Brian,Jon,

we have done several investigations of medical PC in accordance to IEC60601.
The requirements of IEC60601-1  is different from IEC60950  mainly in
regards of spacing, dielectric testing, leakage and the insulation of the
interfaces. This has to be documented well in the IEC60601 report.  The
report should be in the CB -scheme format. (no CB certificate is required)

The EMC testing should be done to IEC60601-1-2.

There is no problem to declare a medical PC and issue a D . of . C
accordingly.
When you like to sell it as medical PC in Europe, then you will also require
D. of. C for a medical components.

With best regards
Horst Haug

INNOVA Product Service GmbH
Ampferweg 6
Germany 87677 Stottwang
Tel: (49) 8345-952727


Von: owner-emc-p...@majordomo.ieee.org
[mailto:owner-emc-p...@majordomo.ieee.org]Im Auftrag von Jon Griver
Gesendet: Mittwoch, 26. Februar 2003 09:39
An: emc-p...@majordomo.ieee.org
Betreff: Re: D of C for medical P.S.


Brian,

You are right not issuing a D of C to the MDD, as your product is not a
medical device or an
accessory ot a medical device.

On the other hand I see no reason not to issue a D of C to the LVD.
Compliance with a standard in
the list of Harmonised Standards is not the only way to demonstrate
compliance with the LVD. It
does not take very much effort to draw up a comparison table showing that
the requirements for a
power supply in IEC 60601-1 are more stringent than say, IEC 60950 or IEC
61010-1. Using
engineering judgment on this basis, I believe that you can issue a D of C to
the LVD, specifying
IEC 60601-1 as the applied standard. You can present your comparison table
to anyone who questions
this, demonstrating the engineering basis for your declaration.

Regards,

Jon Griver
http://www.601help.com


On 25 Feb 2003 at 13:40, boconn...@t-yuden.com wrote:

>
> Good people of PSTC
> One of our custom (component) SMPS is evaluated/certified to
IEC/EN60601-1. The LVD precludes
> medical stuff, and the MDD is ambiguous on "components" or "materials"
being scoped by the MDD.
> The customer has determined that indicated conformance should be the LVD.
If the unit's D of C
> claims conformance to the LVD, then there is no applicable standard,
within the scope of the LVD
> that can be stated.
> So what is the correct content of the unit's Declaration of Conformity?
Can anyone please cite
> normative OJ guidance?
> thanx much.
> R/S,
> Brian




This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
     majord...@ieee.org
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Ron Pickard:              emc-p...@hypercom.com
     Dave Heald:               davehe...@attbi.com

For policy questions, send mail to:
     Richard Nute:           ri...@ieee.org
     Jim Bacher:             j.bac...@ieee.org

Archive is being moved, we will announce when it is back on-line.
All emc-pstc postings are archived and searchable on the web at:
    http://www.ieeecommunities.org/emc-pstc



This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
     majord...@ieee.org
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Ron Pickard:              emc-p...@hypercom.com
     Dave Heald:               davehe...@attbi.com

For policy questions, send mail to:
     Richard Nute:           ri...@ieee.org
     Jim Bacher:             j.bac...@ieee.org

Archive is being moved, we will announce when it is back on-line.
All emc-pstc postings are archived and searchable on the web at:
    http://www.ieeecommunities.org/emc-pstc

Reply via email to