Re: [volt-nuts] Precision resistors

2015-08-11 Thread Tony Holt

On 11/08/2015 17:53, Richard Moore wrote:

True of any resistor that you want to trust to better than 10ppm, including the 
Vishays.
Of course, but Dr Frank's experience with the oil filled Vishay foils 
has been good - from another eevblog post:


'Only the hermetically sealed, oil filled types (e.g. VHP202Z) give a 
big advantage. Their rate is typically 2ppm/6yrs., and therefore will 
add about 0.02ppm/yr only.
In picture 1 you’ll find long-term stability monitoring of 3 EA of my 5 
VHP202Z. After 2 years, they really remain within  0.5ppm of their 
initial value, so that is obviously no fake advertisement.

(Remark: The measurement stability was improved also during that time.)'

That's pretty close to the SR104's typical drift of .2ppm/year, .5ppm 
max (1ppm first 2 years), but the TCR of his parts were much worse than 
the SR104's .1ppm typical ranging from .3ppm to 1ppm. He might well have 
got lucky but I read on a Chinese volt-nut type blog that Vishay 
originally specified  2ppm/10 years but reduced it to 6 years - 
presumably due to complaints/experience. Personally I would be happy to 
trust that they would remain within 10ppm for many years but you would 
have to get them measured periodically to know for certain. Depending on 
how cheaply you could buy one, it might be cheaper to buy new ones for 2 
or 3 years rather than getting one calibrated and using them to 
determine the drift of the earlier parts. And you would have a 
collection of resistors to improve confidence in the secondary standard.


Does anyone know how much it would cost to get a 10k resistor measured 
to  2ppm in the UK by a calibration company? Does anyone know how much 
it would cost to buy a 1% Vishay VH102Z/VHP101 or similar with a  2ppm 
measurement?


Edwin Pettis quoted me $7.28 for one 10k resistor ($5.46 for 11 to 24) 
so they could be a viable way of getting accurately (approx 1ppm) 
measured resistors. The higher TCR, 3ppm/C would increase the 
uncertainties though as it would require them to be used within 1C or so 
of that when measured by Mr Pettis, but he can select for 1ppm TCR or 
less for an extra $2 which would easily be justified for this purpose. 
Obviously the uncertainties due to transport shocks/vibration (although 
Mr Pettis claims they are very rugged) and drift related to the time in 
transport would need to be considered.



On Aug 11, 2015, at 9:00 AM, volt-nuts-requ...@febo.com wrote:

Edwin Pettis states his resistors drift is typically better than 2ppm in
the first year, so pretty good but you'd still need to have them
measured every few years. If you have to get them professionally
calibrated it may be cheaper to buy the Vishay parts. Edwin could
provide the measured values as could Vishay if you bought directly from
them.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.



___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] 3458A reference boards on ebay

2015-01-27 Thread Tony Holt

Orin,

What do you intend doing with it?

I was thinking about getting one of those, with a view to putting it
in a box with a couple of terminals to have something to compare with
my 6.5 digit 3457A.  But what put me off is a lack of knowledge in
knowing how to convert a 3458A reference board into a boxed unit with
a known output voltage at the terminals. How would I avoid / control
thermal EMFs?

I'd be interested to hear what your plans are for it, and how you
intend tackling those issues.

Do you know what the difference in the reference is between a standard
3458A (8 ppm) and the high stability option 002 (4 ppm) model? I'm
guessing the chips for the option 002 might be the top performing
ones. I wonder if there's any way to tell from your board if it came
from a standard 3458A or a 3458A with option 002.

Dave


Take a look at this site for some good info on using the 3458A reference:

http://www.maxmcarter.com/vref/

Tony H

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] LM399 Long term drift specification

2014-09-11 Thread Tony Holt

On 11/09/2014 01:50, Mike S wrote:

On 9/10/2014 7:00 PM, Tony wrote:

I've just noticed that TI and Linear's specs for 'Long Term Stability'
(typical) are different. TI state 20ppm/1000Hr while Linear state
8ppm/SQRT(kHr). That's  a big difference - is this likely to be a real
difference or just specmanship?

I note that Linear (in Note 4) also state that Devices with maximum
guaranteed long-term stability of 20ppm/SQRT(kH) are available.
Presumably they would be a special order as there doesn't appear to be a
unique part no. Would they be likely to be much more expensive?


Isn't 8ppm/SQRT(kHr) better than 20ppm/SQRT(kH)? Why would the latter 
be more expensive? Or is it the difference between typical and 
guaranteed?


I'm guessing that typical in this case means the one sigma value so the 
three sigma value would be 24ppm. In any case three sigma still only 
means 93.32% of parts come within that limit, or 6.7% exceed 24ppm, and 
a few could be considerably worse. That compares to a guarantee that all 
all parts meet 20ppm. This link: 
http://www.gellerlabs.com/LM299AH-20_Case_Study.htm provided in 
Andreas's response is very interesting:


Certified Long Term Drift The National Semiconductor LM199AH-20, 
LM299AH-20, and LM399AH-50 are ultra-stable Zener references specially 
selected from the production runs of LM199AH, LM299AH, LM399AH and 
tested to confirm a long-term stability of 20, 20, or 50 ppm per 1000 
hours, respectively...


So in this case they really do mean a guarantee. And I doubt that 
individual testing came cheap. I say 'came' because I wonder if they 
still 100% test the 20ppm parts or if they select them using some lower 
costs means?


Tony H
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Re: [volt-nuts] LM399 Long term drift specification

2014-09-11 Thread Tony Holt

Andreas,

Thanks for taking the time to respond. Actually I've seen many of your 
postings on eevblog and here - you've clearly done a great deal of work 
in this area and would like to thank you for making it available to us all.


On 11/09/2014 06:07, Andreas Jahn wrote:

Hello,

many questions I will keep it short:

All ageing specs are typical if you want to have guaranteed values 
you will have to measure it over a reasonable time. (I recommend min 6 
months).
Every treatment (soldering, mechanical/temperature shock) of a 
reference may create a new ageing cycle with different slope.


True. I guess that the new ageing cycle from soldering in an LM399 is 
not going to be as bad as that for a surface mount plastic device.




So 100ppm/15 years outside of lab conditions (23 deg , constant 
humidity) is something that I would not guarantee without re-calibration.


I had a feeling that would be the answer - though surely humidity 
shouldn't be a factor as these are hermetic parts. The questions remains 
though, what level might you specify - if you were forced to come up 
with a number (ok a guess!) - for non-selected, non-pre-aged parts after 
15years continuous operation without re-calibration? Obviously this is 
given the context of the presumably limited numbers of samples you've 
tested and I guess you wouldn't have bothered to further test early 
rejects.


Although typical drift of pre-aged + selected references will be in 
the 1-2ppm/year range if properly treated.


What would you classify as pre-aged? Do they need to be powered up or 
can they be maintained at a suitable temperature? How many rejects would 
you expect to get to get one that achieves 1-2ppm?


Is it known if the major instrument manufacturers preselect and burn-in 
LM399s themselves for their middle-range instruments? I'm pretty sure 
the top end kit will be all use carefully tested and selected parts, but 
what about a 34401A for example? The basic accuracy spec for that is 
20ppm for 90 days, 35ppm for 12 months so even a 20ppm guaranteed part 
wouldn't be good enough, especially allowing margin for drift in other 
components. I guess I just answered my own question!




Also its meaningless if you want to have LT or National (TI) parts 
since LT is the only manufacturer which still produces them.

With high demands you will also have to sort out the noisy references.

Some typical LM399 (all from NS) ageing data can be found on web:

http://www.gellerlabs.com/LM299AH-20_Case_Study.htm


That's very interesting. I have to agree that the raw data looks 
suspect. I wonder what the rejection rate is for this 20ppm selection 
and does it mean that non-selected parts have a high probability of 
being worse than 20ppm?


http://www.eevblog.com/forum/projects/lm399-based-10-v-reference/msg478496/#msg478496 



With best regards

Andreas



I just came across another part which looks very interesting given its 
low cost - the automotive qualified REF5050-Q1. Although its only spec'd 
as 3ppm/C typical, 8ppm/C max, that's using the box method over -40 to 
+125C. The 'typical' chart however, figure 4, page 5 shows the gradients 
to be very flat between 25 and 50. Its typical of course, so real parts 
may be very different aka Vishay foil resistors. The 0 to 85C histogram, 
fig 1 on page 5, do show the majority of parts being in the range 
.75ppm/C to 1.75ppm/C which is pretty good, and with luck, in the 25 to 
50C range may well be much better so a crude heating arrangement may be 
worthwhile (made easier by the 5050's temperature output!)


I can't reconcile fig 4 with the histograms though; from the chart I 
reckon the 0-85 typical is approx 65ppm/85C = .76ppm/C and for -40 to 
125C is approx 310ppm/165C = 1.88ppm/C. Figs 1 and 2 though show modal 
values of 1.25 and 2.25/2.5ppm/C. Am I doing something wrong or are 
these specs inconsistent?


Even more surprising is the headline feature on page 1:

EXCELLENT LONG-TERM STABILITY:
 – 5 ppm/1000 hr (typ) after 1000 hours

Unfortunately that seems to be an error as the 'typical' spec on page 4 is:
90ppm (0-1000 hours)
10ppm (1000 to 2000 hours).

The chart (fig 23, page 8) showing 1000 to 2000 hour drift of 96 parts 
show the worst case being +25ppm, with the bulk ending approx between 0 
and 15ppm. I wonder if they carry on improving after 2kHrs?


That's definitely not the SQRT(1kHr) characteristic and is very 
different from the standard REF5050 which quotes 100ppm (1st 100hours), 
50ppm (1000 to 2000 hours).


If you are in a position to pre-age them for 1000 hours that 10ppm spec 
is almost as good as the LM399 and best of all, TI quote a price of 
$1.60 @ 1k parts, compared to $4.65 for LM399s @ 1k from Linear. One off 
prices are rather more at $4.15 from Digikey (part no REF5050AQDRQ1) but 
again is still a lot cheaper than an LM399 at $9.92. At $1.60 and .8mA 
supply current, using 4, 8 or even dozens is a realistic proposition to 
exploit statistical improvements and noise 

Re: [volt-nuts] Switches in integrating ADC

2014-04-14 Thread Tony Holt

On 14/04/2014 10:03, John Devereux wrote:

Jan Fredriksson j...@41hz.com writes:


What kind of switches are used in integrating ADC, ie to switch
between voltage sources (ref and external) and to switch in multisloping
resistors? FETs?

Yes, but I believe they are integrated ones usually. Either ye olde 4066
style or custom integrated circuits in the case of the HP 3458A.

As Jan says, the 3458A switches around the ADC are integrated according 
to the April 1989 HP journal which describes the 3458A design:


   *Because the switches are in series with the resistors, they can
   add to the temperature coefficient of the ADC. A custom chip design
   was chosen so that each switch could be scaled to the size of the
   resistor to which it is connected. This allows the ADC to be
   sensitive to the ratio-tracking temperature coefficient of the
   switches and not to the absolute temperature coefficient.***

I expect that optimising and balancing charge injection would have been 
an important design objective too. It would be interesting to know how 
modern off-the-shelf analogue switches compare - ie. with low enough on 
resistance so that absolute temp coefficient doesn't matter, without 
introducing excessive charge injection. I expect that's a bit of a tall 
order.


The 8 digit Solartron 7081 uses discrete Fets, but it uses a voltage to 
time converter for its ADC. The HP 6 digit 34401A uses a 74HC4053D 2:1 
Mux to switch the ADC integrator.


For interest, the signal switching in the input path of the 3458A, for 
selecting high voltage divider / low voltage input, current sources and 
DC amplifier gains etc. all use Siliconix J2472 J-FETs (N channel 
depletion mode). I guess there were no packaged switches up to the job 
at the time.


Vishay bought Siliconix since and shut down production some while ago so 
good luck finding any parts or even a datasheet. I expect they are very 
low leakage types; no doubt there are suitable alternatives available - 
perhaps ones recommended for electrometer applications?


Tony H
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Some questions to zeners (1N823-1N829)

2013-01-30 Thread Tony Holt

On 28/01/2013 22:47, Andreas Jahn wrote:

After a run in phase of nearly 1 year the ageing of ADC #13 stabilized.
Currently I compare ADC13 nearly every day with 3 heated references (1 
LM399 = LM_2 and 2 LTZ1000A = LTZ_1/2).
The last half year the ageing is about 0.5 to 1.5 ppm for 6 months 
compared to the heated references.

See picture ADC13_longterm:
X-Axis is day
Y-Axis left is drift in ppm with red = LM399#2, green = LTZ1000A #1, 
blue = LTZ1000A #2
Y-axis right is temperature in degree celsius of the temperature 
sensor near ADC13 reference.


By the way: up to now I could not measure any effect which is related 
to thermocouples.
Ok my temperature step noise is still too high. And probably I am 
using the wrong connectors in my tests:
cheap D-Sub connectors where a metal shield is equalizing the 
temperature of 2 relative close neighboured contacts.


With best regards

Andreas

Andreas,

Very interesting results - thanks for sharing your painstaking work. 
Hope you don't mind me asking a few questions though:


How are you dealing with the issue of drift in the thermocouple 
measurements (including the cold junction compensation)? Do you 
calibrate it periodically? Thermocouples aren't noted for high stability 
- but presumably at room temperature its perhaps not much of an issue.


Do you know what temperature the LTZ1000 references are operating at, 
and how long have they been operated for - ie. have they been aged prior 
to starting the long term test? (Presumably the answer to that is the 
fact that you are showing results from day 460 onwards?)


Have you any insight into how stable the ADC's reference (AD586LJ) is? 
I.E. Have you made any occasional or periodic measurements with other 
calibrated instruments during the long term test or is it the long term 
test results themselves which leads you to state: After a run in phase 
of nearly 1 year the ageing of ADC #13 stabilized.?


Thanks, Tony H