hi
the EPROMs are in sockets, no soldering needed.
but again, buying a precision instrument but reprogramming cal data that is 
years old does not make any sense.
unless of course if you are just a collector and do not use its accuracy.
adrian


 Gesendet: Dienstag, 07. Oktober 2014 um 16:35 Uhr
 Von: cfo xne...@luna.dyndns.dk
 An: volt-nuts@febo.com
 Betreff: Re: [volt-nuts] Old HP3458A - SN: 2823A 03939

 I'm now the proud owner of a HP 3458A OPT-002.
 
 I have adapted Mark's 3458A CalRam program to use linux-gpib , and have 
 made backups of the calram , and the dataram.
 
 I will get a new calram : 
 http://tinyurl.com/mggqmos
 
 And 2 x dataram : 
 http://tinyurl.com/otspeek
 
 
 I'll  get 4 x 32k SRAM for opt 001 - hopefully these are correct. 
 http://tinyurl.com/kjbtcla
 
 
 Does anyone know if the Eproms are already in sockets ?
 
 I was planning on installing sockets for the calram , datarams , but have 
 hoped the eeproms was already in sockets.
 
 
 TIA
 CFO - Denmark
 
 ___
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.
 
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Old HP3458A - SN: 2823A 03939

2014-10-08 Thread Mike S

On 10/8/2014 4:23 AM, acb...@gmx.de wrote:

the EPROMs are in sockets, no soldering needed.
but again, buying a precision instrument but reprogramming cal data that is 
years old does not make any sense.
unless of course if you are just a collector and do not use its accuracy.


It makes perfect sense, for the same reason that HP doesn't touch the 
cal if it's in spec - for tracking/characterization. By keeping the same 
cal constants, if and when he does send it in for calibration he'll be 
able to know how much it drifted since it was last cal'd (25 years ago?).


--
Mike
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Old HP3458A - SN: 2823A 03939

2014-10-08 Thread acbern
well, I do that with my references if they cannot be adjusted reasonably such 
as the 732a. but for meters and calibrators I do not. 
achievieng high precison with such an instrument then means you still need to 
characterize it (i.e. calibrate, and not adjust it, doesnt save money 
essentially)
and secondly you always need to apply a transfer factor, i am too lazy for 
that, i rather get it adjusted when i cal it.
also, the reference is stable, true, but other things do drift, and we talk 10+ 
yeras here (last cal sticker i saw). it would certainly be interesting to know 
where its different ranges went inbetween (if it was known that it was adjusted 
to within spec back then, which probably is not known)


 Gesendet: Mittwoch, 08. Oktober 2014 um 10:32 Uhr
 Von: Mike S mi...@flatsurface.com
 An: volt-nuts@febo.com
 Betreff: Re: [volt-nuts] Old HP3458A - SN: 2823A 03939

 On 10/8/2014 4:23 AM, acb...@gmx.de wrote:
  the EPROMs are in sockets, no soldering needed.
  but again, buying a precision instrument but reprogramming cal data that is 
  years old does not make any sense.
  unless of course if you are just a collector and do not use its accuracy.
 
 It makes perfect sense, for the same reason that HP doesn't touch the 
 cal if it's in spec - for tracking/characterization. By keeping the same 
 cal constants, if and when he does send it in for calibration he'll be 
 able to know how much it drifted since it was last cal'd (25 years ago?).
 
 -- 
 Mike
 ___
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.
 
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Old HP3458A - SN: 2823A 03939

2014-10-07 Thread cfo
I'm now the proud owner of a HP 3458A OPT-002.

I have adapted Mark's 3458A CalRam program to use linux-gpib , and have 
made backups of the calram , and the dataram.

I will get a new calram : 
http://tinyurl.com/mggqmos

And 2 x dataram : 
http://tinyurl.com/otspeek


I'll  get 4 x 32k SRAM for opt 001 - hopefully these are correct. 
http://tinyurl.com/kjbtcla


Does anyone know if the Eproms are already in sockets ?

I was planning on installing sockets for the calram , datarams , but have 
hoped the eeproms was already in sockets.


TIA
CFO - Denmark

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Old HP3458A - SN: 2823A 03939

2014-09-24 Thread Poul-Henning Kamp

In message 006401cfd782$031657f0$094307d0$@att.net, J. L. Trantham writes:

I don't know where the '1' firmware resides.

In an 8051 on the isolated side of things.

As far as I know, there are only a '1' and '2' version of that,
but I don't have access to the '2' version, so I don't know what
the difference might be.



-- 
Poul-Henning Kamp   | UNIX since Zilog Zeus 3.20
p...@freebsd.org | TCP/IP since RFC 956
FreeBSD committer   | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Old HP3458A - SN: 2823A 03939

2014-09-23 Thread cfo
On Sun, 21 Sep 2014 15:22:14 +, cfo wrote:

 Gents
 
 I'm looking for a 3458A , and have found one : SN: 2823A 03939 w. opt
 002 Seller says it's passing selftest.

The seller has returned with this info : 

 hello again too,
 rev : 5,1
 i think CAL-Ram's have been changed on 1999 but i m not sure.
 In can accept a 30 days warranty no problem.

So it's fw: 5,1

CFO

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Old HP3458A - SN: 2823A 03939

2014-09-23 Thread J. L. Trantham
I suspect it has the six EPROM version of the A5 board, where the '5' firmware 
resides.  The Ver 8 files are 'out there' and it should be a simple matter of 
burning the six chips if you have a chip burner.  

I don't know where the '1' firmware resides.

Joe

-Original Message-
From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of cfo
Sent: Tuesday, September 23, 2014 1:57 PM
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Old HP3458A - SN: 2823A 03939

On Sun, 21 Sep 2014 15:22:14 +, cfo wrote:

 Gents
 
 I'm looking for a 3458A , and have found one : SN: 2823A 03939 w. opt
 002 Seller says it's passing selftest.

The seller has returned with this info : 

 hello again too,
 rev : 5,1
 i think CAL-Ram's have been changed on 1999 but i m not sure.
 In can accept a 30 days warranty no problem.

So it's fw: 5,1

CFO

___
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Old HP3458A - SN: 2823A 03939

2014-09-22 Thread Todd Micallef
I have replaced a fan in one of my 3458A. I used this one ... 

http://www.digikey.com/product-detail/en/622L/381-2266-ND/1131198

I contacted ebm-papst with the original part number. They were not able to pull 
the original specs to offer a newer sub. I happened to notice it was made in 
West Germany. They told me that there was no information available and that the 
one I selected was close based on the label.

It works fine but it seems to be slightly louder than the other two older 
3458A. The original connecter has to be spliced onto the bare wire of the new 
fan.

I agree with keeping the older one and only change it if the fan fails or is 
too loud.


Todd

Sent from my iPad

 On Sep 22, 2014, at 10:02, Frank Stellmach frank.stellm...@freenet.de wrote:
 
 
 Hello,
 
 to update such an old instrument you need the following components:
 
 1EA DS1220AD-150+, about 8$ from maxim/Dallas or distributor
 2EA DS1230Y-150+, about 30$
 6EA 27C512-100, about 20$, STM
 4EA 62256-80, about 10$
 3 precision DIL sockets for nvSRAMs
 
 I assume, the firmware will be the earliest one, (#2 ??)  and also the 
 calibration will be old, but still valid, i.e. the calibration constants in 
 the nv-SRAMs might not be corrupted yet.
 
 You will have to read out the CALRAM content first over HPIB (described here 
 and elsewhere), then (easily) desolder the 3 nv-SRAMs.
 Replace these by the precision DIL sockets.
 Copy the 2kB CALRAM content into the fresh one, and assemble these 3 RAMs.
 The content of the 2x32KB RAMs will be initialized automatically.
 And the calibration, especially for the higher frequency AC measurements, 
 will be conserved for the next 10..20 years.
 Also plug in these 4 SRAMs for cheap opt. 1. (148K memory)
 
 After these updates - and only if you really need a fresh calibration, i.e. 
 if you don't have access to precise 10V and 10kOhm, you may spend these  500€ 
 for calibration by a HP or Fluke lab.
 
 Firmware 9 you may buy from Keysight, but order the single EPROM (4MBit) 
 version for about  30$, or so, but you have to split that content bytewise 
 into the six EPROMs (all FW versions are obviously compatible concerning the 
 different HW versions).
 They take around 200$ for the old 6 EPROMs version only..
 
 Maybe a friendly volt-nuts is able to assist at that point ;-)
 
 
 Yep, the fan is quite loud, but might still work all right, so keep it.
 Nobody has identified a better (less noisier) part yet.
 
 If the plastic filter is intact, you only have to disassemble and clean it 
 carefully, and that's it.
 (That has to be done on regular basis anyhow, so that Ti = (Ta + 15°C) at 
 any time).
 
 Service Note 18-A,  formally applies to a certain batch of agilent 
 instruments only, S/N US28031400 - US28032927, as it is indicated on that 
 note. On that batch, the LTZ1000 was unstable, and the A/D converter caused 
 short-term out-of-spec.
 
 We have discussed that, and it's not clear, whether the first issue might 
 also apply to all LTZ references in general, if they are not powered 
 continuously.. Remember that the LTZ1000 @ 95°C is at least 8 times less 
 stable than at the regular 65°C, and it is hysteretic at that elevated 
 temperature, so the 1yr. stability of 8ppm or even 4ppm can only be 
 guaranteed by pre-selection of the reference module.
 
 But I strongly assume, that this old HP instrument is not affected at all by 
 SN 18A.
 
 After that, the instrument will be as good as new, best case for 70$ in total.
 
 Frank
 
 
 ___
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Old HP3458A - SN: 2823A 03939

2014-09-21 Thread Bill Gold
TIA:

Push blue key then blue R ( the shifted TRIG key) and ENTER.
Then press the downarrow V key 4 times to get to the REV? on the
display.  Then press ENTER and two numbers will be displayed with a comma
in the middle.  The first number is the multimeter's master processor
firmware revision.  The second number is the slave processor's firmware
revision.  This is if the full menu has been set.  If not, see the MENU
command in the Operator's Manual and set it to FULL.

Mine are both 7,2 and I have no problems.  The latest is 9.

Bill

- Original Message - 
From: cfo xne...@luna.dyndns.dk
To: volt-nuts@febo.com
Sent: Sunday, September 21, 2014 8:22 AM
Subject: [volt-nuts] Old HP3458A - SN: 2823A 03939


 Gents

 I'm looking for a 3458A , and have found one : SN: 2823A 03939 w. opt 002
 Seller says it's passing selftest.

 But it's an old serial# , is there any way one can verify that it has an
 up-to date firmware , and maybe via that see if it has has all the needed
 upgrades or ServiceNotes applied ?

 TIA
 CFO
 Denmark

 ___
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Old HP3458A - SN: 2823A 03939

2014-09-21 Thread Randy Evans
I checked my 3458A and it says that the Rev is 8.2.  What can I infer about
the unit?  I have not opened it up yet to check the status of  the NVRAM
since I am waiting for the posidrive tools.

Randy

On Sun, Sep 21, 2014 at 9:20 AM, J. L. Trantham jlt...@att.net wrote:

 Not in front of my meter right now but IIRC, 'Blue Key' (shift), 'R',
 'Down Arrow' until you find Rev? then hit enter.  It will return something
 like 7,1 or 8,2 etc.

 The first digit is the master processor firmware version and the second
 number is the slave processor firmware version.  The master processor
 firmware is located on the A5 Board, the Outguard Controller.  I'm not sure
 where the slave processor firmware is located.  9,2 is the latest AFAIK.
 I've heard that there is not much difference between 8 and 9.

 I don't know how to check for the Service Note Compliance.  Would probably
 have to read each note then explore the meter to see what is installed.

 Good luck.

 Joe

 -Original Message-
 From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of cfo
 Sent: Sunday, September 21, 2014 10:22 AM
 To: volt-nuts@febo.com
 Subject: [volt-nuts] Old HP3458A - SN: 2823A 03939

 Gents

 I'm looking for a 3458A , and have found one : SN: 2823A 03939 w. opt 002
 Seller says it's passing selftest.

 But it's an old serial# , is there any way one can verify that it has an
 up-to date firmware , and maybe via that see if it has has all the needed
 upgrades or ServiceNotes applied ?

 TIA
 CFO
 Denmark

 ___
 volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.

 ___
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Old HP3458A - SN: 2823A 03939

2014-09-21 Thread cfo
Thank you for the info about the fw. revision.

I'm going to ask the seller what it says, when performing those steps.

CFO

-- 
E-mail:xne...@luna.dyndns.dk

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.