The 1503 was Tek's long-line (10km) TDR analyzer. IIRC, it put haversine pulses of about 5v onto the line under test (~10v open circuit). The 1502 (~600m line length) put pulses of about 200mV onto the line under test (~400mV open circuit).

Best regards,

Charles


Scott wrote:

This is highly dependent on the TDR especially ones designed for long twisted 
pair runs where a high voltage pulse is used to overcome resistive losses

David had written:

The Tektronix 1502 uses a tunnel diode pulser to produce a 50
picosecond output step of about 200 millivolts.  There is a misprint
in the theory section of the service manual which says "400 V" instead
of "400mV".

Scott had written:

Time Domain Reflectrometry is the usual technique for finding cable length but 
even there the cables NVP is an essential parameter if you want to compute 
length but not essential in time nuts application because we are interested in 
delay which a TDR reads directly When using a TDR its best if cable is 
unterminated as the discontinuity at the end is helpful as a marker.   Also 
most TDRs like the Tek 1502 can put 100v or more on the cable which will blow 
most GPS antennas


_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to