Greetings,

I have run some time interval test and it did not quite calculate correctly.  I 
used a TADD2 that is configured for 2 outputs at 100 Hz, 2 outputs at 10 Hz, 
and two outputs at 1 PPS.  The HP5370B clock output was used as the input for 
the TADD2.  The output of the TADD2 was feed to both time interval inputs using 
a tee.  I am using a ProLogix GPIB/USB adapter and hyperterminal in windows to 
capture the data.  The object of the test was to make a baseline noise floor 
measurement.

The 5370B manual shows there must be a minimum of 330 microseconds between 
measurements.  If the unit is put in the time interval binary mode via the 
GPIB, the unit can output up to a 6 Khz rate on measurements.  The 5370B manual 
does not state what the instrument can output on the GPIB, in normal use 
(meaning, not in the time interval binary mode - just plain talk mode).

I made some measurements using 100 hz to both inputs and in one hour, the data 
file shows only 51437 samples.  This equates to 14.288 samples per second.

I made some measurements using 10 Hz to both inputs and in one hour, the data 
file shows 35988 samples.  This basically works out to 10 samples per second, 
as expected.

Knowing the results at 100 hertz, how can one tell a program such as Stable32 
or Plotter about the dead time between measurements ?  Or would it be best to 
keep the inputs at a maximum rate of 10 PPS, and say thats the limits in the 
normal TI mode ?

____________________________________________________________
GET FREE SMILEYS FOR YOUR IM & EMAIL - Learn more at 
http://www.inbox.com/smileys
Works with AIM®, MSN® Messenger, Yahoo!® Messenger, ICQ®, Google Talk™ and most 
webmails


_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to