Re: [volt-nuts] Fluke 732A themistor reading

2014-08-18 Thread acbern
well, re. the resistor, question is how stable it is, not so much absolute 
value. if not very good, your 3458a may be more stable that your reference. vpg 
hermetic foil 4 wire resistors are very stable (1ppm pa) and are cheap (50usd).

re the wire, you should use twisted shielded pair. there are good teflon tsp 
wires with kapton isolation (as used in satellites) on sale on ebay. use copper 
gold plated terminals, around one usd and crimp them. pure copper wire is not 
needed.

cheers

adrian


 Gesendet: Montag, 18. August 2014 um 01:03 Uhr
 Von: Randy Evans randyevans2...@gmail.com
 An: Discussion of precise voltage measurement volt-nuts@febo.com
 Betreff: Re: [volt-nuts] Fluke 732A themistor reading

 I do have an ESI SR1 10Kohm standard, but I'm not sure its accurate
 enough...time to look around I suppose.
 
 BTW, I understand that TV twin lead, the copper wire type, makes a good
 test lead for the 3458 measurements.  It seems reasonable, although the
 leakage resistance might be questionable.
 
 Randy
 
 
 On Sun, Aug 17, 2014 at 1:36 PM, Todd Micallef tmical...@gmail.com wrote:
 
  Randy,
 
  Next thing you know, you will be looking at a 10K standard resistor to go
  with that 3458A and possible 732A.
 
  Todd
 
 
  On Sun, Aug 17, 2014 at 3:51 PM, Randy Evans randyevans2...@gmail.com
  wrote:
 
   Todd and Bill,
  
   Thanks for the input.  The 3.6 K thermistor reading seems viable and i
  will
   pursue it further.
  
   I may have over committed myself to modifying the HP-419A, the Fluke
  845A,
   buying the 3458A and looking at buying a Fluke 732A.
  
   So much to do and so little time.
  
   Thanks again,
  
   Randy
  
  
   On Sun, Aug 17, 2014 at 11:48 AM, Bill Gold wpgold3...@att.net wrote:
  
Randy:
   
I have 6 ea Fluke 732A instruments.  The results of being an over
  the
edge and deeply obsessed Volt Nut, they just kept adding
  (reproducing?)
over the years.  My readings for the themistors range from 3.407K OHMS
  to
4.514K OHMS and all work just fine.  3.6K OHMS would seem just fine to
   me.
It really depends upon the final inside temp of the oven assembly.  The
   one
with 3.407K OHMS measures around 47.5 degrees C while the others are
   around
45 C.  It is the stability of this reading over time that is very
important.
Fluke says that it could change as much as 10 ohms per month but mine
  are
rock solid for years and years.
   
Bill
   
- Original Message -
From: Randy Evans randyevans2...@gmail.com
To: volt-nuts@febo.com
Sent: Sunday, August 17, 2014 11:08 AM
Subject: [volt-nuts] Fluke 732A themistor reading
   
   
 Can anyone tell me what value to expect for a Fluke 732A thermistor
 reading?  I am looking to get a used 732A and one of the things I was
told
 to look for is to measure the thermistor reading after the unit has
heated
 up for at least an hour.  I saw in one of the posts that a typical
 reading was 4950 ohms.  One vendor I asked said he measured about
  3600
ohms
 after a two hour warmup. I was concerned that this is too low.  What
  do
you
 guys think?

 Thanks,

 Randy
 ___
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.
   
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
   
   ___
   volt-nuts mailing list -- volt-nuts@febo.com
   To unsubscribe, go to
   https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
   and follow the instructions there.
  
  ___
  volt-nuts mailing list -- volt-nuts@febo.com
  To unsubscribe, go to
  https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
  and follow the instructions there.
 
 ___
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.
 
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Fluke 732A themistor reading

2014-08-17 Thread Todd Micallef
Randy,

Two of mine read in the 3.7-3.8K range, another reads in the 4.2K range. It
is a good sign if the reading dropped in a few hours after turn-on. I
believe the lower thermistor readings indicate the older 732A oven. Some
minor changes were made and both manuals are available on Didier's site.
Verify that the 1V and 1.018V outputs are functioning as well as the 10V
obviously. There have been a few sold recently on eBay with one or both of
the 1V outputs reading low or dead.

Todd


On Sun, Aug 17, 2014 at 2:08 PM, Randy Evans randyevans2...@gmail.com
wrote:

 Can anyone tell me what value to expect for a Fluke 732A thermistor
 reading?  I am looking to get a used 732A and one of the things I was told
 to look for is to measure the thermistor reading after the unit has heated
 up for at least an hour.  I saw in one of the posts that a typical
 reading was 4950 ohms.  One vendor I asked said he measured about 3600 ohms
 after a two hour warmup. I was concerned that this is too low.  What do you
 guys think?

 Thanks,

 Randy
 ___
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Fluke 732A themistor reading

2014-08-17 Thread Bill Gold
Randy:

I have 6 ea Fluke 732A instruments.  The results of being an over the
edge and deeply obsessed Volt Nut, they just kept adding (reproducing?)
over the years.  My readings for the themistors range from 3.407K OHMS to
4.514K OHMS and all work just fine.  3.6K OHMS would seem just fine to me.
It really depends upon the final inside temp of the oven assembly.  The one
with 3.407K OHMS measures around 47.5 degrees C while the others are around
45 C.  It is the stability of this reading over time that is very important.
Fluke says that it could change as much as 10 ohms per month but mine are
rock solid for years and years.

Bill

- Original Message - 
From: Randy Evans randyevans2...@gmail.com
To: volt-nuts@febo.com
Sent: Sunday, August 17, 2014 11:08 AM
Subject: [volt-nuts] Fluke 732A themistor reading


 Can anyone tell me what value to expect for a Fluke 732A thermistor
 reading?  I am looking to get a used 732A and one of the things I was told
 to look for is to measure the thermistor reading after the unit has heated
 up for at least an hour.  I saw in one of the posts that a typical
 reading was 4950 ohms.  One vendor I asked said he measured about 3600
ohms
 after a two hour warmup. I was concerned that this is too low.  What do
you
 guys think?

 Thanks,

 Randy
 ___
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Fluke 732A themistor reading

2014-08-17 Thread Randy Evans
Todd and Bill,

Thanks for the input.  The 3.6 K thermistor reading seems viable and i will
pursue it further.

I may have over committed myself to modifying the HP-419A, the Fluke 845A,
buying the 3458A and looking at buying a Fluke 732A.

So much to do and so little time.

Thanks again,

Randy


On Sun, Aug 17, 2014 at 11:48 AM, Bill Gold wpgold3...@att.net wrote:

 Randy:

 I have 6 ea Fluke 732A instruments.  The results of being an over the
 edge and deeply obsessed Volt Nut, they just kept adding (reproducing?)
 over the years.  My readings for the themistors range from 3.407K OHMS to
 4.514K OHMS and all work just fine.  3.6K OHMS would seem just fine to me.
 It really depends upon the final inside temp of the oven assembly.  The one
 with 3.407K OHMS measures around 47.5 degrees C while the others are around
 45 C.  It is the stability of this reading over time that is very
 important.
 Fluke says that it could change as much as 10 ohms per month but mine are
 rock solid for years and years.

 Bill

 - Original Message -
 From: Randy Evans randyevans2...@gmail.com
 To: volt-nuts@febo.com
 Sent: Sunday, August 17, 2014 11:08 AM
 Subject: [volt-nuts] Fluke 732A themistor reading


  Can anyone tell me what value to expect for a Fluke 732A thermistor
  reading?  I am looking to get a used 732A and one of the things I was
 told
  to look for is to measure the thermistor reading after the unit has
 heated
  up for at least an hour.  I saw in one of the posts that a typical
  reading was 4950 ohms.  One vendor I asked said he measured about 3600
 ohms
  after a two hour warmup. I was concerned that this is too low.  What do
 you
  guys think?
 
  Thanks,
 
  Randy
  ___
  volt-nuts mailing list -- volt-nuts@febo.com
  To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
  and follow the instructions there.

 ___
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Fluke 732A themistor reading

2014-08-17 Thread Randy Evans
I do have an ESI SR1 10Kohm standard, but I'm not sure its accurate
enough...time to look around I suppose.

BTW, I understand that TV twin lead, the copper wire type, makes a good
test lead for the 3458 measurements.  It seems reasonable, although the
leakage resistance might be questionable.

Randy


On Sun, Aug 17, 2014 at 1:36 PM, Todd Micallef tmical...@gmail.com wrote:

 Randy,

 Next thing you know, you will be looking at a 10K standard resistor to go
 with that 3458A and possible 732A.

 Todd


 On Sun, Aug 17, 2014 at 3:51 PM, Randy Evans randyevans2...@gmail.com
 wrote:

  Todd and Bill,
 
  Thanks for the input.  The 3.6 K thermistor reading seems viable and i
 will
  pursue it further.
 
  I may have over committed myself to modifying the HP-419A, the Fluke
 845A,
  buying the 3458A and looking at buying a Fluke 732A.
 
  So much to do and so little time.
 
  Thanks again,
 
  Randy
 
 
  On Sun, Aug 17, 2014 at 11:48 AM, Bill Gold wpgold3...@att.net wrote:
 
   Randy:
  
   I have 6 ea Fluke 732A instruments.  The results of being an over
 the
   edge and deeply obsessed Volt Nut, they just kept adding
 (reproducing?)
   over the years.  My readings for the themistors range from 3.407K OHMS
 to
   4.514K OHMS and all work just fine.  3.6K OHMS would seem just fine to
  me.
   It really depends upon the final inside temp of the oven assembly.  The
  one
   with 3.407K OHMS measures around 47.5 degrees C while the others are
  around
   45 C.  It is the stability of this reading over time that is very
   important.
   Fluke says that it could change as much as 10 ohms per month but mine
 are
   rock solid for years and years.
  
   Bill
  
   - Original Message -
   From: Randy Evans randyevans2...@gmail.com
   To: volt-nuts@febo.com
   Sent: Sunday, August 17, 2014 11:08 AM
   Subject: [volt-nuts] Fluke 732A themistor reading
  
  
Can anyone tell me what value to expect for a Fluke 732A thermistor
reading?  I am looking to get a used 732A and one of the things I was
   told
to look for is to measure the thermistor reading after the unit has
   heated
up for at least an hour.  I saw in one of the posts that a typical
reading was 4950 ohms.  One vendor I asked said he measured about
 3600
   ohms
after a two hour warmup. I was concerned that this is too low.  What
 do
   you
guys think?
   
Thanks,
   
Randy
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
   https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
  
   ___
   volt-nuts mailing list -- volt-nuts@febo.com
   To unsubscribe, go to
   https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
   and follow the instructions there.
  
  ___
  volt-nuts mailing list -- volt-nuts@febo.com
  To unsubscribe, go to
  https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
  and follow the instructions there.
 
 ___
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.