> BTW: I discovered that Timelab stops processing after 10'000'000 datapoints,
> which is kind inconvenient when doing a long term measurment...
> 
> Attila Kinali

I've collected a day of TimeLab/TimePod data at tau 0.001 which is 86'400'000 
datapoints. Should be no problem.

Note Stable32 has a user-configurable limit (Conf -> Data -> 
Max_Data_File_Size). Or you can decimate during reading.

My command line tools have no sample limit (well, just malloc() limited) and 
can be orders of magnitude faster than Stable32 or Timelab since they are batch 
and not GUI.

But this begs the question -- what are you doing with 10M datapoints? Once you 
get beyond a couple of decades of data it's often better to compute statistics 
in segments and display all the segments of the whole as a time series.

So, for example, don't compute a single stddev or ADEV number from the entire 
10M data set. While this gives an apparently "more precise" measurement due to 
sampling, it will also hide key factors like trends, periodicity, spectral 
components, outliers, and glitches.

I'm not sure if you got your answer on synthetic data, but Stable32 has a data 
noise generator, where you get to specify alpha from -2 to +2. I created 1M 
samples of each of the 5 noise types and use those cached files as a noise 
reference.

See also the 5 noise types in high-res here:

http://www.leapsecond.com/pages/allan/Exploring_Allan_Deviation_v2.pdf
http://www.leapsecond.com/pages/allan/

/tvb