Re: [volt-nuts] Extender Card for Guildline 9975

2014-01-18 Thread Dr. David Kirkby
On 18 Jan 2014 14:38, Joe Hobart n...@npgcable.com wrote:

 Does anyone have or know where I can buy or borrow a 12/24 contact
 extender card
 for the Guildline 9975?  I have a single sided extender, but the signal
 demodulator board requires a double sided extender.  Of course, it is this
 card
 that needs additional testing and possible repair.

 Thank you,
 Joe Hobart
 Flagstaff, Arizona
 ___
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


[volt-nuts] What made a HP3458A so expensive

2014-01-18 Thread Frank Stellmach

Well,

the 3458A now is in production for about 25 years , and only around 50k 
units have been sold.

(That's an estimation by Joe Gellers series number collection)

As the 3458A is a niche product with homeopathic series volume, the 
development, verification/validation, special reliability engineering, 
selection and burn-in measures, QA costs had to be paid per unit in the 
beginning.


To my opinion, especially the 3458A was designed mainly for military 
requirements (Tamb 55°C). The military was willing to pay a premium (HP 
= High Price) at that time. So HP was able to realize that price.


All that development budget is long paid, and after end of the Cold War, 
the military does not order so many devices anymore, what caused the 
problems of the TM business (finally = Keysight, urgh!).


But as that market is tight, competitors are few, so the price is not 
going down, instead it's increasing over the years, from $5900 to around 
$8500.


It would be interesting to calcualte the BOM of the 3458A.
As they use many custom specific components, it should be relatively 
expensive.




There is no parameter in the HP3458A specifications, that the device has 
to be powered constantly to meet the specs.
Those very high 8ppm/yr. drift might apply only during continuous 
operation.
During power down, the LTZ should not drift at all, as the ageing 
mechanism is driven by temperature.

But there might be (there are indeed) considerable hysteresis effects.

I have set the temperature of my HP3458A to ~ 60°C, I shut it down after 
usage, and the periodical comparison to 3 other references shows a drift 
of less than 1ppm/year.


Frank




___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


[volt-nuts] What made a HP3458A so expensive

2014-01-18 Thread Mark Sims
One problem with leaving the 3458A on all the time is the vacuum fluorescent 
display.  These have a rather limited lifetime and leaving them on 24/7 puts a 
lot of hours on the tube.  Someday those displays will become unobtainium...
  
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] What made a HP3458A so expensive

2014-01-18 Thread Tom Knox
Actually in spite of it's high price I feel the 3458A represents a bargain. 
When it was introduced in 1989 it was $5900. A price it held for decades. 
Although designed in 
then 80's it's accuracy is still unsurpassed.  The 8846A is much less 
expensive because it has only about a quarter the parts. With few that 
are hand selected and/or aged. And all though both meter have well 
executed designs the result is the 3458A has about a magnitude greater 
accuracy. And anyone in Metrology can attest to the fact that the cost 
of accuracy is exponential. But the thought I wanted to contribute to the 
dialog is the real value of the 3458A is the body 
of knowledge built around the thousands of 358A's some running 
continuously for nearly three decades. The characteristics of the 3458A 
are perhaps the best documented of any electronic instrument ever made. 
That is priceless. When comparing that body of knowledge to individual 
units I have found every 3458A is a little different and seem to each 
have their own personality. In a side note, recently I owned what appeared to 
be the original 3458A. I cannot remember the serial number, but it had the 
numbered stickers identifying the boards that can be seen in the original 
service manual. And it was still  working flawlessly.

Thomas Knox



 Date: Sat, 18 Jan 2014 17:41:26 +0100
 From: frank.stellm...@freenet.de
 To: volt-nuts@febo.com
 Subject: [volt-nuts] What made a HP3458A so expensive
 
 Well,
 
 the 3458A now is in production for about 25 years , and only around 50k 
 units have been sold.
 (That's an estimation by Joe Gellers series number collection)
 
 As the 3458A is a niche product with homeopathic series volume, the 
 development, verification/validation, special reliability engineering, 
 selection and burn-in measures, QA costs had to be paid per unit in the 
 beginning.
 
 To my opinion, especially the 3458A was designed mainly for military 
 requirements (Tamb 55°C). The military was willing to pay a premium (HP 
 = High Price) at that time. So HP was able to realize that price.
 
 All that development budget is long paid, and after end of the Cold War, 
 the military does not order so many devices anymore, what caused the 
 problems of the TM business (finally = Keysight, urgh!).
 
 But as that market is tight, competitors are few, so the price is not 
 going down, instead it's increasing over the years, from $5900 to around 
 $8500.
 
 It would be interesting to calcualte the BOM of the 3458A.
 As they use many custom specific components, it should be relatively 
 expensive.
 
 
 
 There is no parameter in the HP3458A specifications, that the device has 
 to be powered constantly to meet the specs.
 Those very high 8ppm/yr. drift might apply only during continuous 
 operation.
 During power down, the LTZ should not drift at all, as the ageing 
 mechanism is driven by temperature.
 But there might be (there are indeed) considerable hysteresis effects.
 
 I have set the temperature of my HP3458A to ~ 60°C, I shut it down after 
 usage, and the periodical comparison to 3 other references shows a drift 
 of less than 1ppm/year.
 
 Frank
 
 
 
 
 ___
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.
  
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] What made a HP3458A so expensive

2014-01-18 Thread Bill Gold
In fact if you look at a 1982 HP catalog you will see that the 3456A was
selling for around $3,700 and given inflation the 3458A is still a pretty
good bargain when it was introduced in 1989.  In 1989 the 3456A was selling
for $4,600 while the 3458A was selling for $5,900.  I wonder if there will
ever be a 3459A 9.5 digit meter?  With a super miniture Josehpson Junction
for a reference?
Bill

- Original Message - 
From: Tom Knox act...@hotmail.com
To: Discussion of precise voltage measurement volt-nuts@febo.com
Sent: Saturday, January 18, 2014 1:29 PM
Subject: Re: [volt-nuts] What made a HP3458A so expensive


Actually in spite of it's high price I feel the 3458A represents a bargain.
When it was introduced in 1989 it was $5900. A price it held for decades.
Although designed in
then 80's it's accuracy is still unsurpassed.  The 8846A is much less
expensive because it has only about a quarter the parts. With few that
are hand selected and/or aged. And all though both meter have well
executed designs the result is the 3458A has about a magnitude greater
accuracy. And anyone in Metrology can attest to the fact that the cost
of accuracy is exponential. But the thought I wanted to contribute to the
dialog is the real value of the 3458A is the body
of knowledge built around the thousands of 358A's some running
continuously for nearly three decades. The characteristics of the 3458A
are perhaps the best documented of any electronic instrument ever made.
That is priceless. When comparing that body of knowledge to individual
units I have found every 3458A is a little different and seem to each
have their own personality. In a side note, recently I owned what appeared
to be the original 3458A. I cannot remember the serial number, but it had
the numbered stickers identifying the boards that can be seen in the
original service manual. And it was still  working flawlessly.

Thomas Knox



 Date: Sat, 18 Jan 2014 17:41:26 +0100
 From: frank.stellm...@freenet.de
 To: volt-nuts@febo.com
 Subject: [volt-nuts] What made a HP3458A so expensive

 Well,

 the 3458A now is in production for about 25 years , and only around 50k
 units have been sold.
 (That's an estimation by Joe Gellers series number collection)

 As the 3458A is a niche product with homeopathic series volume, the
 development, verification/validation, special reliability engineering,
 selection and burn-in measures, QA costs had to be paid per unit in the
 beginning.

 To my opinion, especially the 3458A was designed mainly for military
 requirements (Tamb 55°C). The military was willing to pay a premium (HP
 = High Price) at that time. So HP was able to realize that price.

 All that development budget is long paid, and after end of the Cold War,
 the military does not order so many devices anymore, what caused the
 problems of the TM business (finally = Keysight, urgh!).

 But as that market is tight, competitors are few, so the price is not
 going down, instead it's increasing over the years, from $5900 to around
 $8500.

 It would be interesting to calcualte the BOM of the 3458A.
 As they use many custom specific components, it should be relatively
 expensive.



 There is no parameter in the HP3458A specifications, that the device has
 to be powered constantly to meet the specs.
 Those very high 8ppm/yr. drift might apply only during continuous
 operation.
 During power down, the LTZ should not drift at all, as the ageing
 mechanism is driven by temperature.
 But there might be (there are indeed) considerable hysteresis effects.

 I have set the temperature of my HP3458A to ~ 60°C, I shut it down after
 usage, and the periodical comparison to 3 other references shows a drift
 of less than 1ppm/year.

 Frank




 ___
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] What made a HP3458A so expensive

2014-01-18 Thread Joe Hobart
Bill,

Would a Josephson Junction standard need to be calibrated?

Adjusted and maybe compared, yes, but you should not need to calibrate a primary
standard?  Years ago we had HP Cesium Frequency Standards at work.  There were
primary standards and good to 4E-12 with no additional calibration.

Joe


On 1/18/2014 3:37 PM, Tom Miller wrote:
 Can you imagine what it would cost to get that calibrated?
 
 Tom
 - Original Message - From: Bill Gold wpgold3...@att.net
 To: Discussion of precise voltage measurement volt-nuts@febo.com
 Sent: Saturday, January 18, 2014 5:17 PM
 Subject: Re: [volt-nuts] What made a HP3458A so expensive
 
 
 In fact if you look at a 1982 HP catalog you will see that the 3456A was
 selling for around $3,700 and given inflation the 3458A is still a pretty
 good bargain when it was introduced in 1989.  In 1989 the 3456A was selling
 for $4,600 while the 3458A was selling for $5,900.  I wonder if there will
 
 ever be a 3459A 9.5 digit meter?  With a super miniture Josehpson Junction
 for a reference?
 Bill
 
 - Original Message - From: Tom Knox act...@hotmail.com
 To: Discussion of precise voltage measurement volt-nuts@febo.com
 Sent: Saturday, January 18, 2014 1:29 PM
 Subject: Re: [volt-nuts] What made a HP3458A so expensive
 
 
 Actually in spite of it's high price I feel the 3458A represents a bargain.
 When it was introduced in 1989 it was $5900. A price it held for decades.
 Although designed in
 then 80's it's accuracy is still unsurpassed.  The 8846A is much less
 expensive because it has only about a quarter the parts. With few that
 are hand selected and/or aged. And all though both meter have well
 executed designs the result is the 3458A has about a magnitude greater
 accuracy. And anyone in Metrology can attest to the fact that the cost
 of accuracy is exponential. But the thought I wanted to contribute to the
 dialog is the real value of the 3458A is the body
 of knowledge built around the thousands of 358A's some running
 continuously for nearly three decades. The characteristics of the 3458A
 are perhaps the best documented of any electronic instrument ever made.
 That is priceless. When comparing that body of knowledge to individual
 units I have found every 3458A is a little different and seem to each
 have their own personality. In a side note, recently I owned what appeared
 
 to be the original 3458A. I cannot remember the serial number, but it had
 the numbered stickers identifying the boards that can be seen in the
 original service manual. And it was still  working flawlessly.
 
 Thomas Knox
 
 
 
 Date: Sat, 18 Jan 2014 17:41:26 +0100
 From: frank.stellm...@freenet.de
 To: volt-nuts@febo.com
 Subject: [volt-nuts] What made a HP3458A so expensive

 Well,

 the 3458A now is in production for about 25 years , and only around 50k
 units have been sold.
 (That's an estimation by Joe Gellers series number collection)

 As the 3458A is a niche product with homeopathic series volume, the
 development, verification/validation, special reliability engineering,
 selection and burn-in measures, QA costs had to be paid per unit in the
 beginning.

 To my opinion, especially the 3458A was designed mainly for military
 requirements (Tamb 55°C). The military was willing to pay a premium (HP
 = High Price) at that time. So HP was able to realize that price.

 All that development budget is long paid, and after end of the Cold War,
 
 the military does not order so many devices anymore, what caused the
 problems of the TM business (finally = Keysight, urgh!).

 But as that market is tight, competitors are few, so the price is not
 going down, instead it's increasing over the years, from $5900 to around
 
 $8500.

 It would be interesting to calcualte the BOM of the 3458A.
 As they use many custom specific components, it should be relatively
 expensive.



 There is no parameter in the HP3458A specifications, that the device has
 
 to be powered constantly to meet the specs.
 Those very high 8ppm/yr. drift might apply only during continuous
 operation.
 During power down, the LTZ should not drift at all, as the ageing
 mechanism is driven by temperature.
 But there might be (there are indeed) considerable hysteresis effects.

 I have set the temperature of my HP3458A to ~ 60°C, I shut it down after
 usage, and the periodical comparison to 3 other references shows a drift
 
 of less than 1ppm/year.

 Frank

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] What made a HP3458A so expensive

2014-01-18 Thread Chuck Harris

Of course you would need to calibrate a JJ standard.  If you recall,
the JJ works by being cooled to some LN2 like temperature, and then
being fed with a microwave frequency that is part of the standard's
definition.  If the temperature is wrong, or the frequency is wrong,
or ..., the voltage will be wrong.  The whole point behind the JJ is
that it connects the standard voltage to the standard frequency/time
... the most accurately specified of the known standards.

Cs standards, even though they are primary standards, are also in need
of calibration because the frequency they create is dependent on the
magnetic field surrounding the tube, and the construction of the tube.

All Josephson Junction V standards produce a slightly different voltage,
and all Cs standards produce a slightly different frequency.

-Chuck Harris

Joe Hobart wrote:

Bill,

Would a Josephson Junction standard need to be calibrated?

Adjusted and maybe compared, yes, but you should not need to calibrate a primary
standard?  Years ago we had HP Cesium Frequency Standards at work.  There were
primary standards and good to 4E-12 with no additional calibration.

Joe

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] What made a HP3458A so expensive

2014-01-18 Thread Tom Knox
A Josephson Junction is not a Cesium Standard. It is not quite to the Plug and 
Play level. And takes a few racks worth of space.
Operation and Maintenance is more difficult then most calibrations. But big 
advances are being made. 
At some point in the not to distant future I foresee JJ arrays being configured 
as not just a references,
but full function calibrators with direct quantum synthesis of any DC, AC, 
Ohms, and Current. Sam Benz
has lead much of this research. 
http://www.nist.gov/pml/div686/devices/voltage.cfm 
I am amazed that greater advances have not been made in zeners references or 
more specifically their packaging


Thomas Knox



 Date: Sat, 18 Jan 2014 17:44:43 -0700
 From: n...@npgcable.com
 To: tmiller11...@verizon.net; volt-nuts@febo.com
 Subject: Re: [volt-nuts] What made a HP3458A so expensive
 
 Bill,
 
 Would a Josephson Junction standard need to be calibrated?
 
 Adjusted and maybe compared, yes, but you should not need to calibrate a 
 primary
 standard?  Years ago we had HP Cesium Frequency Standards at work.  There were
 primary standards and good to 4E-12 with no additional calibration.
 
 Joe
 
 
 On 1/18/2014 3:37 PM, Tom Miller wrote:
  Can you imagine what it would cost to get that calibrated?
  
  Tom
  - Original Message - From: Bill Gold wpgold3...@att.net
  To: Discussion of precise voltage measurement volt-nuts@febo.com
  Sent: Saturday, January 18, 2014 5:17 PM
  Subject: Re: [volt-nuts] What made a HP3458A so expensive
  
  
  In fact if you look at a 1982 HP catalog you will see that the 3456A was
  selling for around $3,700 and given inflation the 3458A is still a pretty
  good bargain when it was introduced in 1989.  In 1989 the 3456A was selling
  for $4,600 while the 3458A was selling for $5,900.  I wonder if there will
  
  ever be a 3459A 9.5 digit meter?  With a super miniture Josehpson Junction
  for a reference?
  Bill
  
  - Original Message - From: Tom Knox act...@hotmail.com
  To: Discussion of precise voltage measurement volt-nuts@febo.com
  Sent: Saturday, January 18, 2014 1:29 PM
  Subject: Re: [volt-nuts] What made a HP3458A so expensive
  
  
  Actually in spite of it's high price I feel the 3458A represents a bargain.
  When it was introduced in 1989 it was $5900. A price it held for decades.
  Although designed in
  then 80's it's accuracy is still unsurpassed.  The 8846A is much less
  expensive because it has only about a quarter the parts. With few that
  are hand selected and/or aged. And all though both meter have well
  executed designs the result is the 3458A has about a magnitude greater
  accuracy. And anyone in Metrology can attest to the fact that the cost
  of accuracy is exponential. But the thought I wanted to contribute to the
  dialog is the real value of the 3458A is the body
  of knowledge built around the thousands of 358A's some running
  continuously for nearly three decades. The characteristics of the 3458A
  are perhaps the best documented of any electronic instrument ever made.
  That is priceless. When comparing that body of knowledge to individual
  units I have found every 3458A is a little different and seem to each
  have their own personality. In a side note, recently I owned what appeared
  
  to be the original 3458A. I cannot remember the serial number, but it had
  the numbered stickers identifying the boards that can be seen in the
  original service manual. And it was still  working flawlessly.
  
  Thomas Knox
  
  
  
  Date: Sat, 18 Jan 2014 17:41:26 +0100
  From: frank.stellm...@freenet.de
  To: volt-nuts@febo.com
  Subject: [volt-nuts] What made a HP3458A so expensive
 
  Well,
 
  the 3458A now is in production for about 25 years , and only around 50k
  units have been sold.
  (That's an estimation by Joe Gellers series number collection)
 
  As the 3458A is a niche product with homeopathic series volume, the
  development, verification/validation, special reliability engineering,
  selection and burn-in measures, QA costs had to be paid per unit in the
  beginning.
 
  To my opinion, especially the 3458A was designed mainly for military
  requirements (Tamb 55°C). The military was willing to pay a premium (HP
  = High Price) at that time. So HP was able to realize that price.
 
  All that development budget is long paid, and after end of the Cold War,
  
  the military does not order so many devices anymore, what caused the
  problems of the TM business (finally = Keysight, urgh!).
 
  But as that market is tight, competitors are few, so the price is not
  going down, instead it's increasing over the years, from $5900 to around
  
  $8500.
 
  It would be interesting to calcualte the BOM of the 3458A.
  As they use many custom specific components, it should be relatively
  expensive.
 
 
 
  There is no parameter in the HP3458A specifications, that the device has
  
  to be powered constantly to meet the specs.
  Those very high 8ppm/yr. drift might apply only during continuous