A couple of people asked why the GST/His steps. The only way I can expressed 
this protein is by co-expression. Co-protein is GST tagged and my protein of 
interest is N-term His and C-term StrepTag. Even with the coexpression, there 
is a major degradation product, the reason why for the strep tag, but now it is 
not being removed by the column. I have tried size exclusion (140kDa vs 100kDa 
does not work) and MonoQ (pI are near identical). I have been using HABA for 
regeneration. I will try some of the suggestions to try and rescue it. 
Summaries are presented below. 

A summary of answers:

1)
We use a Streptactin resin from IBA and they recommend regenerating with HABA, 
which is non-destructive and I believe can be used many times (we've done 
greater than 7 times I believe?).  They say their linker is more affected than 
GE's version so NaOH is quite destructive and cannot be run more than ~ 3 times 
with their resin.  Perhaps GE resin is also affected more by NaOH? HABA will 
displace any biotin.

2)
I use 5 ml StrepTrap columns from GE and these columns work almost indefinitely 
with reproducible results if the following trick is used .I pre-clear biotin 
from my lysate with a pinch of avidin (Sigma  #A9275-25MG) and find that this 
works really well. You could try adding a small amount of avidin to your 
protein sample after your Nickel column. After each run,  I wash the column 
with 3 CV 0.5 N NaOH,  10 CV water, 10 CV 1 mM HABA (Sigma H5126-25G) 
containing buffer (pH 8) and finally equilibrate with 10 CV binding buffer. In 
the past, I used to see a gradient of pale orange at the top of the column and 
dark orange at the bottom as the column got older, presumably because of sites 
inactivated by biotin. Surprisingly, this phenomenon went away after avidin 
treatment suggesting that it might be possible to rescue old columns 
inactivated by biotin although I haven't rigorously tested this.

3)
We have been using Strep-Tactin SuperFlow (high capacity) columns more than 100 
times, even when run under preparative conditions with protein extracts from E. 
coli fermentations. Biotin binding to the engineered streptavidin is not 
irreversible but can be cured by extended washing. To check the column status I 
recommend our HABA binding procedure described in: Schmidt, T. G. M. & Skerra, 
A. (2007) The Strep-tag system for one-step purification and high affinity 
detection or capturing of proteins. Nat. Protoc. 2, 1528-1535.

4)
i am not the last authority on this, and really, IBA have in the past been very 
helpful (http://www.iba-go.com) with information. But here my 2 cents on this: 
i would give it a 20 times possibly. you CAN even apply lysate directly and iba 
describes on their pages how you avoid biotin contamination with addition of 
streptavidin to your sample. streptavidin catches all the free biotin, but does 
not bind the strep tag. done it and works real well, too ...

5)
I used the Strep-Tactin resin from IBA Bio TAGnologies, 2-1201-010, and would 
regenerate it after each use with 2 mM HABA (Sigma H5126-25G). I did not 
observe a decrease in affinity over multiple uses. Also, I only used the one 
column and obtained 99% pure protein.