Re: Using GPIB to talk to HP4083A Switching Controller.

2004-04-24 Thread ScottieB
The forums seem to be working okay, I wonder if there is a firewall issue. You can upload the files to ftp.ni.com in the Incoming folder. Zip them up and name them BCLioness.zip and post here when you have done that and I'll take a look at them. Sorry for all the trouble. Scott

Re: Using GPIB to talk to HP4083A Switching Controller.

2004-04-24 Thread ScottieB
Lori, Again, this is a very intriguing problem. When the bus is broken and the offending instrument is on it, do you get the error in MAX that says that it could not find any instruments, or does it simply not tell you anything after you hit Scan for instruments? Looking at your SPY file, I see

Re: Using GPIB to talk to HP4083A Switching Controller.

2004-04-24 Thread BCLioness
It simply doesn't tell me anything, looks blank like it would if there were no instruments to find. Thanks for telling me I can still use the instrument in LabVIEW applications, I tried it and it works, which is the main thing I needed. Thanks again, Lori

Re: Using GPIB to talk to HP4083A Switching Controller.

2004-04-20 Thread BCLioness
Sorry about that, the site has been saying it's down for maintenance when I try to attach the files, I have been checking since Friday and it still isn't working ... is there something else I should try to get you these files?

Re: Using GPIB to talk to HP4083A Switching Controller.

2004-04-18 Thread ScottieB
Hi Lori, Your attachments are all showing 0 bytes, could you try them again? Thanks, Scott B. Applications Engineer National Instruments

Re: Using GPIB to talk to HP4083A Switching Controller.

2004-04-13 Thread BCLioness
4 other instruments are on the bus, but I tried taking all them off and just having the 4083, and it didn't get listed in MA. NI spy tracks the commands sent over the GPIB bus, it shows red color when there is an error. When I look at it after searching for instruments, it shows successful

Re: Using GPIB to talk to HP4083A Switching Controller.

2004-04-09 Thread ScottieB
Hello, Could you give me more details about your GPIB setup? How many instruments do you have on the bus? It sounds like the 4083 is breaking the GPIB bus where no other instruments can communicate on the bus when you have the 4083 connected, is this what you're seeing? What do you mean that

Using GPIB to talk to HP4083A Switching Controller.

2004-04-08 Thread BCLioness
idn? gives the status of the front panel (C 0 0 0 ... ), IBIC shows me the 4083 at address 18 (not used on any other equipment). When I have the 4083 hooked up to the network, none of the equipment is listed on Measurement Automation, although NI Spy shows it sees everything, including the 4083.