Dear Manichandra,
in addition to Jens' reply.
Your document shows you the response for the different generations of
the sts2. Ideally you should know which one your instruments are, and
use the corresponding response. All plots seem to show an amplitude of
1500 V/m/s at 1 Hz. In Seisan, your best option is to create gse paz
files. To do this you can prepare an ascii file with the input, see
manual, and then use the resp program to generate the response files.
The paz given are for a velocity response, while the GSE format gives a
displacement response, so you need to add one zero.
Also remember that in Seisan, you can use Seisan, GSE, SAC and SEED
ASCII response files.
Regards,
Lars
Your document gives the paz and gain, this should be all you need.
On 14/02/11 11:07, Jens Havskov wrote:
Manichandra
in tghe seisan manual you find a detailed description on how to use
poles and zeros. In the documentation you sent there is also a constat.
if you generate the responsee using the constat and the poles and
zereos, you should get a response curve similar to the one shown in the
figure . i suspect that the sesmometer gain of 1500 v/m/s is not
included so the gain should be 1. in addtion you have o add the gain of
the digitizer which you find somewhere in the documentation, it depends
on the actual setup of you digitizer. you can disregard the digitizer
filters. The digitizer gain should be in the range 100 000 to 1 000 000
counts per volt.
alterantively you can make a simplified response assuming the sts2 to be
a standard seismometer, using the following:
free period 120 s
damping 0.7
generator constant 1500 v/m/s
gain of digitizer, somtehing like 200 000 counts/V, check you spec sheet
jens
On Sat, 12 Feb 2011, Manichandra Sanoujam wrote:
Dear Sir,
Recently we have been using Q330 with STS-2 for our observatories
located in
the Subduction zone of Indo-Myanmar region. We have some difficulties in
preparing the Seisan response files for these system. I am attaching the
information given by Q330 group to me related with the response file.
They
have sent the poles-zeros information to me however I could not get the
value Ao. The STS-2 that we are using is having 9 poles and 4 zeros.
Please help me in this regard.
With warm regards,
Yours sincerely
Dr. Manichandra Sanoujam
Seismological Observatory
Manipur University, Imphal
INDIA
_______________________________________________
seisan mailing list
[email protected]
http://mailman.uib.no/listinfo/seisan
--
Lars Ottemøller
Associate Professor
Department of Earth Science
University of Bergen
Allegt. 41
5007 Bergen
NORWAY
Phone +47 5558 2616
_______________________________________________
seisan mailing list
[email protected]
http://mailman.uib.no/listinfo/seisan