Hi

> On Jan 18, 2015, at 5:12 PM, Stéphane Rey <steph....@wanadoo.fr> wrote:
> 
> Bonsoir Magnus (Are you in Sweeden ?) 
> 
> Being able to measure high stability and low phase noise is definitely a need 
> for me as I'm trying to design low noise synthesizers and I'm already 
> reaching the limits of my current tools for phase noise and I can't afford an 
> E5052 for my own. At work I've one but I will probably not stay after august. 
> And anyway I need such tools in my lab at home…

If you have tools at work, the best possible thing to do is to get some 
oscillators / standards characterized. If you *know* what this or that 
oscillator is doing in terms of ADEV or phase noise at this Tau or frequency 
offset, it’s much easier to figure a lot of this out. 

The most basic way to do phase noise in the basement is with a single mixer 
setup running into some sort of audio FFT device. A sound card can be used or 
an audio spectrum analyzer. Parts are < $100 to get one setup once you can do 
the audio measurements. 

For ADEV, a DMTD or it’s cousin, the single mixer is the easy way to go. The 
single mixer does not get a lot of discussion these days. It is much easier to 
set up than a DMTD. It does require an offset oscillator. Once you have a 
single mixer phase noise setup, you are about half way to a single mixer ADEV 
setup. Cost for one is < $100 in parts. You already have a counter to collect 
the data out of it.

In both cases you are running a comparison device. Having a characterized OCXO 
to compare to is a really nice thing. 

Bob



> As low-noise and stable synthetizers depends on the standard used, I need as 
> well to measure them as well...
> 
> Let's start with this simple experiments and once I will understand the ins 
> and outs I will try to improve. I know techniques of cross-correlations and 
> you've already talked about DMTD that for sure I will have to come to...
> 
> Good night
> Stephane
> 
> -----Message d'origine-----
> De : Magnus Danielson [mailto:mag...@rubidium.se] 
> Envoyé : dimanche 18 janvier 2015 22:46
> À : Stéphane Rey; 'Discussion of precise time and frequency measurement'
> Cc : mag...@rubidium.se
> Objet : Re: [time-nuts] question Alan deviation measured with Timelab and 
> counters
> 
> Bonsoir Stéphane,
> 
> On 01/18/2015 10:34 PM, Stéphane Rey wrote:
>> Thanks a lot Bob and Magnus for your very helpful comments.
>> 
>> The HP5370a was indeed in TI mode. By the way what is the difference with 
>> +/-TI, the button just aside...
>> 
>> But I guess I understand where I've missed something : I've tried to put the 
>> Rb on channel A and the DUT on channel B but result was always the same but 
>> I do understand now that there is indeed a switch to change from COMmon to 
>> SEParate and it was always on COM meaning I believe that channel B wasn't 
>> used. This explains a lot of things I did not understand. I'm sorry for 
>> these so basic issues that might have been solved if I had read carefully 
>> the HP5370a manual first.
> 
> Good. This confirmation makes sense to be and Bob, now we can relax as the 
> mystery is solved.
> 
>> So possible conclusions until now are that I have actually measured the ADEV 
>> floor of the system rather than my DUT... which is already nice. The second 
>> conclusion from these oscillations seen with the GPSDO under test is that 
>> there is very likely in this GPSDO design a systemic noise added to the 10 
>> MHz output (power supply, PCB coupling, ... I'll make further investigations 
>> on it later on).
> 
> It's a great opportunity to learn the tools, and once you have the tools, you 
> can see if you can't improve things.
> 
>> I will experiment all the suggestions you made and will come back. For 
>> information the 1PPS from the HP58503b has a positive pulse width that is 
>> only few us length.
> 
> This only makes it hard to view on a scope, but long enough to reliably 
> trigger your counter and scope.
> 
>> Now, when considering that the method is to compare the DUT to an other 
>> source, I assume then that the other source shall be at least 1 order of 
>> magnitude better than the DUT. Otherwise this will be impossible to 
>> distinguish who is the instability contributor between the source and DUT, 
>> right ?
> 
> For a simple setup, yes. But then we are the time-nuts, we have ways of 
> handling these things. :) Let's get you started with the basic measurement, 
> it will be a good start.
> 
>> Then the second question is what kind of very stable source can be used to 
>> measure DUT which could be Rb or GPSDO which are already in the range of 
>> 10E-10 to 10E-12 < 100s ?
> 
> Time-nuts tend to spend their time and money getting even more stable clocks 
> and tools. If you have the right tool, you can measure near and
> *under* the noise-level of your reference, but not without running into 
> issues. One such trick is called cross-correlation, while another is to use 
> three-corner hat techniques.
> 
> Cheers,
> Magnus
> 
> 
> 
> ---
> L'absence de virus dans ce courrier électronique a été vérifiée par le 
> logiciel antivirus Avast.
> http://www.avast.com
> 
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.

_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.