he is on the right truck, just look around in your testing environment,
do you have shielded test set up, common ground for all the test gears ?
On 8/7/2014 3:58 PM, Hal Murray wrote:
b...@evoria.net said:
So, I may throw another cap on it, but it seems to be clean down to what I
can measure at the OCXO on my old Tek 455 with an X10 probe.
Another thing to consider when chasing that sort of problem: How much are you
picking up with your scope probe and/or its ground wire?
The classic home brew better-probe is to use a chunk of 50 ohm coax feeding
into scope with a 50 ohm terminator. (You get 2 of them if you cut a readily
available connectorized cable in half.) Insert 950 ohms at the probe end to
get higher input impedance at the cost of a 20:1 divider.
http://www.signalintegrity.com/Pubs/straight/probes.htm
It usually requires soldering at the DUT. For me, it also requires a mental
shift. You have to think of your scope probe as a consumable rather than
something that won't wear out if you take care of it. A chunk of coax will
last a long time. Just cut off another 1/2 inch when the end gets too
mangled.