EN 60601-1-2:2001 was just published in the OJ. The CENELEC web site
has not been updated yet.
Best Regards,
Jim
Jim Conrad, Convenor IEC SC62A/MT23
P.O. Box 25
Hamilton, MA 01936 U.S.A.
E-mail: jc...@shore.net
Phone: 978-468-3909
FAX: 978-468-1034
From: owner-emc-p
Do you know if this bug came through the IEEE fire wall? Any more info
available? I went to the F-sure site but was unsure if I should download
the removal tool. Anyone use it yet?
Jim
Jim
-Original Message-
From: owner-emc-p...@majordomo.ieee.org
Do you know if this bug came through the IEEE fire wall? Any more info
available? I went to the F-sure site but was unsure if I should download
the removal tool. Anyone use it yet?
Jim
-Original Message-
From: owner-emc-p...@majordomo.ieee.org
Have you checked 60601-1?
Jim
-Original Message-
From: owner-emc-p...@majordomo.ieee.org
[mailto:owner-emc-p...@majordomo.ieee.org]On Behalf Of Naftali Shani
Sent: Tuesday, November 05, 2002 1:27 PM
To: 'Mandel, Joel'; 'emc-p...@majordomo.ieee.org'
Subject: RE: Medical Devices
Hi Paul,
Take a look at ISO 7637-1 for 12V systems and -2 for 24V vehicles.
Jim
-Original Message-
From: owner-emc-p...@majordomo.ieee.org
[mailto:owner-emc-p...@majordomo.ieee.org]On Behalf Of Denomme, Paul
S.
Sent: Tuesday, October 29, 2002 9:16 AM
To: 'emc-p...@majordomo.ieee.org'
To: Jim Conrad; m.bushn...@ieee.org; emc-p...@majordomo.ieee.org
Cc: Jim Conrad
Subject: Re: RTCA document vs. IEC 60601-1-02
I think you mean 25 dBuV/m, but regardless of the number, it is not
a direct
comparison to CISPR because
a) the DO-160 EUT-antenna separation is 1 meter, not 3 or 10 meters
I'm not sure if DO-199 or 233 has EMC requirements but I am very
interested if you come with any. IEC 62A/MT 23 is in the process of
amending 60601-1-2 for medical equipment used in the aircraft
environment. We have based our requirements on the environment
specified in DO-160. In general,
John Woodgate wrote:
There was a BIG argument between experts over the test signal to be
used
for assessing immunity to mobile phone emissions. CENELEC people
prematurely chose the keyed-carrier signal, but it presents
inconvenience in both generation and measurement. IEC experts (some
from
happen in November.
Best Regards,
Jim
Jim Conrad, Convenor IEC SC62A/MT23
P.O. Box 25
Hamilton, MA 01936 U.S.A.
E-mail: jc...@shore.net
Phone: 978-468-3909
FAX: 978-468-1034
-Original Message-
From: owner-emc-p...@majordomo.ieee.org
[mailto:owner-emc-p...@majordomo.ieee.org
Hi Peter,
I'm not sure what your wrist device does. Does it have electronics
in it with clock frequencies over 9 kHz? Does it transmit a EMF?
If the device is within CISPR limits then you should be OK for
pacemakers. I think they are tested at 100 V/m. Some pacemakers
have a magnetic switch
Alex wrote:
Do you think it is reasonable to specify in my power supply
specification
that the manufacturers SMPS should meet EN55022 Class B -6dB
Yes. I would ask for 6 dB margin since measurement uncertainty for
a typical conducted measurements system is 5.4 dB. You must also
allow for
are in the area of
w=l then use Zo=sqrt(L/C) where L is the inductance per unit length
and C is the capacitance per unit length. Measure these on your HP
impedance or network analyzer. This gets you very close. Good
luck! Give me a call if you would like more info.
Best Regards,
Jim
Jim Conrad
Hi Chris,
I would try a Silver loaded conductive epoxy or use SS self tapping screws.
Jim
-Original Message-
From: owner-emc-p...@majordomo.ieee.org
[mailto:owner-emc-p...@majordomo.ieee.org]On Behalf Of Chris Wells
Sent: Friday, May 31, 2002 10:23 PM
To: EMC-PSTC Discussion Group
Hi Ron,
I think I am using the same part. Is it the isoloop part?
That part is good for 2.5 kV for one minute. We plan to do
humidity soak and dielectric tests to show that the basic
insulation of the part is adequate for 601-1. I have also
talked to the vendor to get the part UL2601 listed.
Hi Al,
I do not know of a guide but it is usually stated in the directive what
placing a product on the market means. It is my understanding that you
can not ship(place on the market) anything into the EC that carries the CE
mark after the date of withdrawal of a standard that is used to show
Hi Kim,
902-928 MHz is an Amateur Radio band in the US and I don't think it is
available for general use. I am sending you an FCC publication entitled
UNDERSTANDING THE FCC REGULATIONS FOR LOW-POWER, NON-LICENSED
TRANSMITTERS. I hope it helps.
Best regards,
Jim
-Original Message-
Amund,
I am curious why you included ISM in your list of industrial locations. ISM
can be either class A or class B. I thought the distinction was made
according to the mains power distribution system; CISPR 11, subclause 4.2.
JIm
-Original Message-
From:
Emissions Limit?
I read in !emc-pstc that Jim Conrad jc...@shore.net wrote (in
ncbbkfbmfdcdcgaccfbaaeimgmaa.jc...@shore.net) about 'Medical Device
Emissions Limit?', on Tue, 12 Feb 2002:
ESU = Electro Surgical Unit - also know as the Bovie knife
As opposed to Bowie. Or the German version thereof
...@majordomo.ieee.org
[mailto:owner-emc-p...@majordomo.ieee.org]On Behalf Of John Woodgate
Sent: Monday, February 11, 2002 10:47 PM
To: emc-p...@majordomo.ieee.org
Subject: Re: Medical Device Emissions Limit?
I read in !emc-pstc that Jim Conrad jc...@shore.net wrote (in
ncbbkfbmfdcdcgaccfbageihgmaa.jc
Hi Kevin,
Hospitals are generally considered Class A. If the device will be used in
the OR, you should also consider ESU testing even thought it is not
specifically mentioned (at this time) in standards you mentioned below.
Your customers in the OR will be real unhappy when the Bovie goes off
Hi Ned,
I think everyone is asking the same question. You have received the correct
interpretation from the FDA. The 2nd edition is effective as of 10-2001 and
if you claim compliance with it on your 510k, the FDA will not come back and
ask for more information regarding EMC. They are however
Ken, I don't recall your original query but it sounds like you were asking
about BCI testing requirements. I can answer for medical equipment. The
2nd ED of 60601-1-2 requires BCI testing of all patient connected cables.
Best regards,
Jim
-Original Message-
From:
Hi Kaz,
I agree with Jon. However there are battery standards that you may consider
meeting if you plan on CE marking the battery so it can be imported and sold
in Europe as an accessory to your device. I just went through this with a
recent medical product I worked on. There are also
of the 2nd ED if you are going
the standards route of compliance to the MDD. I hope this helps.
Jim
-Original Message-
From: Paolo Peruzzi [mailto:paolo.peru...@esaote.com]
Sent: Wednesday, January 09, 2002 2:53 AM
To: Jim Conrad
Cc: emc-p...@majordomo.ieee.org
Subject: RE: EMC for cardio : which
Hi Pierre,
How is the skiing this year?
Regarding your question, you need to look at the system standard, 60601-1-1,
for interconnecting components of a medical electrical system. Generally, a
computer does not meet the enclosure leakage requirements and thus requires
some kind of an isolation
Sam,
You should also consider the radiation may be from multiple sources on the
EUT. Polarization of the source may not be the same at the substitution
antenna. Try 0, 45, and 90 degree polarization of the bi-con. A tuned
dipole would also give you better results as a bi-con has a very high
Hi George,
I don't think the margin makes any difference in applying the 80/80 rule.
If you have a series produced product you must apply the 80/80 rule using a
sample size of not less than 5 and not more than 12 pieces. See Clause 11,
Assessment of conformity of equipment, of CISPR 11: 1999.
We plow the snow away from around the building. Unfortunately the building
is up on a small rise but it does make it easy to get the snow line below
the ground plane.
Jim
-Original Message-
From: owner-emc-p...@majordomo.ieee.org
[mailto:owner-emc-p...@majordomo.ieee.org]On Behalf Of
Cecil,
We design our products for -10 dB limit, accept -8 dB for a lab prototype
pre-production) and allow -6 dB for production units. We feel this is not
conservative based on past history of measurement uncertainty and production
variations.
Jim
-Original Message-
From:
Brian,
For medical use 60601-1-2. They prefer the 9-2001 version.
Jim
-Original Message-
From: owner-emc-p...@majordomo.ieee.org
[mailto:owner-emc-p...@majordomo.ieee.org]On Behalf Of MCA Compliance
Sent: Saturday, December 01, 2001 12:53 PM
To: Emc-Pstc Post
Cc: Brian McAuliffe
Jon,
If you look at Sec. 18.101 Basis and purpose, part 18 only applies to
medical equipment that emits
electromagnetic energy on frequencies within the radio frequency
spectrum in order to prevent harmful interference to authorized radio
communication services. The Ultrasound equipment that I
Jon, Ultrasound equipment does us use RF directly to treat or diagnose
patients. I don't believe Ultrasound equipment falls under part 18.
The FDA is now recommending the use of IEC 60601-1-2; 2001 for EMC.
Best regards,
Jim
-Original Message-
From: owner-emc-p...@majordomo.ieee.org
Just to give you an idea of how we handled this in 60601-1-2, we said: The
time between discharges shall have an initial value of 1 s. Longer time
between discharges may be required in order to be able to distinguish
between a response caused by a single discharge and a response caused by a
Jim Conrad
Conrad EMC Consulting
P.O. Box 25
Hamilton, MA 01936 USA
Phone: 978-468-3909
E-Mail: jc...@shore.net
-Original Message-
From: owner-emc-p...@majordomo.ieee.org
[mailto:owner-emc-p...@majordomo.ieee.org]On Behalf Of Nick Williams
Sent: Tuesday, July 31, 2001 10:02 PM
To: emc-p
Just to add to Tanias comment, all medical electrical equipment is exempt
from part 15.
Jim
-Original Message-
From: owner-emc-p...@majordomo.ieee.org
[mailto:owner-emc-p...@majordomo.ieee.org]On Behalf Of Tania Grant
Sent: Wednesday, August 01, 2001 12:16 AM
To: Doug McKean; EMC-PSTC
Hi Santo,
You didn't say what type of products you are making but for Medical
Electrical Equipment complying with the 2nd ED of 60601-1-2 it will be 8 kV
air and 6 kV contact.
Best regards,
Jim
-Original Message-
From: owner-emc-p...@majordomo.ieee.org
everywhere else. Hope this helps.
Best regards,
Jim
P.S. I hope the main return path for the ISM transmitter is not back
through the power lines!
Jim Conrad
Conrad EMC Consulting
P.O. Box 25
Hamilton, MA 01936 USA
Phone: 978-468-3909
E-Mail: jc...@shore.net
-Original Message-
From
FYI, IEC 60601-1-2: 2ED, EMC for Medical Electrical Equipment requires 3/10
V/m testing to 2.5 GHz!
Best regards,
Jim
-Original Message-
From: owner-emc-p...@majordomo.ieee.org
[mailto:owner-emc-p...@majordomo.ieee.org] On Behalf Of Pettit, Ghery
Sent: Monday, June 18, 2001 11:19
IEC 61000-4-2 (2001-04) Consolidated Edition 1.2
https://domino.iec.ch/webstore/webstore.nsf/all/3391D5E5625356F8C1256A3B001
1096C?OpenDocument was just published by the IEC CO.
Best regards,
Jim
-Original Message-
From: owner-emc-p...@ieee.org [mailto:owner-emc-p...@ieee.org]On
I don't think my first reply went out. This a second try. Sorry Cyril.
Best regards,
Jim
Cyril, Please see my answers to your questions below:
-Original Message-
From: owner-emc-p...@ieee.org [mailto:owner-emc-p...@ieee.org] On Behalf
Of Binnom, Cyril A
Sent: Friday, April
Hi Chris,
I spoke with Agilent in Rockaway a couple of weeks ago and they assured me they
will be addressing A14 with a software update. That is the good news.
However, it will not be ready for another 4 -6 months!!!
Best regards,
Jim
Colgan, Chris wrote on 2/28/01 8:41 am:
Hi all
Does
:
jim_bac...@mail.monarch.com Michael Garretson:
pstc_ad...@garretson.org
For policy questions, send mail to:
Richard Nute: ri...@ieee.org
Jim Conrad
P.O. Box 25
Hamilton, M.A. 01936
Phone: 978-468-3909
attachment: winmail.dat
as if the
product was drawing 15A(not sure of this exact #). But essentially
anything will pass now. You can purchase A14 on the IEC web site at
www.iec.chThere is also a 75 watt exemption. Products less that 75
watts automatically comply.
Best regards,
Jim'
Jim Conrad
P.O. Box 25
Hamilton, MA
Yes, you can use the A14 for the January 1, 2001 compliance. Just be sure
to add 61000-3-2 A14 to your DOC.
Best regards,
Jim
Jim Conrad
P.O. Box 25
Hamilton, MA 01936-0025
USA
jc...@shore.net
Phone #: 978-468-3909
FAX #: 978-468-3909
-Original Message-
From: owner-emc-p
for domestic purposes. However, as you mentioned, these
standard are subject to change and my understanding is that both harmonics
and flicker requirements will apply to all environments in the future. I
hope this helps.
Jim
Jim Conrad
P.O. Box 25
Hamilton, MA 01936-0025
USA
jc...@shore.net
Phone
Hi Bruce,
Your customer is correct. If the device is classified as a Class A
(according to CISPR 11), you do not need to comply with flicker
requirements.
Jim
Jim Conrad
P.O. Box 25
Hamilton, MA 01936-0025
USA
jc...@shore.net
Phone #: 978-468-3909
FAX #: 978-468-3909
-Original
46 matches
Mail list logo