> On Dec 11, 2014, at 8:20 PM, d...@irtelemetrics.com wrote:
> 
>    Hi Bob,
> Some numbers. Maybe you can double check my math, just to be sure I'm not 
> getting something completely wrong. That is very possible, since I'm new 
> here... ;) 
> The DAC is moving up and down about 7.5 counts with room temp swings. 20 bit 
> resolution at 6.6volts full scale output.  6.6 volts * 7.5counts / (2^20) =  
> +/- 47uV.  (This is verified as reasonable with a 24bit data logger, as it's 
> seeing about +/-50 uV temp swings on EFC. Resolution of about 1uV.)
> Tuning sensitivity of the oscillator is 1Hz/10Volts. Or 47uV * 1Hz/10V = +/- 
> 4.7uHz. 

or 1x10^-8 per volt. If it’s a 10 MHz OCXO.
That would be 1x10^-14 per uV
4.7 x 10^-13 for 47 uV


> The temp swing is +/- 2 degF with ~45 minute period. So, in the ballpark of 
> your +/- 1 Deg C guess. 

That’s pretty normal for a modern HVAC system.

> +/-4.7uHz / 10e6 Hz oscillator = +/-4.7e-13, or near a 1e-12 full swing over 
> 2.2 Deg C.

1x10^-12 for full swing is about right. 

> (Or, am I completely out to lunch here???) 
> I should qualify, there is aging/retrace here. It's in the range of 3e-11 per 
> day right now, and I took the +/- 7.5counts off of what was left after 
> removing the slope of the aging drift.

That’s a very common (and legit) thing to do. 

> The aging looks huge over a day compared to the thermal cycling. 
> Currently the system has ~2ppm/C reference, and .04nV/C opamps. So, Yeah a 
> little overkill.

What makes you believe that the OCXO’s temperature performance it not the issue?

1x10^-12 per 2 C -> 1x10^-10 over 100 C (say -30 to +70C). That’s a very good 
spec on an OCXO. Also consider that gradients could easily amplify the impact 
by 2X or more. 

> But these things are getting cheap nowadays, so why not? Before the 'good' 
> reference and opamps, there was about 10 times as much swing in the PWM DAC 
> over temp cycles. As you have suggested there is probably some room to 'relax 
> the spec', and still be fine…   

I’d guess that the analog stuff is much better than it needs to be.

Bob

> Thanks,
> Dan
>  
>  
>  
>  
>  
>  
> >
> > Hi
> >
> > If your OCXO has a stability of +/-3x10^-10 over 0 to 60C (numbers picked 
> > to make math easy):
> >
> > 6x10^-10 / 60 = 1x10^-11 per C
> >
> > If the OCXO is 10X better than that (unlike) you are at 1x10^-12/C
> >
> > If the room temperature swings 1 C, you get a 1x10^-11 swing in the OCXO.  >
> > Keep in mind that the OCXO also responds to things like gradients as much 
> > as it does to absolute temperature.  >
> > If your EFC range is 1x10^-7 for 5V (pretty common):
> >
> > Each 1 mv change is 2 x10^-11.  >
> > A 78L05 will hold 1 mv over 1C. Roughly 90% of them will hold that over 
> > 20C. That’s a cheap regulator for 2x10^-12.  >
> > A 10 ppm / C reference will get you to 1x10^-13 / C
> >
> > You don’t *need* an EFC at 1x10^-7. Something 1/10 that size is probably 
> > good enough. Knocking it down to that level is just a couple of resistors. 
> > Way less money than fancy references.  >
> > Bob
> > 
> 
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.

_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to