Re: [volt-nuts] Fluke 732A themistor reading

2014-08-18 Thread acbern
well, re. the resistor, question is how stable it is, not so much absolute 
value. if not very good, your 3458a may be more stable that your reference. vpg 
hermetic foil 4 wire resistors are very stable (1ppm pa) and are cheap (<50usd).

re the wire, you should use twisted shielded pair. there are good teflon tsp 
wires with kapton isolation (as used in satellites) on sale on ebay. use copper 
gold plated terminals, around one usd and crimp them. pure copper wire is not 
needed.

cheers

adrian


> Gesendet: Montag, 18. August 2014 um 01:03 Uhr
> Von: "Randy Evans" 
> An: "Discussion of precise voltage measurement" 
> Betreff: Re: [volt-nuts] Fluke 732A themistor reading
>
> I do have an ESI SR1 10Kohm standard, but I'm not sure its accurate
> enough...time to look around I suppose.
> 
> BTW, I understand that TV twin lead, the copper wire type, makes a good
> test lead for the 3458 measurements.  It seems reasonable, although the
> leakage resistance might be questionable.
> 
> Randy
> 
> 
> On Sun, Aug 17, 2014 at 1:36 PM, Todd Micallef  wrote:
> 
> > Randy,
> >
> > Next thing you know, you will be looking at a 10K standard resistor to go
> > with that 3458A and possible 732A.
> >
> > Todd
> >
> >
> > On Sun, Aug 17, 2014 at 3:51 PM, Randy Evans 
> > wrote:
> >
> > > Todd and Bill,
> > >
> > > Thanks for the input.  The 3.6 K thermistor reading seems viable and i
> > will
> > > pursue it further.
> > >
> > > I may have over committed myself to modifying the HP-419A, the Fluke
> > 845A,
> > > buying the 3458A and looking at buying a Fluke 732A.
> > >
> > > So much to do and so little time.
> > >
> > > Thanks again,
> > >
> > > Randy
> > >
> > >
> > > On Sun, Aug 17, 2014 at 11:48 AM, Bill Gold  wrote:
> > >
> > > > Randy:
> > > >
> > > > I have 6 ea Fluke 732A instruments.  The results of being an over
> > the
> > > > edge and deeply obsessed "Volt Nut", they just kept adding
> > (reproducing?)
> > > > over the years.  My readings for the themistors range from 3.407K OHMS
> > to
> > > > 4.514K OHMS and all work just fine.  3.6K OHMS would seem just fine to
> > > me.
> > > > It really depends upon the final inside temp of the oven assembly.  The
> > > one
> > > > with 3.407K OHMS measures around 47.5 degrees C while the others are
> > > around
> > > > 45 C.  It is the stability of this reading over time that is very
> > > > important.
> > > > Fluke says that it could change as much as 10 ohms per month but mine
> > are
> > > > rock solid for years and years.
> > > >
> > > > Bill
> > > >
> > > > - Original Message -
> > > > From: "Randy Evans" 
> > > > To: 
> > > > Sent: Sunday, August 17, 2014 11:08 AM
> > > > Subject: [volt-nuts] Fluke 732A themistor reading
> > > >
> > > >
> > > > > Can anyone tell me what value to expect for a Fluke 732A thermistor
> > > > > reading?  I am looking to get a used 732A and one of the things I was
> > > > told
> > > > > to look for is to measure the thermistor reading after the unit has
> > > > heated
> > > > > up for at least an hour.  I saw in one of the posts that a "typical"
> > > > > reading was 4950 ohms.  One vendor I asked said he measured about
> > 3600
> > > > ohms
> > > > > after a two hour warmup. I was concerned that this is too low.  What
> > do
> > > > you
> > > > > guys think?
> > > > >
> > > > > Thanks,
> > > > >
> > > > > Randy
> > > > > ___
> > > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > > > To unsubscribe, go to
> > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > > and follow the instructions there.
> > > >
> > > > ___
> > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > > To unsubscribe, go to
> > > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > and follow the instructions there.
> > > >
> > > ___
> > > volt-nuts mailing list -- volt-nuts@febo.com
> > > To unsubscribe, go to
> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > and follow the instructions there.
> > >
> > ___
> > volt-nuts mailing list -- volt-nuts@febo.com
> > To unsubscribe, go to
> > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
> >
> ___
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
> 
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


[volt-nuts] Fluke 732A themistor reading

2014-08-17 Thread Mark Sims
Be aware that finding any sort of pure copper in a consumer style cable these 
days is a losing proposition...  even cables sold as "100% oxygen free copper" 
from name brand makers with a reputation to lose.   And then there is that 
magnetic 100% OFC copper from China...  that has 1/3 the conductivity of normal 
copper...  And don't get me started on bogo AWG wire size specs...  ever see 16 
AWG wire with like  5 hair-thin strands of bogo-copper in it?

--

BTW, I understand that TV twin lead, the copper wire type, makes a good
test lead for the 3458 measurements.  
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Fluke 732A themistor reading

2014-08-17 Thread Randy Evans
I do have an ESI SR1 10Kohm standard, but I'm not sure its accurate
enough...time to look around I suppose.

BTW, I understand that TV twin lead, the copper wire type, makes a good
test lead for the 3458 measurements.  It seems reasonable, although the
leakage resistance might be questionable.

Randy


On Sun, Aug 17, 2014 at 1:36 PM, Todd Micallef  wrote:

> Randy,
>
> Next thing you know, you will be looking at a 10K standard resistor to go
> with that 3458A and possible 732A.
>
> Todd
>
>
> On Sun, Aug 17, 2014 at 3:51 PM, Randy Evans 
> wrote:
>
> > Todd and Bill,
> >
> > Thanks for the input.  The 3.6 K thermistor reading seems viable and i
> will
> > pursue it further.
> >
> > I may have over committed myself to modifying the HP-419A, the Fluke
> 845A,
> > buying the 3458A and looking at buying a Fluke 732A.
> >
> > So much to do and so little time.
> >
> > Thanks again,
> >
> > Randy
> >
> >
> > On Sun, Aug 17, 2014 at 11:48 AM, Bill Gold  wrote:
> >
> > > Randy:
> > >
> > > I have 6 ea Fluke 732A instruments.  The results of being an over
> the
> > > edge and deeply obsessed "Volt Nut", they just kept adding
> (reproducing?)
> > > over the years.  My readings for the themistors range from 3.407K OHMS
> to
> > > 4.514K OHMS and all work just fine.  3.6K OHMS would seem just fine to
> > me.
> > > It really depends upon the final inside temp of the oven assembly.  The
> > one
> > > with 3.407K OHMS measures around 47.5 degrees C while the others are
> > around
> > > 45 C.  It is the stability of this reading over time that is very
> > > important.
> > > Fluke says that it could change as much as 10 ohms per month but mine
> are
> > > rock solid for years and years.
> > >
> > > Bill
> > >
> > > - Original Message -
> > > From: "Randy Evans" 
> > > To: 
> > > Sent: Sunday, August 17, 2014 11:08 AM
> > > Subject: [volt-nuts] Fluke 732A themistor reading
> > >
> > >
> > > > Can anyone tell me what value to expect for a Fluke 732A thermistor
> > > > reading?  I am looking to get a used 732A and one of the things I was
> > > told
> > > > to look for is to measure the thermistor reading after the unit has
> > > heated
> > > > up for at least an hour.  I saw in one of the posts that a "typical"
> > > > reading was 4950 ohms.  One vendor I asked said he measured about
> 3600
> > > ohms
> > > > after a two hour warmup. I was concerned that this is too low.  What
> do
> > > you
> > > > guys think?
> > > >
> > > > Thanks,
> > > >
> > > > Randy
> > > > ___
> > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > > To unsubscribe, go to
> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > and follow the instructions there.
> > >
> > > ___
> > > volt-nuts mailing list -- volt-nuts@febo.com
> > > To unsubscribe, go to
> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > and follow the instructions there.
> > >
> > ___
> > volt-nuts mailing list -- volt-nuts@febo.com
> > To unsubscribe, go to
> > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
> >
> ___
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Fluke 732A themistor reading

2014-08-17 Thread Todd Micallef
Randy,

Next thing you know, you will be looking at a 10K standard resistor to go
with that 3458A and possible 732A.

Todd


On Sun, Aug 17, 2014 at 3:51 PM, Randy Evans 
wrote:

> Todd and Bill,
>
> Thanks for the input.  The 3.6 K thermistor reading seems viable and i will
> pursue it further.
>
> I may have over committed myself to modifying the HP-419A, the Fluke 845A,
> buying the 3458A and looking at buying a Fluke 732A.
>
> So much to do and so little time.
>
> Thanks again,
>
> Randy
>
>
> On Sun, Aug 17, 2014 at 11:48 AM, Bill Gold  wrote:
>
> > Randy:
> >
> > I have 6 ea Fluke 732A instruments.  The results of being an over the
> > edge and deeply obsessed "Volt Nut", they just kept adding (reproducing?)
> > over the years.  My readings for the themistors range from 3.407K OHMS to
> > 4.514K OHMS and all work just fine.  3.6K OHMS would seem just fine to
> me.
> > It really depends upon the final inside temp of the oven assembly.  The
> one
> > with 3.407K OHMS measures around 47.5 degrees C while the others are
> around
> > 45 C.  It is the stability of this reading over time that is very
> > important.
> > Fluke says that it could change as much as 10 ohms per month but mine are
> > rock solid for years and years.
> >
> > Bill
> >
> > - Original Message -
> > From: "Randy Evans" 
> > To: 
> > Sent: Sunday, August 17, 2014 11:08 AM
> > Subject: [volt-nuts] Fluke 732A themistor reading
> >
> >
> > > Can anyone tell me what value to expect for a Fluke 732A thermistor
> > > reading?  I am looking to get a used 732A and one of the things I was
> > told
> > > to look for is to measure the thermistor reading after the unit has
> > heated
> > > up for at least an hour.  I saw in one of the posts that a "typical"
> > > reading was 4950 ohms.  One vendor I asked said he measured about 3600
> > ohms
> > > after a two hour warmup. I was concerned that this is too low.  What do
> > you
> > > guys think?
> > >
> > > Thanks,
> > >
> > > Randy
> > > ___
> > > volt-nuts mailing list -- volt-nuts@febo.com
> > > To unsubscribe, go to
> > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > and follow the instructions there.
> >
> > ___
> > volt-nuts mailing list -- volt-nuts@febo.com
> > To unsubscribe, go to
> > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
> >
> ___
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Fluke 732A themistor reading

2014-08-17 Thread R.Phillips

Randy
I have had mine working non-stop for almost 1 year - its current reading is 
4.51902 K ohms, it never deviates more than 0.150 ohms.

Roy


-Original Message- 
From: Randy Evans

Sent: Sunday, August 17, 2014 8:51 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Fluke 732A themistor reading

Todd and Bill,

Thanks for the input.  The 3.6 K thermistor reading seems viable and i will
pursue it further.

I may have over committed myself to modifying the HP-419A, the Fluke 845A,
buying the 3458A and looking at buying a Fluke 732A.

So much to do and so little time.

Thanks again,

Randy


On Sun, Aug 17, 2014 at 11:48 AM, Bill Gold  wrote:


Randy:

I have 6 ea Fluke 732A instruments.  The results of being an over the
edge and deeply obsessed "Volt Nut", they just kept adding (reproducing?)
over the years.  My readings for the themistors range from 3.407K OHMS to
4.514K OHMS and all work just fine.  3.6K OHMS would seem just fine to me.
It really depends upon the final inside temp of the oven assembly.  The 
one
with 3.407K OHMS measures around 47.5 degrees C while the others are 
around

45 C.  It is the stability of this reading over time that is very
important.
Fluke says that it could change as much as 10 ohms per month but mine are
rock solid for years and years.

Bill

- Original Message -
From: "Randy Evans" 
To: 
Sent: Sunday, August 17, 2014 11:08 AM
Subject: [volt-nuts] Fluke 732A themistor reading


> Can anyone tell me what value to expect for a Fluke 732A thermistor
> reading?  I am looking to get a used 732A and one of the things I was
told
> to look for is to measure the thermistor reading after the unit has
heated
> up for at least an hour.  I saw in one of the posts that a "typical"
> reading was 4950 ohms.  One vendor I asked said he measured about 3600
ohms
> after a two hour warmup. I was concerned that this is too low.  What do
you
> guys think?
>
> Thanks,
>
> Randy
> ___
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there. 



___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Fluke 732A themistor reading

2014-08-17 Thread Randy Evans
Todd and Bill,

Thanks for the input.  The 3.6 K thermistor reading seems viable and i will
pursue it further.

I may have over committed myself to modifying the HP-419A, the Fluke 845A,
buying the 3458A and looking at buying a Fluke 732A.

So much to do and so little time.

Thanks again,

Randy


On Sun, Aug 17, 2014 at 11:48 AM, Bill Gold  wrote:

> Randy:
>
> I have 6 ea Fluke 732A instruments.  The results of being an over the
> edge and deeply obsessed "Volt Nut", they just kept adding (reproducing?)
> over the years.  My readings for the themistors range from 3.407K OHMS to
> 4.514K OHMS and all work just fine.  3.6K OHMS would seem just fine to me.
> It really depends upon the final inside temp of the oven assembly.  The one
> with 3.407K OHMS measures around 47.5 degrees C while the others are around
> 45 C.  It is the stability of this reading over time that is very
> important.
> Fluke says that it could change as much as 10 ohms per month but mine are
> rock solid for years and years.
>
> Bill
>
> - Original Message -
> From: "Randy Evans" 
> To: 
> Sent: Sunday, August 17, 2014 11:08 AM
> Subject: [volt-nuts] Fluke 732A themistor reading
>
>
> > Can anyone tell me what value to expect for a Fluke 732A thermistor
> > reading?  I am looking to get a used 732A and one of the things I was
> told
> > to look for is to measure the thermistor reading after the unit has
> heated
> > up for at least an hour.  I saw in one of the posts that a "typical"
> > reading was 4950 ohms.  One vendor I asked said he measured about 3600
> ohms
> > after a two hour warmup. I was concerned that this is too low.  What do
> you
> > guys think?
> >
> > Thanks,
> >
> > Randy
> > ___
> > volt-nuts mailing list -- volt-nuts@febo.com
> > To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
>
> ___
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Fluke 732A themistor reading

2014-08-17 Thread Bill Gold
Randy:

I have 6 ea Fluke 732A instruments.  The results of being an over the
edge and deeply obsessed "Volt Nut", they just kept adding (reproducing?)
over the years.  My readings for the themistors range from 3.407K OHMS to
4.514K OHMS and all work just fine.  3.6K OHMS would seem just fine to me.
It really depends upon the final inside temp of the oven assembly.  The one
with 3.407K OHMS measures around 47.5 degrees C while the others are around
45 C.  It is the stability of this reading over time that is very important.
Fluke says that it could change as much as 10 ohms per month but mine are
rock solid for years and years.

Bill

- Original Message - 
From: "Randy Evans" 
To: 
Sent: Sunday, August 17, 2014 11:08 AM
Subject: [volt-nuts] Fluke 732A themistor reading


> Can anyone tell me what value to expect for a Fluke 732A thermistor
> reading?  I am looking to get a used 732A and one of the things I was told
> to look for is to measure the thermistor reading after the unit has heated
> up for at least an hour.  I saw in one of the posts that a "typical"
> reading was 4950 ohms.  One vendor I asked said he measured about 3600
ohms
> after a two hour warmup. I was concerned that this is too low.  What do
you
> guys think?
>
> Thanks,
>
> Randy
> ___
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Fluke 732A themistor reading

2014-08-17 Thread Todd Micallef
Randy,

Two of mine read in the 3.7-3.8K range, another reads in the 4.2K range. It
is a good sign if the reading dropped in a few hours after turn-on. I
believe the lower thermistor readings indicate the older 732A oven. Some
minor changes were made and both manuals are available on Didier's site.
Verify that the 1V and 1.018V outputs are functioning as well as the 10V
obviously. There have been a few sold recently on eBay with one or both of
the 1V outputs reading low or dead.

Todd


On Sun, Aug 17, 2014 at 2:08 PM, Randy Evans 
wrote:

> Can anyone tell me what value to expect for a Fluke 732A thermistor
> reading?  I am looking to get a used 732A and one of the things I was told
> to look for is to measure the thermistor reading after the unit has heated
> up for at least an hour.  I saw in one of the posts that a "typical"
> reading was 4950 ohms.  One vendor I asked said he measured about 3600 ohms
> after a two hour warmup. I was concerned that this is too low.  What do you
> guys think?
>
> Thanks,
>
> Randy
> ___
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


[volt-nuts] Fluke 732A themistor reading

2014-08-17 Thread Randy Evans
Can anyone tell me what value to expect for a Fluke 732A thermistor
reading?  I am looking to get a used 732A and one of the things I was told
to look for is to measure the thermistor reading after the unit has heated
up for at least an hour.  I saw in one of the posts that a "typical"
reading was 4950 ohms.  One vendor I asked said he measured about 3600 ohms
after a two hour warmup. I was concerned that this is too low.  What do you
guys think?

Thanks,

Randy
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.