Azelio,

In frequency average mode the 53132A makes something like 200,000 measurements 
per second, according to the manual. This is why the instrument does so well as 
a smoothed frequency counter. The resolution "gain" is some fraction of 
sqrt(200000), subject to the normal caveats like the need for independent 
samples. That's why these counters have reduced resolution when the input 
frequency is too close a multiple/fraction of the timebase.

Details on this technique (used also in the Pendulum CNT-9x counters):

"New frequency counting principle improves resolution"
http://tycho.usno.navy.mil/ptti/ptti2005/paper67.pdf

See also this fine presentation by Staffan Johansson:
http://www.npl.co.uk/upload/pdf/20060209_t-f_johansson_1.pdf

Or, google for: continuous time-stamping frequency counter

/tvb