Joe,

The single-shot resolution of the Option 040 series input is 10 ns.
The T.I. interval option adds jitter to make the averaging useful.

If your PPS was instead a burst or a clock, comparing it to your clock would make more sense and you could use that higher resolution.

If I had a GPIB interface for my 5328A, I would play around with it more.

I think I recall a 10 ps resolution being achievable according to fancy specs, with averaging to it's max.

Cheers,
Magnus

On 02/04/2015 09:58 PM, Joseph Gray wrote:
Sorry for the simple questions. Old hands can simply ignore this
message if it bothers you.

I have an HP 5328A that I have pieced together from broken units. As
soon as I make a ribbon cable for the GPIB card, I want to use it for
T.I. measurements. I know it isn't nearly as good as a 5370, but it is
what I have at the moment.

The manual states that the Universal module adds dithering to the
multiplied 100 MHz internal reference. If I use a GPSDO as an external
reference, will this multiplying/dithering allow me to use the same
GPSDO (divided down to PPS) on either the START or STOP inputs? If
not, then how about using a different GPSDO as the external reference?
Or even an FE5680A rubidium?

If I read the specs correctly, using T.I. averaging on the 5328A, a
1000 s average should get me 100 ps resolution? With this particular
instrument, what is the best, useful resolution I can expect?

Joe Gray
W5JG
_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to