Re: [volt-nuts] Fluke 5200A instability... A resolution...

2015-08-27 Thread John Phillips
The HP 3455A had a much better AC section but it was a pain to cal. HP work hard to make cal easy and settled for a weak AC converter so they could do a closed case cal. The 3455A, 3457A and 3458A all integrate at line frequency and all have a better AC converter. Integration is tied to line fre

Re: [volt-nuts] Fluke 5200A instability... A resolution...

2015-08-27 Thread Chuck Harris
Hi David, I don't mind so much the journey, as I do being disappointed in the performance of the HP3456A's AC section. DC and Ohms, it is a wonderful DVM. AC, not so much. Since it times its integrators to the power line frequency, it is dooming itself to work stably only at those exact freque

Re: [volt-nuts] Fluke 5200A instability... A resolution...

2015-08-27 Thread David C. Partridge
ded to turn on the filter for <400Hz signals. Cheers Dave -Original Message- From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of Chuck Harris Sent: 27 August 2015 17:41 To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Fluke 5200A instability... A resol

Re: [volt-nuts] Fluke 5200A instability... A resolution...

2015-08-27 Thread Chuck Harris