understand, it is maybe different here in Germany then, there are a number of
labs with pretty low uncertainties who also support old gear adjustments.
on the other hand, you have to know who you are working with for what gear.
Otherwise you can have unpleasant surprises.
in your case, if you
-a calibration certificate without uncertainsties is totally useless. in is not
even a calibration.
-I have never understood why people are so keen on getting things calibrated at
Keysight.
> Gesendet: Donnerstag, 19. April 2018 um 14:08 Uhr
> Von: "Dr. David Kirkby"
the manual defines a 4 hours warmup time. after tat the meter is supposed to be
within spec.
then the note 18 test can be started (ACAL and reading the cal constant).
then it should be left on for a couple of days before a new acal is done,
ideally as long as the note 18 specifies (7days iirc)
If you are confident that the 732A is stable to 0.1ppm then just run an acal
afer a wek, and determnine what has driifted. Your A/D board, as can be seen
with a drifty cal constant, or, els the A9 ref.
Dont do a checl everyday. this hides information.
> Gesendet: Freitag, 01. Dezember 2017
What you describe (A9 drift) would be the explanation. However question is how
much you see. The A9 should not drift much in the intervals you talk about
(days).
Also, you should run the ACAL to determine the cal constant not several times a
day but with several days inbetween, and then divide
mine, since it was built in 1977, has drifted upwards by 1.6ppm.
There is also a picture showing druft data of several SR104s on eevblog, but
dont have the link handy.
> Gesendet: Mittwoch, 29. November 2017 um 15:25 Uhr
> Von: "Electronics and Books via volt-nuts"
> An:
The question is what accuracy you need.
The classical way to do that (achieving high accuracy) is to apply a known
accurate current (say 10A) and measure the voltage drop accross the rod with a
nanovoltmeter.
As the piece of aluminum is isothermal you should not expect a big
thermovoltage. You
Igor,
I would actually ignore the 1V and 1.018V outputs. You can always determine
their actual value using your 3458A in transfer mode against the 10V. The
purpose of a voltage reference is not to be at exactly the nominal value all
the time (requiring readjustment), but just to be close to
well, certain Rogers materials have an even higher surface resistance than
PTFE, and are the preferred choice in that case.
> Gesendet: Donnerstag, 06. April 2017 um 06:38 Uhr
> Von: "Attila Kinali"
> An: volt-nuts@febo.com
> Betreff: [volt-nuts] PCBs with ceramic substrates
About two years ago I started an exercise to determine what it would take to
build a kind of DIY JJ standard and also looked at certain detail technical
aspects of designing/building what is reasonable doable. Baseline was 10V DC.
It was clear that it would not be cheap, and I also looked at
The value of the HFL is that by spec, it is better. So you can use its improved
values in your uncertainty budget. If you do an upgrade yourself, you cannot
rely on that. So you have to characterize your instrument after that. But you
can as well do that with the standard meter as-is, and
so far so good.
but what do we learn from this? the solder alloy is essentially irrelevant
> Gesendet: Mittwoch, 06. Juli 2016 um 15:42 Uhr
> Von: "Herbert Poetzl"
> An: "Discussion of precise voltage measurement"
> Betreff: Re: [volt-nuts] Thermal
Dave,
great thing, thanks! could you not also post to xdefs or ko4bb?
I guess more popular these days.
cheers
> Gesendet: Dienstag, 14. Juni 2016 um 14:27 Uhr
> Von: "David C. Partridge"
> An: "'Discussion of precise voltage measurement'"
>
the issue of resistor drifts is mainly related to the amplifier (7 to 10V) gain
resistors, the other resistor drifts do have a very limited impact on the
output (see data sheet).
> Gesendet: Freitag, 03. Juni 2016 um 23:47 Uhr
> Von: "Lars Walenius"
> An:
not sure the TI filter circuitry, although very nice, is stable with the
ADA4522.
have not checked, but R2/C2 loop stabilizer may not work with ad4522.
would require analysis/test, but positive result would sure make this a great
reference buffer amp.
if you do, would be nice if you let the
Low frequency noise is very good, but is is high noise at about 800kHz
(chopper) and above (possible artifacts). That should be filtered, depending on
your application.
What did you foresee to get rid of it, or would you accept it? Simple RC filter
as indicated in the manual increases the
well, are you sure that includes the bracket and so? I think I paid more than
that about 1+ years ago.
> Gesendet: Freitag, 18. März 2016 um 03:56 Uhr
> Von: "Pete Lancashire"
> An: "Discussion of precise voltage measurement"
> Betreff:
Charles and group,
another persons opinion:
I guess the reference to the "standards" means those sold on ebay US. If so, I
would think it is a false expectation this would meet 3ppm acc. guaranteed
within a year.
There is a lengthy chat in eevblog about it, in case you are not aware, and
There are many things to factor in. Drift of reference itself over time and
temperature (399 is 1ppm/K worst case alone, over say 10K, i.e. 23C+/-5K).
Drift of the gain setting resistor is critical (assuming you have an amp
generating 10V out of the 399 output voltage). and so on.
Also, you
at keysight directly, e.g. on stock at keysight germany, less than 20 euro.
> Gesendet: Montag, 12. Oktober 2015 um 20:02 Uhr
> Von: starb...@uplink.net
> An: "Discussion of precise voltage measurement"
> Betreff: [volt-nuts] U150 1920-8937 Relay Driver for HP 34401A
>
>
I did a quick and dirty check, using a standard wire DUT adapter with my (not
calibrated) 4275 meter and 2x 100R and 1x 1000pf. According to the guide below,
this should reslt in 10uH.
Checked the parts allone with the meter before, all reasonably close to nominal.
One end of the resistors
Thats a great document, and certainly is a supperior method than doing some DIY
coils, provided very stable resistors and capacitors are used. Resistors is
straightforward. As capacitors, mica glass is recommended. There are some NOS
russian mica high rel capacitors available for little money.
the best is really to use bare copper wires, such as used for wiring bells at
home. does not look fancy, but low EMF.
Re. the low EMF Pomona cables with 4mm plugs, when you insert them in their
(Pomona's) low EMF binding terminals, the spring that activates the safty
protection pushes them out
it is this resistor, in its z-foil version.
they can be ordered from vpg distributors also in small qty (some may have a
min. order value), lead time is usually pretty high.
Gesendet: Montag, 18. Mai 2015 um 02:39 Uhr
Von: Anton Moehammad via volt-nuts volt-nuts@febo.com
An: Discussion of
Hello,
Your last statement actually could be a nice project in one of the well known
blogs...
I have thought about this as well. It should not be too complicated to do that.
The core components the stability are defined through are not that many. ac
current may be an issue though (I have seen
A few things to keep in mind:
- the 720A (and the 752A) are self-calibrating, i.e. you can (easily) calibrate
it yourself before use. The 3458A, using its external artifact self-cal
procedure based on only 10v and 10k, requires a performance verification
therafter (at least every second time,
these instruments are painfull to use, complex measurement procedures and
pretty unstable (if you want to use their ppm resolution). I spend hours and
hours calibrating a set. lots of data available from standards labs.
what I would recommend is a ballantine 1605A. easy to use, very precise,
the 3458A is very well suited as a nullmeter (there e.g. also is an appnote
from Fluke on this). I have checked this against other meters such as the e.g.
the keithley 155 and 34420A, and for my setup (Fluke 732A, Datron 4910 and
others) I have the least problems with noise, common mode issues
I have a smiliar issue, so interesting to read this. Mine is related to older
gear like 4192a and 4275A meters which I am using. All these old meters, if
properly adjusted an calibrated, are very good meters, you can repair them
yourself, and are very cost efficient (what did you pay for our
this would be over-engineered, unless you have a lot of spare time or want to
do this for fun. If you use plastic supports as they were used when these TO
packages were common (these supports should still be available, they surround
all pins and quality plastics were standard) and use any cap
hi
the EPROMs are in sockets, no soldering needed.
but again, buying a precision instrument but reprogramming cal data that is
years old does not make any sense.
unless of course if you are just a collector and do not use its accuracy.
adrian
Gesendet: Dienstag, 07. Oktober 2014 um 16:35 Uhr
well, I do that with my references if they cannot be adjusted reasonably such
as the 732a. but for meters and calibrators I do not.
achievieng high precison with such an instrument then means you still need to
characterize it (i.e. calibrate, and not adjust it, doesnt save money
essentially)
folks,
does anybody know the difference in specs between a 4920M and the standard
4920. No data on the web.
thanks
Gesendet: Sonntag, 31. August 2014 um 23:41 Uhr
Von: Charles Black cbl...@centurytel.net
An: Discussion of precise voltage measurement volt-nuts@febo.com
Betreff: Re:
Resend as it did not go to the group (please also see my question at the end of
my mail):
mitch,
you bring up a good point. I did not know the document, thanks for sharing. it
inspired me to do some measurements, thats why it took a while to respond. the
results were quite surprising to me.
thanks, I would welcome comments on calibration too.
Gesendet: Dienstag, 02. September 2014 um 12:23 Uhr
Von: Stephen Grady grady.st...@gmail.com
An: 'Discussion of precise voltage measurement' volt-nuts@febo.com
Betreff: Re: [volt-nuts] Datron 4920M, AC Voltmeter, for sale on ebay - 1 day
charlie,
you can expand these tvcs by range resistors to work as higher voltage tvcs. if
you do it right (rf-type setup) you can work up to a couple hundred khz with
just minor additional error beyond its error in its orgininal voltage range.
need to calibrate them however. using an attenuator
I do fully agree with you. this has become, to a big extent, an academic
debate. it may be relavant if the resolution needs to be in the nv-range, such
as working with josephson-elements. I guess nobody in this group here is.
to your question, yes, the 34420a and the 2182 nanovolt meters use
what nist means is that a precision meter is not considered a standard. you
always measure against a true standard (732a, esi sr104...). nist does not mean
that as part of doing equipment calibration a 3458a cannot be used as aid. also
keep in mind nist has a different approach than a cal lab,
well, what you do is to measure the stability of the 3458a. 1ppm drift though
sounds much as overall short term average variation if your temperature is
stable and your 3458a is always on. probably the temp is not stable, and that
is what you see, amongst potentially some other smaller drifts
hi randy,
the specified drift of the 3458a over your 38.1 to 40.3 (about 1k) is 1ppm +/-
allone in the 10v range. thats 10uv. in other ranges its worse.
unless your 732a is very bad (very unlikely), you measure mostly the 3458a
temp. drift. 1000nplc and 100 readings average do not make sense in
Dave,
the title is:
Thermal Voltage Converters and Comparator for Very Accurate AC Voltage
Measurements
by E.S.Williams.
Adrian
Gesendet: Montag, 25. August 2014 um 19:32 Uhr
Von: Dave M dgmin...@mediacombb.net
An: Discussion of precise voltage measurement volt-nuts@febo.com
Betreff: Re:
hi randy,
just for curiosity, why doing 100 measurements at nplc 1000. is this to sample
a changing value?
when i am doing 10 measurements from a stable signal at nplc 100 (only there
many subsequent measuremnts with statistics make sense) I am already getting a
stanard deviation below 0.1ppm.
I have used the pomona spades, mainly to interface the low emf pomona banana
cables to binding posts. I have stopped this, reasons being, they are large and
worse, that the pomona spring loaded insulation tube that covers the banana
plug conductor uses such a strong spring that slowly the plug
well, your last point is the issue, how can you have a temperature difference
within a few microns of material in said connections. theory is one thing, but
in reality it does not happen due to the givens of the setup.
therefore in practice it is irrelevant if the wire is silver or gold plated
fred,
generally you raise a good point, I had the same issue of calibrating an ac
voltage to a high level of accuracy. you need this e.g. to validate the
self.cal of a 3458a or other precison stuff like the 8506a0.
what i would recommend to do if you want to keep costs down is:
in a nutshell,
Randy,
I have replaced batteries in two units so far, which I bought on ebay (6V) and
they fitted absolutelly perfect, no need to rework an metal and there was o
isue with shortages. So there certainly are standard 6V batteries that fit
without mods. I always try to keep things in original
Randy,
that is strange. two things come in mind that might be worth checking. first,
is the mains frequency setting of the 3458a set to 60Hz (I assume you are in
the US)? this is important to suppress mains ac disturbing the measurement.
secondly, have you connected guard? tsp cables would be
if you buy a voltage source that is cal'ed to 6ppm you do not end uop with a
factor of 10 (60ppm). the fatcor of 10 is often used to be on the safe side,
but in high percision cals 10 is not achievable anyways. some mil standards
call for 4, but what you should do is to analyze the error
well, re. the resistor, question is how stable it is, not so much absolute
value. if not very good, your 3458a may be more stable that your reference. vpg
hermetic foil 4 wire resistors are very stable (1ppm pa) and are cheap (50usd).
re the wire, you should use twisted shielded pair. there are
I have no personal experience with the 845, however am using the keithley 155
as nullmeter fo precision calibration. I had used the 182-M (a low noise
version of the 182) and other digital gear before such as the 2182, and the 155
performs very well, is easy to use with not much noise and very
yes, charge injection is an issue with all these switches and these also vary,
in other words are somewhat unpredictable. now some of this may be compensated
by a bigger c, but there are natural limits too. so for a production unit to
sell, this would probably be a killer, but also for
and yes, I forgot: only down-dividing of course, so to reach 10V, two LTZ1000
would be needed in series. advantage is that noise statistically is reduced by
factor of about 1.4. formally also applies to drift.
Gesendet: Montag, 21. Juli 2014 um 20:28 Uhr
Von: Bob Smither smit...@c-c-i.com
I have a set of 7 TVCs, from 1 to 100V. having them all calibrated externally
is just too expensive. There is no doubt that having all 7 TVCs calibrated at
NIST or PTB, wherever you are, is much more precise, I will just not spend it
and live with the accuracy I get.
As source I am using a
in a nutshell, what i am doing is that I first establish the dc (+/-) output of
the ref. TVC at nominal and then determine the ac voltage from a precision,
highly linear (datron 4808) ac source that generates that default output
voltage at the key frequencies. that establishes a set of ac
well, some time ago, when I bought my first 3458A, I looked into the data sheet
of the 3458a and I figured that the basic accuracy of their standards is not
terribly good. 2ppm each, you need to add this to all their specs. dont get me
wrong, for most it is really sufficient, but as we are all
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
56 matches