Re: [volt-nuts] 3458A Zero reading

2015-07-29 Thread Poul-Henning Kamp

In message 19942e77-f4fd-4ba5-82cd-75a35bb5a...@gmail.com, Richard Moore writ
es:

Heavy gauge wire is only needed for current measurements, which I seldom do.

Heavy gauge wire has different thermal behaviour than think wire (area deepends
on radius squared, surface only on radius)

It is not always obvious which behaviour is preferable.


-- 
Poul-Henning Kamp   | UNIX since Zilog Zeus 3.20
p...@freebsd.org | TCP/IP since RFC 956
FreeBSD committer   | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] 3458A Zero reading

2015-07-28 Thread Richard Moore
Randy — not sure why you feel the need for shielded wire for DC measurements, 
especially if you turn the LP filter on, unless you live under a transmitter. I 
use twisted pair 22 ga. solid copper bell wire for voltage measurments with no 
difficulties, and the thermals reduce very quickly. Silver plated copper would 
be even better I suppose. Heavy gauge wire is only needed for current 
measurements, which I seldom do.

Dick Moore
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] 3458A Zero reading

2015-07-28 Thread John Phillips
Dick it is all in trying to get the last little bit of accuracy out of the
meter. 1000 NPLC  and or math stat would do better than messing with the
leads... All this gets into the law of diminishing returns. 1 ohm test lead
in series with a 10Meg input resistance that only come into play on the 2
upper ranges will induce an error of 0.1 ppm.

On Tue, Jul 28, 2015 at 1:26 PM, Richard Moore richiem5...@gmail.com
wrote:

 Randy — not sure why you feel the need for shielded wire for DC
 measurements, especially if you turn the LP filter on, unless you live
 under a transmitter. I use twisted pair 22 ga. solid copper bell wire for
 voltage measurments with no difficulties, and the thermals reduce very
 quickly. Silver plated copper would be even better I suppose. Heavy gauge
 wire is only needed for current measurements, which I seldom do.

 Dick Moore
 ___
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.




-- 

*John Phillips*
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] 3458A Zero reading

2015-07-27 Thread Richard Moore

Randy — My 3458A manual specs 12 or 14 gauge solid copper wire for the short, 
bunt in to a U shape. I use 14 gauge. Poul-Henning’s comments about thermals 
are well-taken — even a simple folded up carboard shield helps. But I think 
you’re near the limit.
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] 3458A Zero reading

2015-07-27 Thread J M via volt-nuts
I use a 8 oz styrofoam cup to isolate air currents around the 3458A terminals 
when zeroing with shorting bar


Sent from my 20 Texas sized 
iPhone


 On Jul 27, 2015, at 2:57 PM, Richard Moore richiem5...@gmail.com wrote:
 
 
 Randy — My 3458A manual specs 12 or 14 gauge solid copper wire for the short, 
 bunt in to a U shape. I use 14 gauge. Poul-Henning’s comments about thermals 
 are well-taken — even a simple folded up carboard shield helps. But I think 
 you’re near the limit.
 ___
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] 3458A Zero reading

2015-07-27 Thread Randy Evans
I used a 12 AWG wire formed into a U shape for the short and the zero
voltage measured less than 0.2 uV after a good 2+ hour warmup.  However, I
am seeing drifts of several uV for the 752 calibration.  I was using TV
twin lead to connect the 732 to 752 and 752 to 3458A since that's really
all I have for now.  I suspect thermals are killing me so I need to find
some low thermal shielded differential cable to connect from the 752 to the
3458A.  I did not see anything like that in Pomona's arsenal.  I was
thinking about Bill Gold's idea of making my own but I think it would end
up being \very stiff.  Any other ideas for a good source of shielded copper
wire twinax?

Randy AE6YG

On Mon, Jul 27, 2015 at 4:10 PM, J M via volt-nuts volt-nuts@febo.com
wrote:

 I use a 8 oz styrofoam cup to isolate air currents around the 3458A
 terminals when zeroing with shorting bar


 Sent from my 20 Texas sized
 iPhone


  On Jul 27, 2015, at 2:57 PM, Richard Moore richiem5...@gmail.com
 wrote:
 
 
  Randy — My 3458A manual specs 12 or 14 gauge solid copper wire for the
 short, bunt in to a U shape. I use 14 gauge. Poul-Henning’s comments about
 thermals are well-taken — even a simple folded up carboard shield helps.
 But I think you’re near the limit.
  ___
  volt-nuts mailing list -- volt-nuts@febo.com
  To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
  and follow the instructions there.
 ___
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.