Re: [volt-nuts] What's the probability of a random used 3458A passing a Keysight calibration?

2018-01-15 Thread John Phillips
my experience is that most of the eBay meters that do not give errors are
very close to spec. These old meters do not drift as much as a new meter.
If you have a good 10 volts and 10k resistor calibration is a snap...
verifying cal in not as easy. The high-frequency AC cal is more difficult.

On Mon, Jan 15, 2018 at 7:18 PM, J. L. Trantham  wrote:

> Forgot to add.  All were purchased on theBay.
>
> Joe
>
> -Original Message-
> From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of J. L.
> Trantham
> Sent: Monday, January 15, 2018 9:12 PM
> To: 'Discussion of precise voltage measurement'
> Subject: Re: [volt-nuts] What's the probability of a random used 3458A
> passing a Keysight calibration?
>
> Dave,
>
> I have three of these meters that have all been sent to Agilent/Keysight
> for the 'Agilent/Keysight' Calibration and all have passed.  Two of the
> three were still 'in spec' when arrived.  The third lost its calibration
> when I failed to properly remove the 'CalRam' and read it.  I think I tried
> to read it when still 'hot' from the removal process.  There is a process
> that you can use via HPIB to read the contents of the CalRam but I did not
> know about it at the time.  I wanted to remove the chip, install a socket
> and install a new chip before I sent it back to Agilent/Keysight.
>
> I had previously sent a Solartron 7081 to Ametek in England for
> calibration and I used it to provide a 'transfer standard' to do my own 'in
> house' recalibration of that 3458A and it was 'in spec' in all ranges when
> it went back to Agilent, except a couple of the AC voltage ranges.
> However, it was able to be successfully 'recalibrated' and was 'in spec' on
> its next trip for recalibration.  That was at a time when the calibrations
> were done in Loveland.
>
> Once a 3458A has passed it's 'recalibration', you can purchase the 'Repair
> Agreement' for $192 per year with up to 5 years at a time (for a slight
> discount) that will allow you to send it back to Keysight for repair and
> recalibration for 'no charge' other than the cost of the 'Repair Agreement'
> if it ever fails during the covered time frame.
>
> Hope this helps.
>
> Joe
>
> -Original Message-
> From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of Dr.
> David Kirkby
> Sent: Monday, January 15, 2018 3:56 PM
> To: Discussion of precise voltage measurement
> Subject: [volt-nuts] What's the probability of a random used 3458A passing
> a Keysight calibration?
>
> There's an "interesting" $5400 buy-it-now on eBay at the minute. It's
> described as "Agilent 3458A 8.5 digit Multimeter CALIBRATED".
>
> http://www.ebay.co.uk/itm/Agilent-3458A-8-5-Digit-Multimeter-CALIBRATED/
> 232593692038
>
> When one reads the description, one can see it is an HP (not Agilent)
> 3458A, that was sent to Keysight for calibration, but *failed* on the 3 V 4
> MHz AC voltage. Apparently it passed on everything else. Personally I would
> sell that as "spares or repair" rather than "used" and "CALIBRATED", but
> different people have different ideas of what is good or not.
>
> Out of interest, I asked the seller what Keysight had quoted to repair
> this. He said 2100 Euros, (which about $2600 USD or £1900 GBP). So to get
> it in shape, it is not exactly peanuts.
>
> The seller said to me that he has a *lot* of experience of 3458As, and
> knows that a used one has less than a 50% chance of passing a Keysight
> calibration. I wonder if that's the experience of others that have bought
> used 3458As and sent them to Keysight. If true, and I ever buy a 3458A, I
> think I would only buy one on the condition that the seller  sends it to
> Keysight, I pay the calibration cost, but if it fails, I get a full refund,
> including the calibration cost.
>
> I can understand a meter going out of spec over time, but if it can't be
> sorted out without thousands of dollars spent on it, then it is not such a
> good buy. But are more than 50% of 3458As like this?
>
> Dave
>
>
> Dr David Kirkby Ph.D C.Eng MIET
> Kirkby Microwave Ltd
> Registered office: Stokes Hall Lodge, Burnham Rd, Chelmsford, Essex, CM3
> 6DT, United Kingdom Registered in England and Wales as company number
> 08914892 http://www.kirkbymicrowave.co.uk/ Tel 01621'680100 / +44
> 1621-680100 ___
> volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
> ___
> volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
> ___
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/
> mailman/listinfo/volt-nuts
> and follow the instructions there.
>



-- 

*John Phillips*
__

Re: [volt-nuts] What's the probability of a random used 3458A passing a Keysight calibration?

2018-01-15 Thread J. L. Trantham
Forgot to add.  All were purchased on theBay.

Joe

-Original Message-
From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of J. L. Trantham
Sent: Monday, January 15, 2018 9:12 PM
To: 'Discussion of precise voltage measurement'
Subject: Re: [volt-nuts] What's the probability of a random used 3458A passing 
a Keysight calibration?

Dave,

I have three of these meters that have all been sent to Agilent/Keysight for 
the 'Agilent/Keysight' Calibration and all have passed.  Two of the three were 
still 'in spec' when arrived.  The third lost its calibration when I failed to 
properly remove the 'CalRam' and read it.  I think I tried to read it when 
still 'hot' from the removal process.  There is a process that you can use via 
HPIB to read the contents of the CalRam but I did not know about it at the 
time.  I wanted to remove the chip, install a socket and install a new chip 
before I sent it back to Agilent/Keysight.

I had previously sent a Solartron 7081 to Ametek in England for calibration and 
I used it to provide a 'transfer standard' to do my own 'in house' 
recalibration of that 3458A and it was 'in spec' in all ranges when it went 
back to Agilent, except a couple of the AC voltage ranges.  However, it was 
able to be successfully 'recalibrated' and was 'in spec' on its next trip for 
recalibration.  That was at a time when the calibrations were done in Loveland.

Once a 3458A has passed it's 'recalibration', you can purchase the 'Repair 
Agreement' for $192 per year with up to 5 years at a time (for a slight 
discount) that will allow you to send it back to Keysight for repair and 
recalibration for 'no charge' other than the cost of the 'Repair Agreement' if 
it ever fails during the covered time frame.

Hope this helps.

Joe

-Original Message-
From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of Dr. David 
Kirkby
Sent: Monday, January 15, 2018 3:56 PM
To: Discussion of precise voltage measurement
Subject: [volt-nuts] What's the probability of a random used 3458A passing a 
Keysight calibration?

There's an "interesting" $5400 buy-it-now on eBay at the minute. It's described 
as "Agilent 3458A 8.5 digit Multimeter CALIBRATED".

http://www.ebay.co.uk/itm/Agilent-3458A-8-5-Digit-Multimeter-CALIBRATED/
232593692038

When one reads the description, one can see it is an HP (not Agilent) 3458A, 
that was sent to Keysight for calibration, but *failed* on the 3 V 4 MHz AC 
voltage. Apparently it passed on everything else. Personally I would sell that 
as "spares or repair" rather than "used" and "CALIBRATED", but different people 
have different ideas of what is good or not.

Out of interest, I asked the seller what Keysight had quoted to repair this. He 
said 2100 Euros, (which about $2600 USD or £1900 GBP). So to get it in shape, 
it is not exactly peanuts.

The seller said to me that he has a *lot* of experience of 3458As, and knows 
that a used one has less than a 50% chance of passing a Keysight calibration. I 
wonder if that's the experience of others that have bought used 3458As and sent 
them to Keysight. If true, and I ever buy a 3458A, I think I would only buy one 
on the condition that the seller  sends it to Keysight, I pay the calibration 
cost, but if it fails, I get a full refund, including the calibration cost.

I can understand a meter going out of spec over time, but if it can't be sorted 
out without thousands of dollars spent on it, then it is not such a good buy. 
But are more than 50% of 3458As like this?

Dave


Dr David Kirkby Ph.D C.Eng MIET
Kirkby Microwave Ltd
Registered office: Stokes Hall Lodge, Burnham Rd, Chelmsford, Essex, CM3 6DT, 
United Kingdom Registered in England and Wales as company number 08914892 
http://www.kirkbymicrowave.co.uk/ Tel 01621'680100 / +44 1621-680100 
___
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] What's the probability of a random used 3458A passing a Keysight calibration?

2018-01-15 Thread J. L. Trantham
Dave,

I have three of these meters that have all been sent to Agilent/Keysight for 
the 'Agilent/Keysight' Calibration and all have passed.  Two of the three were 
still 'in spec' when arrived.  The third lost its calibration when I failed to 
properly remove the 'CalRam' and read it.  I think I tried to read it when 
still 'hot' from the removal process.  There is a process that you can use via 
HPIB to read the contents of the CalRam but I did not know about it at the 
time.  I wanted to remove the chip, install a socket and install a new chip 
before I sent it back to Agilent/Keysight.

I had previously sent a Solartron 7081 to Ametek in England for calibration and 
I used it to provide a 'transfer standard' to do my own 'in house' 
recalibration of that 3458A and it was 'in spec' in all ranges when it went 
back to Agilent, except a couple of the AC voltage ranges.  However, it was 
able to be successfully 'recalibrated' and was 'in spec' on its next trip for 
recalibration.  That was at a time when the calibrations were done in Loveland.

Once a 3458A has passed it's 'recalibration', you can purchase the 'Repair 
Agreement' for $192 per year with up to 5 years at a time (for a slight 
discount) that will allow you to send it back to Keysight for repair and 
recalibration for 'no charge' other than the cost of the 'Repair Agreement' if 
it ever fails during the covered time frame.

Hope this helps.

Joe

-Original Message-
From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of Dr. David 
Kirkby
Sent: Monday, January 15, 2018 3:56 PM
To: Discussion of precise voltage measurement
Subject: [volt-nuts] What's the probability of a random used 3458A passing a 
Keysight calibration?

There's an "interesting" $5400 buy-it-now on eBay at the minute. It's described 
as "Agilent 3458A 8.5 digit Multimeter CALIBRATED".

http://www.ebay.co.uk/itm/Agilent-3458A-8-5-Digit-Multimeter-CALIBRATED/
232593692038

When one reads the description, one can see it is an HP (not Agilent) 3458A, 
that was sent to Keysight for calibration, but *failed* on the 3 V 4 MHz AC 
voltage. Apparently it passed on everything else. Personally I would sell that 
as "spares or repair" rather than "used" and "CALIBRATED", but different people 
have different ideas of what is good or not.

Out of interest, I asked the seller what Keysight had quoted to repair this. He 
said 2100 Euros, (which about $2600 USD or £1900 GBP). So to get it in shape, 
it is not exactly peanuts.

The seller said to me that he has a *lot* of experience of 3458As, and knows 
that a used one has less than a 50% chance of passing a Keysight calibration. I 
wonder if that's the experience of others that have bought used 3458As and sent 
them to Keysight. If true, and I ever buy a 3458A, I think I would only buy one 
on the condition that the seller  sends it to Keysight, I pay the calibration 
cost, but if it fails, I get a full refund, including the calibration cost.

I can understand a meter going out of spec over time, but if it can't be sorted 
out without thousands of dollars spent on it, then it is not such a good buy. 
But are more than 50% of 3458As like this?

Dave


Dr David Kirkby Ph.D C.Eng MIET
Kirkby Microwave Ltd
Registered office: Stokes Hall Lodge, Burnham Rd, Chelmsford, Essex, CM3 6DT, 
United Kingdom Registered in England and Wales as company number 08914892 
http://www.kirkbymicrowave.co.uk/ Tel 01621'680100 / +44 1621-680100 
___
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


[volt-nuts] What's the probability of a random used 3458A passing a Keysight calibration?

2018-01-15 Thread Dr. David Kirkby
There's an "interesting" $5400 buy-it-now on eBay at the minute. It's
described as "Agilent 3458A 8.5 digit Multimeter CALIBRATED".

http://www.ebay.co.uk/itm/Agilent-3458A-8-5-Digit-Multimeter-CALIBRATED/
232593692038

When one reads the description, one can see it is an HP (not Agilent) 3458A,
that was sent to Keysight for calibration, but *failed* on the 3 V 4 MHz AC
voltage. Apparently it passed on everything else. Personally I would sell
that as "spares or repair" rather than "used" and "CALIBRATED", but
different people have different ideas of what is good or not.

Out of interest, I asked the seller what Keysight had quoted to repair
this. He said 2100 Euros, (which about $2600 USD or £1900 GBP). So to get
it in shape, it is not exactly peanuts.

The seller said to me that he has a *lot* of experience of 3458As, and
knows that a used one has less than a 50% chance of passing a Keysight
calibration. I wonder if that's the experience of others that have bought
used 3458As and sent them to Keysight. If true, and I ever buy a 3458A, I
think I would only buy one on the condition that the seller  sends it to
Keysight, I pay the calibration cost, but if it fails, I get a full refund,
including the calibration cost.

I can understand a meter going out of spec over time, but if it can't be
sorted out without thousands of dollars spent on it, then it is not such a
good buy. But are more than 50% of 3458As like this?

Dave


Dr David Kirkby Ph.D C.Eng MIET
Kirkby Microwave Ltd
Registered office: Stokes Hall Lodge, Burnham Rd, Chelmsford, Essex, CM3
6DT, United Kingdom
Registered in England and Wales as company number 08914892
http://www.kirkbymicrowave.co.uk/
Tel 01621'680100 / +44 1621-680100
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Any list members in UK who can accurately characterise a 732A and a 10K resistor 4HV.org

2018-01-15 Thread David C. Partridge
Please go right ahead...

-Original Message-
From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of Andre
Sent: 15 January 2018 06:43
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Any list members in UK who can accurately characterise 
a 732A and a 10K resistor 4HV.org

Maybe I can ask on 4HV.org ? 

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.