I just bought an HP 105B Oscillator. It's doing something that may or may not be a problem. I'm hoping that someone can help me out.

It's a newer unit that's based on the 10811 oscillator plus two interface boards to adapt the 10811 to the rest of the 105B circuitry. Does anyone have a soft copy of the manual for this configuration? Mine has a serial # prefix of 2308. I have the manual from Didier's web site, but it's for the older style with a serial # prefix of 808. Or, does anyone have schematics & circuit descriptions for the 5061B Cesium Standard? The more important of the two interface boards is part # 05061-6165. Pictures of the interior of the 5061B show a 10811-style oscillator with two interface boards in the same location that the 5061A had an older-105B style oscillator. It looks like both units had similar upgrades.

My questions concern the performance of the 5 MHz output under different loading situations. The older manual suggests that what I'm seeing isn't normal.

First, all circuit check functions are good (80 +-7) with nothing connected to the outputs.

My first question is whether its okay to use both the front and rear connections simultaneously. The 1 MHz and 100 KHz outputs show the expected drop in level if you have two 50 ohm terminations. The 5 MHz output drops from 76 (3V2 p-p) with a single termination to 60 (2V2 p-p) for the double termination which is fine since the spec is >1V rms i.e. >2V8 p-p. But this drop in level crashes the 1 MHz and 100 KHz signals completely. The older manual includes a stability test with the 5 MHz output shorted and the test equipment connected to the 1 MHz output. This unit would fail that test completely.

Second, if I attach a scope (1 Mohm impedance) via BNC cable to the 5 MHz output, the circuit check level rises from 86 to over 100. The scope shows a clean sine wave of about 4V2 p-p. Even a 10 Mohm probe with a BNC adapter gives a circuit check reading of 96 and the same 4V2 p-p signal. Adding the 1 Mohm scope to the 10 Mohm probe raises the meter from 96 to over 100 and increases the amplitude to about 4V8 p-p. AC or DC coupling has no effect. I checked for ground loops by plugging everything into the same power bar. I ran the 105B on batteries. I also tried a second scope. No change. I repeated the test with a 50 ohm termination on the scope and the circuit check shows 76 with a 3V2 p-p amplitude which is perfect. But adding the 1 Mohm scope to the termination raised the meter to 82 with a 3V4 p-p amplitude. The same tests on either the 1 MHz or 100 KHz outputs make more sense. The high impedance connections don't affect the circuit-check reading and adding a terminator drops the level by about half as measured by either the meter or the oscilloscope.

In summary, should I be able to use both front & rear connectors simultaneously without disrupting the other outputs and does it make sense that high impedance connections cause the 5 MHz level to rise?

Sorry for the rather long message.  Thanks for any help anyone can provide.

Ed