Re: [volt-nuts] HP-3458A questions

2015-06-03 Thread Randy Evans
Frank,

Thanks for the data.  The 100K seems acceptable since I have a Fluke 732A
 to recalibrate the 3458A with after the change and the 3458A is not in a
rack.

Randy

On Tue, Jun 2, 2015 at 2:50 PM, Frank Stellmach frank.stellm...@freenet.de
wrote:

 Hi Randy,

 in advance, this change only improves the stability of the instrument,
 when the 3458A is powered.
 The LTZ1000 will not age (nearly), when the instrument is switched off.
 Therefore, it makes sense mainly, if your instrument is running for an
 extended period of time.

 (Reference: Spreadbury: Ultra Zener..)

 If you change the oven temperature, you will loose the 10V calibration, as
 the LTZ1000A will change its reference voltage.


 Although my instrument is strictly switched off, when not doing
 measurements, I reduced the oven temp. by connecting a 100k precision
 wirewound (3ppm/K) in parallel to the 15k BMF resistor. That gives 13k in
 total, yielding about 65°C like in the LT datasheet.
 An alternative placement is already foreseen on the PCB; in the CLIP, it's
 component designator is X411.
 I'll try to link in a photo of my change.

 lymex on the bbs35.hot forum obviously was more agressive, and paralled
 75k of Vitrohm Thin Film, giving 12.5k for about 55°C.

 I would recommend my solution above the one of lymex, although his one
 gives theoretically more stability.
 Also, the lower stability of TF reistors may affect the stability (T, t)
 of the 15k BMF.

 With 100k/65°C, the oven runs about 30°C lower.
 Therefore, the ambient temperature has to be limited to max. 35°C, the fan
 has ALWAYS to be kept clean, giving a maximum temperature rise of 15°C, and
 the instrument may not be put in a rack (which would give additional
 10..20°C temperature rise). Under these conditions, (35°C ambient + 15°C
 3458A inner heating + 10°C self heating of LTZ1000A + 5°C regulation
 margin), the oven will always be stable.

 These 75k / 55°C would be possible, only if you limit the ambient
 temperature further, to = 25°C, and precisely monitor the 3458A inner
 temperature (TEMP?).

 Although my instrument is kept at 20..23°C throughout the year, i wouldn't
 feel comfortable with 55°C.
 The only possibility would be the exchange of the reference to the non-A
 version, as the self heating effect may vanish!

 Frank

 ___
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


[volt-nuts] HP-3458A questions

2015-06-02 Thread Frank Stellmach

Hi Randy,

in advance, this change only improves the stability of the instrument, 
when the 3458A is powered.

The LTZ1000 will not age (nearly), when the instrument is switched off.
Therefore, it makes sense mainly, if your instrument is running for an 
extended period of time.


(Reference: Spreadbury: Ultra Zener..)

If you change the oven temperature, you will loose the 10V calibration, 
as the LTZ1000A will change its reference voltage.



Although my instrument is strictly switched off, when not doing 
measurements, I reduced the oven temp. by connecting a 100k precision 
wirewound (3ppm/K) in parallel to the 15k BMF resistor. That gives 13k 
in total, yielding about 65°C like in the LT datasheet.
An alternative placement is already foreseen on the PCB; in the CLIP, 
it's component designator is X411.

I'll try to link in a photo of my change.

lymex on the bbs35.hot forum obviously was more agressive, and paralled 
75k of Vitrohm Thin Film, giving 12.5k for about 55°C.


I would recommend my solution above the one of lymex, although his one 
gives theoretically more stability.
Also, the lower stability of TF reistors may affect the stability (T, t) 
of the 15k BMF.


With 100k/65°C, the oven runs about 30°C lower.
Therefore, the ambient temperature has to be limited to max. 35°C, the 
fan has ALWAYS to be kept clean, giving a maximum temperature rise of 
15°C, and the instrument may not be put in a rack (which would give 
additional 10..20°C temperature rise). Under these conditions, (35°C 
ambient + 15°C 3458A inner heating + 10°C self heating of LTZ1000A + 5°C 
regulation margin), the oven will always be stable.


These 75k / 55°C would be possible, only if you limit the ambient 
temperature further, to = 25°C, and precisely monitor the 3458A inner 
temperature (TEMP?).


Although my instrument is kept at 20..23°C throughout the year, i 
wouldn't feel comfortable with 55°C.
The only possibility would be the exchange of the reference to the non-A 
version, as the self heating effect may vanish!


Frank
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Re: [volt-nuts] HP-3458A questions

2015-06-02 Thread J. L. Trantham
Frank,

Thanks for the picture of your board.  

I have been interested in the 'plastic cap' on the LTZ1000A and you provided me 
with information about this in the past.

Do you have a picture of the bottom of the board?  I would love to see how the 
bottom half of the 'cap' looks.

Thanks.

Joe

-Original Message-
From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of Frank Stellmach
Sent: Tuesday, June 02, 2015 4:50 PM
To: volt-nuts@febo.com
Subject: [volt-nuts] HP-3458A questions

Hi Randy,

in advance, this change only improves the stability of the instrument, when the 
3458A is powered.
The LTZ1000 will not age (nearly), when the instrument is switched off.
Therefore, it makes sense mainly, if your instrument is running for an extended 
period of time.

(Reference: Spreadbury: Ultra Zener..)

If you change the oven temperature, you will loose the 10V calibration, as the 
LTZ1000A will change its reference voltage.


Although my instrument is strictly switched off, when not doing 
measurements, I reduced the oven temp. by connecting a 100k precision 
wirewound (3ppm/K) in parallel to the 15k BMF resistor. That gives 13k 
in total, yielding about 65°C like in the LT datasheet.
An alternative placement is already foreseen on the PCB; in the CLIP, 
it's component designator is X411.
I'll try to link in a photo of my change.

lymex on the bbs35.hot forum obviously was more agressive, and paralled 
75k of Vitrohm Thin Film, giving 12.5k for about 55°C.

I would recommend my solution above the one of lymex, although his one 
gives theoretically more stability.
Also, the lower stability of TF reistors may affect the stability (T, t) 
of the 15k BMF.

With 100k/65°C, the oven runs about 30°C lower.
Therefore, the ambient temperature has to be limited to max. 35°C, the 
fan has ALWAYS to be kept clean, giving a maximum temperature rise of 
15°C, and the instrument may not be put in a rack (which would give 
additional 10..20°C temperature rise). Under these conditions, (35°C 
ambient + 15°C 3458A inner heating + 10°C self heating of LTZ1000A + 5°C 
regulation margin), the oven will always be stable.

These 75k / 55°C would be possible, only if you limit the ambient 
temperature further, to = 25°C, and precisely monitor the 3458A inner 
temperature (TEMP?).

Although my instrument is kept at 20..23°C throughout the year, i 
wouldn't feel comfortable with 55°C.
The only possibility would be the exchange of the reference to the non-A 
version, as the self heating effect may vanish!

Frank

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] HP-3458A questions

2015-06-02 Thread Andreas Jahn

Hello,

A bottom picture of the board is currently in a auction (so no need to 
deinstall the board from a life system):

http://www.ebay.com/itm/Agilent-03458-66509-DCV-Reference-PCA-for-3458-PART-/371306381959?pt=LH_DefaultDomain_0hash=item567395c287

With best regards

Andreas

Am 03.06.2015 um 01:00 schrieb J. L. Trantham:

Frank,

Thanks for the picture of your board.

I have been interested in the 'plastic cap' on the LTZ1000A and you provided me 
with information about this in the past.

Do you have a picture of the bottom of the board?  I would love to see how the 
bottom half of the 'cap' looks.

Thanks.

Joe

-Original Message-
From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of Frank Stellmach
Sent: Tuesday, June 02, 2015 4:50 PM
To: volt-nuts@febo.com
Subject: [volt-nuts] HP-3458A questions

Hi Randy,

in advance, this change only improves the stability of the instrument, when the 
3458A is powered.
The LTZ1000 will not age (nearly), when the instrument is switched off.
Therefore, it makes sense mainly, if your instrument is running for an extended 
period of time.

(Reference: Spreadbury: Ultra Zener..)

If you change the oven temperature, you will loose the 10V calibration, as the 
LTZ1000A will change its reference voltage.


Although my instrument is strictly switched off, when not doing
measurements, I reduced the oven temp. by connecting a 100k precision
wirewound (3ppm/K) in parallel to the 15k BMF resistor. That gives 13k
in total, yielding about 65°C like in the LT datasheet.
An alternative placement is already foreseen on the PCB; in the CLIP,
it's component designator is X411.
I'll try to link in a photo of my change.

lymex on the bbs35.hot forum obviously was more agressive, and paralled
75k of Vitrohm Thin Film, giving 12.5k for about 55°C.

I would recommend my solution above the one of lymex, although his one
gives theoretically more stability.
Also, the lower stability of TF reistors may affect the stability (T, t)
of the 15k BMF.

With 100k/65°C, the oven runs about 30°C lower.
Therefore, the ambient temperature has to be limited to max. 35°C, the
fan has ALWAYS to be kept clean, giving a maximum temperature rise of
15°C, and the instrument may not be put in a rack (which would give
additional 10..20°C temperature rise). Under these conditions, (35°C
ambient + 15°C 3458A inner heating + 10°C self heating of LTZ1000A + 5°C
regulation margin), the oven will always be stable.

These 75k / 55°C would be possible, only if you limit the ambient
temperature further, to = 25°C, and precisely monitor the 3458A inner
temperature (TEMP?).

Although my instrument is kept at 20..23°C throughout the year, i
wouldn't feel comfortable with 55°C.
The only possibility would be the exchange of the reference to the non-A
version, as the self heating effect may vanish!

Frank

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


[volt-nuts] HP-3458A questions

2015-06-02 Thread Randy Evans
Does anyone have explicit instructions on how to lower the temperature of
the LTZ1000 for better stability; e.g., what resistors to change and to
what value?

Similarly, step-by-step instructions on how to back up the cal ram in the
HP-3458A?  I now Mark Sims posted his program on Jon Miles site but I am
confused how to use it (I want to back up my HP-3458A before I attempt to
replace the display board).

thanks,

Randy Evans AE6YG