Tom Van Baak wrote:
>> I have been able to find schematics of the 10544A but not a service manual.
>> Is there a service manual available?
>>     
>
> http://www.leapsecond.com/museum/10544/
>
> I'll look for something like a service manual. I know I have some
> interesting info in the 10544. When I find it I'll update the page.
>
> One thing you could do is look at hp op/service manuals of
> instruments (e.g., frequency counters) made in the late 70's
> and see how they integrated the 10544 into their design. This
> will give you hints on both input power and output buffering.
>
> I know *late* model 5245L counters used the 10544A oscillator.
> Perhaps others in the group can provide other examples and
> then you can head off to the online manual sites and grab the
> info you need from the PDF files.
>
> /tvb
>
>
>   
Tom

No 5245 manuals on Agilent site nor the agilent ftp site nor BAMA site.

The HP5370A also used a 10544A as an optional high stability timebase OCXO.
The reference buffer PCB is close to the 10544A socket.
The reference buffer consists of a series string of 3x 1/3 MC10216 ECL
line receivers.
The reference input pin uses a 1k resistor in series with 10pF connected
to the input of the first ECL line receiver.
This buffer is somewhat crude, but nevetheless has adequate performance
for its intended application.

Bruce