Re: [volt-nuts] Wavetek 1281 - Selfcal Module
No problem. I was just looking back at the documents wondering if I missed anything. The transformer solution is interesting and I see that the 8508A has a second low voltage transformer (T1). I have no idea if it performs the same function using a different calibration. On Sat, Oct 31, 2015 at 9:33 AM, Frank Stellmachwrote: > Todd, > > you already have sent me the appropriate SelfCAL circuitry, but I did not > recognize (understand) it as such! > > I also received a hint from another volt-nuts, John, about the complete > documents for the DATRON 1271 on the KO4BB site, which is very similar, and > also contains a SelfCAL module, being a bit different than the one inside > the 1281. > > The schematics for the 1271, anhow reveal how the transformer inside this > module provides different voltage ratios. The transformer there is labelled > +9.93V, +/-7.23V (that's the main references value), 1.83V, 1.00V and 0.19V. > > These ratios were generated with a precision switch mode / chopper > circuit, and as the ratios on a transformer are very stable, you only have > to measure them once during initial calibration. > > Obviously that is good for ppm ratio accuracy in the SelfCAL procedure. > > So thanks again > > Frank > > > > ___ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Wavetek 1281 - Selfcal Module
Ok, I will look at them again. Sent from my iPad > On Oct 30, 2015, at 07:26, Frank Stellmachwrote: > > Hello Todd, > > many thanks for the reference schematics. > > But that's not the SelfCAL module. > > This should be capable of doing precise 10:1 ratio transfers, (I'm guessing), > so recalibrating all ranges and reducing the T.C. You'll find this > improvement (by a factor of nearly two) in the specifications, its key word > is "Enhanced" specification > > Maybe, when you find some time again, please have a deeper look into the > documents. > > Maybe, in the other service manuals, there might be a 'Theory of Operation' > chapter? > > Frank > ___ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Wavetek 1281 - Selfcal Module
I recently picked up a 1281 that is throwing a couple of errors. While researching the problem, I looked for the SelfCAL description and did not see anything conclusive. I will pull out my schematics and see if there are any references to a particular module. I was under the impression they were using the second LTZ1000 reference module for this function but was never certain. I have the full repair manuals of the 1281 from Manuals Plus before they stopped selling. I will look up anything I can find this evening. Todd On Thu, Oct 29, 2015 at 11:59 AM, Frank Stellmach < frank.stellm...@freenet.de> wrote: > Hello, > I'm comparing different "AUTOCAL" techniques, like in the HP3458A, Fluke > 5440A, 5720A, and have recently discussed the 34465A/34470A "ACAL" feature > with Keysight. > > The Datron 1281 had a SelfCal feature, which is briefly described as > relying on a special SelfCal module inside this instrument. This should > enable the instrument to make a complete (?) recalibration possible, > between the annual calibration dates, or during room temperature changes, > maybe comparable to the 3458A ACAL. > > The Fluke 8508A is obviously based upon the Datron 1281, but does not have > this SelfCAL feature & module. > > Now, here's my question: > > Does any volt-nuts here, have an idea, what's hidden behind this 1281 > SelfCAL module, and how its capability is, also regarding its specification > improvement? > Are there any schematics or descriptions available? > > Thanks - Frank > ___ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.