Hi all!  I finally had a chance to do some measurements with the
TDS-1 ppm meter from Gerald Perry.

My wife made up some test solutions of several chemicals at different
dilutions. We've got some initial results...

The de-ionized water she used to make up the solution reads zero ppm.
That's a good start!

We measured several concentrations of NaCl (common salt) as follows:

A solution of 600 ppm read 583 ppm for an error of -3%
A solution of 300 ppm read 294 ppm for an error of -2%
A solution of 60 ppm read 63 ppm for an error of +5%

So, obviously, the evaluation unit is well calibrated for salt. The 
standard Gerald used to pre-calibrate the unit was NaCl, so it shows 
that the unit is stable enough to hold the setting.

We measured a couple concentrations of sucrose (table sugar) and the 
meter didn't read accurately at all:

A solution of 500 ppm reads as 1 ppm
A solution of 100 ppm reads as 1 ppm 

The 1 ppm reading is quite steady, as opposed to the alternating 0
and 1 readings in clean water or air. So it barely detects sugar.

She had some copper sulfate (CuSO4) sitting around the lab and made 
up a couple of samples:

A solution of 360 ppm read as 242 ppm for an error of -33%
A solution of 36 ppm read as 35 ppm for an error of -3%

So the results for CuSO4 are inconclusive.

We're thinking of a few other things to try: 

* A more thorough series of dilutions for CuSO4, possibly, to see if
the error is linear or concentration dependent. 

* Testing and re-testing a couple of CS preps at the environmental
lab for direct comparison. 

* Other materials?

The fact the meter won't do sugar may be because it isn't a polar
molecule (it has no charge) when in solution, unlike salt that
disassociates into Na+ and Cl- ions in water. Or the measurement
technique has a molecular weight limit and won't detect that big of
a molecule? Don't know what to say about that yet!

How does it relate to CS? Won't know until further testing and
cross-checking with other methods. I have some CS  measurement
results that I will report in another post... 

The meter is utter simplicity to use, and the readings are stable and
reproduceable. If I can convince myself that its readings correlate
with actual ppm of CS, I'll be entirely satisfied. 

Thank you!

Mike D.

[Mike Devour, Citizen, Patriot, Libertarian]
[[email protected]                       ]
[Speaking only for myself...              ]


--
The silver-list is a moderated forum for discussion of colloidal silver.

To join or quit silver-list or silver-digest send an e-mail message to: 
[email protected]  -or-  [email protected]
with the word subscribe or unsubscribe in the subject: line.

To post, address your message to: [email protected]

List maintainer: Mike Devour <[email protected]>