Joe,
More data will be required before the group would be able to help with a
diagnosis. You say
that each instrument is on it's own line. Are you referring to an ac power
line? Also:

Is each instrument provided with it's own isolated ground receptacle?
Are they networked? If so, is the connection copper or optical?
Are the instruments microprocessor based?
Is there a central device (PC) controlling them?
Are all devices exhibiting these symptoms, or just a few?
Do they shut down while doing something, or when sitting idle?

I would suggest keeping a log of events, with each machine assigned a
number, in order to
create a matrix showing which machine shut down, when it shut down, what
else was
happening, etc. I understand the need to maintain confidentiality of
proprietary information,
but it should be possible to provide some generic data.

Scott Lacey

        -----Original Message-----
        From:   marti...@pebio.com [SMTP:marti...@pebio.com]
        Sent:   Tuesday, September 07, 1999 5:04 PM
        To:     emc-p...@majordomo.ieee.org
        Cc:     marti...@pebio.com
        Subject:        Immunity Problems ??




        EMC Professionals,

        We have 42 of the same instruments installed in one laboratory.
They are each
        installed on their own dedicated line.  On a regular basis, an
instrument will
        shut down during a run.  The instrument that shuts down is different
each time.

        These instruments are laboratory equipment that meet Class A
radiated limits of
        EN 55011
        and meet the immunity requirements of EN 50082-1.

        Any ideas as to the potential cause of the problem and possible ways
to resolve
        the problem would be greatly appreciated.

        Regards

        Joe Martin
        EMC/Product Safety Engineer
        P.E. Biosystems
        marti...@pebio.com



        ---------
        This message is coming from the emc-pstc discussion list.
        To cancel your subscription, send mail to majord...@ieee.org
        with the single line: "unsubscribe emc-pstc" (without the
        quotes).  For help, send mail to ed.pr...@cubic.com,
        jim_bac...@monarch.com, ri...@sdd.hp.com, or
        roger.volgst...@compaq.com (the list administrators).
        

---------
This message is coming from the emc-pstc discussion list.
To cancel your subscription, send mail to majord...@ieee.org
with the single line: "unsubscribe emc-pstc" (without the
quotes).  For help, send mail to ed.pr...@cubic.com,
jim_bac...@monarch.com, ri...@sdd.hp.com, or
roger.volgst...@compaq.com (the list administrators).

Reply via email to