Hi Magnus,

Am I understanding correctly that the difference between Pi, Delta, and Omega
is basically one of software processing after the sample is made?

If so, how can this be best leveraged by those of us that have 5370A/B counters
that are running John Seamon's enhanced BBB processor in place of the original
5370 Motorola processor?

Thanks!

-Chuck Harris

Magnus Danielson wrote:
Fellow time-nuts,

Since I haven't seen any reports on this, I though I would write down a few 
lines.

While normal counters use a pair of phase-samples to estimate the frequency, now
called Pi counters (big pi, which has the shape of the weighing function of 
frequency
samples), counter vendors have been figuring out how to improve the precision 
of the
frequency estimation for the given observation time. One approach is to overlay
multiple measurements in blocks, which for the frequency estimation looks like a
triangle-shape weighing, so this type of counter is referred to as Delta 
counters
(again to resemble the shape).

Classical counters of the Pi shape is HP5370A, SR620 etc.
Classical counter of the Delta shape is the HP53132A.

However, counters using the Linear Regression methodology does not fit into 
either of
those categories. Enrico Rubiola derived the parabolic shape of the weighing 
function
(which I then independently verified after we spoke during EFTF 2014), and he 
then
passed on the results to Francois Vernotte and other colleagues to continue the
analysis.