Re: [volt-nuts] HP3457

2012-08-22 Thread Marvin E. Gozum
The 3457a is younger, has added functions, such as amps, and better 
long term accuracy; the 3456a is just a DVM.


The previously posted link has a large comparative database:

http://electronicsandbooks.com/eab1/kb/k0zfca/i0zfca7_.htm

3457a tends to be have more unobtanium, compared to the 3456a.  For 
700 euros for a calibrated 3457a, its better to look at new 6.5 digit 
DMMs such as the 34410a which I see farnell lists for 850 euros.


http://uk.farnell.com/agilent-technologies/34410a/multimeter-bench-61-2-digit/dp/1440376

or the cheaper 34401a or equivalents from other makers, Fluke, Keithley etc..

again, electronicsandbooks has a wonderful comparative listing.

At 07:21 PM 8/21/2012, Fred Schneider wrote:
I was aiming for the holy grale of meters and a 3457 turns out to be 
a whole different beast. Someone I know bought a 3456 for 225 euro, 
fresh calibrated, that is 6,5 digits too. So i wondered why is a 
3457 so much more expensive.




Sincerely,



Marv Gozum
Philadelphia, PA  




___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] HP3457

2012-08-22 Thread Fred Schneider
ibes a test  were you run it at 15 degrees, then 
>> at 38 degrees and then adjust diodecurrent until you get the best tempco. 
>> That would imply the thing should not react much on a few degrees.
>> 
>> Would it be an option to ad tempcontrol. The meter itself is a sort of 
>> uncontrolled oven. If I add a temp sensor and find a way to keep the temp 
>> steady the meter should stay more stable. In this case cooling would be the 
>> best option because it gets over 45 - 50 degrees inside the meter. So it can 
>> heats it self pretty well, The problem is temp follows the oudside temp, so 
>> if Measure inside temp at for instance 18 or 20 degrees room temp and cool 
>> it if the inside gets hotter then the inner temp at 20 degrees outside, than 
>> it must be more stable.
>> 
>> Maybe a piezo element from a coolerbox inside (the piezo is inside, a 
>> coolrib and fan on theoutside. I experimented with a coolrib on the cabinet 
>> straight above the transformer, that was allready an improvent. Or some 
>> aluminum tubes through the cabinet. Open on one side, a box on the other 
>> side with a fan in it sucking the air through the pipes and cooling the 
>> inside when needed, or copper tubing where i can solder ribs on. If not 
>> enough I can pomp water through it and keep the water in an isolated box at 
>> 30 degrees using an aquarium heater if needed. That would cover 95% of the 
>> year.
>> 
>> Stupid idea ?
>> 
>> 
>> Fred PA4TIM
>> 
>> Op 22 aug. 2012 om 03:32 heeft John Phillips  het 
>> volgende geschreven:
>> 
>>> To get better performance form you Solarton keep it a constant Hot
>>> temperature all the time. Do not turn it off! Do not let the room
>>> temperature fluctuate more than a few degrees. The more stale the better.
>>> Cooling in most cases will induce error. If you can keep it at 20C all the
>>> time go for it but if you are more stable at 28C then 28 is better.
>>> 
>>> The longer you run it the more stable it will get. We do not turn off any
>>> of our best meters unless we are moving them or the power gos off.
>>> 
>>> 
>>> On Tue, Aug 21, 2012 at 4:21 PM, Fred Schneider  wrote:
>>> 
>>>> I looked at the DC specs of the 3457 and thats the same as the solartron.
>>>> I am looking sub uV like the solartron. The 3457 does not have that. So I
>>>> think, regarding your reactions and the price, i'd better first take a look
>>>> at the solartron. Modify it a bit ( cooling-ribs on the outside, one on top
>>>> above the transformer and one on the back where the regulators should be (
>>>> i'm told, not looked yet) maybe a fan blowing over the ribs.) replace the
>>>> keylock with a switch ( i have not a good key, it works but not smooth)
>>>> then put banana busses instead of the back nr2 input connector on the front
>>>> and replace the mainsswitch to the front. ( instead of climbing on a hair
>>>> to reach the switch)
>>>> 
>>>> I printed the service manual of the 7061 and i think I first do the tests
>>>> ( psu, frequency, input current ect) i also noted my zero problem is
>>>> because I shorted both inputs but i need to short the reference and front
>>>> inputs. That could explain why zero often does not work ( error, zero to
>>>> large or something lile that)
>>>> 
>>>> I bought a Keithley 2000 last year, it was comparable to the Fluke 8046
>>>> but around 400 euro cheaper. I bought it straight from Keithley including
>>>> calibration. Besides that I have a keithley 196 that was just calibrated
>>>> when I bought it. ( and a 199)
>>>> 
>>>> I was aiming for the holy grale of meters and a 3457 turns out to be a
>>>> whole different beast. Someone I know bought a 3456 for 225 euro, fresh
>>>> calibrated, that is 6,5 digits too. So i wondered why is a 3457 so much
>>>> more expensive.
>>>> 
>>>> But I better wait and save more money in the hope to find one day an
>>>> affortable 3458.
>>>> One should always keep something to dream about :-)
>>>> 
>>>> Fred
>>>> 
>>>> 
>>>> 
>>>> 
>>>> Fred PA4TIM
>>>> 
>>>> Op 21 aug. 2012 om 22:06 heeft John Phillips 
>>>> het volgende geschreven:
>>>> 
>>>>> you can put your 3457A into statistics mode and pick up +/- 29,999,999
>>>>> counts from the display.
&g

Re: [volt-nuts] HP3457

2012-08-22 Thread J. L. Trantham
I have never tried this but I had the idea of building an 'enclosed set of
shelves' above my workbench with a Lucite or similar door with thermometers
built in and a way of 'gently' stirring the air inside while having a heat
exchanger and the ability to 'cool' the air inside (since my shop is more
likely to be hot and the equipment inside would generate heat).  It would
need power and other connections (Test Leads, Coax, LAN, HPIB, USB, RS232,
etc.) that would be through some sort of 'sealed' access.

It might take some work and it might be quite cumbersome but it sounds
doable and would permit putting the appropriate instrument under test inside
to do some simple 'environmental testing'.

I wonder if such 'environmental test chambers' of a small size are
commercially available?

Joe

-Original Message-
From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On
Behalf Of Fred Schneider
Sent: Wednesday, August 22, 2012 2:23 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] HP3457


So this is normal, hmm, my shop can be over 30 degrees when it is a hot
summer and 15 degrees in a Cold winter so keeping it powered will not help
in my case ( i am just an amateur voltnut so climatize my room is no option.

But The service manual describes a test  were you run it at 15 degrees, then
at 38 degrees and then adjust diodecurrent until you get the best tempco.
That would imply the thing should not react much on a few degrees. 

Would it be an option to ad tempcontrol. The meter itself is a sort of
uncontrolled oven. If I add a temp sensor and find a way to keep the temp
steady the meter should stay more stable. In this case cooling would be the
best option because it gets over 45 - 50 degrees inside the meter. So it can
heats it self pretty well, The problem is temp follows the oudside temp, so
if Measure inside temp at for instance 18 or 20 degrees room temp and cool
it if the inside gets hotter then the inner temp at 20 degrees outside, than
it must be more stable.

Maybe a piezo element from a coolerbox inside (the piezo is inside, a
coolrib and fan on theoutside. I experimented with a coolrib on the cabinet
straight above the transformer, that was allready an improvent. Or some
aluminum tubes through the cabinet. Open on one side, a box on the other
side with a fan in it sucking the air through the pipes and cooling the
inside when needed, or copper tubing where i can solder ribs on. If not
enough I can pomp water through it and keep the water in an isolated box at
30 degrees using an aquarium heater if needed. That would cover 95% of the
year.

Stupid idea ?


Fred PA4TIM

Op 22 aug. 2012 om 03:32 heeft John Phillips  het
volgende geschreven:

> To get better performance form you Solarton keep it a constant Hot 
> temperature all the time. Do not turn it off! Do not let the room 
> temperature fluctuate more than a few degrees. The more stale the 
> better. Cooling in most cases will induce error. If you can keep it at 
> 20C all the time go for it but if you are more stable at 28C then 28 
> is better.
> 
> The longer you run it the more stable it will get. We do not turn off 
> any of our best meters unless we are moving them or the power gos off.
> 
> 
> On Tue, Aug 21, 2012 at 4:21 PM, Fred Schneider  
> wrote:
> 
>> I looked at the DC specs of the 3457 and thats the same as the 
>> solartron. I am looking sub uV like the solartron. The 3457 does not 
>> have that. So I think, regarding your reactions and the price, i'd 
>> better first take a look at the solartron. Modify it a bit ( 
>> cooling-ribs on the outside, one on top above the transformer and one 
>> on the back where the regulators should be ( i'm told, not looked 
>> yet) maybe a fan blowing over the ribs.) replace the keylock with a 
>> switch ( i have not a good key, it works but not smooth) then put 
>> banana busses instead of the back nr2 input connector on the front 
>> and replace the mainsswitch to the front. ( instead of climbing on a 
>> hair to reach the switch)
>> 
>> I printed the service manual of the 7061 and i think I first do the 
>> tests ( psu, frequency, input current ect) i also noted my zero 
>> problem is because I shorted both inputs but i need to short the 
>> reference and front inputs. That could explain why zero often does 
>> not work ( error, zero to large or something lile that)
>> 
>> I bought a Keithley 2000 last year, it was comparable to the Fluke 
>> 8046 but around 400 euro cheaper. I bought it straight from Keithley 
>> including calibration. Besides that I have a keithley 196 that was 
>> just calibrated when I bought it. ( and a 199)
>> 
>> I was aiming for the holy grale of meters and a 3457 turns out

Re: [volt-nuts] HP3457 - Cal Key for Solartron 7061

2012-08-22 Thread J. L. Trantham
Can't recall if I posted this here or not but the Cal Key for my 7081 also
fits the 7061.

Joe

-Original Message-
From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On
Behalf Of Fabio Eboli
Sent: Wednesday, August 22, 2012 2:58 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] HP3457



As for the key, I can duplicate mine and send it to you,
if you want to try another key.

Fabio.



___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] HP3457

2012-08-22 Thread Fabio Eboli
 instead of the back nr2 input connector on the front
and replace the mainsswitch to the front. ( instead of climbing on a hair
to reach the switch)

I printed the service manual of the 7061 and i think I first do the tests
( psu, frequency, input current ect) i also noted my zero problem is
because I shorted both inputs but i need to short the reference and front
inputs. That could explain why zero often does not work ( error, zero to
large or something lile that)

I bought a Keithley 2000 last year, it was comparable to the Fluke 8046
but around 400 euro cheaper. I bought it straight from Keithley including
calibration. Besides that I have a keithley 196 that was just calibrated
when I bought it. ( and a 199)

I was aiming for the holy grale of meters and a 3457 turns out to be a
whole different beast. Someone I know bought a 3456 for 225 euro, fresh
calibrated, that is 6,5 digits too. So i wondered why is a 3457 so much
more expensive.

But I better wait and save more money in the hope to find one day an
affortable 3458.
One should always keep something to dream about :-)

Fred




Fred PA4TIM

Op 21 aug. 2012 om 22:06 heeft John Phillips 
het volgende geschreven:


you can put your 3457A into statistics mode and pick up +/- 29,999,999
counts from the display.
+/- 2,999,999 counts  is all you can get form the normal /live mode.

On Tue, Aug 21, 2012 at 12:52 PM, J. L. Trantham  wrote:


Fred,

700 euro seems a bit high, at least by US standards, for a 3457A.  It

is a

good meter and I have one.  Though no longer supported, Agilent will

still

calibrate one for $200.22, according to their website.

You might want to consider a Fluke 8846A.  I  have seen them sell for

$900

new.  I don't have one but it was going to be my next addition until I

got

a
3458A.

What is the issue with the 7061?

Joe

-Original Message-
From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On
Behalf Of Fred Schneider
Sent: Tuesday, August 21, 2012 2:00 PM
To: volt-nuts@febo.com
Subject: [volt-nuts] HP3457

I have a solatron 7061 but i am not very impressed with it. Must I go
looking for an alternative or is this 7061 when i gave it a good

checkup /

overhaul a good meter ?
I am offered a 3457 for 700 euro and the owner, will calibrate it (he

has

several calibrtors, standards and a couple of 3458  meters. My intuition
tells me to trusts him ( and that had never let me down before))

But he also told it is only 7,5 digit over hpib, and has 6,5 digit

display.


My Keithly 2000 is 6,5 digits and over RS232 i get two digits extra. I

once

tried that, but that took me a lot of time to get it more ore less

working,

I am not so into computers and afraid I do not get it to work.

Is it really better as my ( one year old 2000 ?) and better as my

Solartron

can be ?
I am afraid a 3458 is outside my budget.

Fred



Fred PA4TIM
___
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.





--
John Phillips
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to

https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts

and follow the instructions there.


___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.





--
John Phillips
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to  
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts

and follow the instructions there.


___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.






This message was sent using IMP, the Internet Messaging Program.



___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] HP3457

2012-08-22 Thread John Phillips
>> het volgende geschreven:
> >>
> >>> you can put your 3457A into statistics mode and pick up +/- 29,999,999
> >>> counts from the display.
> >>> +/- 2,999,999 counts  is all you can get form the normal /live mode.
> >>>
> >>> On Tue, Aug 21, 2012 at 12:52 PM, J. L. Trantham 
> wrote:
> >>>
> >>>> Fred,
> >>>>
> >>>> 700 euro seems a bit high, at least by US standards, for a 3457A.  It
> >> is a
> >>>> good meter and I have one.  Though no longer supported, Agilent will
> >> still
> >>>> calibrate one for $200.22, according to their website.
> >>>>
> >>>> You might want to consider a Fluke 8846A.  I  have seen them sell for
> >> $900
> >>>> new.  I don't have one but it was going to be my next addition until I
> >> got
> >>>> a
> >>>> 3458A.
> >>>>
> >>>> What is the issue with the 7061?
> >>>>
> >>>> Joe
> >>>>
> >>>> -Original Message-
> >>>> From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com]
> On
> >>>> Behalf Of Fred Schneider
> >>>> Sent: Tuesday, August 21, 2012 2:00 PM
> >>>> To: volt-nuts@febo.com
> >>>> Subject: [volt-nuts] HP3457
> >>>>
> >>>> I have a solatron 7061 but i am not very impressed with it. Must I go
> >>>> looking for an alternative or is this 7061 when i gave it a good
> >> checkup /
> >>>> overhaul a good meter ?
> >>>> I am offered a 3457 for 700 euro and the owner, will calibrate it (he
> >> has
> >>>> several calibrtors, standards and a couple of 3458  meters. My
> intuition
> >>>> tells me to trusts him ( and that had never let me down before))
> >>>>
> >>>> But he also told it is only 7,5 digit over hpib, and has 6,5 digit
> >> display.
> >>>>
> >>>> My Keithly 2000 is 6,5 digits and over RS232 i get two digits extra. I
> >> once
> >>>> tried that, but that took me a lot of time to get it more ore less
> >> working,
> >>>> I am not so into computers and afraid I do not get it to work.
> >>>>
> >>>> Is it really better as my ( one year old 2000 ?) and better as my
> >> Solartron
> >>>> can be ?
> >>>> I am afraid a 3458 is outside my budget.
> >>>>
> >>>> Fred
> >>>>
> >>>>
> >>>>
> >>>> Fred PA4TIM
> >>>> ___
> >>>> volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
> >>>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >>>> and follow the instructions there.
> >>>>
> >>>>
> >>>> ___
> >>>> volt-nuts mailing list -- volt-nuts@febo.com
> >>>> To unsubscribe, go to
> >>>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >>>> and follow the instructions there.
> >>>>
> >>>
> >>>
> >>>
> >>> --
> >>> John Phillips
> >>> ___
> >>> volt-nuts mailing list -- volt-nuts@febo.com
> >>> To unsubscribe, go to
> >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >>> and follow the instructions there.
> >>
> >> ___
> >> volt-nuts mailing list -- volt-nuts@febo.com
> >> To unsubscribe, go to
> >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> >> and follow the instructions there.
> >>
> >
> >
> >
> > --
> > John Phillips
> > ___
> > volt-nuts mailing list -- volt-nuts@febo.com
> > To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
>
> ___
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>



-- 
John Phillips
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] HP3457

2012-08-22 Thread Fred Schneider
alibrate one for $200.22, according to their website.
>>>> 
>>>> You might want to consider a Fluke 8846A.  I  have seen them sell for
>> $900
>>>> new.  I don't have one but it was going to be my next addition until I
>> got
>>>> a
>>>> 3458A.
>>>> 
>>>> What is the issue with the 7061?
>>>> 
>>>> Joe
>>>> 
>>>> -Original Message-
>>>> From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On
>>>> Behalf Of Fred Schneider
>>>> Sent: Tuesday, August 21, 2012 2:00 PM
>>>> To: volt-nuts@febo.com
>>>> Subject: [volt-nuts] HP3457
>>>> 
>>>> I have a solatron 7061 but i am not very impressed with it. Must I go
>>>> looking for an alternative or is this 7061 when i gave it a good
>> checkup /
>>>> overhaul a good meter ?
>>>> I am offered a 3457 for 700 euro and the owner, will calibrate it (he
>> has
>>>> several calibrtors, standards and a couple of 3458  meters. My intuition
>>>> tells me to trusts him ( and that had never let me down before))
>>>> 
>>>> But he also told it is only 7,5 digit over hpib, and has 6,5 digit
>> display.
>>>> 
>>>> My Keithly 2000 is 6,5 digits and over RS232 i get two digits extra. I
>> once
>>>> tried that, but that took me a lot of time to get it more ore less
>> working,
>>>> I am not so into computers and afraid I do not get it to work.
>>>> 
>>>> Is it really better as my ( one year old 2000 ?) and better as my
>> Solartron
>>>> can be ?
>>>> I am afraid a 3458 is outside my budget.
>>>> 
>>>> Fred
>>>> 
>>>> 
>>>> 
>>>> Fred PA4TIM
>>>> ___
>>>> volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
>>>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>>>> and follow the instructions there.
>>>> 
>>>> 
>>>> ___
>>>> volt-nuts mailing list -- volt-nuts@febo.com
>>>> To unsubscribe, go to
>>>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>>>> and follow the instructions there.
>>>> 
>>> 
>>> 
>>> 
>>> --
>>> John Phillips
>>> ___
>>> volt-nuts mailing list -- volt-nuts@febo.com
>>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>>> and follow the instructions there.
>> 
>> ___
>> volt-nuts mailing list -- volt-nuts@febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> and follow the instructions there.
>> 
> 
> 
> 
> -- 
> John Phillips
> ___
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] HP3457

2012-08-21 Thread John Phillips
To get better performance form you Solarton keep it a constant Hot
temperature all the time. Do not turn it off! Do not let the room
temperature fluctuate more than a few degrees. The more stale the better.
Cooling in most cases will induce error. If you can keep it at 20C all the
time go for it but if you are more stable at 28C then 28 is better.

The longer you run it the more stable it will get. We do not turn off any
of our best meters unless we are moving them or the power gos off.


On Tue, Aug 21, 2012 at 4:21 PM, Fred Schneider  wrote:

> I looked at the DC specs of the 3457 and thats the same as the solartron.
> I am looking sub uV like the solartron. The 3457 does not have that. So I
> think, regarding your reactions and the price, i'd better first take a look
> at the solartron. Modify it a bit ( cooling-ribs on the outside, one on top
> above the transformer and one on the back where the regulators should be (
> i'm told, not looked yet) maybe a fan blowing over the ribs.) replace the
> keylock with a switch ( i have not a good key, it works but not smooth)
> then put banana busses instead of the back nr2 input connector on the front
> and replace the mainsswitch to the front. ( instead of climbing on a hair
> to reach the switch)
>
> I printed the service manual of the 7061 and i think I first do the tests
> ( psu, frequency, input current ect) i also noted my zero problem is
> because I shorted both inputs but i need to short the reference and front
> inputs. That could explain why zero often does not work ( error, zero to
> large or something lile that)
>
> I bought a Keithley 2000 last year, it was comparable to the Fluke 8046
> but around 400 euro cheaper. I bought it straight from Keithley including
> calibration. Besides that I have a keithley 196 that was just calibrated
> when I bought it. ( and a 199)
>
> I was aiming for the holy grale of meters and a 3457 turns out to be a
> whole different beast. Someone I know bought a 3456 for 225 euro, fresh
> calibrated, that is 6,5 digits too. So i wondered why is a 3457 so much
> more expensive.
>
>  But I better wait and save more money in the hope to find one day an
> affortable 3458.
> One should always keep something to dream about :-)
>
> Fred
>
>
>
>
> Fred PA4TIM
>
> Op 21 aug. 2012 om 22:06 heeft John Phillips 
> het volgende geschreven:
>
> > you can put your 3457A into statistics mode and pick up +/- 29,999,999
> > counts from the display.
> > +/- 2,999,999 counts  is all you can get form the normal /live mode.
> >
> > On Tue, Aug 21, 2012 at 12:52 PM, J. L. Trantham  wrote:
> >
> >> Fred,
> >>
> >> 700 euro seems a bit high, at least by US standards, for a 3457A.  It
> is a
> >> good meter and I have one.  Though no longer supported, Agilent will
> still
> >> calibrate one for $200.22, according to their website.
> >>
> >> You might want to consider a Fluke 8846A.  I  have seen them sell for
> $900
> >> new.  I don't have one but it was going to be my next addition until I
> got
> >> a
> >> 3458A.
> >>
> >> What is the issue with the 7061?
> >>
> >> Joe
> >>
> >> -Original Message-
> >> From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On
> >> Behalf Of Fred Schneider
> >> Sent: Tuesday, August 21, 2012 2:00 PM
> >> To: volt-nuts@febo.com
> >> Subject: [volt-nuts] HP3457
> >>
> >> I have a solatron 7061 but i am not very impressed with it. Must I go
> >> looking for an alternative or is this 7061 when i gave it a good
> checkup /
> >> overhaul a good meter ?
> >> I am offered a 3457 for 700 euro and the owner, will calibrate it (he
> has
> >> several calibrtors, standards and a couple of 3458  meters. My intuition
> >> tells me to trusts him ( and that had never let me down before))
> >>
> >> But he also told it is only 7,5 digit over hpib, and has 6,5 digit
> display.
> >>
> >> My Keithly 2000 is 6,5 digits and over RS232 i get two digits extra. I
> once
> >> tried that, but that took me a lot of time to get it more ore less
> working,
> >> I am not so into computers and afraid I do not get it to work.
> >>
> >> Is it really better as my ( one year old 2000 ?) and better as my
> Solartron
> >> can be ?
> >> I am afraid a 3458 is outside my budget.
> >>
> >> Fred
> >>
> >>
> >>
> >> Fred PA4TIM
> >> ___
> >> volt-nuts mailing list -- volt-nuts@

Re: [volt-nuts] HP3457

2012-08-21 Thread Fred Schneider
I looked at the DC specs of the 3457 and thats the same as the solartron. I am 
looking sub uV like the solartron. The 3457 does not have that. So I think, 
regarding your reactions and the price, i'd better first take a look at the 
solartron. Modify it a bit ( cooling-ribs on the outside, one on top above the 
transformer and one on the back where the regulators should be ( i'm told, not 
looked yet) maybe a fan blowing over the ribs.) replace the keylock with a 
switch ( i have not a good key, it works but not smooth) then put banana busses 
instead of the back nr2 input connector on the front and replace the 
mainsswitch to the front. ( instead of climbing on a hair to reach the switch)

I printed the service manual of the 7061 and i think I first do the tests ( 
psu, frequency, input current ect) i also noted my zero problem is because I 
shorted both inputs but i need to short the reference and front inputs. That 
could explain why zero often does not work ( error, zero to large or something 
lile that)

I bought a Keithley 2000 last year, it was comparable to the Fluke 8046 but 
around 400 euro cheaper. I bought it straight from Keithley including 
calibration. Besides that I have a keithley 196 that was just calibrated when I 
bought it. ( and a 199)

I was aiming for the holy grale of meters and a 3457 turns out to be a whole 
different beast. Someone I know bought a 3456 for 225 euro, fresh calibrated, 
that is 6,5 digits too. So i wondered why is a 3457 so much more expensive.

 But I better wait and save more money in the hope to find one day an 
affortable 3458.
One should always keep something to dream about :-)

Fred




Fred PA4TIM

Op 21 aug. 2012 om 22:06 heeft John Phillips  het 
volgende geschreven:

> you can put your 3457A into statistics mode and pick up +/- 29,999,999
> counts from the display.
> +/- 2,999,999 counts  is all you can get form the normal /live mode.
> 
> On Tue, Aug 21, 2012 at 12:52 PM, J. L. Trantham  wrote:
> 
>> Fred,
>> 
>> 700 euro seems a bit high, at least by US standards, for a 3457A.  It is a
>> good meter and I have one.  Though no longer supported, Agilent will still
>> calibrate one for $200.22, according to their website.
>> 
>> You might want to consider a Fluke 8846A.  I  have seen them sell for $900
>> new.  I don't have one but it was going to be my next addition until I got
>> a
>> 3458A.
>> 
>> What is the issue with the 7061?
>> 
>> Joe
>> 
>> -Original Message-
>> From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On
>> Behalf Of Fred Schneider
>> Sent: Tuesday, August 21, 2012 2:00 PM
>> To: volt-nuts@febo.com
>> Subject: [volt-nuts] HP3457
>> 
>> I have a solatron 7061 but i am not very impressed with it. Must I go
>> looking for an alternative or is this 7061 when i gave it a good checkup /
>> overhaul a good meter ?
>> I am offered a 3457 for 700 euro and the owner, will calibrate it (he has
>> several calibrtors, standards and a couple of 3458  meters. My intuition
>> tells me to trusts him ( and that had never let me down before))
>> 
>> But he also told it is only 7,5 digit over hpib, and has 6,5 digit display.
>> 
>> My Keithly 2000 is 6,5 digits and over RS232 i get two digits extra. I once
>> tried that, but that took me a lot of time to get it more ore less working,
>> I am not so into computers and afraid I do not get it to work.
>> 
>> Is it really better as my ( one year old 2000 ?) and better as my Solartron
>> can be ?
>> I am afraid a 3458 is outside my budget.
>> 
>> Fred
>> 
>> 
>> 
>> Fred PA4TIM
>> ___
>> volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> and follow the instructions there.
>> 
>> 
>> ___
>> volt-nuts mailing list -- volt-nuts@febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> and follow the instructions there.
>> 
> 
> 
> 
> -- 
> John Phillips
> ___
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] HP3457

2012-08-21 Thread John Phillips
you can put your 3457A into statistics mode and pick up +/- 29,999,999
counts from the display.
+/- 2,999,999 counts  is all you can get form the normal /live mode.

On Tue, Aug 21, 2012 at 12:52 PM, J. L. Trantham  wrote:

> Fred,
>
> 700 euro seems a bit high, at least by US standards, for a 3457A.  It is a
> good meter and I have one.  Though no longer supported, Agilent will still
> calibrate one for $200.22, according to their website.
>
> You might want to consider a Fluke 8846A.  I  have seen them sell for $900
> new.  I don't have one but it was going to be my next addition until I got
> a
> 3458A.
>
> What is the issue with the 7061?
>
> Joe
>
> -Original Message-
> From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On
> Behalf Of Fred Schneider
> Sent: Tuesday, August 21, 2012 2:00 PM
> To: volt-nuts@febo.com
> Subject: [volt-nuts] HP3457
>
> I have a solatron 7061 but i am not very impressed with it. Must I go
> looking for an alternative or is this 7061 when i gave it a good checkup /
> overhaul a good meter ?
> I am offered a 3457 for 700 euro and the owner, will calibrate it (he has
> several calibrtors, standards and a couple of 3458  meters. My intuition
> tells me to trusts him ( and that had never let me down before))
>
> But he also told it is only 7,5 digit over hpib, and has 6,5 digit display.
>
> My Keithly 2000 is 6,5 digits and over RS232 i get two digits extra. I once
> tried that, but that took me a lot of time to get it more ore less working,
> I am not so into computers and afraid I do not get it to work.
>
> Is it really better as my ( one year old 2000 ?) and better as my Solartron
> can be ?
> I am afraid a 3458 is outside my budget.
>
> Fred
>
>
>
> Fred PA4TIM
> ___
> volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
>
> ___
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>



-- 
John Phillips
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] HP3457

2012-08-21 Thread J. L. Trantham
Fred,

700 euro seems a bit high, at least by US standards, for a 3457A.  It is a
good meter and I have one.  Though no longer supported, Agilent will still
calibrate one for $200.22, according to their website.

You might want to consider a Fluke 8846A.  I  have seen them sell for $900
new.  I don't have one but it was going to be my next addition until I got a
3458A.

What is the issue with the 7061?

Joe

-Original Message-
From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On
Behalf Of Fred Schneider
Sent: Tuesday, August 21, 2012 2:00 PM
To: volt-nuts@febo.com
Subject: [volt-nuts] HP3457

I have a solatron 7061 but i am not very impressed with it. Must I go
looking for an alternative or is this 7061 when i gave it a good checkup /
overhaul a good meter ?
I am offered a 3457 for 700 euro and the owner, will calibrate it (he has
several calibrtors, standards and a couple of 3458  meters. My intuition
tells me to trusts him ( and that had never let me down before)) 

But he also told it is only 7,5 digit over hpib, and has 6,5 digit display.

My Keithly 2000 is 6,5 digits and over RS232 i get two digits extra. I once
tried that, but that took me a lot of time to get it more ore less working,
I am not so into computers and afraid I do not get it to work.

Is it really better as my ( one year old 2000 ?) and better as my Solartron
can be ?
I am afraid a 3458 is outside my budget.

Fred



Fred PA4TIM
___
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] HP3457

2012-08-21 Thread Rob Klein

Fred,

I think 700 euros is way too much for a 3457. I also have one, as well 
as a Keithley 2000 and if I ever should decide to get rid

of one of them, it will not be the 2000.

I have yet to try the RS-232 connection on the 2000, but I'll get in 
touch with you when I do. A nice, simple program to get at

all the features shouldn't be very hard to write for me.

You know how to reach me if you have any further questions.


Rob Klein.


Op 21-8-2012 20:59, Fred Schneider schreef:

I have a solatron 7061 but i am not very impressed with it. Must I go looking 
for an alternative or is this 7061 when i gave it a good checkup / overhaul a 
good meter ?
I am offered a 3457 for 700 euro and the owner, will calibrate it (he has 
several calibrtors, standards and a couple of 3458  meters. My intuition tells 
me to trusts him ( and that had never let me down before))

But he also told it is only 7,5 digit over hpib, and has 6,5 digit display.

My Keithly 2000 is 6,5 digits and over RS232 i get two digits extra. I once 
tried that, but that took me a lot of time to get it more ore less working, I 
am not so into computers and afraid I do not get it to work.

Is it really better as my ( one year old 2000 ?) and better as my Solartron can 
be ?
I am afraid a 3458 is outside my budget.

Fred



Fred PA4TIM
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.




___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


[volt-nuts] HP3457

2012-08-21 Thread Fred Schneider
I have a solatron 7061 but i am not very impressed with it. Must I go looking 
for an alternative or is this 7061 when i gave it a good checkup / overhaul a 
good meter ?
I am offered a 3457 for 700 euro and the owner, will calibrate it (he has 
several calibrtors, standards and a couple of 3458  meters. My intuition tells 
me to trusts him ( and that had never let me down before)) 

But he also told it is only 7,5 digit over hpib, and has 6,5 digit display.

My Keithly 2000 is 6,5 digits and over RS232 i get two digits extra. I once 
tried that, but that took me a lot of time to get it more ore less working, I 
am not so into computers and afraid I do not get it to work.

Is it really better as my ( one year old 2000 ?) and better as my Solartron can 
be ?
I am afraid a 3458 is outside my budget.

Fred



Fred PA4TIM
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.