Hi Richard,

500-1500 what? Are these the raw ADC values?

If you are using the auto-ranging function turn it off and manually control
the range until you find the optimum one.

If you are already do this I can't think of anything that you could do in
the code to improve resolution.

If you have the necessary skills in SMT rework, you could probably
substitute one of the resistors on the GSR daughter card for a more
appropriate value. To get details on what might be possible/advisable on
this you should contact [email protected].

Mike


On Wed, Aug 10, 2011 at 2:46 PM, <[email protected]> wrote:

> Hi,
> Currently my GSR values fall inside the 500-1500 range, with a lot of
> quantization happening during smaller changes. So I was wondering if there
> is a way to change the resolution of the GSR module by adjusting either the
> resistance or the code, such that I could get a wider range and use a
> different set of the polynomial coefficients to translate it?
> Thanks,
> Richard
>
> ______________________________**_________________
> Shimmer-users mailing list
> [email protected]
> https://lists.eecs.harvard.**edu/mailman/listinfo/shimmer-**users<https://lists.eecs.harvard.edu/mailman/listinfo/shimmer-users>
>
_______________________________________________
Shimmer-users mailing list
[email protected]
https://lists.eecs.harvard.edu/mailman/listinfo/shimmer-users