Re: [volt-nuts] 732A and Prologix received

2014-08-31 Thread Don@True-Cal
Charlie,

Did my reply to your question along with my lab temperature graph link come 
through? Just checking as there has been some issues with the posts.

I don't like the feature where you don't see your own posts.

Don

-Original Message-
From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of Charles Black
Sent: Friday, August 29, 2014 3:44 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] 732A and Prologix received

Hi Don,

I'm a bit of trouble sending you this message for some reason. Maybe the third 
ti me is the charm.

Thanks for sharing the graph a few days ago. My 3458A graphs look about the 
same except I only do voltage so far. I use NPLC 1000 which looks smoother but 
doesn't really have any additional useful info. I have been meaning to have my 
TEMP? servoed  some time in the near future. My TEMP? 
runs about 37C at a room temp of 23C. Is your room temp a bit higher?

Charlie


On 8/29/2014 1:36 PM, Charles Black wrote:
 ___
 volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to 
 https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.





___
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Re: [volt-nuts] 732A and Prologix received

2014-08-31 Thread Charles Black

Hi Don,

Your post and link are fine but I need to get some time today to look at 
it more carefully.


I had two posts recently that were blank but since then all seems back 
to normal.


Charlie


On 8/31/2014 7:47 AM, Don@True-Cal wrote:

Charlie,

Did my reply to your question along with my lab temperature graph link come 
through? Just checking as there has been some issues with the posts.

I don't like the feature where you don't see your own posts.

Don

-Original Message-
From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of Charles Black
Sent: Friday, August 29, 2014 3:44 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] 732A and Prologix received

Hi Don,

I'm a bit of trouble sending you this message for some reason. Maybe the third 
ti me is the charm.

Thanks for sharing the graph a few days ago. My 3458A graphs look about the 
same except I only do voltage so far. I use NPLC 1000 which looks smoother but 
doesn't really have any additional useful info. I have been meaning to have my 
TEMP? servoed  some time in the near future. My TEMP?
runs about 37C at a room temp of 23C. Is your room temp a bit higher?

Charlie


On 8/29/2014 1:36 PM, Charles Black wrote:

___
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.





___
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.






___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Re: [volt-nuts] 732A and Prologix received

2014-08-30 Thread Don@True-Cal
Hi Charlie,

Yes, in the summer time my lab runs between ~24.5  26.5 C with an occasional 
excursion beyond either of those limits on extreme days. It is a basement 
(walkout type) lab with ceramic tile floor which helps account for the muted 
extremes, and also not having an extreme hot or cold attic overhead. See the 
link for a lab temperature graph over several days including the last day of 
the summer voltage run I provided earlier. Since the 3458A accuracy is +/- ~0.5 
PPM over typical lab temp extremes, I maintain a database of lab temp, 3458A 
ref 10.0V readings and 3458A internal temp so I can always refer back if doing 
a critical voltage transfer. The computer program not only does the graphing 
but also logs the raw data. While logging, I use lower (100 or 200) NPLC 
settings so my raw data resolution provides more detail for other types of post 
processing. But even when doing a voltage transfer, I never see an advantage, 
and don't use an NPLC setting over 300 (5-min). I would love to maintain the 
standard 23C lab temp but in the summer months here in the Midwest US, that 
gets too expensive. Winter temperatures are centered fairly close to 23C. The 
lab temperature graph link shows four PRT sensors, one is simply in open air 
thus subject to the same air currents as all instruments while the remaining 
three are in two different temperature dry-wells (powered off of course) that 
contribute considerable thermal mass.

https://www.dropbox.com/s/r5vl3v0k3s9t8n1/Lab%20Temp%208-22%20to%208-30%202014.pdf?dl=0

Don

-Original Message-
From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of Charles Black
Sent: Friday, August 29, 2014 3:44 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] 732A and Prologix received

Hi Don,

I'm a bit of trouble sending you this message for some reason. Maybe the third 
ti me is the charm.

Thanks for sharing the graph a few days ago. My 3458A graphs look about the 
same except I only do voltage so far. I use NPLC 1000 which looks smoother but 
doesn't really have any additional useful info. I have been meaning to have my 
TEMP? servoed  some time in the near future. My TEMP? 
runs about 37C at a room temp of 23C. Is your room temp a bit higher?

Charlie


On 8/29/2014 1:36 PM, Charles Black wrote:
 ___
 volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to 
 https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.





___
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Re: [volt-nuts] 732A and Prologix received

2014-08-30 Thread Don@True-Cal
Woops sorry, meant to type don't use an NPLC setting over 300 (5-sec) not 
5-min. Actually, with AZERO=on, I guess that's 5 seconds of integration over 10 
seconds or time.

Don

-Original Message-
From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of Charles Black
Sent: Friday, August 29, 2014 3:44 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] 732A and Prologix received

Hi Don,

I'm a bit of trouble sending you this message for some reason. Maybe the third 
ti me is the charm.

Thanks for sharing the graph a few days ago. My 3458A graphs look about the 
same except I only do voltage so far. I use NPLC 1000 which looks smoother but 
doesn't really have any additional useful info. I have been meaning to have my 
TEMP? servoed  some time in the near future. My TEMP? 
runs about 37C at a room temp of 23C. Is your room temp a bit higher?

Charlie


On 8/29/2014 1:36 PM, Charles Black wrote:
 ___
 volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to 
 https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.





___
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Re: [volt-nuts] 732A and Prologix received

2014-08-29 Thread Charles Black

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Re: [volt-nuts] 732A and Prologix received

2014-08-29 Thread Charles Black

Hi Don,

I'm a bit of trouble sending you this message for some reason. Maybe the 
third ti me is the charm.


Thanks for sharing the graph a few days ago. My 3458A graphs look about 
the same except I only do voltage so far. I use NPLC 1000 which looks 
smoother but doesn't really have any additional useful info. I have been 
meaning to have my TEMP? servoed  some time in the near future. My TEMP? 
runs about 37C at a room temp of 23C. Is your room temp a bit higher?


Charlie


On 8/29/2014 1:36 PM, Charles Black wrote:

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.






___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Re: [volt-nuts] 732A and Prologix received

2014-08-28 Thread Poul-Henning Kamp

In message CAPmQ2J_Fa1Oo6m=doj-u05f4dqhgpakdvjpsvggyrofe6tj...@mail.gmail.com
, Stan Katz writes:

*real* metrology purists keep their instruments in a pre helium atmosphere
to improve cooling, prevent corrosion and stop people from breathning on
things :-)

-- 
Poul-Henning Kamp   | UNIX since Zilog Zeus 3.20
p...@freebsd.org | TCP/IP since RFC 956
FreeBSD committer   | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Re: [volt-nuts] 732A and Prologix received

2014-08-28 Thread acbern
I do fully agree with you. this has become, to a big extent, an academic 
debate. it may be relavant if the resolution needs to be in the nv-range, such 
as working with josephson-elements. I guess nobody in this group here is.
to your question, yes, the 34420a and the 2182 nanovolt meters use pure copper 
contacts, and from time to time they should be cleaned with deoxit. I have not 
seen any issue with these connectors. would I use unplated (non-gold-surface) 
copper connectors if I had a choice. never. did i ever have an emf problem with 
my gold plated copper spades in use together with the 34420a and 3458a. no. so 
for me at least the answer is simple.
 

 Gesendet: Mittwoch, 27. August 2014 um 23:50 Uhr
 Von: Stan Katz stan.katz...@gmail.com
 An: Discussion of precise voltage measurement volt-nuts@febo.com
 Betreff: Re: [volt-nuts] 732A and Prologix received

 As a non-scholar in metrology, I tend to want to simplify the results of an
 academic debate to make the results of the debate useful to me. One thing
 is clear from my web search, copper alloyed with Tellurium, or Beryllium,
 still oxidizes, only at a slower rate. It appears that a big disadvantage
 of Beryllium oxide is its very hard, a useful industrial characteristic.
 Not a nice property in the metrology lab.
 
  Here goes:
 
 For the purists with lab grade metallurgical abrasives, polishing
 equipment, and oxide removal chemistry, only, copper, whether alloyed with
 Beryllium or Tellurium is all they want to see in their lead terminations,
 and they will try to procure only instruments with copper/copper alloy
 terminals. These purists will clean/deoxidize  all connections in a
 controlled atmosphere, and dry them in completely dry nitrogen, or other
 inert gas. The connections will then be used immediately during the
 measurement and the purists will have assured themselves that all
 instrument to lead connections are tight enough to be oxygen free.
 
 Can anyone mention any precision metrology instrumentation in current
 production with pure copper, or copper alloy connectors? I have a 38 year
 old Hp 740b, and yep, it's got Gold flashed connectors. This Gold flashing
 seems to be a tradition.
 
 The practical metrologist, will accept the fact that precision metrology
 instruments are meant to last many years, and the terminals supplied with
 these instruments must provide a stable thermal emf profile over time.
 Thus, they accept instruments that come with gold plated terminals. The
 Gold may need to be cleansed of debris, and degreased from time-to-time,
 but the procedure is much simpler, and not as time consuming as oxide
 removal.  The leads to these instruments are meant to be used day in, and
 day out, as well. Therefore, the leads also terminate in some form of gold
 plated connector. The practical metrologist is fastidious in controlling
 temperature, and air movement in the lab, to minimize thermal unbalance in
 his/her lash up.
 
 Is this a foolish simplification of the thermals debate, or can I feel
 vindicated using my homemade, gold plated lead terminations with my old
 740b, and 731b?
 
 
 On Wed, Aug 27, 2014 at 3:45 PM, M K m1k...@hotmail.com wrote:
 
  On 26/08/2014 16:05, Mike S wrote:
 
  After some more research, I think I've answered some of my own questions -
 
  Tellurium copper is used for binding posts, not because it has any
  special thermal or EMF mojo, but because it machines much better than pure
  copper. And, I suppose, because it sounds like it's extra special.
 
  The Seebeck coefficients (uV/C, relative to Cu) of some relevant
  materials:
  Cu 0.0
  Ag .2
  Au .5
  Yellow brass 1.5
  Phosphor bronze 2.0
  63/37 solder 3.0
  Sn 3.1
  Stainless steel 3.1
  Beryllium copper 5.0
  Fe -12.3
  Ni  22.3
  Te -49.25
 
  Based on the extreme Seebeck coefficient of pure tellurium vs. copper,
  I'd expect that there might be some coefficient between Cu and CuTe (0.5%
  Te), but I could find no reference. The relatively large number for CuBe is
  interesting, since that's a common material for banana plug springs, where
  one might expect the greatest temperature differential to occur in such a
  connection (between the thermal masses of the binding post/jack and the
  bulk of the banana plug). Heat has to flow a considerable distance through
  the springs, very much more than when it flows through a surface plating.
 
  The Pomona (Fluke) EM5295-48-0# uses CuBe (gold plated) for the spring
  contacts. It seems there might be an improvement to be had by using the
  older style pin plugs, where a solid pin was partially sliced into 4
  sections which were then spread apart a bit to create tension. That could
  eliminate relatively large thermocouples at a thermal gradient, and might
  also be expected to have less thermal resistance, allowing the connection
  to settle quicker.
 
  But maybe not - I'm still not clear on how plated conductors behave in
  this situation. For a high impedance voltage measurement

Re: [volt-nuts] 732A and Prologix received

2014-08-28 Thread Charles Black

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Re: [volt-nuts] 732A and Prologix received

2014-08-27 Thread Randy Evans
I am doing multiple 100 measurements simply to characterize the stability
of the 3458A and 732A units I just bought.  After about 10 measurement sets
over 2 days I am seeing a variance of about .5 uV for the 10V output, or
0.05 ppm.  However, the mean varies over a range of 10 uV, or 1 ppm.  Does
that sound reasonable/

Randy


On Tue, Aug 26, 2014 at 5:41 PM, acb...@gmx.de wrote:

 hi randy,

 just for curiosity, why doing 100 measurements at nplc 1000. is this to
 sample a changing value?
 when i am doing 10 measurements from a stable signal at nplc 100 (only
 there many subsequent measuremnts with statistics make sense) I am already
 getting a stanard deviation below 0.1ppm.
 in a 30 minute test cycle, i would also be concerned about drifts (acal)
 unless the amb. temperature is really very stable (half a degree already
 adds about 0.25ppm at 10v)

 thanks




  Gesendet: Dienstag, 26. August 2014 um 04:23 Uhr
  Von: Randy Evans randyevans2...@gmail.com
  An: Discussion of precise voltage measurement volt-nuts@febo.com
  Betreff: Re: [volt-nuts] 732A and Prologix received
 
  Is there any way to tell when the function key routine is complete?  In
 the
  case of taking multiple readings using the DEFKEY and MATH function, I
  don't see any indication when the routine is complete.  In one particular
  case, I am taking a 100 readings with NLPC set for 1000 so its a long
 while
  before it's complete, but i have to guess when it's done.
 
  Thanks,
 
  Randy
 
 
  On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans randyevans2...@gmail.com
  wrote:
 
   Bill,
  
   I am trying to figure out the MATH function without much success.  I
 input
   the sequence you said (I looked up the instructions to understand what
 you
   did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG
 4;TRIG; and
   it shows up on the display when I input BLUE F1.  I hit ENTER and it
 takes
   the 40 measurements and the MATH symbol shows on the display during the
   measurements.  After the SMPL symbol no longer blinks I hit MATH 2 and
 I
   get a MATH ERR symbol on the display.  I tried it a couple of times
 and the
   same result so I am doing something wrong.  Is there a better source
 for
   explaining how to do front panel masurements than the User Guide, which
   seems oriented at programming automatic rather than manual
 measurements.
  
   Randy
  
  
  
   On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold wpgold3...@att.net wrote:
  
   Randy:
  
   The MATH function is accessible from the keypad.  I don't have an
 IEEE
   interface right now that works.  You can also program the numeric
 keypad
   keys to have preprogrammed functions.  DEFKEY
  
   I have made my own low thermal measurement leads from Pomona
 #4892
   banana plugs and Belden #9272 wire.  Why 9272, because it was handy
 at the
   time.  It is tin plated copper, shielded twisted pair 20 ga.  I have
 plans
   to do custom cables with 16 ga. bare copper wire that I will twist and
   then
   put a braided shield over it.  I simply cannot find what I want so I
 will
   build my own cable.  I have done something like this before and it
 worked
   fine.  When I get a round toit.
  
   I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I
 have
   used
   in the past to verify my homemade low thermal leads as described
 above.
   Frankly I cannot see any difference between using the 1756 cables and
 my
   homemade cables once I give them a few minutes for the thermals to go
   away.
   As far as I can tell and measure the differences, if any, are below
 0.1
   ppm
   at 10 volts.
  
   Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A
 are all
   adjustable you may be seeing a misadjusted 1 volt from the 732A.  As
 far
   as
   the instability of the readings it is hard to determine which is
 causing
   the
   problem.  I have programed (DEFKEY) a numeric keypad key #1 with the
   following code.  MATH 14;NRDGS 40;TRIG 4;TRIG ;   So what this does
 is
   set
   the MATH to Statistics (store high reading/low reading/ and mean of
 the
   readings) in the registers, the number of readings to 40, the
 trigger to
   hold (which keeps the meter from triggering until I press ENTER
 and
   then
   trigger the sequence of 40 readings when I push the ENTER button.
 You
   can
   do all of this manually from the keypads but since I use this
 sequence a
   lot
   I have preprogrammed it.  This is after I set digits to 8 and PLC to
   100.
   Once those 40 readings are finished then you can access the various
 MATH
   statistic registers, using the menu, by entering MATH and then a 2
 for
   low, a 4 for mean, and 13 for high.  Of course you could do all of
 this
   through the IEEE also.  The 3458A has a very rich set of measurement
   commands.  I am still learning all of them.  It depends upon what I am
   trying to accomplish.
  
   Since the 1.018 and 1.0 volt outputs are passive and derived from
   resistive dividers from the 10 volt

Re: [volt-nuts] 732A and Prologix received

2014-08-27 Thread Randy Evans
I forgot to mention that I reduced the number of measurements to 100 per
set since I wasn't seeing much difference in the variance between 100 and
1000 measurements and the 1000 measurement per set takes too long.

Randy


On Wed, Aug 27, 2014 at 5:13 AM, Randy Evans randyevans2...@gmail.com
wrote:

 I am doing multiple 100 measurements simply to characterize the stability
 of the 3458A and 732A units I just bought.  After about 10 measurement sets
 over 2 days I am seeing a variance of about .5 uV for the 10V output, or
 0.05 ppm.  However, the mean varies over a range of 10 uV, or 1 ppm.  Does
 that sound reasonable/

 Randy


 On Tue, Aug 26, 2014 at 5:41 PM, acb...@gmx.de wrote:

 hi randy,

 just for curiosity, why doing 100 measurements at nplc 1000. is this to
 sample a changing value?
 when i am doing 10 measurements from a stable signal at nplc 100 (only
 there many subsequent measuremnts with statistics make sense) I am already
 getting a stanard deviation below 0.1ppm.
 in a 30 minute test cycle, i would also be concerned about drifts (acal)
 unless the amb. temperature is really very stable (half a degree already
 adds about 0.25ppm at 10v)

 thanks




  Gesendet: Dienstag, 26. August 2014 um 04:23 Uhr
  Von: Randy Evans randyevans2...@gmail.com
  An: Discussion of precise voltage measurement volt-nuts@febo.com
  Betreff: Re: [volt-nuts] 732A and Prologix received
 
  Is there any way to tell when the function key routine is complete?  In
 the
  case of taking multiple readings using the DEFKEY and MATH function, I
  don't see any indication when the routine is complete.  In one
 particular
  case, I am taking a 100 readings with NLPC set for 1000 so its a long
 while
  before it's complete, but i have to guess when it's done.
 
  Thanks,
 
  Randy
 
 
  On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans randyevans2...@gmail.com
  wrote:
 
   Bill,
  
   I am trying to figure out the MATH function without much success.  I
 input
   the sequence you said (I looked up the instructions to understand
 what you
   did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG
 4;TRIG; and
   it shows up on the display when I input BLUE F1.  I hit ENTER and it
 takes
   the 40 measurements and the MATH symbol shows on the display during
 the
   measurements.  After the SMPL symbol no longer blinks I hit MATH 2
 and I
   get a MATH ERR symbol on the display.  I tried it a couple of times
 and the
   same result so I am doing something wrong.  Is there a better source
 for
   explaining how to do front panel masurements than the User Guide,
 which
   seems oriented at programming automatic rather than manual
 measurements.
  
   Randy
  
  
  
   On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold wpgold3...@att.net
 wrote:
  
   Randy:
  
   The MATH function is accessible from the keypad.  I don't have
 an IEEE
   interface right now that works.  You can also program the numeric
 keypad
   keys to have preprogrammed functions.  DEFKEY
  
   I have made my own low thermal measurement leads from Pomona
 #4892
   banana plugs and Belden #9272 wire.  Why 9272, because it was handy
 at the
   time.  It is tin plated copper, shielded twisted pair 20 ga.  I have
 plans
   to do custom cables with 16 ga. bare copper wire that I will twist
 and
   then
   put a braided shield over it.  I simply cannot find what I want so I
 will
   build my own cable.  I have done something like this before and it
 worked
   fine.  When I get a round toit.
  
   I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I
 have
   used
   in the past to verify my homemade low thermal leads as described
 above.
   Frankly I cannot see any difference between using the 1756 cables
 and my
   homemade cables once I give them a few minutes for the thermals to go
   away.
   As far as I can tell and measure the differences, if any, are below
 0.1
   ppm
   at 10 volts.
  
   Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A
 are all
   adjustable you may be seeing a misadjusted 1 volt from the 732A.  As
 far
   as
   the instability of the readings it is hard to determine which is
 causing
   the
   problem.  I have programed (DEFKEY) a numeric keypad key #1 with the
   following code.  MATH 14;NRDGS 40;TRIG 4;TRIG ;   So what this
 does is
   set
   the MATH to Statistics (store high reading/low reading/ and mean
 of the
   readings) in the registers, the number of readings to 40, the
 trigger to
   hold (which keeps the meter from triggering until I press ENTER
 and
   then
   trigger the sequence of 40 readings when I push the ENTER button.
 You
   can
   do all of this manually from the keypads but since I use this
 sequence a
   lot
   I have preprogrammed it.  This is after I set digits to 8 and PLC
 to
   100.
   Once those 40 readings are finished then you can access the various
 MATH
   statistic registers, using the menu, by entering MATH and then a 2
 for
   low, a 4 for mean, and 13 for high

Re: [volt-nuts] 732A and Prologix received

2014-08-27 Thread acbern
well, what you do is to measure the stability of the 3458a. 1ppm drift though 
sounds much as overall short term average variation if your temperature is 
stable and your 3458a is always on. probably the temp is not stable, and that 
is what you see, amongst potentially some other smaller drifts like emf. before 
every cycle you should do an acal dcv, if you still see 10uv +/- drifts that 
then seems too much. (one thing I need to add is that one of my 3458a, not yet 
modified to have lower reference temperature, has drifts when switched on and 
off. that is not so much the case with modified reference. but my assumption re 
the above is that your meter is always on, as I said) 
732a references et al do have small short term drifts, you can determine them 
with a josephson element (these guys told me), but the 732a is certainly very 
stable short term compared to a 3458a. my 732a e.g. has a drift of 0.2ppm per 
year.


 Gesendet: Mittwoch, 27. August 2014 um 14:36 Uhr
 Von: Randy Evans randyevans2...@gmail.com
 An: Discussion of precise voltage measurement volt-nuts@febo.com
 Betreff: Re: [volt-nuts] 732A and Prologix received

 I forgot to mention that I reduced the number of measurements to 100 per
 set since I wasn't seeing much difference in the variance between 100 and
 1000 measurements and the 1000 measurement per set takes too long.
 
 Randy
 
 
 On Wed, Aug 27, 2014 at 5:13 AM, Randy Evans randyevans2...@gmail.com
 wrote:
 
  I am doing multiple 100 measurements simply to characterize the stability
  of the 3458A and 732A units I just bought.  After about 10 measurement sets
  over 2 days I am seeing a variance of about .5 uV for the 10V output, or
  0.05 ppm.  However, the mean varies over a range of 10 uV, or 1 ppm.  Does
  that sound reasonable/
 
  Randy
 
 
  On Tue, Aug 26, 2014 at 5:41 PM, acb...@gmx.de wrote:
 
  hi randy,
 
  just for curiosity, why doing 100 measurements at nplc 1000. is this to
  sample a changing value?
  when i am doing 10 measurements from a stable signal at nplc 100 (only
  there many subsequent measuremnts with statistics make sense) I am already
  getting a stanard deviation below 0.1ppm.
  in a 30 minute test cycle, i would also be concerned about drifts (acal)
  unless the amb. temperature is really very stable (half a degree already
  adds about 0.25ppm at 10v)
 
  thanks
 
 
 
 
   Gesendet: Dienstag, 26. August 2014 um 04:23 Uhr
   Von: Randy Evans randyevans2...@gmail.com
   An: Discussion of precise voltage measurement volt-nuts@febo.com
   Betreff: Re: [volt-nuts] 732A and Prologix received
  
   Is there any way to tell when the function key routine is complete?  In
  the
   case of taking multiple readings using the DEFKEY and MATH function, I
   don't see any indication when the routine is complete.  In one
  particular
   case, I am taking a 100 readings with NLPC set for 1000 so its a long
  while
   before it's complete, but i have to guess when it's done.
  
   Thanks,
  
   Randy
  
  
   On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans randyevans2...@gmail.com
   wrote:
  
Bill,
   
I am trying to figure out the MATH function without much success.  I
  input
the sequence you said (I looked up the instructions to understand
  what you
did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG
  4;TRIG; and
it shows up on the display when I input BLUE F1.  I hit ENTER and it
  takes
the 40 measurements and the MATH symbol shows on the display during
  the
measurements.  After the SMPL symbol no longer blinks I hit MATH 2
  and I
get a MATH ERR symbol on the display.  I tried it a couple of times
  and the
same result so I am doing something wrong.  Is there a better source
  for
explaining how to do front panel masurements than the User Guide,
  which
seems oriented at programming automatic rather than manual
  measurements.
   
Randy
   
   
   
On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold wpgold3...@att.net
  wrote:
   
Randy:
   
The MATH function is accessible from the keypad.  I don't have
  an IEEE
interface right now that works.  You can also program the numeric
  keypad
keys to have preprogrammed functions.  DEFKEY
   
I have made my own low thermal measurement leads from Pomona
  #4892
banana plugs and Belden #9272 wire.  Why 9272, because it was handy
  at the
time.  It is tin plated copper, shielded twisted pair 20 ga.  I have
  plans
to do custom cables with 16 ga. bare copper wire that I will twist
  and
then
put a braided shield over it.  I simply cannot find what I want so I
  will
build my own cable.  I have done something like this before and it
  worked
fine.  When I get a round toit.
   
I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I
  have
used
in the past to verify my homemade low thermal leads as described
  above.
Frankly I cannot see any difference between using the 1756 cables

Re: [volt-nuts] 732A and Prologix received

2014-08-27 Thread Bill Gold
Randy:

Something I had never tried to measure.  As I have found out in the past
there is a lot of overhead going on in the meter during and after a
measurement.

In thinking about this I turned OFF the autozero AZERO and the time
for each SMPL was cut in half to around the estimated 16.66 seconds for
1000 PLC.  So it becomes obvious that the meter makes an autozero
measurement for 1000 PLC and then the actual measurement for 1000 PLC which
explains the 33 seconds.  Makes sense.  This is probably why the AZERO menu
gives you ON, OFF and ONCE.  For short measurement sequences you just
autozero ONCE at the start.

Thanks for the observation, this helps me.  Everyday I learn something
new.

Bill

Thanks for the information
- Original Message - 
From: Randy Evans randyevans2...@gmail.com
To: Discussion of precise voltage measurement volt-nuts@febo.com
Sent: Tuesday, August 26, 2014 2:16 PM
Subject: Re: [volt-nuts] 732A and Prologix received


 Bill,

 I measured the time between SMPL symbols with NPLC set to 1000 and it is
 approximately 33 seconds. It takes an hour to complete 100 readings.

 Randy


 On Mon, Aug 25, 2014 at 10:01 PM, Bill Gold wpgold3...@att.net wrote:

  Not that I know of.  Just wait and when you don't see a SMPL on the
  display it is done.  But then with 1000 PLC that is around 16.66 seconds
  per
  reading times 100 readings is somewhere around 28 minutes and there is
  probably some overhead time so around 30 minutes.  Not from the front
panel
  at any rate.
 
  Bill
 
  - Original Message -
  From: Randy Evans randyevans2...@gmail.com
  To: Discussion of precise voltage measurement volt-nuts@febo.com
  Sent: Monday, August 25, 2014 9:23 PM
  Subject: Re: [volt-nuts] 732A and Prologix received
 
 
   Is there any way to tell when the function key routine is complete?
In
  the
   case of taking multiple readings using the DEFKEY and MATH function, I
   don't see any indication when the routine is complete.  In one
particular
   case, I am taking a 100 readings with NLPC set for 1000 so its a long
  while
   before it's complete, but i have to guess when it's done.
  
   Thanks,
  
   Randy
  
  
   On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans
randyevans2...@gmail.com
   wrote:
  
Bill,
   
I am trying to figure out the MATH function without much success.  I
  input
the sequence you said (I looked up the instructions to understand
what
  you
did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG
4;TRIG;
  and
it shows up on the display when I input BLUE F1.  I hit ENTER and it
  takes
the 40 measurements and the MATH symbol shows on the display during
the
measurements.  After the SMPL symbol no longer blinks I hit MATH 2
and
  I
get a MATH ERR symbol on the display.  I tried it a couple of times
and
  the
same result so I am doing something wrong.  Is there a better source
  for
explaining how to do front panel masurements than the User Guide,
which
seems oriented at programming automatic rather than manual
  measurements.
   
Randy
   
   
   
On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold wpgold3...@att.net
wrote:
   
Randy:
   
The MATH function is accessible from the keypad.  I don't have
an
  IEEE
interface right now that works.  You can also program the numeric
  keypad
keys to have preprogrammed functions.  DEFKEY
   
I have made my own low thermal measurement leads from Pomona
  #4892
banana plugs and Belden #9272 wire.  Why 9272, because it was handy
at
  the
time.  It is tin plated copper, shielded twisted pair 20 ga.  I
have
  plans
to do custom cables with 16 ga. bare copper wire that I will twist
and
then
put a braided shield over it.  I simply cannot find what I want so
I
  will
build my own cable.  I have done something like this before and it
  worked
fine.  When I get a round toit.
   
I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I
  have
used
in the past to verify my homemade low thermal leads as described
  above.
Frankly I cannot see any difference between using the 1756 cables
and
  my
homemade cables once I give them a few minutes for the thermals to
go
away.
As far as I can tell and measure the differences, if any, are below
  0.1
ppm
at 10 volts.
   
Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A
are
  all
adjustable you may be seeing a misadjusted 1 volt from the 732A.
As
  far
as
the instability of the readings it is hard to determine which is
  causing
the
problem.  I have programed (DEFKEY) a numeric keypad key #1 with
the
following code.  MATH 14;NRDGS 40;TRIG 4;TRIG ;   So what this
does
  is
set
the MATH to Statistics (store high reading/low reading/ and mean
of
  the
readings) in the registers, the number of readings to 40, the
  trigger
  to
hold (which keeps the meter from triggering until I

Re: [volt-nuts] 732A and Prologix received

2014-08-27 Thread Don@True-Cal
I have included a link to an ~30-hour measurement run that I consistently do 
that should give you some idea of expected measurement drift over time. 
Virtually all of the measurement drift is due to the 3458A internal temperature 
differences around the 7V reference caused by ambient temperature change. The 
drift associated with the 732A is probably about 2-magnitudes less at this 
ambient temp drift. In my situation, the inferred ambient temperature is 
cycling with the home air-conditioning. Note the initial calibration 
temperatures as well as the ACal temperatures and where the 3458A is measuring 
exactly 10V relative to those temperatures. My primary 732A has been powered 
without loss for 4 years and 5 years before that. The 3458A is Agilent 
(Loveland) Cal'ed yearly and is powered 24-7-365 except for the occasional 
mains power loss. The graphical measurements is using a homegrown Agilent VEE 
program. It is very helpful if not essential to get an HPIB interface setup so 
you can do long term graphical analysis. 

The other link was done in winter time when the lab a few degrees cooler.

https://www.dropbox.com/s/8je482i3r0belqn/732A%20gold%20%26%203458A%20gold%2032-hour%20Ref%20Test%208-21-2014.pdf?dl=0

https://www.dropbox.com/s/xv3l9py7tx6hyh7/HP%20732A%20gold%20%26%203458A%20gold%207-day%2010.0v%20Ref%20Test.pdf?dl=0

I hope these Dropbox links work internationally - I'm new to using this sharing 
method.

Don Johnson

-Original Message-
From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of acb...@gmx.de
Sent: Wednesday, August 27, 2014 8:42 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] 732A and Prologix received

well, what you do is to measure the stability of the 3458a. 1ppm drift though 
sounds much as overall short term average variation if your temperature is 
stable and your 3458a is always on. probably the temp is not stable, and that 
is what you see, amongst potentially some other smaller drifts like emf. before 
every cycle you should do an acal dcv, if you still see 10uv +/- drifts that 
then seems too much. (one thing I need to add is that one of my 3458a, not yet 
modified to have lower reference temperature, has drifts when switched on and 
off. that is not so much the case with modified reference. but my assumption re 
the above is that your meter is always on, as I said) 732a references et al do 
have small short term drifts, you can determine them with a josephson element 
(these guys told me), but the 732a is certainly very stable short term compared 
to a 3458a. my 732a e.g. has a drift of 0.2ppm per year.


 Gesendet: Mittwoch, 27. August 2014 um 14:36 Uhr
 Von: Randy Evans randyevans2...@gmail.com
 An: Discussion of precise voltage measurement volt-nuts@febo.com
 Betreff: Re: [volt-nuts] 732A and Prologix received

 I forgot to mention that I reduced the number of measurements to 100 per
 set since I wasn't seeing much difference in the variance between 100 and
 1000 measurements and the 1000 measurement per set takes too long.
 
 Randy
 
 
 On Wed, Aug 27, 2014 at 5:13 AM, Randy Evans randyevans2...@gmail.com
 wrote:
 
  I am doing multiple 100 measurements simply to characterize the stability
  of the 3458A and 732A units I just bought.  After about 10 measurement sets
  over 2 days I am seeing a variance of about .5 uV for the 10V output, or
  0.05 ppm.  However, the mean varies over a range of 10 uV, or 1 ppm.  Does
  that sound reasonable/
 
  Randy
 
 
  On Tue, Aug 26, 2014 at 5:41 PM, acb...@gmx.de wrote:
 
  hi randy,
 
  just for curiosity, why doing 100 measurements at nplc 1000. is this to
  sample a changing value?
  when i am doing 10 measurements from a stable signal at nplc 100 (only
  there many subsequent measuremnts with statistics make sense) I am already
  getting a stanard deviation below 0.1ppm.
  in a 30 minute test cycle, i would also be concerned about drifts (acal)
  unless the amb. temperature is really very stable (half a degree already
  adds about 0.25ppm at 10v)
 
  thanks
 
 
 
 
   Gesendet: Dienstag, 26. August 2014 um 04:23 Uhr
   Von: Randy Evans randyevans2...@gmail.com
   An: Discussion of precise voltage measurement volt-nuts@febo.com
   Betreff: Re: [volt-nuts] 732A and Prologix received
  
   Is there any way to tell when the function key routine is complete?  In
  the
   case of taking multiple readings using the DEFKEY and MATH function, I
   don't see any indication when the routine is complete.  In one
  particular
   case, I am taking a 100 readings with NLPC set for 1000 so its a long
  while
   before it's complete, but i have to guess when it's done.
  
   Thanks,
  
   Randy
  
  
   On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans randyevans2...@gmail.com
   wrote:
  
Bill,
   
I am trying to figure out the MATH function without much success.  I
  input
the sequence you said (I looked up the instructions to understand
  what you
did - seems logical), BLUE DEFKEY

Re: [volt-nuts] 732A and Prologix received

2014-08-27 Thread Randy Evans
The HP3458A and the Fluke 732A are on continuously and I do an ACAL at
least every few hours, or when the room temperature changes by more than 1
degree C. The total range of measurements is 10uV so the drift is +/-5uV,
or 0.5 ppm.  The room temperature is not particularly stable and varies
over a 3C range.  The internal temp of the 3458 varies from 38.1 to 40.3
degrees C over the set of measurements.  The 732A thermistor resistance
measures from 3.6677 Kohms to  3.6686 Kohms.  I am using copper wires
between the 3458A and the 752A to minimize thermals.  At the moment I have
no way to tell which unit is drifting the most.

Randy


On Wed, Aug 27, 2014 at 6:41 AM, acb...@gmx.de wrote:

 well, what you do is to measure the stability of the 3458a. 1ppm drift
 though sounds much as overall short term average variation if your
 temperature is stable and your 3458a is always on. probably the temp is not
 stable, and that is what you see, amongst potentially some other smaller
 drifts like emf. before every cycle you should do an acal dcv, if you still
 see 10uv +/- drifts that then seems too much. (one thing I need to add is
 that one of my 3458a, not yet modified to have lower reference temperature,
 has drifts when switched on and off. that is not so much the case with
 modified reference. but my assumption re the above is that your meter is
 always on, as I said)
 732a references et al do have small short term drifts, you can determine
 them with a josephson element (these guys told me), but the 732a is
 certainly very stable short term compared to a 3458a. my 732a e.g. has a
 drift of 0.2ppm per year.


  Gesendet: Mittwoch, 27. August 2014 um 14:36 Uhr
  Von: Randy Evans randyevans2...@gmail.com
  An: Discussion of precise voltage measurement volt-nuts@febo.com
  Betreff: Re: [volt-nuts] 732A and Prologix received
 
  I forgot to mention that I reduced the number of measurements to 100 per
  set since I wasn't seeing much difference in the variance between 100 and
  1000 measurements and the 1000 measurement per set takes too long.
 
  Randy
 
 
  On Wed, Aug 27, 2014 at 5:13 AM, Randy Evans randyevans2...@gmail.com
  wrote:
 
   I am doing multiple 100 measurements simply to characterize the
 stability
   of the 3458A and 732A units I just bought.  After about 10 measurement
 sets
   over 2 days I am seeing a variance of about .5 uV for the 10V output,
 or
   0.05 ppm.  However, the mean varies over a range of 10 uV, or 1 ppm.
 Does
   that sound reasonable/
  
   Randy
  
  
   On Tue, Aug 26, 2014 at 5:41 PM, acb...@gmx.de wrote:
  
   hi randy,
  
   just for curiosity, why doing 100 measurements at nplc 1000. is this
 to
   sample a changing value?
   when i am doing 10 measurements from a stable signal at nplc 100 (only
   there many subsequent measuremnts with statistics make sense) I am
 already
   getting a stanard deviation below 0.1ppm.
   in a 30 minute test cycle, i would also be concerned about drifts
 (acal)
   unless the amb. temperature is really very stable (half a degree
 already
   adds about 0.25ppm at 10v)
  
   thanks
  
  
  
  
Gesendet: Dienstag, 26. August 2014 um 04:23 Uhr
Von: Randy Evans randyevans2...@gmail.com
An: Discussion of precise voltage measurement volt-nuts@febo.com
 
Betreff: Re: [volt-nuts] 732A and Prologix received
   
Is there any way to tell when the function key routine is
 complete?  In
   the
case of taking multiple readings using the DEFKEY and MATH
 function, I
don't see any indication when the routine is complete.  In one
   particular
case, I am taking a 100 readings with NLPC set for 1000 so its a
 long
   while
before it's complete, but i have to guess when it's done.
   
Thanks,
   
Randy
   
   
On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans 
 randyevans2...@gmail.com
wrote:
   
 Bill,

 I am trying to figure out the MATH function without much
 success.  I
   input
 the sequence you said (I looked up the instructions to understand
   what you
 did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG
   4;TRIG; and
 it shows up on the display when I input BLUE F1.  I hit ENTER and
 it
   takes
 the 40 measurements and the MATH symbol shows on the display
 during
   the
 measurements.  After the SMPL symbol no longer blinks I hit MATH 2
   and I
 get a MATH ERR symbol on the display.  I tried it a couple of
 times
   and the
 same result so I am doing something wrong.  Is there a better
 source
   for
 explaining how to do front panel masurements than the User Guide,
   which
 seems oriented at programming automatic rather than manual
   measurements.

 Randy



 On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold wpgold3...@att.net
   wrote:

 Randy:

 The MATH function is accessible from the keypad.  I don't
 have
   an IEEE
 interface right now that works.  You can also program the numeric
   keypad

Re: [volt-nuts] 732A and Prologix received

2014-08-27 Thread Randy Evans
Don,

If I am reading your charts correctly, it looks like your system is cycling
over an 7 uV range over several days in summer, similar to what I am
seeing.  I don't have air conditioning (Northern California) so I do get
some pretty good temperature variations from day to night, but less inside
the house.

I need to figure out how to share files in Dropbox since I don't have my
own web site.  Good idea.

thanks,

Randy



On Wed, Aug 27, 2014 at 7:40 AM, Don@True-Cal truecalservi...@gmail.com
wrote:

 I have included a link to an ~30-hour measurement run that I consistently
 do that should give you some idea of expected measurement drift over time.
 Virtually all of the measurement drift is due to the 3458A internal
 temperature differences around the 7V reference caused by ambient
 temperature change. The drift associated with the 732A is probably about
 2-magnitudes less at this ambient temp drift. In my situation, the inferred
 ambient temperature is cycling with the home air-conditioning. Note the
 initial calibration temperatures as well as the ACal temperatures and where
 the 3458A is measuring exactly 10V relative to those temperatures. My
 primary 732A has been powered without loss for 4 years and 5 years before
 that. The 3458A is Agilent (Loveland) Cal'ed yearly and is powered 24-7-365
 except for the occasional mains power loss. The graphical measurements is
 using a homegrown Agilent VEE program. It is very helpful if not essential
 to get an HPIB interface setup so you can do long term graphical analysis.

 The other link was done in winter time when the lab a few degrees cooler.


 https://www.dropbox.com/s/8je482i3r0belqn/732A%20gold%20%26%203458A%20gold%2032-hour%20Ref%20Test%208-21-2014.pdf?dl=0


 https://www.dropbox.com/s/xv3l9py7tx6hyh7/HP%20732A%20gold%20%26%203458A%20gold%207-day%2010.0v%20Ref%20Test.pdf?dl=0

 I hope these Dropbox links work internationally - I'm new to using this
 sharing method.

 Don Johnson

 -Original Message-
 From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of
 acb...@gmx.de
 Sent: Wednesday, August 27, 2014 8:42 AM
 To: Discussion of precise voltage measurement
 Subject: Re: [volt-nuts] 732A and Prologix received

 well, what you do is to measure the stability of the 3458a. 1ppm drift
 though sounds much as overall short term average variation if your
 temperature is stable and your 3458a is always on. probably the temp is not
 stable, and that is what you see, amongst potentially some other smaller
 drifts like emf. before every cycle you should do an acal dcv, if you still
 see 10uv +/- drifts that then seems too much. (one thing I need to add is
 that one of my 3458a, not yet modified to have lower reference temperature,
 has drifts when switched on and off. that is not so much the case with
 modified reference. but my assumption re the above is that your meter is
 always on, as I said) 732a references et al do have small short term
 drifts, you can determine them with a josephson element (these guys told
 me), but the 732a is certainly very stable short term compared to a 3458a.
 my 732a e.g. has a drift of 0.2ppm per year.


  Gesendet: Mittwoch, 27. August 2014 um 14:36 Uhr
  Von: Randy Evans randyevans2...@gmail.com
  An: Discussion of precise voltage measurement volt-nuts@febo.com
  Betreff: Re: [volt-nuts] 732A and Prologix received
 
  I forgot to mention that I reduced the number of measurements to 100 per
  set since I wasn't seeing much difference in the variance between 100 and
  1000 measurements and the 1000 measurement per set takes too long.
 
  Randy
 
 
  On Wed, Aug 27, 2014 at 5:13 AM, Randy Evans randyevans2...@gmail.com
  wrote:
 
   I am doing multiple 100 measurements simply to characterize the
 stability
   of the 3458A and 732A units I just bought.  After about 10 measurement
 sets
   over 2 days I am seeing a variance of about .5 uV for the 10V output,
 or
   0.05 ppm.  However, the mean varies over a range of 10 uV, or 1 ppm.
 Does
   that sound reasonable/
  
   Randy
  
  
   On Tue, Aug 26, 2014 at 5:41 PM, acb...@gmx.de wrote:
  
   hi randy,
  
   just for curiosity, why doing 100 measurements at nplc 1000. is this
 to
   sample a changing value?
   when i am doing 10 measurements from a stable signal at nplc 100 (only
   there many subsequent measuremnts with statistics make sense) I am
 already
   getting a stanard deviation below 0.1ppm.
   in a 30 minute test cycle, i would also be concerned about drifts
 (acal)
   unless the amb. temperature is really very stable (half a degree
 already
   adds about 0.25ppm at 10v)
  
   thanks
  
  
  
  
Gesendet: Dienstag, 26. August 2014 um 04:23 Uhr
Von: Randy Evans randyevans2...@gmail.com
An: Discussion of precise voltage measurement volt-nuts@febo.com
 
Betreff: Re: [volt-nuts] 732A and Prologix received
   
Is there any way to tell when the function key routine is
 complete?  In
   the
case of taking multiple readings using

Re: [volt-nuts] 732A and Prologix received

2014-08-27 Thread Don@True-Cal
Randy,

Sorry for the two graphs being at different scales so just be sure to readjust 
your reference. I like to think in PPM terms so the first graph is +/- 5 PPM 
for the whole gray plot area while the second is +/- 1 PPM. The most extreme 
outliers on the first graph is +0.3 to -0.5 PPM so that would be 10.0 uV. The 
winter graph is +0.1 to -0.9 PPM and would be 10.0 uV. All referenced around 
10.0 V.

Don

-Original Message-
From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of Randy Evans
Sent: Wednesday, August 27, 2014 10:50 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] 732A and Prologix received

Don,

If I am reading your charts correctly, it looks like your system is cycling 
over an 7 uV range over several days in summer, similar to what I am seeing.  I 
don't have air conditioning (Northern California) so I do get some pretty good 
temperature variations from day to night, but less inside the house.

I need to figure out how to share files in Dropbox since I don't have my own 
web site.  Good idea.

thanks,

Randy



On Wed, Aug 27, 2014 at 7:40 AM, Don@True-Cal truecalservi...@gmail.com
wrote:

 I have included a link to an ~30-hour measurement run that I 
 consistently do that should give you some idea of expected measurement drift 
 over time.
 Virtually all of the measurement drift is due to the 3458A internal 
 temperature differences around the 7V reference caused by ambient 
 temperature change. The drift associated with the 732A is probably 
 about 2-magnitudes less at this ambient temp drift. In my situation, 
 the inferred ambient temperature is cycling with the home 
 air-conditioning. Note the initial calibration temperatures as well as 
 the ACal temperatures and where the 3458A is measuring exactly 10V 
 relative to those temperatures. My primary 732A has been powered 
 without loss for 4 years and 5 years before that. The 3458A is 
 Agilent (Loveland) Cal'ed yearly and is powered 24-7-365 except for 
 the occasional mains power loss. The graphical measurements is using a 
 homegrown Agilent VEE program. It is very helpful if not essential to get an 
 HPIB interface setup so you can do long term graphical analysis.

 The other link was done in winter time when the lab a few degrees cooler.


 https://www.dropbox.com/s/8je482i3r0belqn/732A%20gold%20%26%203458A%20
 gold%2032-hour%20Ref%20Test%208-21-2014.pdf?dl=0


 https://www.dropbox.com/s/xv3l9py7tx6hyh7/HP%20732A%20gold%20%26%20345
 8A%20gold%207-day%2010.0v%20Ref%20Test.pdf?dl=0

 I hope these Dropbox links work internationally - I'm new to using 
 this sharing method.

 Don Johnson

 -Original Message-
 From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of 
 acb...@gmx.de
 Sent: Wednesday, August 27, 2014 8:42 AM
 To: Discussion of precise voltage measurement
 Subject: Re: [volt-nuts] 732A and Prologix received

 well, what you do is to measure the stability of the 3458a. 1ppm drift 
 though sounds much as overall short term average variation if your 
 temperature is stable and your 3458a is always on. probably the temp 
 is not stable, and that is what you see, amongst potentially some 
 other smaller drifts like emf. before every cycle you should do an 
 acal dcv, if you still see 10uv +/- drifts that then seems too much. 
 (one thing I need to add is that one of my 3458a, not yet modified to 
 have lower reference temperature, has drifts when switched on and off. 
 that is not so much the case with modified reference. but my 
 assumption re the above is that your meter is always on, as I said) 
 732a references et al do have small short term drifts, you can 
 determine them with a josephson element (these guys told me), but the 732a is 
 certainly very stable short term compared to a 3458a.
 my 732a e.g. has a drift of 0.2ppm per year.


  Gesendet: Mittwoch, 27. August 2014 um 14:36 Uhr
  Von: Randy Evans randyevans2...@gmail.com
  An: Discussion of precise voltage measurement volt-nuts@febo.com
  Betreff: Re: [volt-nuts] 732A and Prologix received
 
  I forgot to mention that I reduced the number of measurements to 100 
  per set since I wasn't seeing much difference in the variance 
  between 100 and
  1000 measurements and the 1000 measurement per set takes too long.
 
  Randy
 
 
  On Wed, Aug 27, 2014 at 5:13 AM, Randy Evans 
  randyevans2...@gmail.com
  wrote:
 
   I am doing multiple 100 measurements simply to characterize the
 stability
   of the 3458A and 732A units I just bought.  After about 10 
   measurement
 sets
   over 2 days I am seeing a variance of about .5 uV for the 10V 
   output,
 or
   0.05 ppm.  However, the mean varies over a range of 10 uV, or 1 ppm.
 Does
   that sound reasonable/
  
   Randy
  
  
   On Tue, Aug 26, 2014 at 5:41 PM, acb...@gmx.de wrote:
  
   hi randy,
  
   just for curiosity, why doing 100 measurements at nplc 1000. is 
   this
 to
   sample a changing value?
   when i am doing 10 measurements from a stable

Re: [volt-nuts] 732A and Prologix received

2014-08-27 Thread acbern
hi randy,
the specified drift of the 3458a over your 38.1 to 40.3 (about 1k) is 1ppm +/- 
allone in the 10v range. thats 10uv. in other ranges its worse.
unless your 732a is very bad (very unlikely), you measure mostly the 3458a 
temp. drift. 1000nplc and 100 readings average do not make sense in that 
context. if your goal is to be at 0.01ppm additional gain error by using the 
nplc1000, you need to be sure your temp related drift is even below that. 
0.01ppm of temp drift equates 20mk temp stability! you see that all this does 
not make much sense.


adrian


 Gesendet: Mittwoch, 27. August 2014 um 15:37 Uhr
 Von: Randy Evans randyevans2...@gmail.com
 An: Discussion of precise voltage measurement volt-nuts@febo.com
 Betreff: Re: [volt-nuts] 732A and Prologix received

 The HP3458A and the Fluke 732A are on continuously and I do an ACAL at
 least every few hours, or when the room temperature changes by more than 1
 degree C. The total range of measurements is 10uV so the drift is +/-5uV,
 or 0.5 ppm.  The room temperature is not particularly stable and varies
 over a 3C range.  The internal temp of the 3458 varies from 38.1 to 40.3
 degrees C over the set of measurements.  The 732A thermistor resistance
 measures from 3.6677 Kohms to  3.6686 Kohms.  I am using copper wires
 between the 3458A and the 752A to minimize thermals.  At the moment I have
 no way to tell which unit is drifting the most.
 
 Randy
 
 
 On Wed, Aug 27, 2014 at 6:41 AM, acb...@gmx.de wrote:
 
  well, what you do is to measure the stability of the 3458a. 1ppm drift
  though sounds much as overall short term average variation if your
  temperature is stable and your 3458a is always on. probably the temp is not
  stable, and that is what you see, amongst potentially some other smaller
  drifts like emf. before every cycle you should do an acal dcv, if you still
  see 10uv +/- drifts that then seems too much. (one thing I need to add is
  that one of my 3458a, not yet modified to have lower reference temperature,
  has drifts when switched on and off. that is not so much the case with
  modified reference. but my assumption re the above is that your meter is
  always on, as I said)
  732a references et al do have small short term drifts, you can determine
  them with a josephson element (these guys told me), but the 732a is
  certainly very stable short term compared to a 3458a. my 732a e.g. has a
  drift of 0.2ppm per year.
 
 
   Gesendet: Mittwoch, 27. August 2014 um 14:36 Uhr
   Von: Randy Evans randyevans2...@gmail.com
   An: Discussion of precise voltage measurement volt-nuts@febo.com
   Betreff: Re: [volt-nuts] 732A and Prologix received
  
   I forgot to mention that I reduced the number of measurements to 100 per
   set since I wasn't seeing much difference in the variance between 100 and
   1000 measurements and the 1000 measurement per set takes too long.
  
   Randy
  
  
   On Wed, Aug 27, 2014 at 5:13 AM, Randy Evans randyevans2...@gmail.com
   wrote:
  
I am doing multiple 100 measurements simply to characterize the
  stability
of the 3458A and 732A units I just bought.  After about 10 measurement
  sets
over 2 days I am seeing a variance of about .5 uV for the 10V output,
  or
0.05 ppm.  However, the mean varies over a range of 10 uV, or 1 ppm.
  Does
that sound reasonable/
   
Randy
   
   
On Tue, Aug 26, 2014 at 5:41 PM, acb...@gmx.de wrote:
   
hi randy,
   
just for curiosity, why doing 100 measurements at nplc 1000. is this
  to
sample a changing value?
when i am doing 10 measurements from a stable signal at nplc 100 (only
there many subsequent measuremnts with statistics make sense) I am
  already
getting a stanard deviation below 0.1ppm.
in a 30 minute test cycle, i would also be concerned about drifts
  (acal)
unless the amb. temperature is really very stable (half a degree
  already
adds about 0.25ppm at 10v)
   
thanks
   
   
   
   
 Gesendet: Dienstag, 26. August 2014 um 04:23 Uhr
 Von: Randy Evans randyevans2...@gmail.com
 An: Discussion of precise voltage measurement volt-nuts@febo.com
  
 Betreff: Re: [volt-nuts] 732A and Prologix received

 Is there any way to tell when the function key routine is
  complete?  In
the
 case of taking multiple readings using the DEFKEY and MATH
  function, I
 don't see any indication when the routine is complete.  In one
particular
 case, I am taking a 100 readings with NLPC set for 1000 so its a
  long
while
 before it's complete, but i have to guess when it's done.

 Thanks,

 Randy


 On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans 
  randyevans2...@gmail.com
 wrote:

  Bill,
 
  I am trying to figure out the MATH function without much
  success.  I
input
  the sequence you said (I looked up the instructions to understand
what you
  did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40

Re: [volt-nuts] 732A and Prologix received

2014-08-27 Thread M K

On 26/08/2014 16:05, Mike S wrote:
After some more research, I think I've answered some of my own 
questions -


Tellurium copper is used for binding posts, not because it has any 
special thermal or EMF mojo, but because it machines much better than 
pure copper. And, I suppose, because it sounds like it's extra special.


The Seebeck coefficients (uV/C, relative to Cu) of some relevant 
materials:

Cu 0.0
Ag .2
Au .5
Yellow brass 1.5
Phosphor bronze 2.0
63/37 solder 3.0
Sn 3.1
Stainless steel 3.1
Beryllium copper 5.0
Fe -12.3
Ni  22.3
Te -49.25

Based on the extreme Seebeck coefficient of pure tellurium vs. copper, 
I'd expect that there might be some coefficient between Cu and CuTe 
(0.5% Te), but I could find no reference. The relatively large number 
for CuBe is interesting, since that's a common material for banana 
plug springs, where one might expect the greatest temperature 
differential to occur in such a connection (between the thermal masses 
of the binding post/jack and the bulk of the banana plug). Heat has to 
flow a considerable distance through the springs, very much more than 
when it flows through a surface plating.


The Pomona (Fluke) EM5295-48-0# uses CuBe (gold plated) for the spring 
contacts. It seems there might be an improvement to be had by using 
the older style pin plugs, where a solid pin was partially sliced into 
4 sections which were then spread apart a bit to create tension. That 
could eliminate relatively large thermocouples at a thermal gradient, 
and might also be expected to have less thermal resistance, allowing 
the connection to settle quicker.


But maybe not - I'm still not clear on how plated conductors behave in 
this situation. For a high impedance voltage measurement where almost 
no current flows, the gold plating may carry the signal, so there is 
no real thermocouple (or more correctly, it's entirely contained 
within the connector). But if that's the case, why fool around with 
special copper connectors when common brass ones would be 
easier/cheaper? For current or resistance, the signal would also flow 
through the base metal, so does this have an effect (especially for 
tinned copper test leads, where there may be a larger temperature 
difference between the ends???


Nickle is avoided as a contact material largely because it is subject 
to fretting corrosion. Tests done by AMP 
(http://www.te.com/documentation/whitepapers/pdf/p154-74.pdf) show 
that a Ni to Ni contact can increase from 8 mOhm to 5 Ohms (sic!) in a 
short time due to this, while Ag and Au plated contacts exhibit 
negligible changes.


Cu (with Be for better machining) seems to be used as the base 
material for jacks/plugs to get thermal EMF cancellation to the wiring 
on both sides (i.e. use copper everywhere except where there is a 
minimal thermal gradient, like platings).


Beryllium copper is a springy material, Tellurium is the material added 
to aid machining without adding too much seebeck coefficient.


I remember someone on this list a long time ago saying that NPL used van 
damme star quad cable and bought bulk quantities of spade lugs that they 
strip all the coating off before crimping and clean before each use.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Re: [volt-nuts] 732A and Prologix received

2014-08-27 Thread Stan Katz
As a non-scholar in metrology, I tend to want to simplify the results of an
academic debate to make the results of the debate useful to me. One thing
is clear from my web search, copper alloyed with Tellurium, or Beryllium,
still oxidizes, only at a slower rate. It appears that a big disadvantage
of Beryllium oxide is its very hard, a useful industrial characteristic.
Not a nice property in the metrology lab.

 Here goes:

For the purists with lab grade metallurgical abrasives, polishing
equipment, and oxide removal chemistry, only, copper, whether alloyed with
Beryllium or Tellurium is all they want to see in their lead terminations,
and they will try to procure only instruments with copper/copper alloy
terminals. These purists will clean/deoxidize  all connections in a
controlled atmosphere, and dry them in completely dry nitrogen, or other
inert gas. The connections will then be used immediately during the
measurement and the purists will have assured themselves that all
instrument to lead connections are tight enough to be oxygen free.

Can anyone mention any precision metrology instrumentation in current
production with pure copper, or copper alloy connectors? I have a 38 year
old Hp 740b, and yep, it's got Gold flashed connectors. This Gold flashing
seems to be a tradition.

The practical metrologist, will accept the fact that precision metrology
instruments are meant to last many years, and the terminals supplied with
these instruments must provide a stable thermal emf profile over time.
Thus, they accept instruments that come with gold plated terminals. The
Gold may need to be cleansed of debris, and degreased from time-to-time,
but the procedure is much simpler, and not as time consuming as oxide
removal.  The leads to these instruments are meant to be used day in, and
day out, as well. Therefore, the leads also terminate in some form of gold
plated connector. The practical metrologist is fastidious in controlling
temperature, and air movement in the lab, to minimize thermal unbalance in
his/her lash up.

Is this a foolish simplification of the thermals debate, or can I feel
vindicated using my homemade, gold plated lead terminations with my old
740b, and 731b?


On Wed, Aug 27, 2014 at 3:45 PM, M K m1k...@hotmail.com wrote:

 On 26/08/2014 16:05, Mike S wrote:

 After some more research, I think I've answered some of my own questions -

 Tellurium copper is used for binding posts, not because it has any
 special thermal or EMF mojo, but because it machines much better than pure
 copper. And, I suppose, because it sounds like it's extra special.

 The Seebeck coefficients (uV/C, relative to Cu) of some relevant
 materials:
 Cu 0.0
 Ag .2
 Au .5
 Yellow brass 1.5
 Phosphor bronze 2.0
 63/37 solder 3.0
 Sn 3.1
 Stainless steel 3.1
 Beryllium copper 5.0
 Fe -12.3
 Ni  22.3
 Te -49.25

 Based on the extreme Seebeck coefficient of pure tellurium vs. copper,
 I'd expect that there might be some coefficient between Cu and CuTe (0.5%
 Te), but I could find no reference. The relatively large number for CuBe is
 interesting, since that's a common material for banana plug springs, where
 one might expect the greatest temperature differential to occur in such a
 connection (between the thermal masses of the binding post/jack and the
 bulk of the banana plug). Heat has to flow a considerable distance through
 the springs, very much more than when it flows through a surface plating.

 The Pomona (Fluke) EM5295-48-0# uses CuBe (gold plated) for the spring
 contacts. It seems there might be an improvement to be had by using the
 older style pin plugs, where a solid pin was partially sliced into 4
 sections which were then spread apart a bit to create tension. That could
 eliminate relatively large thermocouples at a thermal gradient, and might
 also be expected to have less thermal resistance, allowing the connection
 to settle quicker.

 But maybe not - I'm still not clear on how plated conductors behave in
 this situation. For a high impedance voltage measurement where almost no
 current flows, the gold plating may carry the signal, so there is no real
 thermocouple (or more correctly, it's entirely contained within the
 connector). But if that's the case, why fool around with special copper
 connectors when common brass ones would be easier/cheaper? For current or
 resistance, the signal would also flow through the base metal, so does this
 have an effect (especially for tinned copper test leads, where there may be
 a larger temperature difference between the ends???

 Nickle is avoided as a contact material largely because it is subject to
 fretting corrosion. Tests done by AMP (http://www.te.com/
 documentation/whitepapers/pdf/p154-74.pdf) show that a Ni to Ni contact
 can increase from 8 mOhm to 5 Ohms (sic!) in a short time due to this,
 while Ag and Au plated contacts exhibit negligible changes.

 Cu (with Be for better machining) seems to be used as the base material
 for jacks/plugs to get thermal EMF 

Re: [volt-nuts] 732A and Prologix received

2014-08-26 Thread Randy Evans
Bill,

I measured the time between SMPL symbols with NPLC set to 1000 and it is
approximately 33 seconds. It takes an hour to complete 100 readings.

Randy


On Mon, Aug 25, 2014 at 10:01 PM, Bill Gold wpgold3...@att.net wrote:

 Not that I know of.  Just wait and when you don't see a SMPL on the
 display it is done.  But then with 1000 PLC that is around 16.66 seconds
 per
 reading times 100 readings is somewhere around 28 minutes and there is
 probably some overhead time so around 30 minutes.  Not from the front panel
 at any rate.

 Bill

 - Original Message -
 From: Randy Evans randyevans2...@gmail.com
 To: Discussion of precise voltage measurement volt-nuts@febo.com
 Sent: Monday, August 25, 2014 9:23 PM
 Subject: Re: [volt-nuts] 732A and Prologix received


  Is there any way to tell when the function key routine is complete?  In
 the
  case of taking multiple readings using the DEFKEY and MATH function, I
  don't see any indication when the routine is complete.  In one particular
  case, I am taking a 100 readings with NLPC set for 1000 so its a long
 while
  before it's complete, but i have to guess when it's done.
 
  Thanks,
 
  Randy
 
 
  On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans randyevans2...@gmail.com
  wrote:
 
   Bill,
  
   I am trying to figure out the MATH function without much success.  I
 input
   the sequence you said (I looked up the instructions to understand what
 you
   did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG;
 and
   it shows up on the display when I input BLUE F1.  I hit ENTER and it
 takes
   the 40 measurements and the MATH symbol shows on the display during the
   measurements.  After the SMPL symbol no longer blinks I hit MATH 2 and
 I
   get a MATH ERR symbol on the display.  I tried it a couple of times and
 the
   same result so I am doing something wrong.  Is there a better source
 for
   explaining how to do front panel masurements than the User Guide, which
   seems oriented at programming automatic rather than manual
 measurements.
  
   Randy
  
  
  
   On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold wpgold3...@att.net wrote:
  
   Randy:
  
   The MATH function is accessible from the keypad.  I don't have an
 IEEE
   interface right now that works.  You can also program the numeric
 keypad
   keys to have preprogrammed functions.  DEFKEY
  
   I have made my own low thermal measurement leads from Pomona
 #4892
   banana plugs and Belden #9272 wire.  Why 9272, because it was handy at
 the
   time.  It is tin plated copper, shielded twisted pair 20 ga.  I have
 plans
   to do custom cables with 16 ga. bare copper wire that I will twist and
   then
   put a braided shield over it.  I simply cannot find what I want so I
 will
   build my own cable.  I have done something like this before and it
 worked
   fine.  When I get a round toit.
  
   I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I
 have
   used
   in the past to verify my homemade low thermal leads as described
 above.
   Frankly I cannot see any difference between using the 1756 cables and
 my
   homemade cables once I give them a few minutes for the thermals to go
   away.
   As far as I can tell and measure the differences, if any, are below
 0.1
   ppm
   at 10 volts.
  
   Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are
 all
   adjustable you may be seeing a misadjusted 1 volt from the 732A.  As
 far
   as
   the instability of the readings it is hard to determine which is
 causing
   the
   problem.  I have programed (DEFKEY) a numeric keypad key #1 with the
   following code.  MATH 14;NRDGS 40;TRIG 4;TRIG ;   So what this does
 is
   set
   the MATH to Statistics (store high reading/low reading/ and mean of
 the
   readings) in the registers, the number of readings to 40, the
 trigger
 to
   hold (which keeps the meter from triggering until I press ENTER
 and
   then
   trigger the sequence of 40 readings when I push the ENTER button.
 You
   can
   do all of this manually from the keypads but since I use this sequence
 a
   lot
   I have preprogrammed it.  This is after I set digits to 8 and PLC to
   100.
   Once those 40 readings are finished then you can access the various
 MATH
   statistic registers, using the menu, by entering MATH and then a 2
 for
   low, a 4 for mean, and 13 for high.  Of course you could do all of
 this
   through the IEEE also.  The 3458A has a very rich set of measurement
   commands.  I am still learning all of them.  It depends upon what I am
   trying to accomplish.
  
   Since the 1.018 and 1.0 volt outputs are passive and derived from
   resistive dividers from the 10 volt, I don't see how they could
 contribute
   to the varying readings you are measuring.  I think I would put a
 short
 on
   the input of the 3458A and manually set the range to 1 volt and then
   observe
   the variations that way without the 732A involved.  When I do this I
 see a
   variation from low

Re: [volt-nuts] 732A and Prologix received

2014-08-26 Thread acbern
hi randy,

just for curiosity, why doing 100 measurements at nplc 1000. is this to sample 
a changing value?
when i am doing 10 measurements from a stable signal at nplc 100 (only there 
many subsequent measuremnts with statistics make sense) I am already getting a 
stanard deviation below 0.1ppm.
in a 30 minute test cycle, i would also be concerned about drifts (acal) unless 
the amb. temperature is really very stable (half a degree already adds about 
0.25ppm at 10v)

thanks




 Gesendet: Dienstag, 26. August 2014 um 04:23 Uhr
 Von: Randy Evans randyevans2...@gmail.com
 An: Discussion of precise voltage measurement volt-nuts@febo.com
 Betreff: Re: [volt-nuts] 732A and Prologix received

 Is there any way to tell when the function key routine is complete?  In the
 case of taking multiple readings using the DEFKEY and MATH function, I
 don't see any indication when the routine is complete.  In one particular
 case, I am taking a 100 readings with NLPC set for 1000 so its a long while
 before it's complete, but i have to guess when it's done.
 
 Thanks,
 
 Randy
 
 
 On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans randyevans2...@gmail.com
 wrote:
 
  Bill,
 
  I am trying to figure out the MATH function without much success.  I input
  the sequence you said (I looked up the instructions to understand what you
  did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG; and
  it shows up on the display when I input BLUE F1.  I hit ENTER and it takes
  the 40 measurements and the MATH symbol shows on the display during the
  measurements.  After the SMPL symbol no longer blinks I hit MATH 2 and I
  get a MATH ERR symbol on the display.  I tried it a couple of times and the
  same result so I am doing something wrong.  Is there a better source for
  explaining how to do front panel masurements than the User Guide, which
  seems oriented at programming automatic rather than manual measurements.
 
  Randy
 
 
 
  On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold wpgold3...@att.net wrote:
 
  Randy:
 
  The MATH function is accessible from the keypad.  I don't have an IEEE
  interface right now that works.  You can also program the numeric keypad
  keys to have preprogrammed functions.  DEFKEY
 
  I have made my own low thermal measurement leads from Pomona #4892
  banana plugs and Belden #9272 wire.  Why 9272, because it was handy at the
  time.  It is tin plated copper, shielded twisted pair 20 ga.  I have plans
  to do custom cables with 16 ga. bare copper wire that I will twist and
  then
  put a braided shield over it.  I simply cannot find what I want so I will
  build my own cable.  I have done something like this before and it worked
  fine.  When I get a round toit.
 
  I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have
  used
  in the past to verify my homemade low thermal leads as described above.
  Frankly I cannot see any difference between using the 1756 cables and my
  homemade cables once I give them a few minutes for the thermals to go
  away.
  As far as I can tell and measure the differences, if any, are below 0.1
  ppm
  at 10 volts.
 
  Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are all
  adjustable you may be seeing a misadjusted 1 volt from the 732A.  As far
  as
  the instability of the readings it is hard to determine which is causing
  the
  problem.  I have programed (DEFKEY) a numeric keypad key #1 with the
  following code.  MATH 14;NRDGS 40;TRIG 4;TRIG ;   So what this does is
  set
  the MATH to Statistics (store high reading/low reading/ and mean of the
  readings) in the registers, the number of readings to 40, the trigger to
  hold (which keeps the meter from triggering until I press ENTER and
  then
  trigger the sequence of 40 readings when I push the ENTER button.  You
  can
  do all of this manually from the keypads but since I use this sequence a
  lot
  I have preprogrammed it.  This is after I set digits to 8 and PLC to
  100.
  Once those 40 readings are finished then you can access the various MATH
  statistic registers, using the menu, by entering MATH and then a 2 for
  low, a 4 for mean, and 13 for high.  Of course you could do all of this
  through the IEEE also.  The 3458A has a very rich set of measurement
  commands.  I am still learning all of them.  It depends upon what I am
  trying to accomplish.
 
  Since the 1.018 and 1.0 volt outputs are passive and derived from
  resistive dividers from the 10 volt, I don't see how they could contribute
  to the varying readings you are measuring.  I think I would put a short on
  the input of the 3458A and manually set the range to 1 volt and then
  observe
  the variations that way without the 732A involved.  When I do this I see a
  variation from low reading to high reading of 0.125 uVolts and then
  another
  40 I get 0.155 uVolts.  This is without the GUARD connected to the low
  side
  of the measurment terminals, GUARD connected doesn't seem to affect

Re: [volt-nuts] 732A and Prologix received

2014-08-25 Thread acbern
I have used the pomona spades, mainly to interface the low emf pomona banana 
cables to binding posts. I have stopped this, reasons being, they are large and 
worse, that the pomona spring loaded insulation tube that covers the banana 
plug conductor uses such a strong spring that slowly the plug works its way out 
of the spade. this btw also happend to me when I used the pomona low emf 
binding posts together with the pomona low emf banana cables. overall I m not 
happy with these. 
so, due to lack of options, I changed to self-made twisted shielded pair of 
high grade teflon/kapton silver plated copper cable with gold plated copper 
spades (crimped). I use them not only with the 3458a but also with nanovolt 
meters. these have higher resolution and accuracy in low level measurements 
than the 3458a. emf voltages were never an issue with these cables if properly 
used. I have posted some results doing 34420a stabilty measurements on the pmel 
forum, and the results are convincing (purpose was actually not to test the 
cables but the stability of the 34420a, but the emf issue is a part of this of 
course. we use the 34420a to do low voltage precision measurements on thermal 
converters where the full scale signal sometimes is 1mV). 
that btw also relates to don's statements below, I do not concurr with his 
comments about copper telurium as cable and spade material and so on. this 
material, as stated here many times, is used because it is machinable, for 
copper spades one would not use it. the 34420a factory cable uses copper cable 
and copper spades, not telurium-copper. if there was a problem, it would be 
worse with the 34420a than with the 3458a because of its low level ranges. and 
again, I have not seen any problems in a chain of (output to input):
1.copper-tellurium post from e.g. 8 digit calibrator
2.crimped copper spade, gold plated
3.silver plated tsp copper cable
4a.crimped copper spade to copper-tellurium post or 
4b.soldered copper connector(34420)
my consistent results over more than a year using them.



 Gesendet: Montag, 25. August 2014 um 06:33 Uhr
 Von: Orin Eman orin.e...@gmail.com
 An: Discussion of precise voltage measurement volt-nuts@febo.com
 Betreff: Re: [volt-nuts] 732A and Prologix received

 On Sun, Aug 24, 2014 at 1:46 PM, Don@True-Cal truecalservi...@gmail.com
 wrote:
 
  Randy  all,
 
  You have correctly concluded that some (maybe not all) of your measurement
  problem is thermal EMF being added or subtracted in series within your
  measurement interconnect. This thermal EMF is generated at the junction of
  dissimilar metals when accompanied with thermal gradients between the test
  lead and device terminals. You have to eliminate both the dissimilarity of
  the metal junctions as well as minimize the thermal differences. The
  terminals of the 3458A as well as the 732A are Beryllium Copper so you want
  to use the same test lead terminals. Forget the typical Tin plated lugs or
  even Gold plated as both are not Beryllium Copper and constitute dissimilar
  metals. The best solution (as usually the most expensive) is to use a set
  of
  Fluke 5440A-7005 (48) cables. I also have just as good results using the
  much more flexible Pomona 11174A (lugs end always stay connected to the
  732A) or 11058A with more convenient shielded banana plugs. The Fluke cable
  has the added Guard built in but be sure to also use a Guard lead with the
  Pomona cabled. The Guard lead does not need to be low thermal EMF. DIY
  cables is usually not a good idea because the lead wire to terminal also
  constitutes just as critical of junction. The above cables use Tellurium
  Copper wire which is usually hard to find and hard to crimp properly and
  NEVER solder.
 
 
 
 11058A and 11174A are discontinued at Keysight.  However, Pomona 5295 spade
 to banana cables are available (5295-36 at Mouser et al) and claim that
 they are designed to minimize thermal EMFs.  Datasheet is here:
 http://www.mouser.com/ds/2/159/d5295_1_01-51722.pdf  Any comments on these
 as an alternative?
 
 Orin.
 ___
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Re: [volt-nuts] 732A and Prologix received

2014-08-25 Thread Bill Gold
Randy:

Sorry my fault.  You have to use the RMATH command to get the various
values stored in the registers.  See the RMATH command in the User's Guide
for a list of what registers you can read.

I sure haven't found any other guides other than the 4 manuals.  User's
Guide, Quick Reference Guide, Calibration Manual, and Assembly Level Repair.
It is just a matter of reading the guides and trying to remember what
commands are available.  It took me a lot of time to figure out what
commands I use now.  I am sure I am missing other commands that might be
useful.

Bill



- Original Message - 
From: Randy Evans randyevans2...@gmail.com
To: Discussion of precise voltage measurement volt-nuts@febo.com
Sent: Sunday, August 24, 2014 6:04 PM
Subject: Re: [volt-nuts] 732A and Prologix received


 Bill,

 I am trying to figure out the MATH function without much success.  I input
 the sequence you said (I looked up the instructions to understand what you
 did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG;
and
 it shows up on the display when I input BLUE F1.  I hit ENTER and it takes
 the 40 measurements and the MATH symbol shows on the display during the
 measurements.  After the SMPL symbol no longer blinks I hit MATH 2 and I
 get a MATH ERR symbol on the display.  I tried it a couple of times and
the
 same result so I am doing something wrong.  Is there a better source for
 explaining how to do front panel masurements than the User Guide, which
 seems oriented at programming automatic rather than manual measurements.

 Randy



 On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold wpgold3...@att.net wrote:

  Randy:
 
  The MATH function is accessible from the keypad.  I don't have an
IEEE
  interface right now that works.  You can also program the numeric keypad
  keys to have preprogrammed functions.  DEFKEY
 
  I have made my own low thermal measurement leads from Pomona #4892
  banana plugs and Belden #9272 wire.  Why 9272, because it was handy at
the
  time.  It is tin plated copper, shielded twisted pair 20 ga.  I have
plans
  to do custom cables with 16 ga. bare copper wire that I will twist and
then
  put a braided shield over it.  I simply cannot find what I want so I
will
  build my own cable.  I have done something like this before and it
worked
  fine.  When I get a round toit.
 
  I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have
  used
  in the past to verify my homemade low thermal leads as described
above.
  Frankly I cannot see any difference between using the 1756 cables and my
  homemade cables once I give them a few minutes for the thermals to go
away.
  As far as I can tell and measure the differences, if any, are below 0.1
ppm
  at 10 volts.
 
  Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are
all
  adjustable you may be seeing a misadjusted 1 volt from the 732A.  As far
as
  the instability of the readings it is hard to determine which is causing
  the
  problem.  I have programed (DEFKEY) a numeric keypad key #1 with the
  following code.  MATH 14;NRDGS 40;TRIG 4;TRIG ;   So what this does is
  set
  the MATH to Statistics (store high reading/low reading/ and mean of
the
  readings) in the registers, the number of readings to 40, the trigger
to
  hold (which keeps the meter from triggering until I press ENTER and
  then
  trigger the sequence of 40 readings when I push the ENTER button.  You
  can
  do all of this manually from the keypads but since I use this sequence a
  lot
  I have preprogrammed it.  This is after I set digits to 8 and PLC to
100.
  Once those 40 readings are finished then you can access the various MATH
  statistic registers, using the menu, by entering MATH and then a 2 for
  low, a 4 for mean, and 13 for high.  Of course you could do all of this
  through the IEEE also.  The 3458A has a very rich set of measurement
  commands.  I am still learning all of them.  It depends upon what I am
  trying to accomplish.
 
  Since the 1.018 and 1.0 volt outputs are passive and derived from
  resistive dividers from the 10 volt, I don't see how they could
contribute
  to the varying readings you are measuring.  I think I would put a short
on
  the input of the 3458A and manually set the range to 1 volt and then
  observe
  the variations that way without the 732A involved.  When I do this I see
a
  variation from low reading to high reading of 0.125 uVolts and then
another
  40 I get 0.155 uVolts.  This is without the GUARD connected to the low
side
  of the measurment terminals, GUARD connected doesn't seem to affect the
  readings.  So that is the base noise of the 3458A without the 732A,
  somewhere below .2uVolts.  When hooked up to the 732A 1.0 volt output I
got
  a variation of 0.159 uVolts using the same 40 reading method above.  I
  would
  use this to determine where your problem might exist.  Just having the
  meter
  input shorted will point you in the right direction.  Meter, cables

Re: [volt-nuts] 732A and Prologix received

2014-08-25 Thread Don@True-Cal
Tellurium Copper is usually not used for a device's terminal posts but used as 
the lead wire due, as you say, for the malleability to crimp well and 
flexibility. The point I was making is to use the same interconnect test lead 
material throughout as the DUT terminal posts. The 3458A and the 732A both use 
Beryllium Copper alloy making that type interconnect lug or plug the best 
choice to minimize the dissimilar metal EMF or Seebeck voltage. The 34420A uses 
pure copper rather than an alloy terminal and for the same reason, minimal 
Seebeck voltage is realized with a pure copper interconnect. Any type of Silver 
or Gold plating on the terminal or wire will introduce the undesirable  
dissimilar metal properties, both at the plating junction and at the plating 
metal to DUT terminal.

The NI website had this chart that quantifies the Seebeck voltage best:
When two, dissimilar metals are joined a voltage is created. This voltage is 
known as the thermal electromotive force (EMF) or the Seebeck voltage. The 
Seebeck voltage is dependent on the temperature of the junction and the 
composition of the metals joined. The specific metal-to-metal junctions result 
in specific temperature coefficients (µV/°C), also known as Seebeck 
coefficients. The following table lists the most common metals and their 
respective Seebeck coefficients.

JunctionµV/°C
Copper-Copper   0.3
Copper-Gold 0.5
Copper-Silver   0.5
Copper-Brass3
Copper-Nickel   10
Copper-Lead-Tin Solder 1-3
Copper-Aluminum 5
Copper-Kovar40
Copper-Copper Oxide 500

Granted, Gold and Silver are the next best choice, and is certainly why they 
are satisfactory, but using either warrants a more critical temperature 
gradient issue. If your measurements were satisfactorily convincing, than you 
probably had no appreciable junction temperature differences. 

Don Johnson

-Original Message-
From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of acb...@gmx.de
Sent: Monday, August 25, 2014 5:37 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] 732A and Prologix received

I have used the pomona spades, mainly to interface the low emf pomona banana 
cables to binding posts. I have stopped this, reasons being, they are large and 
worse, that the pomona spring loaded insulation tube that covers the banana 
plug conductor uses such a strong spring that slowly the plug works its way out 
of the spade. this btw also happend to me when I used the pomona low emf 
binding posts together with the pomona low emf banana cables. overall I m not 
happy with these. 
so, due to lack of options, I changed to self-made twisted shielded pair of 
high grade teflon/kapton silver plated copper cable with gold plated copper 
spades (crimped). I use them not only with the 3458a but also with nanovolt 
meters. these have higher resolution and accuracy in low level measurements 
than the 3458a. emf voltages were never an issue with these cables if properly 
used. I have posted some results doing 34420a stabilty measurements on the pmel 
forum, and the results are convincing (purpose was actually not to test the 
cables but the stability of the 34420a, but the emf issue is a part of this of 
course. we use the 34420a to do low voltage precision measurements on thermal 
converters where the full scale signal sometimes is 1mV). 
that btw also relates to don's statements below, I do not concurr with his 
comments about copper telurium as cable and spade material and so on. this 
material, as stated here many times, is used because it is machinable, for 
copper spades one would not use it. the 34420a factory cable uses copper cable 
and copper spades, not telurium-copper. if there was a problem, it would be 
worse with the 34420a than with the 3458a because of its low level ranges. and 
again, I have not seen any problems in a chain of (output to input):
1.copper-tellurium post from e.g. 8 digit calibrator 2.crimped copper spade, 
gold plated 3.silver plated tsp copper cable 4a.crimped copper spade to 
copper-tellurium post or 4b.soldered copper connector(34420) my consistent 
results over more than a year using them.



 Gesendet: Montag, 25. August 2014 um 06:33 Uhr
 Von: Orin Eman orin.e...@gmail.com
 An: Discussion of precise voltage measurement volt-nuts@febo.com
 Betreff: Re: [volt-nuts] 732A and Prologix received

 On Sun, Aug 24, 2014 at 1:46 PM, Don@True-Cal truecalservi...@gmail.com
 wrote:
 
  Randy  all,
 
  You have correctly concluded that some (maybe not all) of your measurement
  problem is thermal EMF being added or subtracted in series within your
  measurement interconnect. This thermal EMF is generated at the junction of
  dissimilar metals when accompanied with thermal gradients between the test
  lead and device terminals. You have to eliminate both the dissimilarity of
  the metal junctions as well as minimize the thermal differences. The
  terminals

Re: [volt-nuts] 732A and Prologix received

2014-08-25 Thread acbern
well, your last point is the issue, how can you have a temperature difference 
within a few microns of material in said connections. theory is one thing, but 
in reality it does not happen due to the givens of the setup.
therefore in practice it is irrelevant if the wire is silver or gold plated or 
pure copper. otherwise the gold plated spades and tellurium copper posts from 
pomona and others would be nonsense. and other than the mysterious fluke wire I 
have never seen a tellurium-copper wire from any wire manufacturer.


 Gesendet: Montag, 25. August 2014 um 17:02 Uhr
 Von: Don@True-Cal truecalservi...@gmail.com
 An: 'Discussion of precise voltage measurement' volt-nuts@febo.com
 Betreff: Re: [volt-nuts] 732A and Prologix received

 Tellurium Copper is usually not used for a device's terminal posts but used 
 as the lead wire due, as you say, for the malleability to crimp well and 
 flexibility. The point I was making is to use the same interconnect test lead 
 material throughout as the DUT terminal posts. The 3458A and the 732A both 
 use Beryllium Copper alloy making that type interconnect lug or plug the best 
 choice to minimize the dissimilar metal EMF or Seebeck voltage. The 34420A 
 uses pure copper rather than an alloy terminal and for the same reason, 
 minimal Seebeck voltage is realized with a pure copper interconnect. Any type 
 of Silver or Gold plating on the terminal or wire will introduce the 
 undesirable  dissimilar metal properties, both at the plating junction and at 
 the plating metal to DUT terminal.
 
 The NI website had this chart that quantifies the Seebeck voltage best:
 When two, dissimilar metals are joined a voltage is created. This voltage is 
 known as the thermal electromotive force (EMF) or the Seebeck voltage. The 
 Seebeck voltage is dependent on the temperature of the junction and the 
 composition of the metals joined. The specific metal-to-metal junctions 
 result in specific temperature coefficients (µV/°C), also known as Seebeck 
 coefficients. The following table lists the most common metals and their 
 respective Seebeck coefficients.
 
 Junction  µV/°C
 Copper-Copper 0.3
 Copper-Gold   0.5
 Copper-Silver 0.5
 Copper-Brass  3
 Copper-Nickel 10
 Copper-Lead-Tin Solder 1-3
 Copper-Aluminum   5
 Copper-Kovar  40
 Copper-Copper Oxide   500
 
 Granted, Gold and Silver are the next best choice, and is certainly why they 
 are satisfactory, but using either warrants a more critical temperature 
 gradient issue. If your measurements were satisfactorily convincing, than you 
 probably had no appreciable junction temperature differences. 
 
 Don Johnson
 
 -Original Message-
 From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of acb...@gmx.de
 Sent: Monday, August 25, 2014 5:37 AM
 To: Discussion of precise voltage measurement
 Subject: Re: [volt-nuts] 732A and Prologix received
 
 I have used the pomona spades, mainly to interface the low emf pomona banana 
 cables to binding posts. I have stopped this, reasons being, they are large 
 and worse, that the pomona spring loaded insulation tube that covers the 
 banana plug conductor uses such a strong spring that slowly the plug works 
 its way out of the spade. this btw also happend to me when I used the pomona 
 low emf binding posts together with the pomona low emf banana cables. overall 
 I m not happy with these. 
 so, due to lack of options, I changed to self-made twisted shielded pair of 
 high grade teflon/kapton silver plated copper cable with gold plated copper 
 spades (crimped). I use them not only with the 3458a but also with nanovolt 
 meters. these have higher resolution and accuracy in low level measurements 
 than the 3458a. emf voltages were never an issue with these cables if 
 properly used. I have posted some results doing 34420a stabilty measurements 
 on the pmel forum, and the results are convincing (purpose was actually not 
 to test the cables but the stability of the 34420a, but the emf issue is a 
 part of this of course. we use the 34420a to do low voltage precision 
 measurements on thermal converters where the full scale signal sometimes is 
 1mV). 
 that btw also relates to don's statements below, I do not concurr with his 
 comments about copper telurium as cable and spade material and so on. this 
 material, as stated here many times, is used because it is machinable, for 
 copper spades one would not use it. the 34420a factory cable uses copper 
 cable and copper spades, not telurium-copper. if there was a problem, it 
 would be worse with the 34420a than with the 3458a because of its low level 
 ranges. and again, I have not seen any problems in a chain of (output to 
 input):
 1.copper-tellurium post from e.g. 8 digit calibrator 2.crimped copper spade, 
 gold plated 3.silver plated tsp copper cable 4a.crimped copper spade to 
 copper-tellurium post or 4b.soldered copper connector(34420) my consistent

Re: [volt-nuts] 732A and Prologix received

2014-08-25 Thread Mike S

On 8/25/2014 11:02 AM, Don@True-Cal wrote:

Silver or Gold plating on the terminal or wire will introduce the
undesirable  dissimilar metal properties, both at the plating
junction and at the plating metal to DUT terminal.


Why?

Any Seebeck effect is immediately offset in the opposite direction,
since both junctions are (under normal conditions) at essentially the
same temperature (e.g. there's a copper-gold thermocouple, the minimal
thermal resistance of a micron of gold on the contact(s), then a
gold-copper thermocouple). It seems to me that the improved consistency
of the contact outweighs any loss from the thermocouples.

A more typical contact would be copper-nickel plate-gold plate, but the
concept is the same. Unless there is heat flowing through the entire
assembly so one thermocouple is warmer than the offsetting one (e.g.
shortly after plugging in a banana plug warmed by body heat), they
simply cancel.

Even if connecting gold plated to nickel plated contacts, it works out 
the same - a copper-nickel-gold-nickel-copper connection is completely 
offset. It's when the offsetting thermocouples occur across a 
temperature gradient that you have problems.


--
Mike
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Re: [volt-nuts] 732A and Prologix received

2014-08-25 Thread Don@True-Cal
Why?

Let me count the ways. You can never count on any Seebeck voltage to be 
immediately offset, there are far too many variables. Best example I can think 
of...why is there an Ohms Offset Compensation feature on any good high 
resolution DMM. 1) Try measuring a 1 or 10 Ohm resistor with your 3458A in 
4-wire mode using inexpensive nickel-plated leads and even allow plenty of time 
for everything to thermally stabilize. Using Ohms Offset Compensation, enable 
and disable it and observe the difference. If the Seebeck voltages were all 
immediately offset, as you say, there would be no difference. But there most 
certainly is. Or simply, why is there a need for ohms offset compensation 
feature if all Seebeck voltages cancel each other out. Sure, nickel-plated is a 
horrible choice but if it all canceled, what difference would it make how bad 
is. 2) The cal lab workhorse calibrator is the 5700A/5720A. In between trips 
back to Fluke for full calibration, there is an interim external calibration 
procedure using the 732B, 742A-1  742A-10k. If someone used a set of 
gold-plated interconnects for this procedure, they would be laughed out of the 
lab and the calibrator would be useless until recalibrated properly. A set of 
5440A-7002 (banana plug) cables comes with this calibrator (5440A-7003 spade 
lugs for 5720A) and recommended for the calibration procedure but other 
Beryllium Copper or pure Copper cables are also acceptable. 3) Lab air drafts 
will never allow true thermal symmetry around the DMM or DUT terminals. To 
convince yourself, place an oscillating fan several feet back from the DMM and 
DUT terminals and using the 1 or 10 Ohm setup from above, again with the 
nickel-plated leads, watch the variations. Sure the fan and the nickel-plated 
exaggerates the issue but it quickly dispels the notion that all the Seebeck 
voltages are canceled out.

BTW, the plating layer temperature on a plated terminal will be somewhere 
between the temperature of the base metal and mating terminal it's connected to.

This is not just theory, my 40+ years in the cal lab is driving my arguments 
but it never hurts to have physical-science on your side.

Don Johnson

-Original Message-
From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of Mike S
Sent: Monday, August 25, 2014 11:03 AM
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] 732A and Prologix received

On 8/25/2014 11:02 AM, Don@True-Cal wrote:
 Silver or Gold plating on the terminal or wire will introduce the 
 undesirable  dissimilar metal properties, both at the plating junction 
 and at the plating metal to DUT terminal.

Why?

Any Seebeck effect is immediately offset in the opposite direction, since both 
junctions are (under normal conditions) at essentially the same temperature 
(e.g. there's a copper-gold thermocouple, the minimal thermal resistance of a 
micron of gold on the contact(s), then a gold-copper thermocouple). It seems to 
me that the improved consistency of the contact outweighs any loss from the 
thermocouples.

A more typical contact would be copper-nickel plate-gold plate, but the concept 
is the same. Unless there is heat flowing through the entire assembly so one 
thermocouple is warmer than the offsetting one (e.g.
shortly after plugging in a banana plug warmed by body heat), they simply 
cancel.

Even if connecting gold plated to nickel plated contacts, it works out the same 
- a copper-nickel-gold-nickel-copper connection is completely offset. It's when 
the offsetting thermocouples occur across a temperature gradient that you have 
problems.

--
Mike
___
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Re: [volt-nuts] 732A and Prologix received

2014-08-25 Thread Bill Gold
Not that I know of.  Just wait and when you don't see a SMPL on the
display it is done.  But then with 1000 PLC that is around 16.66 seconds per
reading times 100 readings is somewhere around 28 minutes and there is
probably some overhead time so around 30 minutes.  Not from the front panel
at any rate.

Bill

- Original Message - 
From: Randy Evans randyevans2...@gmail.com
To: Discussion of precise voltage measurement volt-nuts@febo.com
Sent: Monday, August 25, 2014 9:23 PM
Subject: Re: [volt-nuts] 732A and Prologix received


 Is there any way to tell when the function key routine is complete?  In
the
 case of taking multiple readings using the DEFKEY and MATH function, I
 don't see any indication when the routine is complete.  In one particular
 case, I am taking a 100 readings with NLPC set for 1000 so its a long
while
 before it's complete, but i have to guess when it's done.

 Thanks,

 Randy


 On Sun, Aug 24, 2014 at 6:04 PM, Randy Evans randyevans2...@gmail.com
 wrote:

  Bill,
 
  I am trying to figure out the MATH function without much success.  I
input
  the sequence you said (I looked up the instructions to understand what
you
  did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG;
and
  it shows up on the display when I input BLUE F1.  I hit ENTER and it
takes
  the 40 measurements and the MATH symbol shows on the display during the
  measurements.  After the SMPL symbol no longer blinks I hit MATH 2 and I
  get a MATH ERR symbol on the display.  I tried it a couple of times and
the
  same result so I am doing something wrong.  Is there a better source for
  explaining how to do front panel masurements than the User Guide, which
  seems oriented at programming automatic rather than manual measurements.
 
  Randy
 
 
 
  On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold wpgold3...@att.net wrote:
 
  Randy:
 
  The MATH function is accessible from the keypad.  I don't have an
IEEE
  interface right now that works.  You can also program the numeric
keypad
  keys to have preprogrammed functions.  DEFKEY
 
  I have made my own low thermal measurement leads from Pomona
#4892
  banana plugs and Belden #9272 wire.  Why 9272, because it was handy at
the
  time.  It is tin plated copper, shielded twisted pair 20 ga.  I have
plans
  to do custom cables with 16 ga. bare copper wire that I will twist and
  then
  put a braided shield over it.  I simply cannot find what I want so I
will
  build my own cable.  I have done something like this before and it
worked
  fine.  When I get a round toit.
 
  I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have
  used
  in the past to verify my homemade low thermal leads as described
above.
  Frankly I cannot see any difference between using the 1756 cables and
my
  homemade cables once I give them a few minutes for the thermals to go
  away.
  As far as I can tell and measure the differences, if any, are below 0.1
  ppm
  at 10 volts.
 
  Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are
all
  adjustable you may be seeing a misadjusted 1 volt from the 732A.  As
far
  as
  the instability of the readings it is hard to determine which is
causing
  the
  problem.  I have programed (DEFKEY) a numeric keypad key #1 with the
  following code.  MATH 14;NRDGS 40;TRIG 4;TRIG ;   So what this does
is
  set
  the MATH to Statistics (store high reading/low reading/ and mean of
the
  readings) in the registers, the number of readings to 40, the trigger
to
  hold (which keeps the meter from triggering until I press ENTER and
  then
  trigger the sequence of 40 readings when I push the ENTER button.
You
  can
  do all of this manually from the keypads but since I use this sequence
a
  lot
  I have preprogrammed it.  This is after I set digits to 8 and PLC to
  100.
  Once those 40 readings are finished then you can access the various
MATH
  statistic registers, using the menu, by entering MATH and then a 2
for
  low, a 4 for mean, and 13 for high.  Of course you could do all of this
  through the IEEE also.  The 3458A has a very rich set of measurement
  commands.  I am still learning all of them.  It depends upon what I am
  trying to accomplish.
 
  Since the 1.018 and 1.0 volt outputs are passive and derived from
  resistive dividers from the 10 volt, I don't see how they could
contribute
  to the varying readings you are measuring.  I think I would put a short
on
  the input of the 3458A and manually set the range to 1 volt and then
  observe
  the variations that way without the 732A involved.  When I do this I
see a
  variation from low reading to high reading of 0.125 uVolts and then
  another
  40 I get 0.155 uVolts.  This is without the GUARD connected to the low
  side
  of the measurment terminals, GUARD connected doesn't seem to affect the
  readings.  So that is the base noise of the 3458A without the 732A,
  somewhere below .2uVolts.  When hooked up to the 732A 1.0 volt output I
  got

Re: [volt-nuts] 732A and Prologix received

2014-08-24 Thread Randy Evans
: Friday, August 22, 2014 9:03 PM
 Subject: Re: [volt-nuts] 732A and Prologix received


  Todd,
 
  Thanks for the info.  I have several Panasonic 12V 7 AH batteries that I
  keep topped off and they have very low current draw (~2 to 3 mA at 13.5
  VDC) when charged and at their float voltage, so I am pretty sure they
 are
  in good condition.  I will look at getting those in the units after I
  ascertain the condition of the 732.
 
  So now I have a what appears to be a functioning 3458A and a 732A but
 they
  slightly disagree.  I am like the man with two watches that disagree on
 the
  time  - which is correct?  For the moment, i am only concerned with
  stability.  The need for absolute accuracy will come later.
 
  Randy
 
 
  On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef tmical...@gmail.com
 wrote:
 
   Randy,
  
   You have two possible choices. It can be configured with 4 x 6v 4Ah
   batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner has
   modified the battery pack already. A couple of mine needed a nibbler
 tool
   to remove enough of the aluminum cover that fits over the tops of the
   batteries. The original cover will short out to the battery tabs
 regardless
   of the battery configuration if this is not done.
  
   You can find larger capacity batteries that will give you slightly more
   battery life. I lost a couple sets of mail-order batteries after a few
   extended outages. I would recommend going with locally bought batteries
   instead of the cheaper mail order. My local Batteries Plus will
 typically
   have some warranty if I remember correctly. Moving forward I will only
 use
   2 12v batteries and pre-charge them on a battery charger to equalize
 them
   before putting them in the 732A. I think the cheap batteries did not
   discharge equally, and would not recover when power was applied.
  
   Inspect the back plane for damaged traces and look at the capacitors. I
 had
   a few that looked questionable. So far, I have replaced all the big
 caps
 on
   the pre-regulator and regulator boards. My feeling is that once these
 go
   online, they should run as long as possible between repairs.
  
   The battery charger circuit may need adjusting. I tweaked mine and it
   seemed to work fine.
  
   Todd
  
  
   On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans 
 randyevans2...@gmail.com
   wrote:
  
I received my Fluke 732A today.  Just powered it up but it needs new
batteries.  Any suggestions for sources (I haven't opened up the unit
   yet -
I want to make sure it works before doing that).  Also received the
ProLogix USB-GPIB adapter.  I plan on using Mark Sims' CAL ran data
   dumper
program to get the CAL data from my 3458A.  Should be a busy weekend.
   
Randy
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
   
   ___
   volt-nuts mailing list -- volt-nuts@febo.com
   To unsubscribe, go to
   https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
   and follow the instructions there.
  
  ___
  volt-nuts mailing list -- volt-nuts@febo.com
  To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
  and follow the instructions there.

 ___
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] 732A and Prologix received

2014-08-24 Thread Todd Micallef
Bill,

I use a lot of the 12v 7ah batteries for my UPS backups and 732B. But yes,
I was mistaken. The 12V 5ah batteries are the ones that I am using in one
of my 732A. There is more play inside the tray with the 12v batteries by
several mm as compared to the 6v which only has a 2-3mm. The 12v conversion
is not difficult, but it is easier if the battery tray has already been
machined for the 12v batteries. The battery tabs don't line up well with
the existing holes and need to be widened. Two additional holes must also
be added.

Sourcing the batteries locally is more of a convenience than waiting for
the delivery. The el-cheapo 6v batteries are a waste of money and I have 8
batteries so far to prove it. They died the first time I had a 6hr outage
and would not hold a charge after that.

Todd




On Sat, Aug 23, 2014 at 1:21 PM, Bill Gold wpgold3...@att.net wrote:

 Randy:

 I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
 perfect.  Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and
 others
 that are in this size and package.  Order from one of the usual electronics
 distributors like Allied, Mouser, Digikey.  This is a very common battery
 as
 it is used in a lot of EXIT signs so they are lighted when the power goes
 out.  I don't see how the 12V 7AH will fit as they are too large.  I guess
 you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt 4 AH
 but the terminals are in the wrong place so you will have to nibble out
 the aluminum plate that holds them in the 732A battery pack.  You have to
 be
 careful if you use the 12v 5AH as you will have 4 extra battery connection
 leads to deal with and connect correctly.  I would stick with the 6V 4AH.
 New batteries will last around 12 to 14 hours before the CAL light goes
 out when AC power is not applied.  So shipping to Cal Lab can be a problem
 if it is a distance away, or you have to use a shipper like UPS or FEDEX
 and
 you ship the night before and then use their Morning delivery and the Cal
 Lab is expecting your 732A.  Same on the way back to you.  Of course you
 could always strap another battery on the 732A and hook it up to the ext
 power plug to last longer.  I have seen it done.  The issue is to get the
 Cal Lab to charge the extra battery before they ship the 732A back to you.

 When you remove and work on the battery pack always have the AC power
 plugged in.  The CAL led will stay on because the 18.6 v regulated supply
 is working.

 The CAL light is to indicate that power has not been lost to the
 Reference Amp or other associated circuits.  When the raw supply (battery)
 voltage drops below about 21 volts the CAL light will go out.  Below that
 voltage the heater circuits will not work correctly and the 18.6 volt
 regulated supply will not regulate.  The requirement is that the Reference
 Amp be kept alive at all times to maintain the output voltage that was
 measured at the time of the most recent Calibration or Certification.  When
 the semiconductor junctions are unbiased and cool off when power is lost,
 and then power is restored the result will be a different 10 volts than
 before the power failure.  My experience is that after all of the years
 that
 these units have been powered up, this won't happen and when power is lost
 and then restored, even months later, the 732A will come back to almost
 exactly the same 10 volts as when they lost power, usually with in 0.2 PPM
 after 24 hours of warm up.

 What type of hookup leads are you using when measuring the 1 volt
 output?  What is the PLC set to?  I always use 100 PLC to measure this.  If
 you don't have low thermal connection leads you can experience uV changes
 for a minute or more after plugging in the leads due to the thermals
 generated because of the difference in temperature between the banana jacks
 on the 732A and the banana plugs of the leads.  I have found that even just
 plugging in the lead will generate a thermal difference because of
 difference of temps and some heating due to the physical act of just
 inserting the plug because of friction between the jack and plug (my theory
 at any rate).  You have to allow at least a minute or more before being
 able
 to make a measurement after plugging in the leads.  I just measured the
 variation of the 1 volt output of my 732A and using my 3458A and I got a
 total difference of  0.159 uV over 40 measurements using 100 PLC on the 1
 volt range of the 3458A.  Using the MATH function and all of the data you
 can collect.  That was after waiting for several minutes after plugging in
 the leads.

 I hope all of this helps.

 Bill



 - Original Message -
 From: Randy Evans randyevans2...@gmail.com
 To: Discussion of precise voltage measurement volt-nuts@febo.com
 Sent: Friday, August 22, 2014 9:03 PM
 Subject: Re: [volt-nuts] 732A and Prologix received


  Todd,
 
  Thanks for the info.  I have several Panasonic 12V 7 AH batteries that I
  keep topped off

Re: [volt-nuts] 732A and Prologix received

2014-08-24 Thread Charles Black
 experience uV changes
for a minute or more after plugging in the leads due to the thermals
generated because of the difference in temperature between the banana jacks
on the 732A and the banana plugs of the leads.  I have found that even just
plugging in the lead will generate a thermal difference because of
difference of temps and some heating due to the physical act of just
inserting the plug because of friction between the jack and plug (my theory
at any rate).  You have to allow at least a minute or more before being
able
to make a measurement after plugging in the leads.  I just measured the
variation of the 1 volt output of my 732A and using my 3458A and I got a
total difference of  0.159 uV over 40 measurements using 100 PLC on the 1
volt range of the 3458A.  Using the MATH function and all of the data you
can collect.  That was after waiting for several minutes after plugging in
the leads.

 I hope all of this helps.

Bill



- Original Message -
From: Randy Evans randyevans2...@gmail.com
To: Discussion of precise voltage measurement volt-nuts@febo.com
Sent: Friday, August 22, 2014 9:03 PM
Subject: Re: [volt-nuts] 732A and Prologix received



Todd,

Thanks for the info.  I have several Panasonic 12V 7 AH batteries that I
keep topped off and they have very low current draw (~2 to 3 mA at 13.5
VDC) when charged and at their float voltage, so I am pretty sure they

are

in good condition.  I will look at getting those in the units after I
ascertain the condition of the 732.

So now I have a what appears to be a functioning 3458A and a 732A but

they

slightly disagree.  I am like the man with two watches that disagree on

the

time  - which is correct?  For the moment, i am only concerned with
stability.  The need for absolute accuracy will come later.

Randy


On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef tmical...@gmail.com

wrote:

Randy,

You have two possible choices. It can be configured with 4 x 6v 4Ah
batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner has
modified the battery pack already. A couple of mine needed a nibbler

tool

to remove enough of the aluminum cover that fits over the tops of the
batteries. The original cover will short out to the battery tabs

regardless

of the battery configuration if this is not done.

You can find larger capacity batteries that will give you slightly more
battery life. I lost a couple sets of mail-order batteries after a few
extended outages. I would recommend going with locally bought batteries
instead of the cheaper mail order. My local Batteries Plus will

typically

have some warranty if I remember correctly. Moving forward I will only

use

2 12v batteries and pre-charge them on a battery charger to equalize

them

before putting them in the 732A. I think the cheap batteries did not
discharge equally, and would not recover when power was applied.

Inspect the back plane for damaged traces and look at the capacitors. I

had

a few that looked questionable. So far, I have replaced all the big

caps
on

the pre-regulator and regulator boards. My feeling is that once these

go

online, they should run as long as possible between repairs.

The battery charger circuit may need adjusting. I tweaked mine and it
seemed to work fine.

Todd


On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans 

randyevans2...@gmail.com

wrote:


I received my Fluke 732A today.  Just powered it up but it needs new
batteries.  Any suggestions for sources (I haven't opened up the unit

yet -

I want to make sure it works before doing that).  Also received the
ProLogix USB-GPIB adapter.  I plan on using Mark Sims' CAL ran data

dumper

program to get the CAL data from my 3458A.  Should be a busy weekend.

Randy
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to

https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts

and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.






___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Re: [volt-nuts] 732A and Prologix received

2014-08-24 Thread Don@True-Cal
Randy  all,

You have correctly concluded that some (maybe not all) of your measurement
problem is thermal EMF being added or subtracted in series within your
measurement interconnect. This thermal EMF is generated at the junction of
dissimilar metals when accompanied with thermal gradients between the test
lead and device terminals. You have to eliminate both the dissimilarity of
the metal junctions as well as minimize the thermal differences. The
terminals of the 3458A as well as the 732A are Beryllium Copper so you want
to use the same test lead terminals. Forget the typical Tin plated lugs or
even Gold plated as both are not Beryllium Copper and constitute dissimilar
metals. The best solution (as usually the most expensive) is to use a set of
Fluke 5440A-7005 (48) cables. I also have just as good results using the
much more flexible Pomona 11174A (lugs end always stay connected to the
732A) or 11058A with more convenient shielded banana plugs. The Fluke cable
has the added Guard built in but be sure to also use a Guard lead with the
Pomona cabled. The Guard lead does not need to be low thermal EMF. DIY
cables is usually not a good idea because the lead wire to terminal also
constitutes just as critical of junction. The above cables use Tellurium
Copper wire which is usually hard to find and hard to crimp properly and
NEVER solder.

As you observed, merely the friction of plugging in the spring banana plug
and heat transfer from your fingers will require ~3-5 minutes to stabilize.
Low mass terminals help with the time to initially stabilize but is not best
if dealing with local air current drafts that upset the instantaneous
thermal gradients between the higher mass device terminal and the lower mass
lead terminal. This issue is a trade-off and is specifically considered in
the Fluke cables but being careful with localized drafts around the
terminals during critical measurements can virtually eliminate this error
source.

I have attached an ~30-hour measurement run that I consistently do that
should give you some idea of expected measurement drift over time. Virtually
all of the measurement drift is due to the 3458A internal temperature
differences around the 7V reference caused by ambient temperature change.
The drift associated with the 732A is probably about 2-magnitudes less at
this ambient temp drift. In my situation, the inferred ambient temperature
is cycling with the home air-conditioning. Note the initial calibration
temperatures as well as the ACal temperatures and where the 3458A is
measuring exactly 10V. My primary 732A has been powered without loss for 4
years and 5 years before that. The 3458A is Agilent (Loveland) Cal'ed
yearly and is powered 24-7-365 except for the occasional mains power loss.
The graphical measurements is using a homegrown Agilent VEE program.

Don Johnson

-Original Message-
From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On
Behalf Of Randy Evans
Sent: Sunday, August 24, 2014 9:22 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] 732A and Prologix received

Bill,

I have convinced myself that the problem I an seeing is due to thermals.
 If I move the cables (with gold-plated banana plugs) using a small towel
rather than letting my hand touch the plugs, it is much more stable.  If I
then hold the banana plug with my hand after the reading has stabilized, the
reading drifts rapidly upward.  I am trying to check the stability of the
reading but I haven't figured out the MATH function yet.  I assume this is a
programmed function using GPIB only?

The stability I am seeing by manually recording readings (using NLPC of 100
and 1000) is much greater than what you are measuring on your system.  Not
sure how to ascertain if it's the 3458A or the 732.  The value of the
readings are very different between the two - the 3458 reads about 50 uV
high on the 10 V output and about 12 uV low on the 1V output.  Rather large
differences (this is after an ACAL).  I need to find some better cables to
make sure the errors are not due to thermals again.

Randy


On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold wpgold3...@att.net wrote:

 Randy:

 I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit 
 perfect.  Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and 
 others that are in this size and package.  Order from one of the usual 
 electronics distributors like Allied, Mouser, Digikey.  This is a very 
 common battery as it is used in a lot of EXIT signs so they are 
 lighted when the power goes out.  I don't see how the 12V 7AH will fit 
 as they are too large.  I guess you could use a 12V 5AH (PS1250) as it 
 is the same size as 2 x 6 volt 4 AH but the terminals are in the wrong 
 place so you will have to nibble out the aluminum plate that holds 
 them in the 732A battery pack.  You have to be careful if you use the 
 12v 5AH as you will have 4 extra battery connection leads to deal with 
 and connect correctly

Re: [volt-nuts] 732A and Prologix received

2014-08-24 Thread Bill Gold
Todd:

I check the batteries every month per the 732A manual.  Usually I keep a
few used cells from previous packs around that seem to be ok after checking
them for capacity just like checking the battery pack.  I can usually spot
cells going bad with that method.  If I don't have a spare set then I just
substitute a used cell until I receive a new set.

I agree with you that the cheap ones are crap.  I have used Power Sonic
a lot with some a few EaglePicher and one set of Enersys.  The EaglePicher
are crap so I will never use them.  EaglePicher have had a couple out of 8
that have failed very early.  I have never tried Panasonic.  So I stick with
the Power Sonic's.  I am not sure why the 12 volt ones would be better than
the 6 volts because 12 volt is just 6 lead acid cells while the 6volt are 3
lead acid cells.  If one of the 6 cells in the 12 volt battery go bad then
the whole battery has to be changed.  So you better buy good quality 12 volt
batteries.

I just can't seem to source batteries locally easily at a reasonable
price.  So I order them from Allied usually.

Bill

- Original Message - 
From: Todd Micallef tmical...@gmail.com
To: Discussion of precise voltage measurement volt-nuts@febo.com
Sent: Sunday, August 24, 2014 8:13 AM
Subject: Re: [volt-nuts] 732A and Prologix received


 Bill,

 I use a lot of the 12v 7ah batteries for my UPS backups and 732B. But yes,
 I was mistaken. The 12V 5ah batteries are the ones that I am using in one
 of my 732A. There is more play inside the tray with the 12v batteries by
 several mm as compared to the 6v which only has a 2-3mm. The 12v
conversion
 is not difficult, but it is easier if the battery tray has already been
 machined for the 12v batteries. The battery tabs don't line up well with
 the existing holes and need to be widened. Two additional holes must also
 be added.

 Sourcing the batteries locally is more of a convenience than waiting for
 the delivery. The el-cheapo 6v batteries are a waste of money and I have 8
 batteries so far to prove it. They died the first time I had a 6hr outage
 and would not hold a charge after that.

 Todd




 On Sat, Aug 23, 2014 at 1:21 PM, Bill Gold wpgold3...@att.net wrote:

  Randy:
 
  I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
  perfect.  Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and
  others
  that are in this size and package.  Order from one of the usual
electronics
  distributors like Allied, Mouser, Digikey.  This is a very common
battery
  as
  it is used in a lot of EXIT signs so they are lighted when the power
goes
  out.  I don't see how the 12V 7AH will fit as they are too large.  I
guess
  you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt 4
AH
  but the terminals are in the wrong place so you will have to nibble
out
  the aluminum plate that holds them in the 732A battery pack.  You have
to
  be
  careful if you use the 12v 5AH as you will have 4 extra battery
connection
  leads to deal with and connect correctly.  I would stick with the 6V
4AH.
  New batteries will last around 12 to 14 hours before the CAL light
goes
  out when AC power is not applied.  So shipping to Cal Lab can be a
problem
  if it is a distance away, or you have to use a shipper like UPS or FEDEX
  and
  you ship the night before and then use their Morning delivery and the
Cal
  Lab is expecting your 732A.  Same on the way back to you.  Of course you
  could always strap another battery on the 732A and hook it up to the
ext
  power plug to last longer.  I have seen it done.  The issue is to get
the
  Cal Lab to charge the extra battery before they ship the 732A back to
you.
 
  When you remove and work on the battery pack always have the AC
power
  plugged in.  The CAL led will stay on because the 18.6 v regulated
supply
  is working.
 
  The CAL light is to indicate that power has not been lost to the
  Reference Amp or other associated circuits.  When the raw supply
(battery)
  voltage drops below about 21 volts the CAL light will go out.  Below
that
  voltage the heater circuits will not work correctly and the 18.6 volt
  regulated supply will not regulate.  The requirement is that the
Reference
  Amp be kept alive at all times to maintain the output voltage that was
  measured at the time of the most recent Calibration or Certification.
When
  the semiconductor junctions are unbiased and cool off when power is
lost,
  and then power is restored the result will be a different 10 volts than
  before the power failure.  My experience is that after all of the years
  that
  these units have been powered up, this won't happen and when power is
lost
  and then restored, even months later, the 732A will come back to almost
  exactly the same 10 volts as when they lost power, usually with in 0.2
PPM
  after 24 hours of warm up.
 
  What type of hookup leads are you using when measuring the 1 volt
  output?  What is the PLC set to?  I always

Re: [volt-nuts] 732A and Prologix received

2014-08-24 Thread Bill Gold
Randy:

The MATH function is accessible from the keypad.  I don't have an IEEE
interface right now that works.  You can also program the numeric keypad
keys to have preprogrammed functions.  DEFKEY

I have made my own low thermal measurement leads from Pomona #4892
banana plugs and Belden #9272 wire.  Why 9272, because it was handy at the
time.  It is tin plated copper, shielded twisted pair 20 ga.  I have plans
to do custom cables with 16 ga. bare copper wire that I will twist and then
put a braided shield over it.  I simply cannot find what I want so I will
build my own cable.  I have done something like this before and it worked
fine.  When I get a round toit.

I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have used
in the past to verify my homemade low thermal leads as described above.
Frankly I cannot see any difference between using the 1756 cables and my
homemade cables once I give them a few minutes for the thermals to go away.
As far as I can tell and measure the differences, if any, are below 0.1 ppm
at 10 volts.

Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are all
adjustable you may be seeing a misadjusted 1 volt from the 732A.  As far as
the instability of the readings it is hard to determine which is causing the
problem.  I have programed (DEFKEY) a numeric keypad key #1 with the
following code.  MATH 14;NRDGS 40;TRIG 4;TRIG ;   So what this does is set
the MATH to Statistics (store high reading/low reading/ and mean of the
readings) in the registers, the number of readings to 40, the trigger to
hold (which keeps the meter from triggering until I press ENTER and then
trigger the sequence of 40 readings when I push the ENTER button.  You can
do all of this manually from the keypads but since I use this sequence a lot
I have preprogrammed it.  This is after I set digits to 8 and PLC to 100.
Once those 40 readings are finished then you can access the various MATH
statistic registers, using the menu, by entering MATH and then a 2 for
low, a 4 for mean, and 13 for high.  Of course you could do all of this
through the IEEE also.  The 3458A has a very rich set of measurement
commands.  I am still learning all of them.  It depends upon what I am
trying to accomplish.

Since the 1.018 and 1.0 volt outputs are passive and derived from
resistive dividers from the 10 volt, I don't see how they could contribute
to the varying readings you are measuring.  I think I would put a short on
the input of the 3458A and manually set the range to 1 volt and then observe
the variations that way without the 732A involved.  When I do this I see a
variation from low reading to high reading of 0.125 uVolts and then another
40 I get 0.155 uVolts.  This is without the GUARD connected to the low side
of the measurment terminals, GUARD connected doesn't seem to affect the
readings.  So that is the base noise of the 3458A without the 732A,
somewhere below .2uVolts.  When hooked up to the 732A 1.0 volt output I got
a variation of 0.159 uVolts using the same 40 reading method above.  I would
use this to determine where your problem might exist.  Just having the meter
input shorted will point you in the right direction.  Meter, cables or 732A.

Sorry for the long dissertation.  Friends get mad at me for being so
detailed sometimes.

Bill

- Original Message - 
From: Randy Evans randyevans2...@gmail.com
To: Discussion of precise voltage measurement volt-nuts@febo.com
Sent: Sunday, August 24, 2014 7:22 AM
Subject: Re: [volt-nuts] 732A and Prologix received


 Bill,

 I have convinced myself that the problem I an seeing is due to thermals.
  If I move the cables (with gold-plated banana plugs) using a small towel
 rather than letting my hand touch the plugs, it is much more stable.  If I
 then hold the banana plug with my hand after the reading has stabilized,
 the reading drifts rapidly upward.  I am trying to check the stability of
 the reading but I haven't figured out the MATH function yet.  I assume
this
 is a programmed function using GPIB only?

 The stability I am seeing by manually recording readings (using NLPC of
100
 and 1000) is much greater than what you are measuring on your system.  Not
 sure how to ascertain if it's the 3458A or the 732.  The value of the
 readings are very different between the two - the 3458 reads about 50 uV
 high on the 10 V output and about 12 uV low on the 1V output.  Rather
large
 differences (this is after an ACAL).  I need to find some better cables to
 make sure the errors are not due to thermals again.

 Randy


 On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold wpgold3...@att.net wrote:

  Randy:
 
  I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
  perfect.  Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and
  others
  that are in this size and package.  Order from one of the usual
electronics
  distributors like Allied, Mouser, Digikey.  This is a very common
battery
  as
  it is used

Re: [volt-nuts] 732A and Prologix received

2014-08-24 Thread Randy Evans
Bill,

Thanks for the MATH overview.  That could be very helpful.  I definately
will give it a try.

Randy


On Sun, Aug 24, 2014 at 2:44 PM, Bill Gold wpgold3...@att.net wrote:

 Todd:

 I check the batteries every month per the 732A manual.  Usually I keep
 a
 few used cells from previous packs around that seem to be ok after checking
 them for capacity just like checking the battery pack.  I can usually spot
 cells going bad with that method.  If I don't have a spare set then I just
 substitute a used cell until I receive a new set.

 I agree with you that the cheap ones are crap.  I have used Power Sonic
 a lot with some a few EaglePicher and one set of Enersys.  The EaglePicher
 are crap so I will never use them.  EaglePicher have had a couple out of 8
 that have failed very early.  I have never tried Panasonic.  So I stick
 with
 the Power Sonic's.  I am not sure why the 12 volt ones would be better than
 the 6 volts because 12 volt is just 6 lead acid cells while the 6volt are 3
 lead acid cells.  If one of the 6 cells in the 12 volt battery go bad then
 the whole battery has to be changed.  So you better buy good quality 12
 volt
 batteries.

 I just can't seem to source batteries locally easily at a reasonable
 price.  So I order them from Allied usually.

 Bill

 - Original Message -
 From: Todd Micallef tmical...@gmail.com
 To: Discussion of precise voltage measurement volt-nuts@febo.com
 Sent: Sunday, August 24, 2014 8:13 AM
 Subject: Re: [volt-nuts] 732A and Prologix received


  Bill,
 
  I use a lot of the 12v 7ah batteries for my UPS backups and 732B. But
 yes,
  I was mistaken. The 12V 5ah batteries are the ones that I am using in one
  of my 732A. There is more play inside the tray with the 12v batteries by
  several mm as compared to the 6v which only has a 2-3mm. The 12v
 conversion
  is not difficult, but it is easier if the battery tray has already been
  machined for the 12v batteries. The battery tabs don't line up well with
  the existing holes and need to be widened. Two additional holes must also
  be added.
 
  Sourcing the batteries locally is more of a convenience than waiting for
  the delivery. The el-cheapo 6v batteries are a waste of money and I have
 8
  batteries so far to prove it. They died the first time I had a 6hr outage
  and would not hold a charge after that.
 
  Todd
 
 
 
 
  On Sat, Aug 23, 2014 at 1:21 PM, Bill Gold wpgold3...@att.net wrote:
 
   Randy:
  
   I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
   perfect.  Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and
   others
   that are in this size and package.  Order from one of the usual
 electronics
   distributors like Allied, Mouser, Digikey.  This is a very common
 battery
   as
   it is used in a lot of EXIT signs so they are lighted when the power
 goes
   out.  I don't see how the 12V 7AH will fit as they are too large.  I
 guess
   you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt 4
 AH
   but the terminals are in the wrong place so you will have to nibble
 out
   the aluminum plate that holds them in the 732A battery pack.  You have
 to
   be
   careful if you use the 12v 5AH as you will have 4 extra battery
 connection
   leads to deal with and connect correctly.  I would stick with the 6V
 4AH.
   New batteries will last around 12 to 14 hours before the CAL light
 goes
   out when AC power is not applied.  So shipping to Cal Lab can be a
 problem
   if it is a distance away, or you have to use a shipper like UPS or
 FEDEX
   and
   you ship the night before and then use their Morning delivery and the
 Cal
   Lab is expecting your 732A.  Same on the way back to you.  Of course
 you
   could always strap another battery on the 732A and hook it up to the
 ext
   power plug to last longer.  I have seen it done.  The issue is to get
 the
   Cal Lab to charge the extra battery before they ship the 732A back to
 you.
  
   When you remove and work on the battery pack always have the AC
 power
   plugged in.  The CAL led will stay on because the 18.6 v regulated
 supply
   is working.
  
   The CAL light is to indicate that power has not been lost to the
   Reference Amp or other associated circuits.  When the raw supply
 (battery)
   voltage drops below about 21 volts the CAL light will go out.  Below
 that
   voltage the heater circuits will not work correctly and the 18.6 volt
   regulated supply will not regulate.  The requirement is that the
 Reference
   Amp be kept alive at all times to maintain the output voltage that
 was
   measured at the time of the most recent Calibration or Certification.
 When
   the semiconductor junctions are unbiased and cool off when power is
 lost,
   and then power is restored the result will be a different 10 volts than
   before the power failure.  My experience is that after all of the years
   that
   these units have been powered up, this won't happen and when power is
 lost

Re: [volt-nuts] 732A and Prologix received

2014-08-24 Thread Randy Evans
Bill,

I am trying to figure out the MATH function without much success.  The 3458


On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold wpgold3...@att.net wrote:

 Randy:

 The MATH function is accessible from the keypad.  I don't have an IEEE
 interface right now that works.  You can also program the numeric keypad
 keys to have preprogrammed functions.  DEFKEY

 I have made my own low thermal measurement leads from Pomona #4892
 banana plugs and Belden #9272 wire.  Why 9272, because it was handy at the
 time.  It is tin plated copper, shielded twisted pair 20 ga.  I have plans
 to do custom cables with 16 ga. bare copper wire that I will twist and then
 put a braided shield over it.  I simply cannot find what I want so I will
 build my own cable.  I have done something like this before and it worked
 fine.  When I get a round toit.

 I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have
 used
 in the past to verify my homemade low thermal leads as described above.
 Frankly I cannot see any difference between using the 1756 cables and my
 homemade cables once I give them a few minutes for the thermals to go away.
 As far as I can tell and measure the differences, if any, are below 0.1 ppm
 at 10 volts.

 Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are all
 adjustable you may be seeing a misadjusted 1 volt from the 732A.  As far as
 the instability of the readings it is hard to determine which is causing
 the
 problem.  I have programed (DEFKEY) a numeric keypad key #1 with the
 following code.  MATH 14;NRDGS 40;TRIG 4;TRIG ;   So what this does is
 set
 the MATH to Statistics (store high reading/low reading/ and mean of the
 readings) in the registers, the number of readings to 40, the trigger to
 hold (which keeps the meter from triggering until I press ENTER and
 then
 trigger the sequence of 40 readings when I push the ENTER button.  You
 can
 do all of this manually from the keypads but since I use this sequence a
 lot
 I have preprogrammed it.  This is after I set digits to 8 and PLC to 100.
 Once those 40 readings are finished then you can access the various MATH
 statistic registers, using the menu, by entering MATH and then a 2 for
 low, a 4 for mean, and 13 for high.  Of course you could do all of this
 through the IEEE also.  The 3458A has a very rich set of measurement
 commands.  I am still learning all of them.  It depends upon what I am
 trying to accomplish.

 Since the 1.018 and 1.0 volt outputs are passive and derived from
 resistive dividers from the 10 volt, I don't see how they could contribute
 to the varying readings you are measuring.  I think I would put a short on
 the input of the 3458A and manually set the range to 1 volt and then
 observe
 the variations that way without the 732A involved.  When I do this I see a
 variation from low reading to high reading of 0.125 uVolts and then another
 40 I get 0.155 uVolts.  This is without the GUARD connected to the low side
 of the measurment terminals, GUARD connected doesn't seem to affect the
 readings.  So that is the base noise of the 3458A without the 732A,
 somewhere below .2uVolts.  When hooked up to the 732A 1.0 volt output I got
 a variation of 0.159 uVolts using the same 40 reading method above.  I
 would
 use this to determine where your problem might exist.  Just having the
 meter
 input shorted will point you in the right direction.  Meter, cables or
 732A.

 Sorry for the long dissertation.  Friends get mad at me for being so
 detailed sometimes.

 Bill

 - Original Message -
 From: Randy Evans randyevans2...@gmail.com
 To: Discussion of precise voltage measurement volt-nuts@febo.com
 Sent: Sunday, August 24, 2014 7:22 AM
 Subject: Re: [volt-nuts] 732A and Prologix received


  Bill,
 
  I have convinced myself that the problem I an seeing is due to thermals.
   If I move the cables (with gold-plated banana plugs) using a small towel
  rather than letting my hand touch the plugs, it is much more stable.  If
 I
  then hold the banana plug with my hand after the reading has stabilized,
  the reading drifts rapidly upward.  I am trying to check the stability of
  the reading but I haven't figured out the MATH function yet.  I assume
 this
  is a programmed function using GPIB only?
 
  The stability I am seeing by manually recording readings (using NLPC of
 100
  and 1000) is much greater than what you are measuring on your system.
 Not
  sure how to ascertain if it's the 3458A or the 732.  The value of the
  readings are very different between the two - the 3458 reads about 50 uV
  high on the 10 V output and about 12 uV low on the 1V output.  Rather
 large
  differences (this is after an ACAL).  I need to find some better cables
 to
  make sure the errors are not due to thermals again.
 
  Randy
 
 
  On Sat, Aug 23, 2014 at 10:21 AM, Bill Gold wpgold3...@att.net wrote:
 
   Randy:
  
   I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
   perfect

Re: [volt-nuts] 732A and Prologix received

2014-08-24 Thread Randy Evans
Bill,

I am trying to figure out the MATH function without much success.  I input
the sequence you said (I looked up the instructions to understand what you
did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG; and
it shows up on the display when I input BLUE F1.  I hit ENTER and it takes
the 40 measurements and the MATH symbol shows on the display during the
measurements.  After the SMPL symbol no longer blinks I hit MATH 2 and I
get a MATH ERR symbol on the display.  I tried it a couple of times and the
same result so I am doing something wrong.  Is there a better source for
explaining how to do front panel masurements than the User Guide, which
seems oriented at programming automatic rather than manual measurements.

Randy



On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold wpgold3...@att.net wrote:

 Randy:

 The MATH function is accessible from the keypad.  I don't have an IEEE
 interface right now that works.  You can also program the numeric keypad
 keys to have preprogrammed functions.  DEFKEY

 I have made my own low thermal measurement leads from Pomona #4892
 banana plugs and Belden #9272 wire.  Why 9272, because it was handy at the
 time.  It is tin plated copper, shielded twisted pair 20 ga.  I have plans
 to do custom cables with 16 ga. bare copper wire that I will twist and then
 put a braided shield over it.  I simply cannot find what I want so I will
 build my own cable.  I have done something like this before and it worked
 fine.  When I get a round toit.

 I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have
 used
 in the past to verify my homemade low thermal leads as described above.
 Frankly I cannot see any difference between using the 1756 cables and my
 homemade cables once I give them a few minutes for the thermals to go away.
 As far as I can tell and measure the differences, if any, are below 0.1 ppm
 at 10 volts.

 Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are all
 adjustable you may be seeing a misadjusted 1 volt from the 732A.  As far as
 the instability of the readings it is hard to determine which is causing
 the
 problem.  I have programed (DEFKEY) a numeric keypad key #1 with the
 following code.  MATH 14;NRDGS 40;TRIG 4;TRIG ;   So what this does is
 set
 the MATH to Statistics (store high reading/low reading/ and mean of the
 readings) in the registers, the number of readings to 40, the trigger to
 hold (which keeps the meter from triggering until I press ENTER and
 then
 trigger the sequence of 40 readings when I push the ENTER button.  You
 can
 do all of this manually from the keypads but since I use this sequence a
 lot
 I have preprogrammed it.  This is after I set digits to 8 and PLC to 100.
 Once those 40 readings are finished then you can access the various MATH
 statistic registers, using the menu, by entering MATH and then a 2 for
 low, a 4 for mean, and 13 for high.  Of course you could do all of this
 through the IEEE also.  The 3458A has a very rich set of measurement
 commands.  I am still learning all of them.  It depends upon what I am
 trying to accomplish.

 Since the 1.018 and 1.0 volt outputs are passive and derived from
 resistive dividers from the 10 volt, I don't see how they could contribute
 to the varying readings you are measuring.  I think I would put a short on
 the input of the 3458A and manually set the range to 1 volt and then
 observe
 the variations that way without the 732A involved.  When I do this I see a
 variation from low reading to high reading of 0.125 uVolts and then another
 40 I get 0.155 uVolts.  This is without the GUARD connected to the low side
 of the measurment terminals, GUARD connected doesn't seem to affect the
 readings.  So that is the base noise of the 3458A without the 732A,
 somewhere below .2uVolts.  When hooked up to the 732A 1.0 volt output I got
 a variation of 0.159 uVolts using the same 40 reading method above.  I
 would
 use this to determine where your problem might exist.  Just having the
 meter
 input shorted will point you in the right direction.  Meter, cables or
 732A.

 Sorry for the long dissertation.  Friends get mad at me for being so
 detailed sometimes.

 Bill

 - Original Message -
 From: Randy Evans randyevans2...@gmail.com
 To: Discussion of precise voltage measurement volt-nuts@febo.com
 Sent: Sunday, August 24, 2014 7:22 AM
 Subject: Re: [volt-nuts] 732A and Prologix received


  Bill,
 
  I have convinced myself that the problem I an seeing is due to thermals.
   If I move the cables (with gold-plated banana plugs) using a small towel
  rather than letting my hand touch the plugs, it is much more stable.  If
 I
  then hold the banana plug with my hand after the reading has stabilized,
  the reading drifts rapidly upward.  I am trying to check the stability of
  the reading but I haven't figured out the MATH function yet.  I assume
 this
  is a programmed function using GPIB only?
 
  The stability I am

Re: [volt-nuts] 732A and Prologix received

2014-08-24 Thread Charles Black

Hi Randy,

You can use the NPLC command to change the A/D converter's'integration 
time. Set it to 1000 if you want full resolution. It sets the hp3458a's NMR.


Charlie


On 8/24/2014 6:04 PM, Randy Evans wrote:

Bill,

I am trying to figure out the MATH function without much success.  I input
the sequence you said (I looked up the instructions to understand what you
did - seems logical), BLUE DEFKEY BLUE F1 MATH 14;NRDGS 40;TRIG 4;TRIG; and
it shows up on the display when I input BLUE F1.  I hit ENTER and it takes
the 40 measurements and the MATH symbol shows on the display during the
measurements.  After the SMPL symbol no longer blinks I hit MATH 2 and I
get a MATH ERR symbol on the display.  I tried it a couple of times and the
same result so I am doing something wrong.  Is there a better source for
explaining how to do front panel masurements than the User Guide, which
seems oriented at programming automatic rather than manual measurements.

Randy



On Sun, Aug 24, 2014 at 4:46 PM, Bill Gold wpgold3...@att.net wrote:


Randy:

 The MATH function is accessible from the keypad.  I don't have an IEEE
interface right now that works.  You can also program the numeric keypad
keys to have preprogrammed functions.  DEFKEY

 I have made my own low thermal measurement leads from Pomona #4892
banana plugs and Belden #9272 wire.  Why 9272, because it was handy at the
time.  It is tin plated copper, shielded twisted pair 20 ga.  I have plans
to do custom cables with 16 ga. bare copper wire that I will twist and then
put a braided shield over it.  I simply cannot find what I want so I will
build my own cable.  I have done something like this before and it worked
fine.  When I get a round toit.

 I have 6 ea. Pomona 1756-48 spade lug low thermal leads that I have
used
in the past to verify my homemade low thermal leads as described above.
Frankly I cannot see any difference between using the 1756 cables and my
homemade cables once I give them a few minutes for the thermals to go away.
As far as I can tell and measure the differences, if any, are below 0.1 ppm
at 10 volts.

 Since the 10 volt, 1.0 volt and 1.018 volt outputs on the 732A are all
adjustable you may be seeing a misadjusted 1 volt from the 732A.  As far as
the instability of the readings it is hard to determine which is causing
the
problem.  I have programed (DEFKEY) a numeric keypad key #1 with the
following code.  MATH 14;NRDGS 40;TRIG 4;TRIG ;   So what this does is
set
the MATH to Statistics (store high reading/low reading/ and mean of the
readings) in the registers, the number of readings to 40, the trigger to
hold (which keeps the meter from triggering until I press ENTER and
then
trigger the sequence of 40 readings when I push the ENTER button.  You
can
do all of this manually from the keypads but since I use this sequence a
lot
I have preprogrammed it.  This is after I set digits to 8 and PLC to 100.
Once those 40 readings are finished then you can access the various MATH
statistic registers, using the menu, by entering MATH and then a 2 for
low, a 4 for mean, and 13 for high.  Of course you could do all of this
through the IEEE also.  The 3458A has a very rich set of measurement
commands.  I am still learning all of them.  It depends upon what I am
trying to accomplish.

 Since the 1.018 and 1.0 volt outputs are passive and derived from
resistive dividers from the 10 volt, I don't see how they could contribute
to the varying readings you are measuring.  I think I would put a short on
the input of the 3458A and manually set the range to 1 volt and then
observe
the variations that way without the 732A involved.  When I do this I see a
variation from low reading to high reading of 0.125 uVolts and then another
40 I get 0.155 uVolts.  This is without the GUARD connected to the low side
of the measurment terminals, GUARD connected doesn't seem to affect the
readings.  So that is the base noise of the 3458A without the 732A,
somewhere below .2uVolts.  When hooked up to the 732A 1.0 volt output I got
a variation of 0.159 uVolts using the same 40 reading method above.  I
would
use this to determine where your problem might exist.  Just having the
meter
input shorted will point you in the right direction.  Meter, cables or
732A.

 Sorry for the long dissertation.  Friends get mad at me for being so
detailed sometimes.

Bill

- Original Message -
From: Randy Evans randyevans2...@gmail.com
To: Discussion of precise voltage measurement volt-nuts@febo.com
Sent: Sunday, August 24, 2014 7:22 AM
Subject: Re: [volt-nuts] 732A and Prologix received



Bill,

I have convinced myself that the problem I an seeing is due to thermals.
  If I move the cables (with gold-plated banana plugs) using a small towel
rather than letting my hand touch the plugs, it is much more stable.  If

I

then hold the banana plug with my hand after the reading has stabilized,
the reading drifts rapidly upward.  I am trying to check the stability

Re: [volt-nuts] 732A and Prologix received

2014-08-24 Thread Orin Eman
On Sun, Aug 24, 2014 at 1:46 PM, Don@True-Cal truecalservi...@gmail.com
wrote:

 Randy  all,

 You have correctly concluded that some (maybe not all) of your measurement
 problem is thermal EMF being added or subtracted in series within your
 measurement interconnect. This thermal EMF is generated at the junction of
 dissimilar metals when accompanied with thermal gradients between the test
 lead and device terminals. You have to eliminate both the dissimilarity of
 the metal junctions as well as minimize the thermal differences. The
 terminals of the 3458A as well as the 732A are Beryllium Copper so you want
 to use the same test lead terminals. Forget the typical Tin plated lugs or
 even Gold plated as both are not Beryllium Copper and constitute dissimilar
 metals. The best solution (as usually the most expensive) is to use a set
 of
 Fluke 5440A-7005 (48) cables. I also have just as good results using the
 much more flexible Pomona 11174A (lugs end always stay connected to the
 732A) or 11058A with more convenient shielded banana plugs. The Fluke cable
 has the added Guard built in but be sure to also use a Guard lead with the
 Pomona cabled. The Guard lead does not need to be low thermal EMF. DIY
 cables is usually not a good idea because the lead wire to terminal also
 constitutes just as critical of junction. The above cables use Tellurium
 Copper wire which is usually hard to find and hard to crimp properly and
 NEVER solder.



11058A and 11174A are discontinued at Keysight.  However, Pomona 5295 spade
to banana cables are available (5295-36 at Mouser et al) and claim that
they are designed to minimize thermal EMFs.  Datasheet is here:
http://www.mouser.com/ds/2/159/d5295_1_01-51722.pdf  Any comments on these
as an alternative?

Orin.
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Re: [volt-nuts] 732A and Prologix received

2014-08-23 Thread acbern
Randy,

I have replaced batteries in two units so far, which I bought on ebay (6V) and 
they fitted absolutelly perfect, no need to rework an metal and there was o 
isue with shortages. So there certainly are standard 6V batteries that fit 
without mods. I always try to keep things in original config, if possible.
Also, you do not want to adjust the 732a trimmers. It only gets unstable, if 
you are unlucky. I discussed this with some cal labs, same answer. What you 
could do is to adjust the jumper settings for coarse adjustment, if need be.
Adrian



 Gesendet: Samstag, 23. August 2014 um 04:03 Uhr
 Von: Randy Evans randyevans2...@gmail.com
 An: Discussion of precise voltage measurement volt-nuts@febo.com
 Betreff: Re: [volt-nuts] 732A and Prologix received

 Todd,
 
 Thanks for the info.  I have several Panasonic 12V 7 AH batteries that I
 keep topped off and they have very low current draw (~2 to 3 mA at 13.5
 VDC) when charged and at their float voltage, so I am pretty sure they are
 in good condition.  I will look at getting those in the units after I
 ascertain the condition of the 732.
 
 So now I have a what appears to be a functioning 3458A and a 732A but they
 slightly disagree.  I am like the man with two watches that disagree on the
 time  - which is correct?  For the moment, i am only concerned with
 stability.  The need for absolute accuracy will come later.
 
 Randy
 
 
 On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef tmical...@gmail.com wrote:
 
  Randy,
 
  You have two possible choices. It can be configured with 4 x 6v 4Ah
  batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner has
  modified the battery pack already. A couple of mine needed a nibbler tool
  to remove enough of the aluminum cover that fits over the tops of the
  batteries. The original cover will short out to the battery tabs regardless
  of the battery configuration if this is not done.
 
  You can find larger capacity batteries that will give you slightly more
  battery life. I lost a couple sets of mail-order batteries after a few
  extended outages. I would recommend going with locally bought batteries
  instead of the cheaper mail order. My local Batteries Plus will typically
  have some warranty if I remember correctly. Moving forward I will only use
  2 12v batteries and pre-charge them on a battery charger to equalize them
  before putting them in the 732A. I think the cheap batteries did not
  discharge equally, and would not recover when power was applied.
 
  Inspect the back plane for damaged traces and look at the capacitors. I had
  a few that looked questionable. So far, I have replaced all the big caps on
  the pre-regulator and regulator boards. My feeling is that once these go
  online, they should run as long as possible between repairs.
 
  The battery charger circuit may need adjusting. I tweaked mine and it
  seemed to work fine.
 
  Todd
 
 
  On Fri, Aug 22, 2014 at 10:42 PM, Randy Evans randyevans2...@gmail.com
  wrote:
 
   I received my Fluke 732A today.  Just powered it up but it needs new
   batteries.  Any suggestions for sources (I haven't opened up the unit
  yet -
   I want to make sure it works before doing that).  Also received the
   ProLogix USB-GPIB adapter.  I plan on using Mark Sims' CAL ran data
  dumper
   program to get the CAL data from my 3458A.  Should be a busy weekend.
  
   Randy
   ___
   volt-nuts mailing list -- volt-nuts@febo.com
   To unsubscribe, go to
   https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
   and follow the instructions there.
  
  ___
  volt-nuts mailing list -- volt-nuts@febo.com
  To unsubscribe, go to
  https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
  and follow the instructions there.
 
 ___
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.
 
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] 732A and Prologix received

2014-08-23 Thread Bill Gold
Randy:

I get 6 volt 4 amp/hr (or 4.5 amp/hr) batteries and they will fit
perfect.  Power Sonic PS-640, Genesis NP4-6, Panasonic LC-R064R5C and others
that are in this size and package.  Order from one of the usual electronics
distributors like Allied, Mouser, Digikey.  This is a very common battery as
it is used in a lot of EXIT signs so they are lighted when the power goes
out.  I don't see how the 12V 7AH will fit as they are too large.  I guess
you could use a 12V 5AH (PS1250) as it is the same size as 2 x 6 volt 4 AH
but the terminals are in the wrong place so you will have to nibble out
the aluminum plate that holds them in the 732A battery pack.  You have to be
careful if you use the 12v 5AH as you will have 4 extra battery connection
leads to deal with and connect correctly.  I would stick with the 6V 4AH.
New batteries will last around 12 to 14 hours before the CAL light goes
out when AC power is not applied.  So shipping to Cal Lab can be a problem
if it is a distance away, or you have to use a shipper like UPS or FEDEX and
you ship the night before and then use their Morning delivery and the Cal
Lab is expecting your 732A.  Same on the way back to you.  Of course you
could always strap another battery on the 732A and hook it up to the ext
power plug to last longer.  I have seen it done.  The issue is to get the
Cal Lab to charge the extra battery before they ship the 732A back to you.

When you remove and work on the battery pack always have the AC power
plugged in.  The CAL led will stay on because the 18.6 v regulated supply
is working.

The CAL light is to indicate that power has not been lost to the
Reference Amp or other associated circuits.  When the raw supply (battery)
voltage drops below about 21 volts the CAL light will go out.  Below that
voltage the heater circuits will not work correctly and the 18.6 volt
regulated supply will not regulate.  The requirement is that the Reference
Amp be kept alive at all times to maintain the output voltage that was
measured at the time of the most recent Calibration or Certification.  When
the semiconductor junctions are unbiased and cool off when power is lost,
and then power is restored the result will be a different 10 volts than
before the power failure.  My experience is that after all of the years that
these units have been powered up, this won't happen and when power is lost
and then restored, even months later, the 732A will come back to almost
exactly the same 10 volts as when they lost power, usually with in 0.2 PPM
after 24 hours of warm up.

What type of hookup leads are you using when measuring the 1 volt
output?  What is the PLC set to?  I always use 100 PLC to measure this.  If
you don't have low thermal connection leads you can experience uV changes
for a minute or more after plugging in the leads due to the thermals
generated because of the difference in temperature between the banana jacks
on the 732A and the banana plugs of the leads.  I have found that even just
plugging in the lead will generate a thermal difference because of
difference of temps and some heating due to the physical act of just
inserting the plug because of friction between the jack and plug (my theory
at any rate).  You have to allow at least a minute or more before being able
to make a measurement after plugging in the leads.  I just measured the
variation of the 1 volt output of my 732A and using my 3458A and I got a
total difference of  0.159 uV over 40 measurements using 100 PLC on the 1
volt range of the 3458A.  Using the MATH function and all of the data you
can collect.  That was after waiting for several minutes after plugging in
the leads.

I hope all of this helps.

Bill



- Original Message - 
From: Randy Evans randyevans2...@gmail.com
To: Discussion of precise voltage measurement volt-nuts@febo.com
Sent: Friday, August 22, 2014 9:03 PM
Subject: Re: [volt-nuts] 732A and Prologix received


 Todd,

 Thanks for the info.  I have several Panasonic 12V 7 AH batteries that I
 keep topped off and they have very low current draw (~2 to 3 mA at 13.5
 VDC) when charged and at their float voltage, so I am pretty sure they are
 in good condition.  I will look at getting those in the units after I
 ascertain the condition of the 732.

 So now I have a what appears to be a functioning 3458A and a 732A but they
 slightly disagree.  I am like the man with two watches that disagree on
the
 time  - which is correct?  For the moment, i am only concerned with
 stability.  The need for absolute accuracy will come later.

 Randy


 On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef tmical...@gmail.com
wrote:

  Randy,
 
  You have two possible choices. It can be configured with 4 x 6v 4Ah
  batteries or 2 x 12v 7Ah batteries. Hopefully the previous owner has
  modified the battery pack already. A couple of mine needed a nibbler
tool
  to remove enough of the aluminum cover that fits over the tops of the
  batteries. The original cover will short