Re: [volt-nuts] Datron 4920M, AC Voltmeter, for sale on ebay - 1 day left

2014-09-01 Thread ben
Hello Stephen,
I agree that the 4920's will often drift more than their spec after 1 year. 
I find it more convenient to derate them slightly - and use them wisely. 
Note the manual states their absolute specs are valid for 6 months, but I 
still maintain them on a 1 year cycle. Even when de-rated they still have 
very good short term stability over 10 minutes and for weeks after.

On the discussion of inputs at low points in ranges, I find this is where 
the 4920M's will most often cause calibration failures against their specs. 
Perhaps optimistically, Datron has specified their absolute reading spec 
at, say, 3V range, to be +-30ppm of reading (40 to 20kHz) for any input 
from 30% to 100% of range. No floor or percentage of range limiting spec is 
stated.

When the sad day finally arrives and I need to repair units I would resort 
to cannibilising parts from these 4920M's just to keep my more capable 4920 
going.

regards,   
ben.






From: Stephen Grady grady.st...@gmail.com
Sent: Monday, September 01, 2014 5:37 AM
To: Discussion of precise voltage measurement volt-nuts@febo.com
Subject: Re: [volt-nuts] Datron 4920M, AC Voltmeter, for sale on ebay - 1 
day left 

Charlie,


The Error Ur is an under-range error; you have to apply an input above 
10% of the range before it will display a reading. The 4920's are a very 
nice instrument. Their only problem is that they are all so old that they 
are reaching the stage where some components are drifting excessively or 
failing. I have come across 4920's that are drifting a little (more than 
there spec) between cals (and yes they need to be cal'd annually), another 
4920 I came across had an intermittent failure in one of its power 
supplies.

Kind Regards,

Stephen Grady
Sydney Australia

-Original Message-
From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of Charles 
Black
Sent: Monday, 1 September 2014 3:19 AM
To: b...@veritechmeasurements.com.au; Discussion of precise voltage 
measurement
Subject: Re: [volt-nuts] Datron 4920M, AC Voltmeter, for sale on ebay - 1 
day left

Hi Ben,

I am going to check this out today to find if it might be helpfull around 
here for my AC calibration. It has an error message: Error Ur 
so some fixing is going to be needed. On the back it says it's got Option 
80, whatever that is.

Charlie


On 8/31/2014 5:37 AM, ben wrote:
 Hello all,
 Just a heads up. I noted there is a lonely Datron / Wavetek 4920M AC 
 voltmeter for sale on ebay, starting at US$650 (ending in 1 day !). I 
 have little idea of its true working state or not (picture shows it 
powered on).
 If fully working it would be a good buy for an AC voltmeter that is, I 
 reckon, better than an HP 3458A. Only drawback with these models is 
 the voltage input shell is always earthed, not floating. I have two of 
 these 4920M's already, not really tempted by a third. I have a paper 
 copies of operator, calibration, and service manual if anyone interested.

 regards,   ben.
 ___
 volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to 
 https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.





___
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


---
This email is free from viruses and malware because avast! Antivirus 
protection is active.
http://www.avast.com

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
 
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Re: [volt-nuts] Datron 4920M, AC Voltmeter, for sale on ebay - 1 day left

2014-09-01 Thread acbern
charlie,

you can expand these tvcs by range resistors to work as higher voltage tvcs. if 
you do it right (rf-type setup) you can work up to a couple hundred khz with 
just minor additional error beyond its error in its orgininal voltage range. 
need to calibrate them however. using an attenuator is not a good idea, too 
unprecise. still tvcs are the most precise way to measure, if done right. 
pretty tme consuming though. but you need them e.g. to calibrate a 4920. 

adrian

 Gesendet: Sonntag, 31. August 2014 um 21:41 Uhr
 Von: Charles Black cbl...@centurytel.net
 An: Discussion of precise voltage measurement volt-nuts@febo.com
 Betreff: Re: [volt-nuts] Datron 4920M, AC Voltmeter, for sale on ebay - 1 day 
 left

 Hi Stephen,
 
 Thanks for the info on the 4920M. It might be that I should pass on this 
 but it is nice to know that it is probably working. If I don't have the 
 necessary items needed for calibration it might still be a good source 
 uncalibrated.
 
 I was hoping that the 4920 had better minimum voltage out steps than my 
 present Rigol 4062 for 3458a AC calibration. I have two EL 1400 0.25V 
 voltage thermal converters that I was trying to use in a test setup but 
 they had such a low voltage range that I was forced to use my ATV-60 
 attenuator. It all kind of worked but a 0.1 db minimum step  is a little 
 course for easy use. The Rigol also suffers from this malady.
 
 Charlie
 
 On 8/31/2014 1:06 PM, Stephen Grady wrote:
  Charlie,
 
 
  The Error Ur is an under-range error; you have to apply an input above 
  10% of the range before it will display a reading. The 4920's are a very 
  nice instrument. Their only problem is that they are all so old that they 
  are reaching the stage where some components are drifting excessively or 
  failing. I have come across 4920's that are drifting a little (more than 
  there spec) between cals (and yes they need to be cal'd annually), another 
  4920 I came across had an intermittent failure in one of its power supplies.
 
  Kind Regards,
 
  Stephen Grady
  Sydney Australia
 
  -Original Message-
  From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of Charles 
  Black
  Sent: Monday, 1 September 2014 3:19 AM
  To: b...@veritechmeasurements.com.au; Discussion of precise voltage 
  measurement
  Subject: Re: [volt-nuts] Datron 4920M, AC Voltmeter, for sale on ebay - 1 
  day left
 
  Hi Ben,
 
  I am going to check this out today to find if it might be helpfull around 
  here for my AC calibration. It has an error message: Error Ur
  so some fixing is going to be needed. On the back it says it's got Option 
  80, whatever that is.
 
  Charlie
 
 
  On 8/31/2014 5:37 AM, ben wrote:
  Hello all,
  Just a heads up. I noted there is a lonely Datron / Wavetek 4920M AC
  voltmeter for sale on ebay, starting at US$650 (ending in 1 day !). I
  have little idea of its true working state or not (picture shows it 
  powered on).
  If fully working it would be a good buy for an AC voltmeter that is, I
  reckon, better than an HP 3458A. Only drawback with these models is
  the voltage input shell is always earthed, not floating. I have two of
  these 4920M's already, not really tempted by a third. I have a paper
  copies of operator, calibration, and service manual if anyone interested.
 
  regards,   ben.
  ___
  volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
  https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
  and follow the instructions there.
 
 
 
 
  ___
  volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to 
  https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
  and follow the instructions there.
 
 
  ---
  This email is free from viruses and malware because avast! Antivirus 
  protection is active.
  http://www.avast.com
 
  ___
  volt-nuts mailing list -- volt-nuts@febo.com
  To unsubscribe, go to 
  https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
  and follow the instructions there.
 
 
 
 
 
 ___
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Re: [volt-nuts] Datron 4920M, AC Voltmeter, for sale on ebay - 1 day left

2014-09-01 Thread Charles Black

Hi Adrian,

I had some trouble testing my EL 1400's with my precsion attenuator 
(1.2Ghz B/W) not having a small enough minimum step for easy setting. It 
was fairly stable though. I checked everything with my scope and it was 
fairly flat to 20mhz with a gentle falloff to 30mhz if I remember 
correctly. The EL 1400's are useable to 300mhz with corrections that I 
don't have.


I will keep an eye out for a low priced 4920. It will have to be broken 
bad enough to be low priced though.


Charlie

On 9/1/2014 3:29 PM, acb...@gmx.de wrote:

charlie,

you can expand these tvcs by range resistors to work as higher voltage tvcs. if 
you do it right (rf-type setup) you can work up to a couple hundred khz with 
just minor additional error beyond its error in its orgininal voltage range. 
need to calibrate them however. using an attenuator is not a good idea, too 
unprecise. still tvcs are the most precise way to measure, if done right. 
pretty tme consuming though. but you need them e.g. to calibrate a 4920.

adrian


Gesendet: Sonntag, 31. August 2014 um 21:41 Uhr
Von: Charles Black cbl...@centurytel.net
An: Discussion of precise voltage measurement volt-nuts@febo.com
Betreff: Re: [volt-nuts] Datron 4920M, AC Voltmeter, for sale on ebay - 1 day 
left

Hi Stephen,

Thanks for the info on the 4920M. It might be that I should pass on this
but it is nice to know that it is probably working. If I don't have the
necessary items needed for calibration it might still be a good source
uncalibrated.

I was hoping that the 4920 had better minimum voltage out steps than my
present Rigol 4062 for 3458a AC calibration. I have two EL 1400 0.25V
voltage thermal converters that I was trying to use in a test setup but
they had such a low voltage range that I was forced to use my ATV-60
attenuator. It all kind of worked but a 0.1 db minimum step  is a little
course for easy use. The Rigol also suffers from this malady.

Charlie

On 8/31/2014 1:06 PM, Stephen Grady wrote:

Charlie,


The Error Ur is an under-range error; you have to apply an input above 10% of 
the range before it will display a reading. The 4920's are a very nice instrument. Their 
only problem is that they are all so old that they are reaching the stage where some 
components are drifting excessively or failing. I have come across 4920's that are 
drifting a little (more than there spec) between cals (and yes they need to be cal'd 
annually), another 4920 I came across had an intermittent failure in one of its power 
supplies.

Kind Regards,

Stephen Grady
Sydney Australia

-Original Message-
From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of Charles Black
Sent: Monday, 1 September 2014 3:19 AM
To: b...@veritechmeasurements.com.au; Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Datron 4920M, AC Voltmeter, for sale on ebay - 1 day 
left

Hi Ben,

I am going to check this out today to find if it might be helpfull around here for my AC 
calibration. It has an error message: Error Ur
so some fixing is going to be needed. On the back it says it's got Option 80, 
whatever that is.

Charlie


On 8/31/2014 5:37 AM, ben wrote:

Hello all,
Just a heads up. I noted there is a lonely Datron / Wavetek 4920M AC
voltmeter for sale on ebay, starting at US$650 (ending in 1 day !). I
have little idea of its true working state or not (picture shows it powered on).
If fully working it would be a good buy for an AC voltmeter that is, I
reckon, better than an HP 3458A. Only drawback with these models is
the voltage input shell is always earthed, not floating. I have two of
these 4920M's already, not really tempted by a third. I have a paper
copies of operator, calibration, and service manual if anyone interested.

regards,   ben.
___
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.




___
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


---
This email is free from viruses and malware because avast! Antivirus protection 
is active.
http://www.avast.com

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.





___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.






___
volt-nuts mailing