I powered up the second unit. It has been running just fine for an hour now. The only difference I noted is that the initial current was only 1.4 amp vs 1.8 on the other one. However, I set my power supply to 16.0 volts; no other difference. After warmup the current settled at ~700 ma and the 5vdc current at ~80 ma.

On 1/19/2012 1:15 PM, Paul F. Sehorne wrote:
I received my two FE-5680As late yesterday. This morning I hooked one of them up. It locked on in a couple of minutes and at about five minutes quick working. There is a smell of something burnt inside. I have not opened the case to investigate the damage.

I had the unit attached to two current limited bench power supplies. One power supply was set at +16.1 vdc and the other at +5.0 vdc. When the trace on the scope went away I noticed the +15 vdc power supply had gone into current limit mode at 2 amps. The +5 vdc current had dropped from ~80ma to about ~60. Before the failure the +15 vdc supply showed ~1.8 amp during the first minute or two and then dropped to ~700 ma.

While it was working it worked fine and was very near my GPSDO in frequency. The scope trace was moving 5ns in about 20 seconds. I got no further than this when the unit stopped working. I have not powered up the second unit yet; preferring instead to contact nichegeek first.

Paul
_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to