Hi,

We have already hacked a DC-DC converter to feed the output board with separate power. It was relatively easy.

Haven't had the time too look at doing the same to the external trimmer.

Cheers,
Magnus


On 09/30/2014 02:09 AM, Björn wrote:
Hi Chris,

Nice work!

I can't help you with the faulty unit. There are others on the list with deeper 
insight in that OSA Cs core. Hope they have time to chime in.

But I do have a telco variant of the same Cs core running in the basement. Its 
running continously (ocxo and ion pump) but its not burning the last fumes of 
Cs in the tube. This is since I don't see access or money available for a 
replacement tube.

As you noted. With ocxo warm and cosy and the ion pump running it does not take 
long for Cs lock when needed.

A mode where you can use the unit as premium ocxo (bva :-) ) standard that can 
be calibrated against the cs tube periodically as needed would suit my 
situation.

However there are (at least) two mods needed. Because unless the cs tube 
burnes, there is no power to the

  1)  output buffer card. And thus no output signal.
  2) potentiometer adjusting the bva frequency. Thus without cs tube running 
the bva is off freq.

This design might have been a safe one in the original market.

But I think modding the unit to let it run mostly as a very stable crystal 
standard is better for hobbyist use.

Have others done these kind of mods to prolong the Cs tube life?

Kind regards,

           Björn



<div>-------- Originalmeddelande --------</div><div>Från: Chris <syseng.greenfi...@btconnect.com> 
</div><div>Datum:2014-09-29  22:43  (GMT+01:00) </div><div>Till: time-nuts@febo.com </div><div>Rubrik: [time-nuts]  
Oscilloquartz 3210 Cesium Standard </div><div>
</div>Another update on progress:
--------------------------
Now at the stage where two of the three 3210's are working. The fault
that was blowing the line fuse turned out to be a shorted reservoir cap.
Managed to find an Ebay seller in Greece with exactly the same voltage,
capacity and outline, though different manufacturer. The caps have a
stud mount at the base of the can, so not easy to find, but bought four
at just over 3 ukp each and replaced both caps in the psu, leaving two
for spares.

 From cold, the spec is up to 90 minutes warmup time, but if the OCXO
and pump have been running for some time, both units lock up within 4
minutes. Have done some setup. For example, normalised the gains and
offsets in the preamps as per the manual and both units show the
expected preamp level of ~165mV and 2nd harmonic amplitude of 9-10 on
the meter, so assume both tubes are in good condition.  None of the
settings were very far out.

Took a set of measurements from one of the good units and the faulty for
comparison, with similar results, Good unit as follows, but faulty unit
more or less the same:

Ion pump = EHT ok, meter = 0
Electron multiplier EHT voltage = 1800 volts, (100Mohm hv probe on o/p
wire)
Cesium oven = Not open circuit, voltage = 5.5 volts
Ioniser = not open circuit, 1 volt p-p square wave at ~26KHz
Synthesiser output = 12.6317715, follows variation in OCXO and in lock
Backplane test points, tpb, tpc = 8.2 volts p-p square wave, ~137 Hz.
Multiplier 12.6... MHz input = 2v p-p (Scope)
Multiplier 137Hz input = 2.25v p-p (Scope)
Multiplier 180MHz output = +26DBm (HP 3406A rf voltmeter + 20DB attenuator)
Microwave tap on waveguide = -14.5DBm, threaded attenuator works
(HP432A, 478A)

The 3406 and 432 are quite old and the sensor head is even older, but
should be reasonably accurate. The 478A head was coupled to the
microwave tap using a ~3 inch length of rigid coax, so not much
attenuation. Hopefully, these figures may help others trying to debug
these standards, but if there's anything i've missed, please let me know.

For the faulty unit, even with all the levels as expected, there is
still no signal, nor 2nd harmonic, even with the preamp gain turned up
to max. Connected a Fluke electrometer in series with the preamp input
socket, but the current is at least 100 times down on 1nA. Looks like
the tube is completely lifed, or has an internal fault. /FX:  Have
visions of a sad, faithful and slowly dying 3210 left for decades in a
rack,  in the dark, powered up, but long forgotten after the accurate
reference it provided ceased to be used anywhere in the organisation. A
ship adrift for thousands of years, but all systems still at least
partially working and waiting to be discovered :-).

So what else am I missing in this puzzle ?...

Regards,

Chris



_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to