I saw in passing a question about preserving cal constants and calibration in general.

These scopes have NO externally-loaded cal constants, and no way I know of to either export or import them. They're (almost) completely self-calibrating. The details are covered in the service manual.

The only thing you have to check (or can change) is the +5.2 volt supply accuracy. That's used to derive internal voltages for calibration.

There is one other adjustable setting, but HP says it does not need to be adjusted routinely and should not be changed unless you replace an attenuator, acquisition hybrid, or the main board. It's to set the HF response for the channel in question. But, it's a real pain to do because you have to move the power supply and you'd need to make up extension cables to do so.

There is a reference to 'default constants', but those are not specific to an individual scope, they just adjust everything to known values for the self-calibration.

BTW, I'm impressed with the triggering on this scope. I decided to see how it handles 1Ghz. Triggers perfectly. Response is 7 dbV down from 500 Mhz but perfectly usable. Triggering was rock-stable. There are also an amazing number of ways to do triggering.

--