Claude,

On 22/03/14 12:48, Claude Fender wrote:
Hi,

I have two generators 3324A and 33120A and one counter 5334B. All the 
instruments are locked to a GPSDO.
I've  measures a 1 kHz output  of the generators with the counter  set with a 
gate time of 1 second, and repeated the measure 3600 times.

The average for the 3324A is 999.9999996 Hz with a standard deviation of 6.79E-5
The average for the 33120A is 999.9999986 Hz with a standard deviation of 
1.04E-5. (see joined graph)

The resolution of the 3324A is 1 mHz, the resolution of the 33120A is 0.1 mHz

So here is my question ! Do you think if I find a generator with a resolution 
less than 0.1 mHz, the repetability will be better ? If so, do you have any 
advice for some stuff less than 1000 USD ?

Thanks in advance !

The 3325A/B has 1 uHz resolution.
Be careful when making frequency measurements, as some counters can create a small measurement bias. Using TI-mode and TimeLab you can avoid this.

Cheers,
Magnus

_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to