Thanks All for your comments and clarifications.

The next OCXOs have been tested today:
        * Morion MV89A 10.000MHz  #1
        * Morion MV89A 10.000MHz  #2
        * Morion MV89A 10.000MHz  #3
        * Isotemp 131-191 10.000MHz #1
        * Pilot C3Z 10.000MHz 
        * Bliley NV26R891 100.000MHz
        * Giacint-M 5.000MHz 

Measurement set up:
        * Pendulum CNT-91 with TimeBase opt.19 and 10
        * Prologix GPIB-USB 
        * TimeLab V1.2 (x64)
        * Windows7 Pro (x64)  
        * PC i5 /3.4GHz / 16GB 

ADEV results 
https://dl.dropboxusercontent.com/u/21338179/hamradio/OCXO_Adev/ocxos_adev_test_20141031.jpg
TimeLab configuration 
https://dl.dropboxusercontent.com/u/21338179/hamradio/OCXO_Adev/timelab_adev_config.jpg
All OCXO's TimeLab TIM files can be download 
https://dl.dropboxusercontent.com/u/21338179/Hamradio/OCXO_Adev/OCXO_Adev.zip

> A good MV89A can be below 1E-12.
As you can see ADEV of all my tested OCXOs are not better than 4E(-10) for tau 
=1s.
Question: can I be more or less happy with this result or something wrong in my 
measurement?
Karen
 
> Date: Thu, 30 Oct 2014 23:32:48 +0100
> From: Adrian <rfn...@arcor.de>
> 
> Karen,
> 
> as by the datasheet, the time interval resolution of your counter is 50 ps.
> This translates into an ADEV value of 5E-11 at tau = 1 sec, 5E-12 at 
> tau = 10 sec etc,  and represents the theoretical measurement limit of the 
> counter.
> It can be measured by feeding the same signal into both inputs (start 
> and
> stop) at the same time. A short delay (additional length of coaxial
> cable) at the stop input keeps the counter from producing phase wraps 
> (jumping between a full input signal period and zero).
> 
> A good MV89A can be below 1E-12. For measuring two good MV89A's 
> against each other, a TI resolution of 50 ps is insufficient because 
> up to 100 sec, to the most part of it, the actual ADEV of the 
> oscillators can be below the measurement limit of the counter.
> 
> Adrian

_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to