I cannot say exactly, as I measured it as 59 Hz.  VA7OJ measured it as 60 Hz.  
I was using an HP lab supply, and Adam was using some other lab supply.  I put 
a 500 uF capacitor across the Vcc pin and it had no effect.  I don't think it 
has anything to do with the power supply since both of us saw the same thing.  
If an attachment will go through this reflector, I can post a spectrum analyzer 
screen shot.  

-----Original Message-----
From: time-nuts [mailto:time-nuts-boun...@febo.com] On Behalf Of jimlux
Sent: Saturday, February 27, 2016 12:07 PM
To: time-nuts@febo.com
Subject: Re: [time-nuts] Inside a CTS 1960017 OCXO



On 2/27/16 9:35 AM, Rob Sherwood. wrote:
> Hi Dan,
>
> I purchased two of these, and VA7OJ one.  One of mine is defective, and the 
> seller is going to replace it.  What was disappointing was all the spurious 
> at 60 Hz intervals relatively close-in to the 10 MHz carrier.  I suppose if 
> all one wants is a reasonably accurate 10 MHz, then they work OK.  My good 
> one and Adam's unit were 1x10^7 low in frequency once warmed up, and there 
> are pins for frequency adjustment and sync to another standard.  However if 
> one wanted it for an oscillator with good close-in phase noise, the spurs 
> ruin that hope.
>
> Do you know if there is a chopper at 60 Hz that runs the proportional oven 
> that is the source of all the spurious?
>


Is it right at line frequency, or is it "close" to 60 Hz?

_______________________________________________
time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.



--
If this email is spam, report it to
https://support.onlymyemail.com/view/report_spam/ODExMjI6MTg1NzM3NjQ1Mzpyb2JAbmMwYi5jb206ZGVsaXZlcmVk

_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to