At 09:22 AM 5/10/2008, Didier Juges wrote... >Most microwave network analyzers have amplitude >resolution of 0.01dB, while their accuracy is just around 1dB in most >cases. > >I have had to argue too many times that a piece of equipment with a >2dB p-p >requirement on flatness was just fine when it measured 2.01dB on the >HP >network analyzer. I would not have gotten in that argument if the data >had >been 1.99dB. Go figure.
I figure you're both wrong. If the measuring instrument is no better than 1 dB absolute, you can't expect to do it - a flat line measurement could represent an actual + or - 1 db, the limit of the spec. _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.