samples for some period of time ranging from
about 2 to 4 hours before the GPIB controller stops responding. I
suspect this one might be the 5370A -488 controller bug I've heard of.
However, there is no way to recover without power-cycling the Prologix.
Even if this is the 5370A, the Prologix should really have an
asynchronous reset capability.

With the 5334A, I can get samples for as long as I want, no problem.
However, if I set the 5334A to 'WA1' mode, which is supposed to wait
until the counter is addressed before taking each sample, what happens
is that I can read one value. Every successive read returns the same
value; the 5334A never triggers for a new sample. Either I don't
understand the behavior of WA1, although HP's sample program indicates
that what I've described is proper, or there is some incompatibility
between the Prologix and the 5334A.

Unfortunately, the documentation for the Prologix is scanty at best.
There is no description of the actual IEEE-488 handshake process that's
going on. For example, does the Prologix do an unlisten after a read?
The HP docs seem to indicate that in WA1 mode, a sample starts when the
5334A is addressed. So, if it's never unlistened, perhaps it never takes
a new sample?

Any thoughts or suggestions would be appreciated.


Bill, what commands and queries are you using with the 5370A?  I have a
couple of those new Prologix LAN dongles here and could try to repro it here
on my 5370B, if it turns out that Abdul needs to see it happen in person.

I don't have a 5334A, but it might help to send it a ++trg trigger command.
If not, you'd probably need to switch auto-read mode off for that one, and
use manual ++read commands.

-- john, KE5FX


___
time-nuts mailing list -- [EMAIL PROTECTED]
To unsubscribe, go to [2]https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

References

   1. mailto:time-nuts@febo.com
   2. https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Prologix GPIB-NET, HP 5370A, HP5334A problem?

2008-08-27 Thread Prologix

You could try explicitly triggering with ++trg followed by ++read 10.


-Original Message-
From: [EMAIL PROTECTED] [mailto:[EMAIL PROTECTED] On
Behalf Of wje
Sent: Wednesday, August 27, 2008 4:14 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Prologix GPIB-NET, HP 5370A, HP5334A problem?

YI don't use read-after-write, and I explicitly set it off before 
reading. However, I did try it once just for fun, and it does work as 
expected.

The 5334 doesn't seem to assert eoi, so I do a '++read 10', which works 
just fine except for the WA1 strangeness.

Bill Ezell
--
They said 'Windows or better'
so I used Linux.



Prologix wrote:
 Bill,

 How are you reading from 5334A? Is the read-after-write mode OFF?
 I recommend the following sequence:

 ++auto 0  -- turn off read-after-write
 ++read eoi-- read until EOI is asserted by the instrument
 ++read eoi
 

 (If your instrument does not assert EOI, specify a character to terminate
 ++read command. See manual for details.)

 Yes, ++read will address the instrument to talk, read any output, and then
 address it to listen.

 We are testing an update that resets the controller if it detects a hung
 GPIB bus. I can send that to you offline.

 Regards,
 Abdul



 -Original Message-
 From: [EMAIL PROTECTED] [mailto:[EMAIL PROTECTED] On
 Behalf Of wje
 Sent: Wednesday, August 27, 2008 3:12 PM
 To: Discussion of precise time and frequency measurement
 Subject: [time-nuts] Prologix GPIB-NET, HP 5370A, HP5334A problem?

 I just got one of the new networked GPIB controllers, and I've been 
 having some issues. I'm not sure if it's the Prologix or the 
 instruments, or both.

 With the 5370A, I can get samples for some period of time ranging from 
 about 2 to 4 hours before the GPIB controller stops responding. I 
 suspect this one might be the 5370A -488 controller bug I've heard of. 
 However, there is no way to recover without power-cycling the Prologix. 
 Even if this is the 5370A, the Prologix should really have an 
 asynchronous reset capability.

 With the 5334A, I can get samples for as long as I want, no problem. 
 However, if I set the 5334A to 'WA1' mode, which is supposed to wait 
 until the counter is addressed before taking each sample, what happens 
 is that I can read one value. Every successive read returns the same 
 value; the 5334A never triggers for a new sample. Either I don't 
 understand the behavior of WA1, although HP's sample program indicates 
 that what I've described is proper, or there is some incompatibility 
 between the Prologix and the 5334A.

 Unfortunately, the documentation for the Prologix is scanty at best. 
 There is no description of the actual IEEE-488 handshake process that's 
 going on. For example, does the Prologix do an unlisten after a read? 
 The HP docs seem to indicate that in WA1 mode, a sample starts when the 
 5334A is addressed. So, if it's never unlistened, perhaps it never takes 
 a new sample?

 Any thoughts or suggestions would be appreciated.

   

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Prologix GPIB-NET, HP 5370A, HP5334A problem?

2008-08-27 Thread wje
   That does the trick, thanks. Interestingly, the HP manual doesn't
   discuss group-execute-trigger at all. It's possible that the HP-85
   programmable controllers would do a trigger before a read; all of the
   examples in the 5334A manual are written for that controller.
Bill Ezell
--
They said 'Windows or better'
so I used Linux.

   Prologix wrote:

You could try explicitly triggering with ++trg followed by ++read 10.


-Original Message-
From: [EMAIL PROTECTED] [[2]mailto:[EMAIL PROTECTED] On
Behalf Of wje
Sent: Wednesday, August 27, 2008 4:14 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Prologix GPIB-NET, HP 5370A, HP5334A problem?

YI don't use read-after-write, and I explicitly set it off before
reading. However, I did try it once just for fun, and it does work as
expected.

The 5334 doesn't seem to assert eoi, so I do a '++read 10', which works
just fine except for the WA1 strangeness.

Bill Ezell
--
They said 'Windows or better'
so I used Linux.



Prologix wrote:

Bill,

How are you reading from 5334A? Is the read-after-write mode OFF?
I recommend the following sequence:

++auto 0-- turn off read-after-write
++read eoi  -- read until EOI is asserted by the instrument
++read eoi


(If your instrument does not assert EOI, specify a character to terminate
++read command. See manual for details.)

Yes, ++read will address the instrument to talk, read any output, and then
address it to listen.

We are testing an update that resets the controller if it detects a hung
GPIB bus. I can send that to you offline.

Regards,
Abdul



-Original Message-
From: [EMAIL PROTECTED] [[4]mailto:[EMAIL PROTECTED] On
Behalf Of wje
Sent: Wednesday, August 27, 2008 3:12 PM
To: Discussion of precise time and frequency measurement
Subject: [time-nuts] Prologix GPIB-NET, HP 5370A, HP5334A problem?

I just got one of the new networked GPIB controllers, and I've been
having some issues. I'm not sure if it's the Prologix or the
instruments, or both.

With the 5370A, I can get samples for some period of time ranging from
about 2 to 4 hours before the GPIB controller stops responding. I
suspect this one might be the 5370A -488 controller bug I've heard of.
However, there is no way to recover without power-cycling the Prologix.
Even if this is the 5370A, the Prologix should really have an
asynchronous reset capability.

With the 5334A, I can get samples for as long as I want, no problem.
However, if I set the 5334A to 'WA1' mode, which is supposed to wait
until the counter is addressed before taking each sample, what happens
is that I can read one value. Every successive read returns the same
value; the 5334A never triggers for a new sample. Either I don't
understand the behavior of WA1, although HP's sample program indicates
that what I've described is proper, or there is some incompatibility
between the Prologix and the 5334A.

Unfortunately, the documentation for the Prologix is scanty at best.
There is no description of the actual IEEE-488 handshake process that's
going on. For example, does the Prologix do an unlisten after a read?
The HP docs seem to indicate that in WA1 mode, a sample starts when the
5334A is addressed. So, if it's never unlistened, perhaps it never takes
a new sample?

Any thoughts or suggestions would be appreciated.



___
time-nuts mailing list -- [EMAIL PROTECTED]
To unsubscribe, go to
[6]https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- [EMAIL PROTECTED]
To unsubscribe, go to [8]https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

References

   1. mailto:[EMAIL PROTECTED]
   2. mailto:[EMAIL PROTECTED]
   3. mailto:[EMAIL PROTECTED]
   4. mailto:[EMAIL PROTECTED]
   5. mailto:time-nuts@febo.com
   6. https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
   7. mailto:time-nuts@febo.com
   8. https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Prologix GPIB-NET, HP 5370A, HP5334A problem?

2008-08-27 Thread wje
Also, if it's of any use, the HP-85 controller does an unlisten, 
my-talk-addr, listen-address-x for each read. There's no mention of the 
HP-85 doing an automatic group-execute-trigger, and none of the examples 
issue an explicit trigger.

There is one sentence in the 5334 manual that says a g-e-t will trigger 
a new measurement, which is what I would expect.

Bill Ezell
--
They said 'Windows or better'
so I used Linux.



wje wrote:
That does the trick, thanks. Interestingly, the HP manual doesn't
discuss group-execute-trigger at all. It's possible that the HP-85
programmable controllers would do a trigger before a read; all of the
examples in the 5334A manual are written for that controller.
 Bill Ezell
 --
 They said 'Windows or better'
 so I used Linux.

Prologix wrote:

 You could try explicitly triggering with ++trg followed by ++read 10.


 -Original Message-
 From: [EMAIL PROTECTED] [[2]mailto:[EMAIL PROTECTED] On
 Behalf Of wje
 Sent: Wednesday, August 27, 2008 4:14 PM
 To: Discussion of precise time and frequency measurement
 Subject: Re: [time-nuts] Prologix GPIB-NET, HP 5370A, HP5334A problem?

 YI don't use read-after-write, and I explicitly set it off before
 reading. However, I did try it once just for fun, and it does work as
 expected.

 The 5334 doesn't seem to assert eoi, so I do a '++read 10', which works
 just fine except for the WA1 strangeness.

 Bill Ezell
 --
 They said 'Windows or better'
 so I used Linux.



 Prologix wrote:

 Bill,

 How are you reading from 5334A? Is the read-after-write mode OFF?
 I recommend the following sequence:

 ++auto 0-- turn off read-after-write
 ++read eoi  -- read until EOI is asserted by the instrument
 ++read eoi
 

 (If your instrument does not assert EOI, specify a character to terminate
 ++read command. See manual for details.)

 Yes, ++read will address the instrument to talk, read any output, and then
 address it to listen.

 We are testing an update that resets the controller if it detects a hung
 GPIB bus. I can send that to you offline.

 Regards,
 Abdul



 -Original Message-
 From: [EMAIL PROTECTED] [[4]mailto:[EMAIL PROTECTED] On
 Behalf Of wje
 Sent: Wednesday, August 27, 2008 3:12 PM
 To: Discussion of precise time and frequency measurement
 Subject: [time-nuts] Prologix GPIB-NET, HP 5370A, HP5334A problem?

 I just got one of the new networked GPIB controllers, and I've been
 having some issues. I'm not sure if it's the Prologix or the
 instruments, or both.

 With the 5370A, I can get samples for some period of time ranging from
 about 2 to 4 hours before the GPIB controller stops responding. I
 suspect this one might be the 5370A -488 controller bug I've heard of.
 However, there is no way to recover without power-cycling the Prologix.
 Even if this is the 5370A, the Prologix should really have an
 asynchronous reset capability.

 With the 5334A, I can get samples for as long as I want, no problem.
 However, if I set the 5334A to 'WA1' mode, which is supposed to wait
 until the counter is addressed before taking each sample, what happens
 is that I can read one value. Every successive read returns the same
 value; the 5334A never triggers for a new sample. Either I don't
 understand the behavior of WA1, although HP's sample program indicates
 that what I've described is proper, or there is some incompatibility
 between the Prologix and the 5334A.

 Unfortunately, the documentation for the Prologix is scanty at best.
 There is no description of the actual IEEE-488 handshake process that's
 going on. For example, does the Prologix do an unlisten after a read?
 The HP docs seem to indicate that in WA1 mode, a sample starts when the
 5334A is addressed. So, if it's never unlistened, perhaps it never takes
 a new sample?

 Any thoughts or suggestions would be appreciated.



 ___
 time-nuts mailing list -- [EMAIL PROTECTED]
 To unsubscribe, go to
 [6]https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.


 ___
 time-nuts mailing list -- [EMAIL PROTECTED]
 To unsubscribe, go to 
 [8]https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

 References

1. mailto:[EMAIL PROTECTED]
2. mailto:[EMAIL PROTECTED]
3. mailto:[EMAIL PROTECTED]
4. mailto:[EMAIL PROTECTED]
5. mailto:time-nuts@febo.com
6. https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
7. mailto:time-nuts@febo.com
8. https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.


   

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https

Re: [time-nuts] Prologix GPIB-NET, HP 5370A, HP5334A problem?

2008-08-27 Thread Gerry Glauser
Bill,
I recall there was some question on the Prologix unlisten, but I suspect
you're seeing a 5334A issue. While testing my driver at max 5334A
throughput, I found found you cannot trust a ++read until the status byte
returns data ready, else you may get the previous reading. That would happen
even at reduced throughput, depending on gate time. So I send a trigger,
then loop on status byte (with timeout of gate time+), then read. Because
that is necessary, I didn't see any advantage to the WA1 mode.
Gerry

wje [EMAIL PROTECTED]
With the 5334A, I can get samples for as long as I want, no problem. 
However, if I set the 5334A to 'WA1' mode, which is supposed to wait 
until the counter is addressed before taking each sample, what happens 
is that I can read one value. Every successive read returns the same 
value; the 5334A never triggers for a new sample. Either I don't 
understand the behavior of WA1, although HP's sample program indicates 
that what I've described is proper, or there is some incompatibility 
between the Prologix and the 5334A.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.