What you can do is calibrate the AD537/sensor(if used)/freq counter as a unit.  
Things like temperature offsets,  chip dissipations,  oscillator shifts,  
component tempcos,  widget wiggles, etc. get compensated for.  Ideally in an 
ocxo you would want the sensor and freq counter in the oven (or at least the 
freq counter reference oscillator).   Hmmm,  could get wickedly interesting if 
you clocked the freq counter with the OCXO output you were controlling.

In one system we used a 20th order polynomial to do the conversion from reading 
to temperature.  After a couple of years in the field (actually under the 
field) they retrieved,  the data was dumped,  and a new calibrations were done. 
 The calibrations had typically shifted less than 100 microdegrees...  and that 
was with a crappy oscillator in the freq counter.

As far as sensitivity...  a lower res unit (around 0.01 degree resolution) 
could easily detect the body heat from people in the lab.  From temperature 
gradients you could reliably count people entering and leaving (a couple of 
particularly ample people counted as two bodies).

----------------------------------------
The AD537 has too much drift and too little sensitivity to reliably 
maintain the oven temperature (at the sensor) to within a few millidegrees.
The internal dissipation of other circuitry within the chip will 
increase the sensor temperature offset from the temperature being measured.
A thermistor bridge is a better idea as it can have high stability and 
sensitivity (about 10x that of an RTD).


_________________________________________________________________
Get more from your digital life.  Find out how.
http://www.windowslive.com/default.html?ocid=TXT_TAGLM_WL_Home2_082008
_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to