Hi

There are entire (big and heavy) books written on quantization errors…..

In a counter, there are a number of different sub-systems contributing 
to the error. Depending on the design, each may ( or may not) be a bit better 
than 
absolutely needed. Toss in things like 10 MHz reference feedthrough 
(which is decidedly weird statistically) and you have a real mess. 

The simple answer is that there is no single answer for a full blown counter.
For an ADC sampling system, there may indeed be a somewhat more
tractable answer to the question (unless feed through is an issue with 
your setup …).

Bob



> On Oct 30, 2016, at 8:32 AM, jimlux <jim...@earthlink.net> wrote:
> 
> On 10/29/16 10:14 PM, Tom Van Baak wrote:
>>> One might expect that the actual ADEV value in this situation would be
>>> exactly 1 ns at tau = 1 second.  Values of 0.5 ns or sqrt(2)/2 ns might not
>>> be surprising. My actual measured value is about 0.65 ns, which does not
>>> seem to have an obvious explanation.  This brings to mind various questions:
>>> 
>>> What is the theoretical ADEV value of a perfect time-interval measurement
>>> quantized at 1 ns? What's the effect of an imperfect measurement
>>> (instrument errors)? Can one use this technique in reverse to sort
>>> instruments by their error contributions, or to tune up an instrument
>>> calibration?
>> 
>> Hi Stu,
>> 
>> If you have white phase noise with standard deviation of 1 then the ADEV 
>> will be sqrt(3). This is because each term in the ADEV formula is based on 
>> the addition/subtraction of 3 phase samples. And the variance of normally 
>> distributed random variables is the sum of the variances. So if your 
>> standard deviation is 0.5 ns, then the AVAR should be 1.5 ns and the ADEV 
>> should be 0.87 ns, which is sqrt(3)/2 ns. You can check this with a quick 
>> simulation [1].
>> 
>> Note this assumes that 1 ns quantization error has a normal distribution 
>> with standard deviation of +/- 0.5 ns. Someone who's actually measured the 
>> hp 5334B quantization noise can correct this assumption.
>> 
> 
> isn't the distribution of quantization more like a rectangular distribution 
> (e.g. like an ADC).  so variance of 1/12th?
> 
> 
> 
> 
> 
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts