Re: [volt-nuts] Datron 1082

2014-04-21 Thread Chuck Harris

Just in case you haven't yet: Always start any investigation with
checking the power supply for voltage and ripple.  Test the main
electrolytic capacitors for capacitance, and ESR.

Nothing in the unit's self testing ability can be trusted if the
power supply is putting out large amounts of ripple, or is at the
wrong voltage.

-Chuck Harris

Randy Evans wrote:

I have a Datron 1082 DMM that just recently stopped working.  When I power
up, it gives a FAIL message on the front panel.  I found a flowchart in
an abbreviated servicing manual that indicates it's from a failure of the
Check for uncorrupted calibration store.  I Assume this likely means the
cal tables are corrupted?  Unfortunately, I don't have sufficient
information (e.g., schematics) to do any further troubleshooting.  Can
anyone help with a complete servicing manual and/or further diagnostic help?


I did download the Datron 1071DMM  OS manual from K04BB but I don't know
how similar it is to the 1082.  Can anyone comment on what the differences
are?



Thanks,



Randy Evans

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Datron 1082

2014-04-21 Thread Randy Evans
I believe the voltages are correct but I have no schematics or
troubleshooting instructions to confirm.  I do see what looks like a
battery in the memory section of the 1082 and it measures 0.08 VDC. Not a
good sign.  I am really hoping someone has a Datron 1082 Service manual
with schematics that they can steer me to, preferably a soft copy so I can
better troubleshoot the unit.  It might just need a new battery (plus a
re-cal) to get it going again but I hate going blind.

Randy


On Mon, Apr 21, 2014 at 8:15 AM, Tom Knox act...@hotmail.com wrote:

 As with many products of this age the tantalum caps are a common point of
 failure in the 1082.
 Although this current problem sounds like a memory battery.

 Thomas Knox



  Date: Mon, 21 Apr 2014 11:03:56 -0400
  From: cfhar...@erols.com
  To: volt-nuts@febo.com
  Subject: Re: [volt-nuts] Datron 1082
 
  Just in case you haven't yet: Always start any investigation with
  checking the power supply for voltage and ripple.  Test the main
  electrolytic capacitors for capacitance, and ESR.
 
  Nothing in the unit's self testing ability can be trusted if the
  power supply is putting out large amounts of ripple, or is at the
  wrong voltage.
 
  -Chuck Harris
 
  Randy Evans wrote:
   I have a Datron 1082 DMM that just recently stopped working.  When I
 power
   up, it gives a FAIL message on the front panel.  I found a flowchart
 in
   an abbreviated servicing manual that indicates it's from a failure of
 the
   Check for uncorrupted calibration store.  I Assume this likely means
 the
   cal tables are corrupted?  Unfortunately, I don't have sufficient
   information (e.g., schematics) to do any further troubleshooting.  Can
   anyone help with a complete servicing manual and/or further diagnostic
 help?
  
  
   I did download the Datron 1071DMM  OS manual from K04BB but I don't
 know
   how similar it is to the 1082.  Can anyone comment on what the
 differences
   are?
  
  
  
   Thanks,
  
  
  
   Randy Evans
  ___
  volt-nuts mailing list -- volt-nuts@febo.com
  To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
  and follow the instructions there.

 ___
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.