On 4/17/12 7:55 AM, John Lofgren wrote:
One feature of the Agilent and Rohde scopes (maybe Tek, too?) that can help in 
some situations is segmented memory.  It allows you to capture periodic or 
random events with the full sample rate but to ignore all the dead time between 
events.  For each trigger it stores one sweep with a time stamp.  When you want 
to look at the record you can roll back through memory and look at each 
individual event with full resolution.

This isn't a cure-all because the time stamps will have limited resolution and 
some amount of jitter, but it can be helpful in some applications.  It also 
assumes that you know what you're looking for and can trigger on it :)



Yes, this was a tek..it does the same thing (called "fast frame" in their manual) and the trigger time stamps were actually high resolution (higher than the sample rate).