WarrenS wrote:
>> Bruce said:
>> The critical requirement is that the 2 standards being compared are 
>> statistically independent.
>> Comparing a pair of Thunderbolts GPSDOs with similar time constants and
>> damping will give optimistic results for Tau comparable with or greater than 
>> the loop time constant.
>> Its is even better is to use 3 or more similar standards simultaneously
>> logging phase differences between the various pairs (0.5*N(N-1) pairs for N 
>> standards).
>> It is then possible to obtain estimates for ADEV, MDEV etc for each standard.
>>     
>
> The optimistic results at and above the loop time constant, that results even 
> when 3 or more units are used, 
> is because the noise is then mostly due to the GPS signal itself and NOT the 
> local oscillators in the GPSDO.
> In effect you are then using the same 1PPS signal into each unit, and any 
> common noise on the 
> GPS 1PPS signal will cancel and not be seen.
> So I think what Bruce is saying is that you can not (or is it should not?) 
> use the GPS signal to 
> measure the GPS's noise.
> But the point is, if you want to compare your GPSDO with different settings, 
> or compare it to 
> another OCXO, It can be done this way, if you do not have a better ref to use.
> You could then add the noise of the GPS nose above the control loop time to 
> your 
> optimistic results if you want true results at high Tau values.
>
> Also note that having the GPS noise cancle is not necessary a bad thing,  It 
> can be a good thing 
> especially if the GPS noise is not what it is that you want to measure. 
>
>   
>> Like all digital phase detectors its best to avoid, if possible, the 
>> nonlinearity inherent at the ends of the range.
>>     
>
> Using a phase detector near its end point (or at its crossover point if there 
> is any deadband) 
> is something that needs to be avoided. 
> The two basic standard ways to insure that just the center of the phase 
> detector's range is use:
> 1) Divide the signals down just enough before sending them to the phase 
> detector so that 
> the end points is not an issue.  This works when both signals are from 
> devices that are 
> locked to a common signal such as the GPS.
>
> 2) When one of signals is from a non locked source such as a OCXO whose phase 
> can drift 
> any amount overtime, One of ways to limit phase detector issues, and use just 
> the very accurate zero phase point, is to use the Phase detector's output to 
> lock the OCXO in a fast control loop and then by knowing the gain of the EFC 
> input, the filtered EFC voltage can be use as freq drift information to find 
> the ADEV's.
>
> WarrenS
>
> *************:
What I am doing to ovoid the "end of range" problem is;
First I divide the signal by two to get a 50% duty cycle. Then when the 
phase difference gets to
10% or 90% of the full scale value I switch the phase detector (or 
counter) to respond to
to the opposite zero crossing. I keep track of those switches in 
software. I use a computer to control things and to keep a log of the 
phase difference.

Bill K7NOM


_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to