Mike, here is the effect of the A/C cycling on and off during a warm spring day 
on the delay through a piece of RG-8 cable maybe 3 feet long:

http://www.ko4bb.com/Test_Equipment/data/HP10811-raw.png

The full scale (screen height) is about 5x10 to the -10, the length of the 
record was about one hour.

The cable was very short, I did not record the temperature variation but it was 
maybe 2F.

You can imagine what a 100W light bulb (or a warm body) might do if you try to 
measure at 10 to the -12, and/or you have longer cables.

Didier KO4BB

Sent from my BlackBerry Wireless thingy while I do other things...

-----Original Message-----
From: Mike S <mi...@flatsurface.com>
Sender: time-nuts-boun...@febo.com
Date: Tue, 15 May 2012 17:23:41 
To: <time-nuts@febo.com>
Reply-To: Discussion of precise time and frequency measurement
        <time-nuts@febo.com>
Subject: Re: [time-nuts] Why are 1PPS signals so skinny?

On 5/15/2012 5:14 PM, saidj...@aol.com wrote:
> it is the effect of what follows after that leading edge, and  propagates
> down the power supplies to cause side effects that is being discussed  here.

I'm asking "What side effects?" I haven't seen any mentioned. And 
really, if an increase in power draw of 10 watts for an entire lab 
environment causes any problems, I'd question the quality of the power 
supplies, and ask what happens when you simply turn on the light?


_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to