Fred,
The 3458A is 8.5 digits, which puts it into the standards category. They are
used as lab standards in many, many labs.
73,
Steve
WB0DBS
On Aug 27, 2014, at 12:08 AM, pa4...@gmail.com wrote:
Thanks, I did not new NIST has so much interesting information on their site.
I read the following on their site :
multirange instruments with up to eight decimal digits of adjustability, are
not considered by NIST to be standards
Does that mean a HP3458 ? Fluke in the Netherlands used a HP-3458 and a 732
to calibrate the Fluke 5101 that I'm working on at this moment (it failed
calibration due to some hardware faults)
Everybody thanks for the information. Turns out, a friend has a GR 1455AH for
me, that is some kind a AC KV divider.
I will test if the output of my HP 3400 is usable to connect a DMM. I have
one I restored a few years ago. Besides that someone mentioned to look at the
LT-1088.
But that has to wait until I have some more time. I only repaired
calibration, percission gear and RF stuff for my hobby, but, not planned, I
started to do this on a commercial base too. See my (non commercial) site
about my collection calibration and other gear and projects www.pa4tim.nl .
Today the 5101 goes back to the custommer but he brings 2 other instruments
in need of some TLC and a precheck before they are shipped to Fluke for
calibration. And yesterday an other company asked if I want/can repair a GM
safety tester (a sort of megger on steroids) that died during calibration.
Fred
Verzonden met Windows Mail
Van: acb...@gmx.de
Verzonden: dinsdag 26 augustus 2014 18:56
Aan: volt-nuts
Dave,
the title is:
Thermal Voltage Converters and Comparator for Very Accurate AC Voltage
Measurements
by E.S.Williams.
Adrian
Gesendet: Montag, 25. August 2014 um 19:32 Uhr
Von: Dave M dgmin...@mediacombb.net
An: Discussion of precise voltage measurement volt-nuts@febo.com
Betreff: Re: [volt-nuts] AC calibration
Adrian,
Do you have a link or title for the NIST paper that you mentioned?
Dave M
acb...@gmx.de wrote:
fred,
generally you raise a good point, I had the same issue of calibrating
an ac voltage to a high level of accuracy. you need this e.g. to
validate the self.cal of a 3458a or other precison stuff like the
8506a0.
what i would recommend to do if you want to keep costs down is:
in a nutshell, get a thermal converter in the lowest range you need
and a second one on range above. build a set of resistor range
extenders (rf type with appropriate connectors and housings) to
expand the range to where you need to be max. get one of the thermal
converter calibrated (the higher one usually, and you need to havr
good cal lab, should be 10ppm accuracy) and use it to calibrate the
rest. generally, up to 20khz, the accuracy is some 20 ppm anyway for
thermal converters! at higher frequencies, due to reflections and
stray capacitance/inductance influences, the accuracy decreases. the
resistor range extenders though, if build up correctly, only have a
few ppm impact (there is a paper from nist on that, but this is only
typical). you can calibrate all converters to the one you got
externally calibrated. do some research in the web, when you do the
calibration, you need to determine the so-called constant N. then do
an ac, dc+, ac, dc-, ac measurement between the the two and establish
the deviation, also establish the error propagation. the end result
will be a set of highly precise (low inaccuracies9 thermal converters
good enough to calibrate a 3458a an better devices. if you want to
spend the money, you could also buy a set of converters/range
resistors (with/without a 540), that typically is a few k altogether,
while a single device sometimes is available for below 100 bucks. you
need to have a stable 7.5 digit nanovoltmeter though for the
measurements of the tvcs (34420a or 2182 typically ) and precision
(stable) dc and ac sources. but in the end, all you need is a single
calibrated thermal converter.
adrian
Gesendet: Montag, 25. August 2014 um 18:38 Uhr
Von: Dave M dgmin...@mediacombb.net
An: Discussion of precise voltage measurement volt-nuts@febo.com
Betreff: Re: [volt-nuts] AC calibration
Well, you sort of answered your own question. The equipment is
called a Thermal Transfer Standard, but instead of thermistors, it
uses a thermocouple. Look at the manual for the Fluke 540B
(http://bama.edebris.com/manuals/fluke/540b/) and you'll see how
it's done. Basically, the AC source is input into the transfer
standard, and the standard's internal reference voltage is adjusted
for a null on the galvanometer. Leaving the reference voltage
setting alone, a DC voltage is input into the unit, and the DC
source is adjusted for a null on the galvanometer. At that point,
the AC voltage source is equal to