Re: [volt-nuts] Old Weston cells

2013-08-27 Thread pa4tim
Are you sure they are unsaturated ? Because most unsaturated cells last only 
10-20 years (AFIK). Saturated cells last much longer but they should be kept at 
30 degrees celcius for the correct value. On the other hand, as far as I know 
they do not survive if they are not kept vertical and most of the time they are 
not when they are kept outside their oven. 


I have 4 weston cells in a guildline cabinet with oven. I have the cal data for 
the first 10 years they have been used. After this time the oven was turned off 
, the lab was closed but all gear  stayed put on that same place over 30 years 
(the lab was turned into a server room but all the gear just stayed unused ) I 
made a graph from the data. Took a ruler and extended the lines. After a few 
months I had a change to measure them with a calibrated 6,5 digit meter and a 
year later a 7,5 digit meter and the values I found where plus/min a few uV on 
the lines. But more important the difference between the individual cells was 
spot on with my graph. So I use them as my house standard (I do not measure 
them direct. I set my 332 at 10V. A fluke 720 of ESI KV at the value the cell 
must have. Connect them both to a Fluke 845 and set the 332 so the meter reads 
zero. Then I check my 731 against the 332 the same way and leave them both 
connected this way for a few hours. If the meter is still at zero after that 
time I use the 332 for calibrating a meter.


The only problem is the high Rout of a KV. So I use a buffer made of a LTC1052 
and LT1010 after the KV  if the multimeter is a 10 M version. But I still 
wonder if that is needed. I tested it with some meters but did not notice a 
difference that was within the meters resolution.


Fred, pa4tim 




Verzonden met Windows



Van: Joseph Gray
Verzonden: ‎dinsdag‎ ‎27‎ ‎augustus‎ ‎2013 ‎06‎:‎26
Aan: Discussion of precise voltage measurement

I suspected that there was nothing to do to rejuvenate these old cells. I
thought I'd ask anyway.

Thanks.

Joe Gray
W5JG



On Mon, Aug 26, 2013 at 9:16 PM, zbigniew169 zbigniew...@gmail.com wrote:

 All my knowledge regarding Weston Cells tells me that is not possible in
 any way :-((


 2013/8/27 Joseph Gray jg...@zianet.com

  I got my hands on a few very old, unsaturated cells. I figured that they
  would be no good after all this time. Just out of curiosity, I measured
  them on my 3457A, which hasn't been sent off for calibration yet.
 
  DMM was set to 3 V scale to keep the input impedance high. Room
 temperature
  = 26 C.
 
  Cell #1 = 1.018288
  Cell #2 = 1.018236
  Cell #3 = 1.018224
 
  I read that when they get below 1.0183 V, they are no good. Is there any
  (easy) way to rejuvenate these cells?
 
  Joe Gray
  W5JG
  ___
  volt-nuts mailing list -- volt-nuts@febo.com
  To unsubscribe, go to
  https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
  and follow the instructions there.
 



 --
 Zbyszek
 ___
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.


___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Re: [volt-nuts] Old Weston cells

2013-08-27 Thread Bill Ezell

pa4...@gmail.com writes:

I have 4 weston cells in a guildline cabinet with oven. I have the cal data for 
the first 10 years they have been used. After this time the oven was turned off 
, the lab was closed but all gear  stayed put on that same place over 30 years 
(the lab was turned into a server room but all the gear just stayed unused ) I 
made a graph from the data. Took a ruler and extended the lines. After a few 
months I had a change to measure them with a calibrated 6,5 digit meter and a 
year later a 7,5 digit meter and the values I found where plus/min a few uV on 
the lines. But more important the difference between the individual cells was 
spot on with my graph.


I have that same unit myself. For some years I kept it on and would 
periodically compare the cells to my (calibrated) Datron 4910's. I kept 
logs of the voltage sampled weekly using an 8.5 digit meter for 
differential comparison for 2 or 3 years. I was surprised at how stable 
and predictable they were. But, I stopped doing that probably 5 years 
ago. Nostalgia strikes, I'll plug it back in and see what's changed. 
Unfortunately, it takes weeks for the voltage to stabilize from a cold 
start.


--
Bill Ezell

They said 'Windows or better'
so I used Linux.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Old Weston cells

2013-08-26 Thread J. L. Trantham
Joe,

Do you have any data about prior measurements of these cells or their
history?

I read somewhere there is a predicted decline rate based on time and
temperature IIRC.

Once you get the meter calibrated, you can 'track' their measurements.

Might also be time for a 'differential voltmeter' to add to the 'stable'.

Joe

-Original Message-
From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On
Behalf Of Joseph Gray
Sent: Monday, August 26, 2013 9:59 PM
To: Discussion of precise voltage measurement
Subject: [volt-nuts] Old Weston cells

I got my hands on a few very old, unsaturated cells. I figured that they
would be no good after all this time. Just out of curiosity, I measured them
on my 3457A, which hasn't been sent off for calibration yet.

DMM was set to 3 V scale to keep the input impedance high. Room temperature
= 26 C.

Cell #1 = 1.018288
Cell #2 = 1.018236
Cell #3 = 1.018224

I read that when they get below 1.0183 V, they are no good. Is there any
(easy) way to rejuvenate these cells?

Joe Gray
W5JG
___
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Old Weston cells

2013-08-26 Thread Joseph Gray
Sorry, no history at all. They were salvaged from old equipment. From what
I read, they are already below their voltage threshold and should be
tossed. I just thought it would be interesting to measure them anyway.

I think the decline rate was supposed to be around 0.08 uV per year, or
something like that.

After the 3457A is calibrated, I will try to track these three cells, even
though they are supposed to be no longer stable. It will be an interesting
exercise and will only cost me a bit of time. If the cells prove to be
unstable, I can always toss them later. I certainly won't be using them as
any type of standard.

I already acquired a Fluke 845A. It needs some attention, but seems to
mostly work.

Joe Gray
W5JG



On Mon, Aug 26, 2013 at 9:10 PM, J. L. Trantham jlt...@att.net wrote:

 Joe,

 Do you have any data about prior measurements of these cells or their
 history?

 I read somewhere there is a predicted decline rate based on time and
 temperature IIRC.

 Once you get the meter calibrated, you can 'track' their measurements.

 Might also be time for a 'differential voltmeter' to add to the 'stable'.

 Joe

 -Original Message-
 From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On
 Behalf Of Joseph Gray
 Sent: Monday, August 26, 2013 9:59 PM
 To: Discussion of precise voltage measurement
 Subject: [volt-nuts] Old Weston cells

 I got my hands on a few very old, unsaturated cells. I figured that they
 would be no good after all this time. Just out of curiosity, I measured
 them
 on my 3457A, which hasn't been sent off for calibration yet.

 DMM was set to 3 V scale to keep the input impedance high. Room temperature
 = 26 C.

 Cell #1 = 1.018288
 Cell #2 = 1.018236
 Cell #3 = 1.018224

 I read that when they get below 1.0183 V, they are no good. Is there any
 (easy) way to rejuvenate these cells?

 Joe Gray
 W5JG
 ___
 volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.

 ___
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.


___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Old Weston cells

2013-08-26 Thread Joseph Gray
I suspected that there was nothing to do to rejuvenate these old cells. I
thought I'd ask anyway.

Thanks.

Joe Gray
W5JG



On Mon, Aug 26, 2013 at 9:16 PM, zbigniew169 zbigniew...@gmail.com wrote:

 All my knowledge regarding Weston Cells tells me that is not possible in
 any way :-((


 2013/8/27 Joseph Gray jg...@zianet.com

  I got my hands on a few very old, unsaturated cells. I figured that they
  would be no good after all this time. Just out of curiosity, I measured
  them on my 3457A, which hasn't been sent off for calibration yet.
 
  DMM was set to 3 V scale to keep the input impedance high. Room
 temperature
  = 26 C.
 
  Cell #1 = 1.018288
  Cell #2 = 1.018236
  Cell #3 = 1.018224
 
  I read that when they get below 1.0183 V, they are no good. Is there any
  (easy) way to rejuvenate these cells?
 
  Joe Gray
  W5JG
  ___
  volt-nuts mailing list -- volt-nuts@febo.com
  To unsubscribe, go to
  https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
  and follow the instructions there.
 



 --
 Zbyszek
 ___
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.


___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.