The standard deviation is good and consistent with others measurements which 
seem to be in the range 10 to 15 psecs.

I think the channel skew is one half of 112psecs = 56 psecs (as it adds one way 
and subtracts the other) and this is close to the Keysight datasheet claim of 
50 psecs typically.

It is interesting to get some practical readings - when I was trying to decide 
on my counter it was very hard to get information - different datasheets give 
different parameters eg Pendulum CNT91/Tek fca 3100 has skew as < 500 psecs 
whilst Keysight put 50 psecs typical.

I ended up getting an fca3100 as I got a very good discount and I found that 
the skew I actually measured was (5.1282 nsecs - 5.0972 nsecs)/2 = 15psecs 
which is very good.

The std dev for the fca3100 on my unit was 30psecs which is 2 to 3 times worse 
than the 53230A which is consistent with the ratio of 50 psecs (fca3100) and 20 
psecs (53230A).

It seems that the typical one-shot resolution is around 2.5 times better than 
the datasheet (your measurement of 12psecs/sqrt(2) is 2.4 times better than the 
nominal 20 psecs whilst my fca 3100 measurement of 30psecs/sqrt(2) is also 2.4 
times better than the nominal 50 psecs.

Interestingly, the SR620 data sheet says the one shot is typically 20 psecs and 
is 50 psecs max which is also a similar ratio.

Perhaps there is some sort of industry standard that states datasheet one shot 
resolution should be expressed as 2.5 times the typical one shot resolution :).

James


 

 

 

-----Original Message-----
From: Anders Wallin <anders.e.e.wal...@gmail.com>
To: Discussion <time-nuts@febo.com>
Sent: Fri, 3 Apr 2015 12:04
Subject: [time-nuts] 53230A noise floor


Hi all,

I ran some tests on a new 53230A counter that just
arrived:
http://www.anderswallin.net/2015/04/keysight-53230a-noise-floor-test/

The
time-interval single-shot noise and resulting ADEV seems easy enough
to
explain.

I didn't yet have time to look closely at Enrico Rubiola's notes
over here:
http://rubiola.org/pdf-slides/2012T-IFCS-Counters.pdf
There is
probably a good explanation for the ADEV-level in standard
(pi-counting?)
reciprocal frequency counter mode as well as the roughly
1/sqrt(10) enhancement
in ADEV when increasing the gate-time 10-fold.

on another note:
The
discussion on phase-noise <-> ADEV calculations earlier was
interesting.
Datasets (synthetic or real) with known correct ADEVs and
phase-noise-spectra
would be a good addition to allantools and a motivation
to add code for the
phase-noise <-> ADEV
conversion.

cheers,
Anders
_______________________________________________
time-nuts
mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the
instructions there.

 
_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to